I started CPAP last July and my AHI has been slowly improving since. I also full time RV, meaning that I am constantly traveling around the US. When I first started I was in Illinois at about 600 feet above sea level. I then traveled west crossing the US toward Oregon. On the trip across my AHI numbers skyrocketed while at altitudes of 3-4000 feet. I spent three months on the Oregon coast at sea level then most of the winter near Yuma, AZ under 1000 feet and my numbers slowly came down from an average of 10 per hour in Aug. to just over 5 by the end of January, dropping approx. 1 per month. In February I spent a couple of weeks recrossing the US eastward going through higher elevations in AZ, NM and Texas (in the 3-4000 ft range) where my numbers again went up. I am now in the Rio Grande Valley, well under 1000 feet and my numbers are returning to the just over 5 range.

I am assuming that altitude causes this increase. Most of my Apneas are CAs. I am on straight CPAP at 16 lbs pressure. Is there something I can do to improve AHIs while at altitude? I would love to spend more time over 3000 feet.
02-27-2016 12:49 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
HermannTheGerman Offline

Preferred Members

Posts: 36
Joined: Feb 2016

Machine: Resmed S9 Autoset-CS
Mask Type: Hybrid
Mask Make & Model: Resmed Liberty
Humidifier: i5
CPAP Pressure: 6-24
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location:

Post: #2
RE: Altitude
(02-27-2016 12:49 PM)dzwiss Wrote:  I am on straight CPAP at 16 lbs pressure.

First of all I doubt the 16 lbs.
16 lbs is equal to 1124 cmH20.

My Resmed S9 ASV can go up to 30 cmH2O.
Even with a Binfort 4000 CPAP you won´t get that big amount of pressure.

And I believe that your CA´s come because the cpap pressure is still the same as at sealevel.

When you are at sealevel, the pressure inside your body is sealevel also wich is 1013 HPA or 1013 cmH2O in normal athmosphere.

The pressure of your CPAP is 16 HPA or 16 cmH2O
So the pressureratio is 16:1013 at sealevel.

At 3000 ft the atmospheric pressure is 100 HPA less than sealevel = 913 HPA.
The CPAP pressure is still the same.
So the pressure ratio is 16:913 now. Thats 10% less.

The relative CPAPpressure is higher than normal and brings you CA´s.

You must adjust the CPAP pressure for a sleep at 3000 ft to:
16/1013*913= 14,4 HPA

So with 14,4 HPA the ca´s will get back to normal.

Hermann
(This post was last modified: 02-27-2016 01:38 PM by HermannTheGerman.)
02-27-2016 01:37 PM
Find all posts by this user Post Reply Quote this message in a reply
justMongo Offline

Monitors

Posts: 3,854
Joined: Sep 2013

Machine: ResMed VPAP Auto (S9)
Mask Type: Full face mask
Mask Make & Model: ResMed Mirage Quattro
Humidifier: ResMed H5i
CPAP Pressure: 18 IPAPmax - 11 EPAPmin, PS=5
CPAP Software: Other Software

Other Comments: μολὼν λαβέ

Sex: Male
Location: Калифорния

Post: #3
RE: Altitude
These machines are supposed to be self compensating up to about 7500 feet above mean sea level.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
02-27-2016 05:45 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
HermannTheGerman Offline

Preferred Members

Posts: 36
Joined: Feb 2016

Machine: Resmed S9 Autoset-CS
Mask Type: Hybrid
Mask Make & Model: Resmed Liberty
Humidifier: i5
CPAP Pressure: 6-24
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location:

Post: #4
RE: Altitude
(02-27-2016 05:45 PM)justMongo Wrote:  These machines are supposed to be self compensating up to about 7500 feet above mean sea level.

Quote:The amount of pressure generated by the CPAP fan decreases with altitude due to lower gas density. For this reason, an increase in fan speed is needed to compensate for the modest drop in pressure at altitude. Older machines have adjustable "altitude" settings in them. Newer machines have internal correction for altitude so that no adjustments are necessary.

They compensate to keep the pressure constant.
But one doesn´t need the same pressure at altitude.
If they would compensate to a lower pressure, dzwiss,
wouldn´t have any differences in his/her AHI.

Hermann
02-27-2016 06:27 PM
Find all posts by this user Post Reply Quote this message in a reply
wolson Offline

Preferred Members

Posts: 92
Joined: Dec 2015

Machine: PRS1 C-Flex
Mask Type: Other
Mask Make & Model: Dreamwear
Humidifier: Remstar
CPAP Pressure: 10 cm H20 C-flex x3
CPAP Software: SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Red Canyon, Lander, Wyoming

Post: #5
RE: Altitude
(02-27-2016 12:49 PM)dzwiss Wrote:  I am assuming that altitude causes this increase. Most of my Apneas are CAs. I am on straight CPAP at 16 lbs pressure. Is there something I can do to improve AHIs while at altitude? I would love to spend more time over 3000 feet.

There doesn't seem to be much that you can do at higher elevations other than switch over to ASV. This is impractical for most people.

I suspect you have your CPAP set at 16 cmH2O a common measurement of pressure in ventilation, not 16 lbs pressure. First pounds is not a pressure unit. S condly, if you meant 16 pounds per square inch (psi) you would not have any lungs left as they would have exploded!

Walt

Walter W. Olson, Ph.D., P.E.
Professor Emeritus
02-27-2016 07:16 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
vsheline Offline

Advisory Members

Posts: 1,908
Joined: Jul 2012

Machine: S9 VPAP Adapt (USA Model# 36007, not better 36037 or 36067)
Mask Type: Full face mask
Mask Make & Model: F10 or SimPlus w/ 2Liners. MirageQuatro & Gecko gel pad
Humidifier: H5i
CPAP Pressure: 15 EPAP, PS 5-10
CPAP Software: ResScan

Other Comments: Marfan Syndrome, chronic bradycardia, occasional Cheyne-Stokes Respiration

Sex: Male
Location: California, USA

Post: #6
RE: Altitude
(02-27-2016 12:49 PM)dzwiss Wrote:  I am assuming that altitude causes this increase. Most of my Apneas are CAs. I am on straight CPAP at 16 lbs pressure. Is there something I can do to improve AHIs while at altitude? I would love to spend more time over 3000 feet.

Yes, this is a common effect of higher elevation for those of us who are susceptible to CA events.

Usually, higher pressure tends to raise CAs, and lower pressures tend to raise Obstructive events.

Personally, I am usually more concerned about having obstructive events, because I think obstructive events often stress our body worse than central events.

Also, obstructive sleep apnea is usually strongly positional, with flat on our back usually being the worst position.

If we can totally prevent rolling onto our back while asleep, we can usually get by with significantly less pressure.

Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment. The Advisory Member group provides advice and suggestions to Apnea Board administrators and staff on matters concerning Apnea Board operation and administrative policies - not on matters concerning treatment for Sleep Apnea. I think it is now too late to change the name of the group but I think Voting Member group would perhaps have been a more descriptive name for the group.
(This post was last modified: 02-28-2016 02:28 AM by vsheline.)
02-28-2016 02:23 AM
Find all posts by this user Post Reply Quote this message in a reply
dzwiss Online

Preferred Members

Posts: 27
Joined: Jul 2015

Machine: ResMed Air Sense 10
Mask Type: Nasal pillows
Mask Make & Model: AirFit P-10
Humidifier: ResMed Air Sense 10 Autoset
CPAP Pressure: 16
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Continuously travel throughout the US- Full Time RVer

Post: #7
RE: Altitude
Thank you all for your replies. I feel silly referring to my pressure lbs and will do a little more to educate myself as to the proper terms. I gather that the things I should try to solve this problem are to lower my pressure when traveling at altitude and/or sleep on my side. I will try those on my next trip. I would like to get my centrals down and will be seeing my doctor in a few months and bring this up.
02-28-2016 02:52 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Rick_V Offline

Preferred Members

Posts: 51
Joined: Jan 2016

Machine: Resmed Airsense 10 Autoset
Mask Type: Nasal pillows
Mask Make & Model: Airfit P10
Humidifier: Unsure
CPAP Pressure: 5-9
CPAP Software: ResScan SleepyHead

Other Comments: CPAPing since 2006

Sex: Male
Location: Fort White, Fl.

Post: #8
RE: Altitude
(02-28-2016 02:52 PM)dzwiss Wrote:  Thank you all for your replies. I feel silly referring to my pressure lbs and will do a little more to educate myself as to the proper terms. I gather that the things I should try to solve this problem are to lower my pressure when traveling at altitude and/or sleep on my side. I will try those on my next trip. I would like to get my centrals down and will be seeing my doctor in a few months and bring this up.
Have you tried changing your machine ipressures? you could experiment with different high and lows and keep a record of exactly what you need at altitudes.
(This post was last modified: 02-28-2016 03:11 PM by Rick_V.)
02-28-2016 03:08 PM
Find all posts by this user Post Reply Quote this message in a reply
dzwiss Online

Preferred Members

Posts: 27
Joined: Jul 2015

Machine: ResMed Air Sense 10
Mask Type: Nasal pillows
Mask Make & Model: AirFit P-10
Humidifier: ResMed Air Sense 10 Autoset
CPAP Pressure: 16
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Continuously travel throughout the US- Full Time RVer

Post: #9
RE: Altitude
Quote:Have you tried changing your machine ipressures? you could experiment with different high and lows and keep a record of exactly what you need at altitudes.

I think that is exactly what I will try on my next cross country venture.
02-28-2016 05:03 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  High Altitude experience with Dreamstation APAP. NorthernGuy 6 517 08-31-2016 08:52 PM
Last Post: NorthernGuy
  1) AHI by altitude? 2) Elevated AHI but decent SpO2 glaxelson 5 599 03-09-2016 12:06 PM
Last Post: PaytonA
  Anecdotal experience regarding CPAP and altitude wolson 5 738 02-27-2016 03:42 PM
Last Post: SideSleeper
  [CPAP] Altitude Lyndaloofa 5 773 09-28-2015 02:35 PM
Last Post: surferdude2
  Change in altitude - AHI went up Rastur 5 1,239 09-20-2015 04:46 PM
Last Post: vsheline
  New to CPAP/BiPAP - living at High Altitude forJO 5 1,597 08-15-2015 03:25 PM
Last Post: vsheline
  High Altitude and Central Apneas nicholb 21 3,156 08-13-2015 03:55 AM
Last Post: quiescence at last

Forum Jump:

Who's Online (Complete List)