AirSense10AutoSet & AmaraView mismatch?
#11
(01-19-2016, 04:52 PM)justMongo Wrote: Yes, but for the machine to know the vent rate, it has a table in firmware for vent rate versus pressure. And, it has one for each mask type.
excuse me, I get confused with numbers

Amara view mask vents at 12 about 35 L/m
ResMed full face mask at 12 about 41 L/m

ResMed nasal mask at 12 about 34 L/m
ResMed nasal pillows mask at 12 about 37 L/m

Maybe select mask type ... Nasal, as its the closest to Amara view vents leak rate

Post Reply Post Reply
#12
(01-19-2016, 11:28 PM)zonk Wrote: excuse me, I get confused with numbers

Amara view mask vents at 12 about 35 50 L/m
ResMed full face mask at 12 about 41 L/m

ResMed nasal mask at 12 about 34 L/m
ResMed nasal pillows mask at 12 about 37 L/m

Maybe select mask type ... Nasal, as its the closest to Amara view vents leak rate Won't work because Amara View vents at 50 LPM @ pressure =12.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#13
(01-19-2016, 12:21 PM)bazder Wrote: Phillips Respironics has a .pdf giving intentional leak rates for different masks but I'm not allowed to post links here yet. If I understood it correctly I think they estimate about 28L/min for the Amara View mask.
Do you mind copy intentional leak rates at different pressures from the pdf, thanks

I've an old pdf download shows intentional leak rates for different masks, including Amara which shows 33 L/m at 15
Also, I find Philips Respironics pdf from Supplier #26 for Amara View, shows 38 L/m at 15

Sorry, justMongo, I'm trying to get to the bottom of it Smile


Post Reply Post Reply


#14
Me too. I am looking at the top vent rate graph in my post #4 that shows 50 LPM at pressure = 12.
I'll have to use some Google-fu to see if there is contrary information on the vent rate for the Amara View.
INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#15
this is the .pdf about Amara View intentional leakage rates

https://www.continuedcare.com/wp-content...-ports.pdf
Post Reply Post Reply
#16
(01-20-2016, 02:22 PM)bazder Wrote: this is the .pdf about Amara View intentional leakage rates

https://www.continuedcare.com/wp-content...-ports.pdf
Thanks, I updated the old thread, links disappear after a while
http://www.apneaboard.com/forums/Thread-...#pid147166
Post Reply Post Reply




Possibly Related Threads...
Thread Author Replies Views Last Post
  S9 Serial number mismatch DariaVader 10 2,533 12-16-2014, 08:54 PM
Last Post: Evpraxia

Forum Jump:

New Posts   Today's Posts




About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.

For any more information, please use our contact form.