I have migrated from the AS10 to AirCurve Vauto bipap.  My AHI is fine, less than 1.0  - only a few OA events. I am still struggling with aerophagia and would like to optimize my levels. 

I am trying to have the lowest possible IPAP that works for me - and the EPAP that makes it easiest to exhale. Exhalation back-pressure difficulty has been my other problem.

My current settings are:

Vpap Auto
Epap   6.00
Ipap    8.80
PS      2.80

If I increase Ipap, will my aerophasia worsen?


To make exhalation easier, do I increase Epap, or decrease Epap?

How low can I set PS? - I'm at 2.8. Is this too low?

I'd like to achieve the lowest functional Ipap for me, The easiest Epap level  (I think)

Obviously, I need help Smile thank you    Big Grin
Post Reply Post Reply
#2
RE: Bipap Pressure Levels
Aerophagia: the strategy is to lower average pressure on a temporary basis untill you can tolerate and return to your therapeutic pressures. How?
Increase PS,. Higher PS can decrease average pressure.
Lowering IPAP
Lowering EPAP

Keep in mind that it is EPAP That treats obstructive apnea. This is the first thing to be set.
Pressure support treats hypopnoea, flow limitations, RERAs.

PS=0 is the lowest possible setting.
Post Reply Post Reply
#3
RE: Bipap Pressure Levels
Thank you for your response

I'm not sure I understand exactly what settings to try

I was at 

Vpap Auto
Epap   6.00
Ipap    8.80
PS      2.80

Last night I changed to:

Vpap Auto

Epap   5.8

Ipap    8.80
PS      3.0

My AHI was still good at 0.66 with just 4 OA events. Pressures seemed comfortable - less aerophagia

I would appreciate exact settings, since I don't quite grasp the concept - I must be a slow learner Smile Thank you
Post Reply Post Reply
#4
RE: Bipap Pressure Levels
Abletree, could you post an Oscar detail chart? We usually suggest setting changes to target a specific objective, and other than aerophagia, we really don't have much to work with in your case. Your recent change to PS 3 seems to have improved your sleep. With OA events, we will normally increase EPAP to stabilized the airway, which would take you to EPAP 6..0, IPAP 9.0 contingent on comfort with aerophagia.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#5
RE: Bipap Pressure Levels
Yes - Thank you.    This is my 2nd week on the Aircurve Vauto.  My chief complaints have been aerophagia and difficulty breating against expiratory pressure, which caused me to unconsciously remove my headgear during the evening

Both the aerophagia and exhalation has improved with the Aircurve (AS 10 prior machine). Also, my AHI has improved Smile

Still, I would like to eliminate the aerophagia completely and enjoy the comfort of a nice and easy EPAP setting, while keeping breathing happening

My sleep doc put me at 11 IPAP and 7 EPAP - at that setting I experienced extreme aerophagia and had some CA events, which I don't normally have

* I have been losing weight, getting fit - so I think that is playing a helpful role as well.

Here's the OSCAR - Thank you!


Attached Files Thumbnail(s)
   
Post Reply Post Reply
#6
RE: Bipap Pressure Levels
Thank you. You have no idea how much the charts help us.
Your numbers are excellent,
Based on that try EPAP min of 4.

We may suggest increasing it after you post the results.
Definitely let us know how your aerophagia fares.
Post Reply Post Reply
#7
RE: Bipap Pressure Levels
It's easy to agree with Gideon since he is pretty darn good at this. I think using EPAP min 4.0, IPAP max 7.2 will give you the ease of expiration and comfort you are seeking. That will be PS 3.2 which you should tolerate well, and protect against aerophagia. Thanks for posting the chart. The way your OA events occur doesn't seem to require the increase in EPAP pressure, and it's really easier to dial you back and move forward in pressure methodically, as I'm sure Gideon is intending.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#8
RE: Bipap Pressure Levels
Thank you so much! I will make the change tonight and report back!
Post Reply Post Reply
#9
RE: Bipap Pressure Levels
Perfect! Thank you so much! I will make the exact changes tonight and report back!  I'm excited Smile
Post Reply Post Reply
#10
RE: Bipap Pressure Levels
Update - I had a very nice night with the new settings as suggested. Also, I maintained a low AHI of 0.47 with no CA events

The EPAP pressure was extremely comfortable. 

I felt like I could use a little boost of more IPAP. It felt like I needed to suck in the air. It wasn't intolerable, but a slight boost might make it nicer - if it won't muck up everything Smile

The aerophgia was gone - completely corrected!

What would you think about boosting the IPAP from 7.2 to 7.4 and the PS from 3.2 to 3.4?   *OSCAR screenshot is below Smile

Thank you so much - I think this is going to be a great success story. You all helped me navigate starting off with the AS10 - and then making a strong argument for the Aircurve Vauto to my providers. Making the pitch for the Aircurve got me a 2nd sleep study, which really dialed in my sleep assessment

The Aircurve truly is a much better machine for me. It has been a journey!

Thanks again ~ on to the next level

*I'm working on getting a Muse S EEG sleep monitor and O2 recorder to find out more about deep sleep status


Attached Files Thumbnail(s)
   
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  BiPAP Pressure for Possible UARS bertchintus 20 978 Yesterday, 03:24 AM
Last Post: SingleH
  flow limitations acceptable levels yankees123 27 1,613 03-23-2024, 02:41 AM
Last Post: CPAPfriend
  Which biPAP? HalfAsleep 9 257 03-14-2024, 09:04 AM
Last Post: HalfAsleep
  Oxygen levels below 90% for 7% of night and below 94% for 34% inspite of Bipap bhavintu79 14 681 03-08-2024, 05:06 AM
Last Post: UnicornRider
  OSCAR pressure readings versus BIPAP display smallhandsfeeding 0 155 02-07-2024, 05:33 PM
Last Post: smallhandsfeeding
  UARS, bipap pressure, I or E Wcsleep 3 439 01-12-2024, 03:21 PM
Last Post: Gideon
  Low O2 Levels During Sleep with Low AHI cmpman1974 60 5,097 01-01-2024, 09:24 PM
Last Post: cmpman1974


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.