Breaking up OA clusters- PR System One Remstar Auto w/ Aflex
I noticed lately that machine not increasing pressure fast enough to open airway. The result seen on the graph is OA clustering, occurring close together. I see no unintentional leaks on graph. Is this typical with my machine? Pressure range set 10- 20.
08-27-2014 08:26 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Terry Offline

Advisory Members

Posts: 553
Joined: May 2014

Machine: Philips Respironics RemStar Auto with A-Flex (560P)
Mask Type: Nasal pillows
Mask Make & Model: P/R Nuance & Pap Cap
Humidifier: Integrated
CPAP Pressure: Auto
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location:

Post: #2
RE: Breaking up OA clusters- PR System One Remstar Auto w/ Aflex
(08-27-2014 08:26 AM)coolbranch Wrote:  I noticed lately that machine not increasing pressure fast enough to open airway. The result seen on the graph is OA clustering, occurring close together. I see no unintentional leaks on graph. Is this typical with my machine? Pressure range set 10- 20.

Mine did that too. The machine seems to be too conservative when raising pressure, and too aggressive in dropping it back down.

I was much happier after I raised my minimum pressure up to the level that simply prevented the OAs.
08-27-2014 11:11 AM
Find all posts by this user Post Reply Quote this message in a reply
coolbranch Offline

Preferred Members

Posts: 42
Joined: May 2014

Machine: AirSense 10 AutoSet
Mask Type: Full face mask
Mask Make & Model: ResMed Quattro Air
Humidifier: HumidAir Heated Humidifier
CPAP Pressure: 8.6- 12.6
CPAP Software: SleepyHead

Other Comments: SleepyHead v0.9.8-1 (testing) Build Date: Oct 3 2014 17:17:35 Revision: 4595153 (master branch) Gra

Sex: Male
Location: SE Alabama

Post: #3
RE: Breaking up OA clusters- PR System One Remstar Auto w/ Aflex
Thanks Terry. I have lowest pressure at 10 now. After further study, I now see where my last week average pressure was 10.69 and last 30 days was 10.84. I plan to bump to 11 on the low side tonight.
08-27-2014 12:39 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Terry Offline

Advisory Members

Posts: 553
Joined: May 2014

Machine: Philips Respironics RemStar Auto with A-Flex (560P)
Mask Type: Nasal pillows
Mask Make & Model: P/R Nuance & Pap Cap
Humidifier: Integrated
CPAP Pressure: Auto
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location:

Post: #4
RE: Breaking up OA clusters- PR System One Remstar Auto w/ Aflex
(08-27-2014 12:39 PM)coolbranch Wrote:  Thanks Terry. I have lowest pressure at 10 now. After further study, I now see where my last week average pressure was 10.69 and last 30 days was 10.84. I plan to bump to 11 on the low side tonight.

It will be interesting to see how that works out.

Terry
08-29-2014 11:39 PM
Find all posts by this user Post Reply Quote this message in a reply
coolbranch Offline

Preferred Members

Posts: 42
Joined: May 2014

Machine: AirSense 10 AutoSet
Mask Type: Full face mask
Mask Make & Model: ResMed Quattro Air
Humidifier: HumidAir Heated Humidifier
CPAP Pressure: 8.6- 12.6
CPAP Software: SleepyHead

Other Comments: SleepyHead v0.9.8-1 (testing) Build Date: Oct 3 2014 17:17:35 Revision: 4595153 (master branch) Gra

Sex: Male
Location: SE Alabama

Post: #5
RE: Breaking up OA clusters- PR System One Remstar Auto w/ Aflex
Last night I was hosed up for 7hrs 44min. Went 7 hrs without an event, then I had 5 OA's in a 4 min period of time. The Remstar Auto bumped pressure up one cm then bumped one more cm 2 min later. No unintentional leaks during the 4 min period. One desat to 89. I would determine bumping the low side pressure to 11 to be somewhat of a successful adjustment. The Remstar rocked on for 7 hrs with it's hunt and peck feature which I like. The machine is what it is when it comes to breaking up clusters. I may have rolled over on my back during that time. I am thinking about doing a video, with timestamp, on myself one night just to see how I sleep.
08-30-2014 08:57 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
coolbranch Offline

Preferred Members

Posts: 42
Joined: May 2014

Machine: AirSense 10 AutoSet
Mask Type: Full face mask
Mask Make & Model: ResMed Quattro Air
Humidifier: HumidAir Heated Humidifier
CPAP Pressure: 8.6- 12.6
CPAP Software: SleepyHead

Other Comments: SleepyHead v0.9.8-1 (testing) Build Date: Oct 3 2014 17:17:35 Revision: 4595153 (master branch) Gra

Sex: Male
Location: SE Alabama

Post: #6
RE: Breaking up OA clusters- PR System One Remstar Auto w/ Aflex
https://dl.dropboxusercontent.com/u/10675309/Pressure.jpg

Since I moved the low side pressure from 10 to 11, the machine has been able to anticipate pressure needs and keep the airway open. AHI has been maintained around 1 or below.

The graph link indicates a pattern for the last couple of weeks.

Thanks Terry. I hope to have found the sweet spot.
09-15-2014 11:52 AM
Find all posts by this user Post Reply Quote this message in a reply
archangle Offline
Wiki Editor
Advisory Members

Posts: 3,159
Joined: Feb 2012

Machine: ResMed S9 AutoSet
Mask Type: Nasal pillows
Mask Make & Model: ResMed Swift FX
Humidifier: ResMed S9 H5i
CPAP Pressure: 16-20
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments: Happy PAPper

Sex: Undisclosed
Location: USA

Post: #7
RE: Breaking up OA clusters- PR System One Remstar Auto w/ Aflex
The current APAP machines aren't intended to react that quickly to apneas.

You need to set the minimum pressure to be high enough to prevent most of your apneas. Yet another reason to have a fully data capable machine, and to review the results and make adjustments.

The tradeoffs are complex, and I don't think we understand all that goes into it. I believe that some of the modern machines adjust upward based on things like "flow limitations," snoring, and the shape of the airflow waveforms.

You might think the machines should react more quickly, but some people are disturbed by the pressure changes. There's also some concern about the machine "running away" to higher pressure with some people. Some of the side effects like aerophagia are helped by lower pressures.

The A10 AutoSet for Her has a "for her" mode that does a more "gentle" auto algorithm. I think the issue of adjustable auto sensitivity/speed is an interesting idea that should be explored.

Get the free SleepyHead software here.
Useful links.
Click here for information on the main alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check it yourself.
09-15-2014 04:12 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Terry Offline

Advisory Members

Posts: 553
Joined: May 2014

Machine: Philips Respironics RemStar Auto with A-Flex (560P)
Mask Type: Nasal pillows
Mask Make & Model: P/R Nuance & Pap Cap
Humidifier: Integrated
CPAP Pressure: Auto
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location:

Post: #8
RE: Breaking up OA clusters- PR System One Remstar Auto w/ Aflex
(09-15-2014 11:52 AM)coolbranch Wrote:  https://dl.dropboxusercontent.com/u/10675309/Pressure.jpg

Since I moved the low side pressure from 10 to 11, the machine has been able to anticipate pressure needs and keep the airway open. AHI has been maintained around 1 or below.

The graph link indicates a pattern for the last couple of weeks.

Thanks Terry. I hope to have found the sweet spot.

Congratulations! Try it where it is for a while.

Later, you may find that bumping it up another notch will keep you right around "0".

Terry
09-15-2014 06:02 PM
Find all posts by this user Post Reply Quote this message in a reply
coolbranch Offline

Preferred Members

Posts: 42
Joined: May 2014

Machine: AirSense 10 AutoSet
Mask Type: Full face mask
Mask Make & Model: ResMed Quattro Air
Humidifier: HumidAir Heated Humidifier
CPAP Pressure: 8.6- 12.6
CPAP Software: SleepyHead

Other Comments: SleepyHead v0.9.8-1 (testing) Build Date: Oct 3 2014 17:17:35 Revision: 4595153 (master branch) Gra

Sex: Male
Location: SE Alabama

Post: #9
RE: Breaking up OA clusters- PR System One Remstar Auto w/ Aflex
Roger that... I will.
09-15-2014 06:28 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Evpraxia Offline

Advisory Members

Posts: 313
Joined: Aug 2014

Machine: ResMed S9 AutoSet
Mask Type: Nasal pillows
Mask Make & Model: ResMed Swift LT for Her - Nasal Pillows System
Humidifier: ResMed H5i
CPAP Pressure: 7-17
CPAP Software: ResScan SleepyHead

Other Comments: Cat Staff, bibliophile, love Classical & Byzantine music

Sex: Female
Location: Pacific Northwest USA

Post: #10
RE: Breaking up OA clusters- PR System One Remstar Auto w/ Aflex
Thank You, THANK YOU for this thread!

I think this is the answer for my hubby's issues with HIS PR System One, his settings are at 12-18 and he has severe SA; OA and CNS. I will bet you he should be at a higher lower number, like 13 or 14 to 18.

You folks are SO great! Go pat yourselves on the back - each of you deserves it!

Evpraxia in the Pacific Northwest USA
Diagnosed: 44 AHI when supine, O2 down to 82%
Treated since 20 Sept 2014:: 0.7 AHI, Settings 7-15, EPR on Full Time at Level 3
Better living through CPAP/APAP machines!
09-15-2014 07:43 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  REMStar SE with DOT Compliance musicman85 11 361 12-01-2016 09:08 PM
Last Post: musicman85
  philips remstar auto a-flex skylersdad 2 154 11-29-2016 05:47 PM
Last Post: trish6hundred
  PR System One: C-Check vs Auto? Ailu 8 2,983 11-29-2016 02:04 PM
Last Post: TonyJ
  Question: sound level on Resmed S9 WITHOUT humidifier, compared to Remstar 560 kowlooner 8 216 11-29-2016 10:51 AM
Last Post: kowlooner
  Philips system one noisy jpsantell 2 263 10-29-2016 10:54 AM
Last Post: robysue
  CPAP cleaning system Tommy C 97 62,532 10-20-2016 05:35 PM
Last Post: PaytonA
  airsense10 auto cpap(for him) VS DreamStation Auto CPAP DSX500T11 w/HT humid rookiepom 5 851 10-09-2016 07:52 AM
Last Post: stephennic

Forum Jump:

Who's Online (Complete List)