Makes more sense. I would use Ti of 1.8 for what I'm seeing here. That will support your normal respiratory inspiration of 2.4 and might help.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#12
RE: Clusters of Hypopneas at beginning of sessions
Thanks.  I'll try it tonight and post tomorrow.
Post Reply Post Reply
#13
RE: Clusters of Hypopneas at beginning of sessions
Changed the Ti to 1.8 and results are attached.  I am seeing more RERA than I am used to, is that an issue?  Or should I give it a few more nights to get a good feel for what's happening.

Thanks again.


Attached Files Thumbnail(s)
   
Post Reply Post Reply
#14
RE: Clusters of Hypopneas at beginning of sessions
I don't think that level of RERA is a problem. I note that most, if not all of your hypopnea is during ramp. I think if you can shorten the time, increase the starting pressure, or turn off ramp all together, you will see some pretty surprising results. This chart looks excellent if we disregard the events during ramp.

The change in Ti has significantly increased your tidal volume from 500 to 800 mL. I'm not certain if your doctor wants to target Vt but if this is too large of a volume increase, it is very easy to back off Ti to 1.4 or 1.5 to fine-tune this aspect. The change to 1.8 did eliminate hypopnea during therapy. Other things to note, your respiration rate is high at 38 bpm. This respiration rate does not match the reported inspiratory time of 1.84 + expiratory time of 3.04, which is 4.88 seconds, or 12.3 BPM. So something is in error on the statistics and we need to look closer.

Your inspiratory/expiratory time has normalized with a shorter inspiratory time and longer expiratory time. So something has improved your respiration to a more normal cycle. With the shorter Ti it was exactly the opposite. To interpret any of this, we would need a close-up view of the respiratory flow chart. That would be a zoomed in view to a 2-minute segment. It would be interesting to see the before and after shots for the Ti change. Just looking at the flow rate chart, you have much more consistent looking respiration with the increased Ti. The "hairbrush" appearance is gone, and all inhale and exhale flow now looks much more "solid", meaning breath to breath flow rate is not changing as much. I think all of this points to improvement.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#15
RE: Clusters of Hypopneas at beginning of sessions
I think that I have the charts correct.  Ti10 is before and Ti18 is the after.

I tried to turn off the ramp feature, but the machine just started with a high pressure and tapered down.  Was not comfortable and still had the clusters of hypopneas.  I then turned the ramp start up as far as it would go to 7.0. I will try to shorten the ramp time to see if that helps.


Attached Files Thumbnail(s)
       
Post Reply Post Reply
#16
RE: Clusters of Hypopneas at beginning of sessions
I think just shortening ramp time should do the trick. The close-up shows that respiration has not changed much, but inspiration, lasts longer if anything. The section shown for the new setting immediately precedes a RERA and that is related to the downward sloping inspiration flow (flow limitation) we see, however I don't see an arousal which should indicate RERA. Results here are better, so I think keeping the Ti higher than 1.0 is key, and you can try settings like 1.5 to see if it's more comfortable. I like the absence of hypopnea other than ramp. I'd have to say the sample size is too small to make more conclusions, so you might try keeping things at these settings a couple days before making changes.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#17
RE: Clusters of Hypopneas at beginning of sessions
I also agree with the sleep/wake junk. I know that the MyAir app algorithm made for a night's sleep seems to adjust for it too. It came to my attention when I would take the occasional nap which always surfaces a higher AHI than compared to a full night's sleep.
Post Reply Post Reply
#18
RE: Clusters of Hypopneas at beginning of sessions
An update to the changes made on my settings.

Original -> New settings are:
Mode. ST -> (ST)
IPAP. 14.0 -> (14.0)
EPAP. 7.0 -> (7.0)
BPM. 12 -> (12)
Ti. 1.0 -> (1.7)
Ramp 4.0cmH2O for 20 min -> ( 7.0cmH2O for 5 minutes)

Things seem pretty good; a nice improvement from where we started.  Last couple of nights AHIs were .84 and 1.18 per SleepyHead and 1.1 and 1.2 per DreamMapper.  Large leaks were high last night and so I need to work on the mask adjustments (AirFit F10 with a comfort cover). I think also that I dream more now and move around just before awakening.

Thanks again to all.  The ApneaBoard is a great resource with knowledgable members willing to help us new to the Pap journey.


Attached Files Thumbnail(s)
   
Post Reply Post Reply
#19
RE: Clusters of Hypopneas at beginning of sessions
Looking good milesarcher!
The hypopneas are usually an indication of too low of a pressure. Because they are occurring while you are in the ramp mode, I would suggest that you bump up the starting pressure to around 7cm. . . If you can tolerate it.
Crimson Nape
Apnea Board Moderator
www.ApneaBoard.com
___________________________________
Useful Links -or- When All Else Fails:
The Guide to Understanding OSCAR
OSCAR Chart Organization
Attaching Images and Files on Apnea Board
Apnea Helpful Tips

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#20
RE: Clusters of Hypopneas at beginning of sessions
Smile   It has been a month since my last post.  Things seem to be going well.  I am grateful for all the help, advice, knowledge and opinions that the Board has provided.  I consistently get 6-7 hours of good sleep with an AHI around 1.0.  I started this journey with an AFib diagnosis that led to the sleep study.

Things that work for me that I am grateful that members have shared.
  • Sleepyhead
  • Dr Dakota anti snore collar. (I find more comfortable and cooler than a full collar.)
  • Xylimelts  (Dry mouth pretty much gone.)
  • Dr Bronner's Unscented baby soap. 
  • Melatonin
  • Mask covers
  • Magnesium 
Again thanks to all.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  [Diagnosis] Help with Hypopneas! strugglingzombie 0 137 04-03-2024, 11:05 AM
Last Post: strugglingzombie
  Worrying session last night - hypopneas increased tenfold! BigWing 12 404 03-23-2024, 06:23 AM
Last Post: BigWing
  [Symptoms] Could These OSA Clusters Be Positional? Kaluni 0 129 03-18-2024, 03:42 PM
Last Post: Kaluni
  Does EPAP or IPAP control hypopneas? MrIvanDrago 3 332 02-29-2024, 05:08 PM
Last Post: Sleeprider
  [Treatment] Help Please - Hypopneas remain with BIPAP Newby 0 213 02-25-2024, 03:52 PM
Last Post: Newby
  Help Interpreting Oscar Charts - Huge Clusters of OA Heaviside 3 423 02-22-2024, 09:58 PM
Last Post: Sleeprider
  [CPAP] Cheyne Stokes Respiration and Central Apnea Clusters Aldiorg 8 877 02-19-2024, 04:11 PM
Last Post: SarcasticDave94


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.