Auto CPAP settings vs Auto BiPAP settings
#1
Question 
Hello everyone,

I have a settings questions for DreamStation CPAP Auto and DreamStation BiPAP Auto.

If your DreamStation CPAP Auto settings are:
Min Pressure: 5
Max Preseure: 15
A-Flex: 1

What would be an "equivalent" settings for DreamStation BiPAP Auto?
Are the following BiPAP Auto settings would work the same way as the CPAP Auto settings mentioned above?

Min EPAP: 5
Max IPAP: 15
PS Min: 0
PS MAX: 1


Post Reply Post Reply
#2
(05-06-2016, 11:05 AM)thasan26 Wrote: Hello everyone,

I have a settings questions for DreamStation CPAP Auto and DreamStation BiPAP Auto.

If your DreamStation CPAP Auto settings are:
Min Pressure: 5
Max Preseure: 15
A-Flex: 1

What would be an "equivalent" settings for DreamStation BiPAP Auto?
Are the following BiPAP Auto settings would work the same way as the CPAP Auto settings mentioned above?

Min EPAP: 5
Max IPAP: 15
PS Min: 0
PS MAX: 1
Not quite, but close. I would say these settings for the DreamStation BiPAP Auto will mimic the CPAP Auto more precisely:

Min EPAP = 5
Max IPAP = 15
PS Min = 0
PS Max = 0
Bi-Flex = 1

Bi-Flex mimics the way A-Flex works much better than max PS=1 does.

Here's why:

A-Flex does NOT reduce the pressure by a fixed amount on each exhalation. The amount it reduces the pressure by can vary anywhere from about 0.5cm to 2.5cm based on the force of the exhalation as well as the Flex setting. In general if the force of the exhalation is the same, the decrease will be greater with a higher flex setting. But it's possible that a very strong exhalation with A-Flex=1 will still result in a drop of more than 1cm of pressure. A very small exhalation, however, will result in a minimal drop of 0.5 cm.

Bi-Flex works the same way on the BiPAP: It reduces the pressure to below the EPAP pressure. The drop pressure by can vary anywhere from about 0.5cm to 2.5cm below EPAP and it is based on the force of the exhalation as well as the Bi-Flex setting. In general if the force of the exhalation is the same, the decrease will be greater with a higher flex setting. But it's possible that a very strong exhalation with Bi-Flex=1 will still result in a drop of more than 1cm of pressure below EPAP. A very small exhalation will result in a minimal drop of 0.5 cm below EPAP.

By setting PS min = PS max = 0, you're making the BiPAP effectively work as a CPAP Auto. By setting Bi-Flex = 1, you're mimicking the A-Flex=1 setting.
Questions about SleepyHead?
See my Guide to SleepyHead
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  New to Bipap, 4 years on Cpap Fabby 27 760 1 hour ago
Last Post: Fabby
  Can't seem to dial my settings in C0mbe 2 98 Yesterday, 01:32 PM
Last Post: C0mbe
Sad [CPAP] Res Med S9 help! Error: Settings reset to default values, please contact service pr dexter020281 14 10,762 Yesterday, 01:22 AM
Last Post: vsheline
  How to access Airsense 10 Expiratory pressure relief (EPR™) settings CCDrogan 3 105 02-17-2017, 11:30 PM
Last Post: trish6hundred
  Auto machine help Sidney 15 508 02-15-2017, 09:36 PM
Last Post: Sidney
Question Why Auto Bipap instead of Bipap? hopefulsleeper 2 158 02-15-2017, 10:26 AM
Last Post: Sleeprider
Surprised Worse centrals without auto, but fewer OA's. Where to go from here? hopefulsleeper 1 106 02-15-2017, 09:50 AM
Last Post: Sleeprider

Forum Jump:

New Posts   Today's Posts




About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.

For any more information, please use our contact form.