RE: Anyone else experience this?
What is your minimum EPAP set at? It looks like the machine's rapid response to OAs is what triggered the 2 clusters of CAs in this data. If your min EPAP is 4 or 5 and if it were me, I would raise the min EPAP by 1 and reduce the Pressure Support by 1 and see if it helps.

I think that you should talk to your sleep doc and present some of this data to him/her as soon as possible. I think that it is ultimately going to require a bilevel machine with a back-up rate to get your CAs down to a good level but that is just my opinion.

Question: Is your Pressure Support currently set at 4?

Good luck with whatever you decide to do.

Question for others: What would cause Sleepyhead to have only one pressure trace for a VPAP? PS set to 0? VPAP in CPAP mode? Anything else?

Best Regards,

PaytonA
07-20-2015 10:33 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
richb Online

Advisory Members

Posts: 534
Joined: Jun 2015

Machine: Resmed AirCurve 10 ASV
Mask Type: Nasal mask
Mask Make & Model: ResMed Mirage FX
Humidifier: Built in
CPAP Pressure: Epap 6 Ipap 8 Auto ASV
CPAP Software: ResScan SleepyHead

Other Comments: Idiopathic Central Apnea

Sex: Male
Location: Vermont

Post: #42
RE: Anyone else experience this?
(07-20-2015 10:15 AM)Sleeprider Wrote:  Rich, that's interesting, particularly that you have essentially zero OA and H. On the theory that pressure support may aggravate CA, what is the minimum CPAP pressure you can tolerate before those begin to reappear?

I tried straight 9 CPAP and 5 9 Bipap with similar results. I have only had 4 nights with AHI below 20 in the past 6 weeks. I am thinking of trying straight 5 this week.

Apnea Board Member RobySue has posted a Beginners Guide to Sleepyhead Software here: http://www.apneaboard.com/wiki/index.php...SleepyHead
07-20-2015 04:30 PM
Find all posts by this user Post Reply Quote this message in a reply
richb Online

Advisory Members

Posts: 534
Joined: Jun 2015

Machine: Resmed AirCurve 10 ASV
Mask Type: Nasal mask
Mask Make & Model: ResMed Mirage FX
Humidifier: Built in
CPAP Pressure: Epap 6 Ipap 8 Auto ASV
CPAP Software: ResScan SleepyHead

Other Comments: Idiopathic Central Apnea

Sex: Male
Location: Vermont

Post: #43
RE: Anyone else experience this?
[attachment=1615]
(07-20-2015 10:15 AM)Sleeprider Wrote:  Rich, that's interesting, particularly that you have essentially zero OA and H. On the theory that pressure support may aggravate CA, what is the minimum CPAP pressure you can tolerate before those begin to reappear?

Here is a snippet at CPAP 9. The HAs have returned big time. I am going to try a straight CPAP 5 which is probably the lowest pressure I can tolerate. I am pretty sure that I have lots of Periodic Breathing without the machine.

Apnea Board Member RobySue has posted a Beginners Guide to Sleepyhead Software here: http://www.apneaboard.com/wiki/index.php...SleepyHead
(This post was last modified: 07-20-2015 06:54 PM by richb.)
07-20-2015 06:44 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
wxmcpo Offline

Preferred Members

Posts: 77
Joined: Apr 2015

Machine: ResMed AirCurve 10 VAuto
Mask Type: Nasal mask
Mask Make & Model: ResMed Mirage FX
Humidifier: Heated (attached to machine)
CPAP Pressure: VAuto
CPAP Software: ResScan SleepyHead

Other Comments: Looking forward to a good night's sleep

Sex: Male
Location: Texas

Post: #44
RE: Anyone else experience this?
(07-20-2015 10:33 AM)PaytonA Wrote:  What is your minimum EPAP set at? It looks like the machine's rapid response to OAs is what triggered the 2 clusters of CAs in this data. If your min EPAP is 4 or 5 and if it were me, I would raise the min EPAP by 1 and reduce the Pressure Support by 1 and see if it helps.

I think that you should talk to your sleep doc and present some of this data to him/her as soon as possible. I think that it is ultimately going to require a bilevel machine with a back-up rate to get your CAs down to a good level but that is just my opinion.

Question: Is your Pressure Support currently set at 4?

Good luck with whatever you decide to do.

Question for others: What would cause Sleepyhead to have only one pressure trace for a VPAP? PS set to 0? VPAP in CPAP mode? Anything else?

Best Regards,

PaytonA

Not really sure who this is directed at since it seems both Rich and I are posting our Sleepyhead data looking for answers. If directed at me, prior to last night my min EPAP was currently set to while my Pressure Support was also set to 4.

Last night I lowered my Pressure Support to 3.0 (min EPAP is still 4.0) to see what that would do for me. Attached is my Sleepyhead report;

[Image: du5rX74l.png]

Once again I had a AHI spike early in my sleep, but for well over 2 hours I had an AHI of 1.0 or less. It's this early spike of 35 that is making my AHI avg out at over 9 for the night. Wish I knew what I was doing when I got the lower AHI so I could do that all night. Any ideas?
07-21-2015 04:37 AM
Find all posts by this user Post Reply Quote this message in a reply
richb Online

Advisory Members

Posts: 534
Joined: Jun 2015

Machine: Resmed AirCurve 10 ASV
Mask Type: Nasal mask
Mask Make & Model: ResMed Mirage FX
Humidifier: Built in
CPAP Pressure: Epap 6 Ipap 8 Auto ASV
CPAP Software: ResScan SleepyHead

Other Comments: Idiopathic Central Apnea

Sex: Male
Location: Vermont

Post: #45
RE: Anyone else experience this?
(07-21-2015 04:37 AM)wxmcpo Wrote:  
(07-20-2015 10:33 AM)PaytonA Wrote:  What is your minimum EPAP set at? It looks like the machine's rapid response to OAs is what triggered the 2 clusters of CAs in this data. If your min EPAP is 4 or 5 and if it were me, I would raise the min EPAP by 1 and reduce the Pressure Support by 1 and see if it helps.

I think that you should talk to your sleep doc and present some of this data to him/her as soon as possible. I think that it is ultimately going to require a bilevel machine with a back-up rate to get your CAs down to a good level but that is just my opinion.

Question: Is your Pressure Support currently set at 4?

Good luck with whatever you decide to do.

Question for others: What would cause Sleepyhead to have only one pressure trace for a VPAP? PS set to 0? VPAP in CPAP mode? Anything else?

Best Regards,

PaytonA

Not really sure who this is directed at since it seems both Rich and I are posting our Sleepyhead data looking for answers. If directed at me, prior to last night my min EPAP was currently set to while my Pressure Support was also set to 4.

Last night I lowered my Pressure Support to 3.0 (min EPAP is still 4.0) to see what that would do for me. Attached is my Sleepyhead report;

[Image: du5rX74l.png]

Once again I had a AHI spike early in my sleep, but for well over 2 hours I had an AHI of 1.0 or less. It's this early spike of 35 that is making my AHI avg out at over 9 for the night. Wish I knew what I was doing when I got the lower AHI so I could do that all night. Any ideas?

Hi wxmcpo,

Take a closer look at your data. There is a section between 20:00 and 21:00 that looks suspicious. Highlight 5 or 10 minute sections of your flow rate graph to see how you are breathing breath to breath. I suspect that your are experiencing periodic breathing in that section as well as many other sections.

richb

Apnea Board Member RobySue has posted a Beginners Guide to Sleepyhead Software here: http://www.apneaboard.com/wiki/index.php...SleepyHead
07-21-2015 06:48 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
richb Online

Advisory Members

Posts: 534
Joined: Jun 2015

Machine: Resmed AirCurve 10 ASV
Mask Type: Nasal mask
Mask Make & Model: ResMed Mirage FX
Humidifier: Built in
CPAP Pressure: Epap 6 Ipap 8 Auto ASV
CPAP Software: ResScan SleepyHead

Other Comments: Idiopathic Central Apnea

Sex: Male
Location: Vermont

Post: #46
RE: Anyone else experience this?
I have had a lot of trouble with my machine since it was re-programed with a new prescription by my provider. I checked my settings and found the machine was set for Vauto min 4 and max 15. The big problem is the PS was set for 8. Previous setting was Vauto min 5 max 16 PS 4. I was unable to get to sleep and so I checked the setting and re adjusted the PS to 4. Slept better but got an AHI of 20. My settings were supposed to be Bi-pap 11 15. This setting produced lower AHI during my 2nd titration. I am going to try that setting tonight and then try CPAP straight 5 the next night. PaytonA suggested I look for a low setting where the events minimize. I think I will see more Hypos at the lowest setting just as I did when I went to straight CPAP 9.
07-21-2015 07:00 AM
Find all posts by this user Post Reply Quote this message in a reply
DeepBreathing Offline
Wiki Editor
Moderators

Posts: 2,296
Joined: Sep 2013

Machine: Resmed S9 VPAP Adapt
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: Resmed H5i
CPAP Pressure: EPAP: 9 - 15 PS: 3 - 10
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Perth, Western Australia

Post: #47
RE: Anyone else experience this?
Rich, I've been watching this thread without intervening until now. However I agree with paytonA that you should be looking at a bilevel with backup. Putting it bluntly, you are on the wrong machine and all the experimenting with pressure modes and settings is wasting time. You need an ASV machine, which will suppress periodic breathing and central apneas. I'm convinced you will see an immediate and dramatic improvement with an ASV.

The only caveat on that is that there have been some adverse outcomes for patients with a narrowly defined congestive heart condition - you need to talk to your doctor if that affects you. See this: http://www.resmed.com/us/en/consumer/new...erapy.html
and this: http://www.sleep.org.au/documents/item/1759

DeepBreathing
Apnea Board Moderator
www.ApneaBoard.com


Bed

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
07-21-2015 07:00 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
richb Online

Advisory Members

Posts: 534
Joined: Jun 2015

Machine: Resmed AirCurve 10 ASV
Mask Type: Nasal mask
Mask Make & Model: ResMed Mirage FX
Humidifier: Built in
CPAP Pressure: Epap 6 Ipap 8 Auto ASV
CPAP Software: ResScan SleepyHead

Other Comments: Idiopathic Central Apnea

Sex: Male
Location: Vermont

Post: #48
RE: Anyone else experience this?
(07-21-2015 07:00 AM)DeepBreathing Wrote:  Rich, I've been watching this thread without intervening until now. However I agree with paytonA that you should be looking at a bilevel with backup. Putting it bluntly, you are on the wrong machine and all the experimenting with pressure modes and settings is wasting time. You need an ASV machine, which will suppress periodic breathing and central apneas. I'm convinced you will see an immediate and dramatic improvement with an ASV.

The only caveat on that is that there have been some adverse outcomes for patients with a narrowly defined congestive heart condition - you need to talk to your doctor if that affects you. See this: http://www.resmed.com/us/en/consumer/new...erapy.html
and this: http://www.sleep.org.au/documents/item/1759

I need 2 more days of data to bring to my sleep doc to show that the machine is actually bad for my health. My BP is elevated above that when I am not on a machine. My sleep is horrible. I have been saying that I need an ASV machine, medication or both for a while now.

richb

Apnea Board Member RobySue has posted a Beginners Guide to Sleepyhead Software here: http://www.apneaboard.com/wiki/index.php...SleepyHead
07-21-2015 07:54 AM
Find all posts by this user Post Reply Quote this message in a reply
wxmcpo Offline

Preferred Members

Posts: 77
Joined: Apr 2015

Machine: ResMed AirCurve 10 VAuto
Mask Type: Nasal mask
Mask Make & Model: ResMed Mirage FX
Humidifier: Heated (attached to machine)
CPAP Pressure: VAuto
CPAP Software: ResScan SleepyHead

Other Comments: Looking forward to a good night's sleep

Sex: Male
Location: Texas

Post: #49
RE: Anyone else experience this?
(07-21-2015 06:48 AM)richb Wrote:  Hi wxmcpo,

Take a closer look at your data. There is a section between 20:00 and 21:00 that looks suspicious. Highlight 5 or 10 minute sections of your flow rate graph to see how you are breathing breath to breath. I suspect that your are experiencing periodic breathing in that section as well as many other sections.

richb

Richb,

So I took a look at the section you were mentioning. Below is 5 minute sections starting at 2100 and going for almost 24 minutes. I will admit I am not 100% sure what I am looking at there, but I can see signs of where it looks like my breathing may have slowed or stop at times. Is that what you were looking for?

[Image: kpDRLYgl.png]

Quote:I have had a lot of trouble with my machine since it was re-programed with a new prescription by my provider. I checked my settings and found the machine was set for Vauto min 4 and max 15. The big problem is the PS was set for 8. Previous setting was Vauto min 5 max 16 PS 4. I was unable to get to sleep and so I checked the setting and re adjusted the PS to 4. Slept better but got an AHI of 20. My settings were supposed to be Bi-pap 11 15. This setting produced lower AHI during my 2nd titration. I am going to try that setting tonight and then try CPAP straight 5 the next night. PaytonA suggested I look for a low setting where the events minimize. I think I will see more Hypos at the lowest setting just as I did when I went to straight CPAP 9.

So my machine is also set to VAuto but with min 4 and max 25. A couple of nights ago I turned my PS down from 4 to 3 and my AHI went from 20+ to 9. So last night I left PS at 3, but turned my TiMax down from 3 to 2 and my AHI dropped to 3.2. This is one of my best numbers ever. Here's a screenshot of last nights data;

[Image: zpcio9Cl.png]

I am going to leave settings where they are for now and see how that does me over the next couple of nights. One big difference I noticed from previous nights is I didn't have an AHI spike early in my sleep, but rather late in my sleep.
(This post was last modified: 07-21-2015 02:18 PM by wxmcpo.)
07-21-2015 02:07 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
kaiasgram Offline

Advisory Members

Posts: 241
Joined: Aug 2013

Machine: S9 VPAP Auto
Mask Type: Nasal pillows
Mask Make & Model: Aloha
Humidifier: H5i
CPAP Pressure: 8-10
CPAP Software: SleepyHead

Other Comments:

Sex: Female
Location: Northern California

Post: #50
RE: Anyone else experience this?
(07-20-2015 10:33 AM)PaytonA Wrote:  Question for others: What would cause Sleepyhead to have only one pressure trace for a VPAP? PS set to 0? VPAP in CPAP mode? Anything else?

Hi PaytonA -- This question came up on another forum. Apparently SleepyHead is only showing the IPAP trace (though you might be able to see both IPAP and EPAP lines in the Mask Pressure graph). Just a little bug -- SH hasn't caught all the way up with the AirCurve machines yet.
(This post was last modified: 07-21-2015 02:39 PM by kaiasgram.)
07-21-2015 02:38 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  [CPAP] New to the CPAP experience - questions DynamicMaroonPenguin1961 11 501 10-18-2016 08:32 PM
Last Post: eseedhouse
  My Experience and Some Questions Lourens1190 7 372 10-14-2016 04:19 PM
Last Post: sonicboom
  UARS - ENT Recommended Septoplasty/Turbinate Reduction....experience? LB1982 11 456 09-30-2016 09:09 AM
Last Post: richb
  High Altitude experience with Dreamstation APAP. NorthernGuy 6 515 08-31-2016 08:52 PM
Last Post: NorthernGuy
  [Equipment] Experience with lighter CPAP machines for travel? jrolli1509 2 1,427 07-15-2016 03:43 PM
Last Post: Grover64
  [Equipment] Battery power experience for Dreamstation Auto CPAP NorthernGuy 9 1,234 07-10-2016 09:03 AM
Last Post: NorthernGuy
  Noob with Humana/Aprea Experience FrankNichols 9 907 05-22-2016 04:39 PM
Last Post: FrankNichols

Forum Jump:

Who's Online (Complete List)