You are using a BiPAP Auto ASV in with EPAP 15 to 18, PS 4 to 8 and IPAP max 23. I said before, these settings are not ideal. Your therapy starts off at a pressure of 19/15. EPAP never or rarely rises in response to OA events, but PS is increasing to the maximum of 23 cm. Looking at your events your CAI and OAI are certainly acceptable, but your hypopnea is very high as are flow limitations.

I recommend the following changes:
Decrease EPAP min to 13.0 cm, no change to max EPAP at 18.
increase the minimum PS from 4 to 5 cm (we may go higher)
Increase maximum PS from 8 to 10.

These settings may allow additional OA, and if that happens we will increase EPAP min, but I'd like to see how you respond to the lower EPAP pressure. The higher minimum PS will cut back on flow limits, snores and hypopnea. We may need to increase that further, if those don't resolve. The hypopnea and flow limitation are from restriction in your airway, and the higher minimum PS should help.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Good AHI but Low Tidal Volume and Minute Ventilation erolalper 1 146 03-15-2024, 07:57 AM
Last Post: BoxcarPete
  15-minute periods of massive events (OSCAR pics included) nicholasb600 2 176 03-14-2024, 09:42 AM
Last Post: BoxcarPete
  Help with self-titrating BiPAP for UARS - low % of spontaneously triggered breaths? Bibbbooker 21 856 01-29-2024, 10:08 AM
Last Post: Sleeprider
  What is optimal respiration rate, minute vent, tidal volume, etc? ATeam91 4 463 01-23-2024, 04:19 PM
Last Post: ATeam91
  Minute Vent, Tidal Volume, etc. MrIvanDrago 5 698 11-20-2023, 11:49 AM
Last Post: Jay51
Sad [Treatment] Weird Flow Rate/Minute Vent OK Laptop 0 402 10-28-2023, 12:40 AM
Last Post: OK Laptop
  Is this normal - no breath for over a minute? Pizzafoundry 12 848 10-16-2023, 02:29 PM
Last Post: Pizzafoundry


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.