1st night with new ResMed S10 Auto (defective machine?)
Author Message
Dagmar Offline

Preferred Members

Posts: 97
Joined: Jul 2015

Machine: Resmed Airsense 10 Autoset
Mask Type: Nasal mask
Mask Make & Model: Wisp by Respironics
Humidifier: Resmed
CPAP Pressure: 10
CPAP Software: ResScan

Other Comments:

Sex: Male
Location: Michigan

Post: #1
1st night with new ResMed S10 Auto (defective machine?)
Hi Everyone,

My first night with my brand new machine didn't go well, I didn't feel like I was getting enough pressure. I checked SleepyHead this morning and found my graphs showed quite a different graph from my previous rental machine. With my rental, the graph stayed right at 10 cmh20 without fluctuating like this one. My new machine is set at 10 cmh20. I double checked the settings. Your thoughts appreciated. I also checked my leak rate which seemed fine to me, both pressure and leak rate graphs are attached.

Thanks, Dagmar


Attached File(s) Thumbnail(s)
       
11-12-2015 09:57 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Possum Offline

Advisory Members

Posts: 254
Joined: Oct 2015

Machine: Resmed Airsense Autoset A10
Mask Type: Nasal pillows
Mask Make & Model: Airfit P10
Humidifier: (included in CPAP model)
CPAP Pressure: 8-12
CPAP Software: ResScan SleepyHead

Other Comments: Innovo CMS-50F+ pulse oximeter

Sex: Male
Location: San Diego, California, USA

Post: #2
RE: 1st night with new ResMed S10 Auto (defective machine?)
(11-12-2015 09:57 AM)Dagmar Wrote:  Hi Everyone,

My first night with my brand new machine didn't go well, I didn't feel like I was getting enough pressure. I checked SleepyHead this morning and found my graphs showed quite a different graph from my previous rental machine. With my rental, the graph stayed right at 10 cmh20 without fluctuating like this one. My new machine is set at 10 cmh20. I double checked the settings. Your thoughts appreciated. I also checked my leak rate which seemed fine to me, both pressure and leak rate graphs are attached.

Thanks, Dagmar

I have the same machine, and yes, the pressure graph should be fairly smooth. Granted, I use ResScan to look at my data, not SleepyHead, so I'm not sure if the settings are different. I was prescribed a range of pressure, 7-12, so the machine auto adjusts all night depending on my needs; however, even the adjustments are fairly smooth and gentle, not noisy like your graph depicts. If you don't feel like you're getting enough air, you might have the Ramp turned on. If you turn it off, you'll go to your prescribed pressure almost instantly upon turning the machine on.

I suggest contacting your supplier, as it looks like there is an issue either with data recording or the machine itself.
11-12-2015 10:11 AM
Find all posts by this user Post Reply Quote this message in a reply
Dagmar Offline

Preferred Members

Posts: 97
Joined: Jul 2015

Machine: Resmed Airsense 10 Autoset
Mask Type: Nasal mask
Mask Make & Model: Wisp by Respironics
Humidifier: Resmed
CPAP Pressure: 10
CPAP Software: ResScan

Other Comments:

Sex: Male
Location: Michigan

Post: #3
RE: 1st night with new ResMed S10 Auto (defective machine?)
(11-12-2015 10:11 AM)Possum Wrote:  
(11-12-2015 09:57 AM)Dagmar Wrote:  Hi Everyone,

My first night with my brand new machine didn't go well, I didn't feel like I was getting enough pressure. I checked SleepyHead this morning and found my graphs showed quite a different graph from my previous rental machine. With my rental, the graph stayed right at 10 cmh20 without fluctuating like this one. My new machine is set at 10 cmh20. I double checked the settings. Your thoughts appreciated. I also checked my leak rate which seemed fine to me, both pressure and leak rate graphs are attached.

Thanks, Dagmar

I have the same machine, and yes, the pressure graph should be fairly smooth. Granted, I use ResScan to look at my data, not SleepyHead, so I'm not sure if the settings are different. I was prescribed a range of pressure, 7-12, so the machine auto adjusts all night depending on my needs; however, even the adjustments are fairly smooth and gentle, not noisy like your graph depicts. If you don't feel like you're getting enough air, you might have the Ramp turned on. If you turn it off, you'll go to your prescribed pressure almost instantly upon turning the machine on.

I suggest contacting your supplier, as it looks like there is an issue either with data recording or the machine itself.

Thanks for the reply, I do have my ramp set at 6 but it is only a five minute ramp. I have attached a typical graph from my previous machine where all settings were the same as the new one, straight 10 cmH20, five minute ramp starting at 6. I will do a walk-in at the DME today. But, I don't want them telling me everything looks normal with the machine when I know it didn't feel at all right to me.

Dagmar


Attached File(s) Thumbnail(s)
   
11-12-2015 10:49 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
MobileBasset Offline

Advisory Members

Posts: 198
Joined: May 2015

Machine: ResMed AIRCURVE 10 V Auto
Mask Type: Nasal mask
Mask Make & Model: Fisher Paykel Zest
Humidifier: Integrated
CPAP Pressure: 10 (minE 6,maxI 12, PS 3)
CPAP Software: ResScan SleepyHead Other Software

Other Comments:

Sex: Female
Location: Florida

Post: #4
RE: 1st night with new ResMed S10 Auto (defective machine?)
Do you have the EPR on? It looks like the pressure is responding to a max EPR setting as you breath.

if you can't decide then you don't have enough data.
11-12-2015 11:05 AM
Find all posts by this user Post Reply Quote this message in a reply
PaytonA Offline
Wiki Editor
Monitors

Posts: 3,005
Joined: Dec 2013

Machine: ResMed S9 VPAP Auto
Mask Type: Full face mask
Mask Make & Model: Resmed Mirage Quattro
Humidifier: H5i(distilled-top up)
CPAP Pressure: VAuto MinE14.0 MaxI 20.6 PS4.0
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Orange County,California

Post: #5
RE: 1st night with new ResMed S10 Auto (defective machine?)
Two reasons for the graph looking different. One is the scale of the graphs ore different. The other is that it looks like you had EPR set at 1 on the old machine and it is set at 3 on the new one.
11-12-2015 11:27 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Dagmar Offline

Preferred Members

Posts: 97
Joined: Jul 2015

Machine: Resmed Airsense 10 Autoset
Mask Type: Nasal mask
Mask Make & Model: Wisp by Respironics
Humidifier: Resmed
CPAP Pressure: 10
CPAP Software: ResScan

Other Comments:

Sex: Male
Location: Michigan

Post: #6
RE: 1st night with new ResMed S10 Auto (defective machine?)
(11-12-2015 11:27 AM)PaytonA Wrote:  Two reasons for the graph looking different. One is the scale of the graphs ore different. The other is that it looks like you had EPR set at 1 on the old machine and it is set at 3 on the new one.


I'm 99% sure EPR was set at 3 on the old machine as well. I never changed the EPR setting myself so I can't be 100% sure but that's what I remember seeing on the old machine. The new machine does have an EPR setting of 3. I think what I am really trying to figure out is how does (if it does) the pressure graph correlate to how my pressure felt during the night. I wouldn't have checked the graph this morning if something hadn't felt wrong in the night. By wrong meaning that the machine didn't seem to provide the usual amount of pressure. It was enough of a difference that I found it both refreshing and concerning. It was nice feeling I had a lower pressure but concerning that it wasn't my normal therapy. I guess I will find out more when I take the machine in for evaluation. I appreciate the comments.
11-12-2015 12:13 PM
Find all posts by this user Post Reply Quote this message in a reply
PaytonA Offline
Wiki Editor
Monitors

Posts: 3,005
Joined: Dec 2013

Machine: ResMed S9 VPAP Auto
Mask Type: Full face mask
Mask Make & Model: Resmed Mirage Quattro
Humidifier: H5i(distilled-top up)
CPAP Pressure: VAuto MinE14.0 MaxI 20.6 PS4.0
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Orange County,California

Post: #7
RE: 1st night with new ResMed S10 Auto (defective machine?)
Frankly, I have looked closely at the mask pressure graph from the old machine and it is quite certain that you were getting 1 cm/H2O of expiratory pressure relief regardless of the setting. With the new one it is quite obvious from the graph that you are getting 3cm/H2O expiratory pressure relief.

Best Regards,

PaytonA
11-12-2015 04:33 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Dagmar Offline

Preferred Members

Posts: 97
Joined: Jul 2015

Machine: Resmed Airsense 10 Autoset
Mask Type: Nasal mask
Mask Make & Model: Wisp by Respironics
Humidifier: Resmed
CPAP Pressure: 10
CPAP Software: ResScan

Other Comments:

Sex: Male
Location: Michigan

Post: #8
RE: 1st night with new ResMed S10 Auto (defective machine?)
Thanks PaytonA, the DME found no problem with the machine and did not catch the difference in pressure relief settings. Your analysis was really helpful to me. Now, I know why the new machine was so different. Thanks for the help! I had never touched or played with that setting so didn't know how the differences would feel.
(This post was last modified: 11-12-2015 08:17 PM by Dagmar.)
11-12-2015 07:17 PM
Find all posts by this user Post Reply Quote this message in a reply
Dagmar Offline

Preferred Members

Posts: 97
Joined: Jul 2015

Machine: Resmed Airsense 10 Autoset
Mask Type: Nasal mask
Mask Make & Model: Wisp by Respironics
Humidifier: Resmed
CPAP Pressure: 10
CPAP Software: ResScan

Other Comments:

Sex: Male
Location: Michigan

Post: #9
RE: 1st night with new ResMed S10 Auto (defective machine?)
I tested my EPR settings a little last night and found quite a large difference between Off, 1, 2, and 3. For now, I'm keeping my machine on an EPR setting of 1. A setting of one seems to provide me with the most comfortable breathing on the CPAP machine.
11-13-2015 03:05 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
PaytonA Offline
Wiki Editor
Monitors

Posts: 3,005
Joined: Dec 2013

Machine: ResMed S9 VPAP Auto
Mask Type: Full face mask
Mask Make & Model: Resmed Mirage Quattro
Humidifier: H5i(distilled-top up)
CPAP Pressure: VAuto MinE14.0 MaxI 20.6 PS4.0
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Orange County,California

Post: #10
RE: 1st night with new ResMed S10 Auto (defective machine?)
Absolutely use what works best for you.
11-13-2015 06:10 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  ResMed Airsense 10 Auto Set...Runaway Pressure Abnormal86 14 528 11-22-2016 11:39 PM
Last Post: sonicboom
Question First Machine, First Night, First results..Help please? SJGino 9 387 11-18-2016 11:55 PM
Last Post: SJGino
  [CPAP] ResMed Air Sense 10 Auto Off Problem Buddie43 16 3,234 11-12-2016 02:51 PM
Last Post: sonicboom
  My first night with DreamStation Auto Perchas 6 705 11-07-2016 02:51 PM
Last Post: Sleeprider
Exclaimation CPAP Machine Choices - read this before you accept a new machine SuperSleeper 2 6,066 10-12-2016 11:37 AM
Last Post: DariaVader
  difference between Apex XT Auto and Resmed S9 Autoset Lanco 11 441 10-11-2016 03:29 PM
Last Post: Sleeprider
  airsense10 auto cpap(for him) VS DreamStation Auto CPAP DSX500T11 w/HT humid rookiepom 5 854 10-09-2016 07:52 AM
Last Post: stephennic

Forum Jump:

Who's Online (Complete List)