My S9 Autoset seems to behave in ways I do not understand.

What I thought I understood was that it begins at the lowest pressure in the range that is set, and only raises pressure after a obstructive event or in case of a leakage event, and I am not even sure it is designed to respond to leakage by raising pressure.

But according to SH it seems to be raising pressure a cm or two at a time randomly. IOW, there seems to be no concurrent events associated with the changes (raises) it makes to pressure.

So I must be missing something about how this is supposed to work. Can someone clear up for me what might be going on here?

TIA
12-26-2014 02:10 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Ghost1958 Offline

Advisory Members

Posts: 957
Joined: Aug 2014

Machine: PRS1 Series Sixty Auto Aflex
Mask Type: Nasal mask
Mask Make & Model: F&P Eson
Humidifier: PRS1 with Heated Hose
CPAP Pressure: 10.5 min 16 max
CPAP Software: SleepyHead EncoreBasic

Other Comments: Moderate to Severe SA Obstructive. Rapid O2 desat off the machine.

Sex: Male
Location: KY

Post: #2
RE: APAP behavior puzzling to me
Never had a Resmed machine but autos pretty much either respond to events or to changes in the wave form they are programmed to see as event precursors and adjust a bit to prevent the event in the first place which is ideal if the machine can accomplish that.

It cant always do that though and reponds after an event that did slip thru in an attempt to prevent further ones. Some Autos respond faster some slower. But something in the waveform is a trigger to the machines programming to respond.
12-26-2014 02:19 AM
Find all posts by this user Post Reply Quote this message in a reply
Galactus Offline

Advisory Members

Posts: 1,384
Joined: Jan 2014

Machine: Phillips Respironics System One PRS1 DS760P Auto BiPAP
Mask Type: Nasal pillows
Mask Make & Model: ResMed Airfit P10 Nasal Pillow
Humidifier: System One Heated Humidifier + Climate Line
CPAP Pressure: E/I 13/20 PS 4/7 BiFlex 3
CPAP Software: SleepyHead EncoreBasic

Other Comments: Sleep can be a good thing.

Sex: Male
Location: USA

Post: #3
RE: APAP behavior puzzling to me
The unit works in the be background to prevent events, it does not really respond to them. It responds to what it sees in the flow graph and uses an algorithm to raise or lower pressure in advance of events. It does it's best to keep a pattern of breathing so that it is proactive rather than reactive. Hope that helps.

If everyone thinks alike, then someone isn't thinking.
Everyone knows something, together we could know everything.
12-26-2014 02:24 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
DeepBreathing Offline
Wiki Editor
Moderators

Posts: 2,294
Joined: Sep 2013

Machine: Resmed S9 VPAP Adapt
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: Resmed H5i
CPAP Pressure: EPAP: 9 - 15 PS: 3 - 10
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Perth, Western Australia

Post: #4
RE: APAP behavior puzzling to me
It also reacts to flow limitations which could be precursors to an apnea (see Galactus' post above). If you turn on the flow limits graph in Sleepyhead you might see a correlation.

DeepBreathing
Apnea Board Moderator
www.ApneaBoard.com


Bed

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
12-26-2014 02:38 AM
Find all posts by this user Post Reply Quote this message in a reply
Crimson Nape Offline

Monitors

Posts: 1,325
Joined: Oct 2014

Machine: ResMed S9 Autoset
Mask Type: Other
Mask Make & Model: P-10 / F&P Simplus / DreamWear
Humidifier: H5i w/Climateline
CPAP Pressure: 8 - EPR 3
CPAP Software: SleepyHead

Other Comments: CMS-50F & 50IW

Sex: Male
Location: Georgia

Post: #5
RE: APAP behavior puzzling to me
I just changed my S9 to a VPAP with a setting of 8-12 from a CPAP 10cmH2O. It turns on at 8 and ramps right up to 12 in about 30 seconds. It seems to ignore the minimum setting and go straight for the maximum. It has done this for the past 5 nights. I was going to post about this problem until I saw this thread. I'm going to do a Cold Boot and see if that helps.

Statistics prove that people who have more birthdays live longer.
12-26-2014 09:54 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
TyroneShoes Offline

Preferred Members-2

Posts: 505
Joined: Oct 2014

Machine: Resmed S9 Autoset
Mask Type: Nasal pillows
Mask Make & Model: AirFit P10
Humidifier: Resmed H5i
CPAP Pressure: 10-14
CPAP Software: SleepyHead

Other Comments:

Sex: Undisclosed
Location:

Post: #6
RE: APAP behavior puzzling to me
(12-26-2014 02:24 AM)Galactus Wrote:  The unit works in the be background to prevent events, it does not really respond to them. It responds to what it sees in the flow graph and uses an algorithm to raise or lower pressure in advance of events. It does it's best to keep a pattern of breathing so that it is proactive rather than reactive. Hope that helps.

It does, and thanks.

I want to clarify something where I mis-spoke. The term "randomly" implies that there is no reason for pressure to rise, and obviously there are reasons.

As far as lowering pressure, I think I understand that better than I understand why it raises pressure. My understanding is that in the absence of events or precursors, it lowers naturally on a percentage-per-minute basis, which seems pretty reasonable. That also seems to be borne out by a "ski-slope" response as it comes back down in the SH graph.

Your explanation of raising pressure seems pretty reasonable too. It is I guess pretty comforting that the APAP can do this (which is why I found it puzzling that my sleep doc told me to brick this thing at straight 8).

It seems like what I have heard so far is that it does not respond to CAs, because raising pressure would not really help, and that it does not respond to OAs until after one passes (assuming it could not predict via a precursor and was somewhat blindsided by that happening).

Its all very interesting stuff. I did notice that leak events do not seem to make any difference, but I did see it go up about 1.5 cm in response to a single large snore event. I will have to take a closer look at whether there is a correlation between auto-pressure and flow limitation.
(This post was last modified: 12-26-2014 04:00 PM by TyroneShoes.)
12-26-2014 03:54 PM
Find all posts by this user Post Reply Quote this message in a reply
Terry Offline

Advisory Members

Posts: 553
Joined: May 2014

Machine: Philips Respironics RemStar Auto with A-Flex (560P)
Mask Type: Nasal pillows
Mask Make & Model: P/R Nuance & Pap Cap
Humidifier: Integrated
CPAP Pressure: Auto
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location:

Post: #7
RE: APAP behavior puzzling to me
(12-26-2014 03:54 PM)TyroneShoes Wrote:  It seems like what I have heard so far is that it does not respond to CAs, because raising pressure would not really help, and that it does not respond to OAs until after one passes (assuming it could not predict via a precursor and was somewhat blindsided by that happening).

Its all very interesting stuff. I did notice that leak events do not seem to make any difference, but I did see it go up about 1.5 cm in response to a single large snore event. I will have to take a closer look at whether there is a correlation between auto-pressure and flow limitation.

I have no idea if your particular machine does this, however there are flow patterns that can be detected before apneas occur, and the pressure adjusted to stop them from ever happening.
12-26-2014 04:12 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Sleeprider Offline
Wiki Editor
Advisory Members

Posts: 3,317
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #8
RE: APAP behavior puzzling to me
(12-26-2014 09:54 AM)sgearhart Wrote:  I just changed my S9 to a VPAP with a setting of 8-12 from a CPAP 10cmH2O. It turns on at 8 and ramps right up to 12 in about 30 seconds. It seems to ignore the minimum setting and go straight for the maximum. It has done this for the past 5 nights. I was going to post about this problem until I saw this thread. I'm going to do a Cold Boot and see if that helps.

We don't know from your post what AHI or other events are being recorded on the baseline, so have you observed an improvement? What happens if you open the top end to 14? The machine programming is seeing something in your breathing patterns that cause the pressure increase. If it wasn't required, it stays at the lowest setting. I doubt a power-off cycle will change that.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
12-26-2014 05:28 PM
Find all posts by this user Post Reply Quote this message in a reply
TyroneShoes Offline

Preferred Members-2

Posts: 505
Joined: Oct 2014

Machine: Resmed S9 Autoset
Mask Type: Nasal pillows
Mask Make & Model: AirFit P10
Humidifier: Resmed H5i
CPAP Pressure: 10-14
CPAP Software: SleepyHead

Other Comments:

Sex: Undisclosed
Location:

Post: #9
RE: APAP behavior puzzling to me
(12-26-2014 09:54 AM)sgearhart Wrote:  I just changed my S9 to a VPAP with a setting of 8-12 from a CPAP 10cmH2O. It turns on at 8 and ramps right up to 12 in about 30 seconds. It seems to ignore the minimum setting and go straight for the maximum. It has done this for the past 5 nights. I was going to post about this problem until I saw this thread. I'm going to do a Cold Boot and see if that helps.

There are two ways of looking at this.

First, it may be that this does not sound like it is working correctly. It seems that an APAP starts at the minimum pressure in the set range, and only goes up if it has a reason to, such as responding to breathing feedback and SA events. If the machine climbs up to the top of the range this quickly, it may be operating differently than what should be expected, as it does not seem that it even has time to respond to breathing feedback. And if little else has changed, why would the behavior of the machine change abruptly, unless it was not working properly?

On the other hand, it may be that your min pressure should be 12 or above, and there was indeed an abrupt change (in setting an APAP range over operating it as a brick). The straight 10 may be an educated guess based on a sleep study, and was too low in the first place.

8-12 is not a significantly large range, and not significantly different than straight 10. So maybe the machine really is seeing something immediately that is telling it to go to the top of its current set range. This is the advantage of APAP over CPAP in that it has the intelligence to titrate your pressure automatically based on feedback. IOW, it may be operating absolutely correctly.

Regardless, and speaking as an Engineer, my experience tells me that a cold boot seems like a reasonable thing to try, and it is non-invasive and can't really make anything worse. I am curious to see if it helps (odds are not that good, but it's worth a shot), so please post back your results.

I would also take this up with your sleep doc and your DME.
(This post was last modified: 12-26-2014 06:03 PM by TyroneShoes.)
12-26-2014 05:39 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
vsheline Offline

Advisory Members

Posts: 1,907
Joined: Jul 2012

Machine: S9 VPAP Adapt (USA Model# 36007, not better 36037 or 36067)
Mask Type: Full face mask
Mask Make & Model: F10 or SimPlus w/ 2Liners. MirageQuatro & Gecko gel pad
Humidifier: H5i
CPAP Pressure: 15 EPAP, PS 5-10
CPAP Software: ResScan

Other Comments: Marfan Syndrome, chronic bradycardia, occasional Cheyne-Stokes Respiration

Sex: Male
Location: California, USA

Post: #10
RE: APAP behavior puzzling to me
(12-26-2014 09:54 AM)sgearhart Wrote:  I just changed my S9 to a VPAP with a setting of 8-12 from a CPAP 10cmH2O. It turns on at 8 and ramps right up to 12 in about 30 seconds. It seems to ignore the minimum setting and go straight for the maximum. It has done this for the past 5 nights. I was going to post about this problem until I saw this thread. I'm going to do a Cold Boot and see if that helps.

Hi sgearhart,

You no longer have an S9 AutoSet? You now have the S9 VPAP Auto?

Obstructive Sleep Apnea is usually highly positional, with flat on our back usually the worst.

Can you try sleeping only on your side?

If using the Ramp, what is the Start a Pressure and Ramp Time?

What therapy mode is your machine set to?

If therapy mode is Auto, what are the settings for EPAP, Max IPAP, and Pressure Support?

If therapy mode is S (Synchronous), what are the settings for EPAP and IPAP?

If therapy mode is CPAP, what are the settings for Pressure and EPR?

Take care,
--- Vaughn

Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment. The Advisory Member group provides advice and suggestions to Apnea Board administrators and staff on matters concerning Apnea Board operation and administrative policies - not on matters concerning treatment for Sleep Apnea. I think it is now too late to change the name of the group but I think Voting Member group would perhaps have been a more descriptive name for the group.
(This post was last modified: 12-26-2014 06:09 PM by vsheline.)
12-26-2014 05:48 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  Medicare Advantage Coverage for APAP Evie 3 180 11-30-2016 08:22 PM
Last Post: Mike1953
  cpap and apap not effictive Ezil71 109 3,078 11-05-2016 11:19 AM
Last Post: sonicboom
  Unable to clear throat while on APAP NorthernGuy 5 304 10-17-2016 08:18 PM
Last Post: Sleeprider
  Apap I_will_never_sleep_again 2 275 10-14-2016 02:58 PM
Last Post: trish6hundred
  Travel to Australia with Apap. Newbee2016 10 529 09-24-2016 12:06 PM
Last Post: SharonN
  Looking for complete list of AUTO apap's compatible with Sleepyhead Lanco 9 471 09-21-2016 08:08 PM
Last Post: Sleepster
  Had to STOP APAP Roy289 13 1,038 09-18-2016 08:52 AM
Last Post: Roy289

Forum Jump:

Who's Online (Complete List)