[CPAP] Machine turning self off and back on
#31
I replied in the other thread. The most common reason for this failure is when the flow sensor in the upper-right corner of the main board fails to detect breathing. This was fixed in later generation Respironics machines (Series 60) by adding a screw to the main-board to maintain contact of the sensor with the breathing circuit. The 550 can be repaired by adding a #8 X 1/2 sheet metal screw with a plastic washer to the empty mounting hole about 1-inch down from the upper right corner of the main-board to hold it down. The 60 series has a screw in this location.

[Image: Jfu8x.jpg]
______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
Reply
#32
Hi Sleeprider, could you please clarify which hole you're referring to (i.e. mark it on the photo) ?. I think I know which one you mean, but don't want to make a wrong assumption.

I appreciate the information and realize I probably just jostled the rubber gasket back into place during my assembly/disassembly getting at the motor/blower. Some of the plastic screw bosses on the lower side of my blower assembly were cracked, but I'm pretty sure that was self inflicted. I bought the CPAP used from Craigslist and while the unit didn't have many hours on it, it accumulated the smells of wet dog and cigarette smoke. Needless to say, I did a complete teardown (as carefully as I could) cleaning every surface with vinegar & water -- probably over-torqued the screws when I reassembled it a few years ago.

Thanks for sharing this information!

DRC
Reply
#33
See the Service manual available from the forum. The screw location is shown in the attachment below. For odor, try Ozium. Spray cotton ball, place it near the inlet and let the machine run 20 minutes. It won't ever be there again. Also note there are 2 silicone seals in the humidifier dry box where the blower exhausts into the humidifier, and one silicone seal on the humidifier lid. These need replaced if you have smoke or other stubborn odor problems. They are available from Supplier #1.

[Image: screw_zpsfqjtani3.jpg]

This is what is on the back-side of that screw:

[Image: th011.jpg]

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
Reply


#34
(10-01-2016, 05:16 PM)Sleeprider Wrote: See the Service manual available from the forum.

Just for reference, members can read about how to get to the service manuals in this thread:

http://www.apneaboard.com/forums/Thread-...Only-FILES

Which reminds me, I think I've got some more to upload to that page. Thinking-about

EDIT:
Page updated with a bunch of new service manuals... see here:

http://www.apneaboard.com/forums/Thread-...#pid178176
SuperSleeper
Apnea Board Administrator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.



Reply
#35
Thank you (belatedly) for the great information! Unfortunately in my case, I'm not sure it is the only problem. I did put a screw into the place you're showing and after disassembly/reassembly things seem to work OK for awhile, but sooner or later the cut-out problem shows up again. I do travel with my CPAP and that probably aggravates the situation.

Maybe in my previous attempts to fix the problem I caused some other issue -- will have a look at the service manual. I really love the auto unit and would like to keep using it. In the meantime I'm back to my old Bipap S/T machine.
Reply
#36
I know this is an old thread, but I just wanted to give you a gigantic thank you for the info! First time my machine has run through the night in quite a long time. Thanks again!
Reply


#37
(12-05-2016, 08:31 AM)Toofattosleep Wrote: I know this is an old thread, but I just wanted to give you a gigantic thank you for the info! First time my machine has run through the night in quite a long time. Thanks again!

It's great to hear new members are finding these results, and they continue to be useful. We rarely hear about this. Thanks back at-cha.
______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Just diagnosed but headed back for 2nd study. Advice? Piggles 56 1,529 Today, 12:27 AM
Last Post: tuckman
Exclaimation CPAP Machine Choices - read this before you accept a new machine SuperSleeper 10 7,631 Yesterday, 06:38 PM
Last Post: Mosquitobait
  Battery Back up for Airsense 10 Matzaball2016 6 394 12-23-2016, 05:45 PM
Last Post: sonicboom
Question New CPAP user, getting your sleep back to normal? PerspexAvenger 9 598 12-16-2016, 05:33 PM
Last Post: msteiner38
  Turn my pressure back up? duece12345 5 281 12-07-2016, 12:47 PM
Last Post: duece12345
  Can't Find Device Number (DN) on Back of Machine gamegrrl 5 303 11-14-2016, 05:32 PM
Last Post: gamegrrl
Gross I AM BACK !! Lost53 15 1,230 10-31-2016, 09:14 AM
Last Post: Sleeprider

Forum Jump:

New Posts   Today's Posts




About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.

For any more information, please use our contact form.