RE: Adjusting my pressure down, advice
generally treatment starts at lowest EPAP (min EPAP) and the IPAP is the EPAP plus the lowest PS setting.
--- example min EPAP=5 PS is 2 min 4 max. start is EPAP 5 IPAP 7.
as treatment continues, IPAP and EPAP may increase, but only until the IPAP gets to max IPAP.
--- example min EPAP=5, max IPAP=12, PS=2 min 2 max. while you start at EPAP 5 IPAP 7, you can end up as high as EPAP 10 IPAP 12, but no further.

it just gets more complicated from there.

suffice it to say, as #vsheline states, raise maxIPAP to allow room to resolve OA, raise maxPS to allow the lowest EPAP to minimize treatment induced CA, increase minIPAP to reduce the amount the machine has to climb in order to help you.

hope this helps. please set your ramp above 8 or turn it off...

QAL



.
Dedicated to QALity sleep.
Post Reply Post Reply
#12
RE: Adjusting my pressure down, advice
Thank you sir I'm going to absorb this and make my changes. Hopefully I can seek your advise again.
Post Reply Post Reply
#13
RE: Adjusting my pressure down, advice
(02-03-2015, 12:01 AM)quiescence at last Wrote: as #vsheline states, raise maxIPAP to allow room to resolve OA, raise maxPS to allow the lowest EPAP to minimize treatment induced CA, increase minIPAP to reduce the amount the machine has to climb in order to help you.

Yes, except I wouldn't recommend increasing Max PS because that could increase the number of centrals.
The Advisory Member group provides advice and suggestions to Apnea Board administrators and staff on matters concerning Apnea Board operation and administrative policies.  Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment.
Post Reply Post Reply
#14
RE: Adjusting my pressure down, advice
@vsheline - would you try to keep PS between 2 and 3 in that situation?
Dedicated to QALity sleep.
Post Reply Post Reply
#15
RE: Adjusting my pressure down, advice
(02-03-2015, 09:52 PM)quiescence at last Wrote: @vsheline - would you try to keep PS between 2 and 3 in that situation?

Increasing Max PS may allow PS to occasionally go higher. A higher PS would be expected to be helpful toward avoiding Flow Limitation and RERA events, but in some patients may also cause more central apneas.

Decreasing the PS may worsen how many RERA events we get but in some patients may also reduce the number of central apneas we get.

Original poster often has too many centrals, but at least sometimes has more RERA events than any other type. So hard to know which whether to increase or decrease improved sleep position might be most important.

If centrals start acting up strongly again, may be helpful to lower PS but raise Min EPAP by an equal amount.
The Advisory Member group provides advice and suggestions to Apnea Board administrators and staff on matters concerning Apnea Board operation and administrative policies.  Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment.
Post Reply Post Reply
#16
RE: Adjusting my pressure down, advice
So I received prescription based on titration. The Dr ordered a new machine ResMed Aircurve10 VAuto with settings of IPAP max17 and EPAP min12. PS of 4. This means a pressure of at least 16 IPAP? What is this pressure going to do to my Centrals? I don't understand the reasoning other than I didn't exhibit Centrals during titration.
Post Reply Post Reply
#17
RE: Adjusting my pressure down, advice
(02-08-2015, 08:59 PM)sumzzzs Wrote: So I received prescription based on titration. The Dr ordered a new machine ResMed Aircurve10 VAuto with settings of IPAP max17 and EPAP min12. PS of 4. This means a pressure of at least 16 IPAP? What is this pressure going to do to my Centrals? I don't understand the reasoning other than I didn't exhibit Centrals during titration.

Hi sumzzzs,

You already have an Auto bilevel device. I would not expect the AirCurve 10 VAuto to work any better than the machine you have.

Request by email the Set-up Manual (clinician guide) for the AirCurve 10 VAuto to see whether it even reports RERA events. I am pretty sure it does not.

I suggest not accepting the precription unless your doctor can explain how the AirCurve 10 VAuto would be better than your present machine. I suspect it would be worse, since I think it does not report RERAs, which in the data you posted were your dominant cause of arousals and fragmented sleep.


The Advisory Member group provides advice and suggestions to Apnea Board administrators and staff on matters concerning Apnea Board operation and administrative policies.  Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment.
Post Reply Post Reply
#18
RE: Adjusting my pressure down, advice
both the titration study and our observations are consistent that you should be seeking (1) a higher start pressure, and (2) a higher max pressure. will you slowly adjust your pressure up to the prescribed pressure to see what that does?

an example adjustment plan might be -
assuming your current setting is max IPAP 14 min EPAP 8, and ramp on:

2/9 -> turn ramp off, max IPAP 15.5 min EPAP 9. use for 4 days.
2/13 -> max IPAP 17 min EPAP 10. use for 4 days.
2/17 -> max IPAP 17 min EPAP 11. use for 4 days.
2/21 -> max IPAP 17 min EPAP 12

in doing so, you might find by 2/17 your treatment seems optimal and may not see the value in making the last two adjustments.

The important thing to note is you can use your present machine to accomplish this.

Good luck.

QAL

ps. would you consider posting an overview similar to below on 2/17 for last 21 to 30 days?

If you do, please set your preference to show RDI versus AHI. [select File Preferences and select the CPAP tab, and you'll see this:]
[Image: zUPgIzU.png]

Then your Overview Tab Chart should look like this.
[Image: QodtmZQ.png]

This will help us gauge your progress on RERA and AHI.


.
Dedicated to QALity sleep.
Post Reply Post Reply
#19
RE: Adjusting my pressure down, advice
(02-09-2015, 05:54 AM)quiescence at last Wrote: both the titration study and our observations are consistent that you should be seeking (1) a higher start pressure, and (2) a higher max pressure. will you slowly adjust your pressure up to the prescribed pressure to see what that does?

an example adjustment plan might be -
assuming your current setting is max IPAP 14 min EPAP 8, and ramp on:

2/9 -> turn ramp off, max IPAP 15.5 min EPAP 9. use for 4 days.
2/13 -> max IPAP 17 min EPAP 10. use for 4 days.
2/17 -> max IPAP 17 min EPAP 11. use for 4 days.
2/21 -> max IPAP 17 min EPAP 12

in doing so, you might find by 2/17 your treatment seems optimal and may not see the value in making the last two adjustments.

The important thing to note is you can use your present machine to accomplish this.

Good luck.

QAL

ps. would you consider posting an overview similar to below on 2/17 for last 21 to 30 days?

If you do, please set your preference to show RDI versus AHI. [select File Preferences and select the CPAP tab, and you'll see this:]
[Image: zUPgIzU.png]

Then your Overview Tab Chart should look like this.
[Image: QodtmZQ.png]

This will help us gauge your progress on RERA and AHI.


.

Ok I'm going to raise Max IPAP tonight my min EPAP is now 9. You know I've noticed a good majority of CA is actually in periods of being semi awake such as this morning when I was anticipating waking to my alarm.

Post Reply Post Reply
#20
RE: Adjusting my pressure down, advice
[Image: l4aTHM0l.png]


Amazing how much better this looks after a little tweak of my my pressure settings. First time under 5, wow, and I feel good also.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  New to CPAP, looking for guidance and advice iamJB 16 295 5 hours ago
Last Post: Expat31
  New- CA CA-Tired help with settings advice jenjabba 8 172 Yesterday, 01:03 PM
Last Post: SarcasticDave94
  New Here, Been using OSCAR, Need Advice Regarding Adjustments FrankReality 26 868 03-26-2024, 01:49 PM
Last Post: FrankReality
  [Pressure] Advice on pressure adjustments akarifur 43 1,801 03-25-2024, 08:58 PM
Last Post: Uzupan
  [CPAP] Adjusting for flow limitations jane2911 10 307 03-25-2024, 01:24 PM
Last Post: jane2911
  [Pressure] Looking for advice about ASV settings rosebladepat 7 144 03-15-2024, 07:12 PM
Last Post: rosebladepat
  [Treatment] Looking for Advice jlacassin 0 79 03-15-2024, 03:51 PM
Last Post: jlacassin


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.