#11
Back up rate is set for Auto.
Post Reply Post Reply
#12
Also forgot to update settings.
Post Reply Post Reply
#13
Ah, okay. Previously you had indicated off or zero.
Post Reply Post Reply


#14
If you check, my machine is set for Pmax 14, APAPmax 11, APAPmin 9. PAmax 4, PSmin 0, BPM Auto. Last night with theses settings my AHI was between 1.3 and 3.5 for the whole night of 7.5 hrs.
Post Reply Post Reply
#15
(Yesterday, 09:02 AM)RLSdss Wrote: If you check, my machine is set for Pmax 14, APAPmax 11, APAPmin 9. PAmax 4, PSmin 0, BPM Auto. Last night with theses settings my AHI was between 1.3 and 3.5 for the whole night of 7.5 hrs.

Some of your terminology is a bit confusing to me:
Pmax  14 (maximum pressure or IPAP max)
APAP max 11 (EPAP max maximum expiratory pressure?)
APAPmin 9 (EPAP min minimum expiratorypressure?)
PAmax  4 (PSmax?)
PSmin 0

If my understanding is correct, your minimum pressure is 9/9 (PS 0 over 9 EPAP).
Your EPAP may rise to 13/9 to 14/11 (PS 0-4 over 14/11 pressure where IPAP max is limited.  
If you have a central apnea, you will need at least 8-10 cm of pressure to cause a breath, so at no time are you getting the benefit of ASV pressure support because it is limited by PSmax and IPAPmax or Pmax.  To let your machine actually help you, you need a PS max of 10 to 12, not 4.0, and a Pmax (IPAPmax) of 21.0, which is 10-cm above your EPAP max.  What you are currently getting is like a BiPAP Auto with very limited pressure support. Although you have a backup rate enabled, the most it is able to provide as inspiratory assist is 3-4 cm of pressure support.

For ASV you need:
EPAP min 9.0
EPAP max 11.0
PS min 0
PS max 10
IPAP max (Pmax) 21.0
Post Reply Post Reply
#16
I meant epap and that is what is on my side bar. If I change to the settings you recommend what would you expect my AHI to be? I'm currently getting an AHI of 1.3 to 3.4.
Post Reply Post Reply


#17
OK I'm trying to follow this. I went back to another thread and you said you were misdiagnosed with Central's and that they were going to put you on a Auto Bipap. Now your asking about settings for the ASV and it sounds like you do get Centrals. Can you please clear up my confusion?
Post Reply Post Reply
#18
(Yesterday, 09:41 AM)RLSdss Wrote: I meant epap and that is what is on my side bar. If I change to the settings you recommend what would you expect my AHI to be? I'm currently getting an AHI of 1.3 to 3.4.

Your AHI is in an acceptable range and we don't know if those events are obstructive, or CA and H which would be targeted by ASV.  Your objectives for the therapy and sense of well-being are yours to define.  If you want to try the full functionality of ASV, then the settings above would work; otherwise, you are using auto-bilevel.  ASV can provide benefits beyond AHI reduction, but since you can't post the data, it's very hard for us to judge.  Hopefully you got copies of your sleep test and have a good sense of why ASV was prescribed, or the reasons it is not needed. At the current settings, you have not experienced ASV.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  [Equipment] DreamStation Auto Humidification problem Giasdaddy2010 16 7,726 10-26-2017, 04:09 PM
Last Post: pknowles
  ResMed Airsense 10 Setting Date Problem Reuven Gruber 10 1,117 10-13-2017, 10:46 AM
Last Post: jaswilliams
  How do you open the humidifier? Problem solved. HalfAsleep 4 247 09-29-2017, 12:10 PM
Last Post: Timur
  [Treatment] Flow Limitations, a problem or not? storywizard 9 360 09-23-2017, 08:34 AM
Last Post: Sleeprider
Question [Pressure] New ResMed Airsense 10 Pressure Problem spookyac47 5 574 09-16-2017, 03:44 PM
Last Post: trish6hundred
  PR System One Humidifier Problem gludden 7 245 09-16-2017, 03:17 PM
Last Post: trish6hundred
  Advice Needed- Problem With Recording on the CMS 50D + Oximeter nico08 4 280 09-07-2017, 12:08 AM
Last Post: aquanaut20

Forum Jump:

New Posts   Today's Posts




About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.

For any more information, please use our contact form.