Comparative Analysis of 8 APAP Machines
I found an article that may be worth reproducing or summarizing on the main forum. I thought I'd run it by you guys first. I don't think there is a commercial issue here. http://openres.ersjournals.com/content/1/1/00031-2015

Eight current machines were tested, AirSense 10 Autoset Standard and Response (ResMed), Dreamstar (Sefam), Icon (Fisher & Paykel), Resmart (BMC), Somnobalance (Weinmann), System One (Respironics) and XT-Auto (Apex). The differences in how the machines respond to obstructive apnea is pretty amazing. We have known for some time that the Resmed Airsense 10 responds fastest to OA, and this is confirmed in this study. The other reasonably fast responders were the F&P Icon and Somnobalance. In order of finish, Respironics, Resmart, Apex and Dreamstar failed to respond fast enough or strongly enough to normalize breathing when set to their minimum starting pressure of 4.0.

This reflects much of our advise in the forum to aggressively increase minimum pressure, particularly with the slower responding machines like Respironics. You can visulaize the pressure response of the machines in the figure below from the article (In order of best to worst finish):
A1 = Resmed Airsense 10 Autoset (purple)
A2 = Resmed Airsense 10 Autoset in Response Setting?? (Dark Blue)
C= F&P Icon (Yellow)
E = Sonomnobalance (Dark Grey)
F = PRS1 Remstar (Green)
D = Resmart (Light Blue)
G = Apex XT-Auto (Light Blue)
B = Dreamstar (Red)

[Image: F3.large.jpg?width=800&height=600&carousel=1]

Table 1 in the article has the best summary of quantitative results from the two test runs, but I'm not sure how to reproduce it here.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
09-09-2016 02:51 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
KSMatthew Offline

Advisory Members

Posts: 179
Joined: Jul 2016

Machine: PR DreamStation Auto
Mask Type: Nasal pillows
Mask Make & Model: ResMed Swift FX
Humidifier: none
CPAP Pressure: 7
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Kansas, United States

Post: #2
RE: Comparative Analysis of 8 APAP Machines
Interesting.

I just read the article. The TL;DR version: different APAPs connected to a simulated patient. The measurements show how quickly the APAPs reached a pressure great enough to resume normal breathing. The dashed black line shows the pressure required to resume normal breathing.

I'm not sure how to interpret that graph. It seems that the two AirSense 10 and the Icon machine were the only ones to reach that pressure?
09-09-2016 03:17 PM
Find all posts by this user Post Reply Quote this message in a reply
justMongo Offline

Monitors

Posts: 3,848
Joined: Sep 2013

Machine: ResMed VPAP Auto (S9)
Mask Type: Full face mask
Mask Make & Model: ResMed Mirage Quattro
Humidifier: ResMed H5i
CPAP Pressure: 18 IPAPmax - 11 EPAPmin, PS=5
CPAP Software: Other Software

Other Comments: μολὼν λαβέ

Sex: Male
Location: Калифорния

Post: #3
RE: Comparative Analysis of 8 APAP Machines
Good article. (I don't see any commercial issues with it either)
There may be some bias since funding seems to have come from one of the manufacturers.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
09-09-2016 03:21 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,342
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #4
RE: Comparative Analysis of 8 APAP Machines
Keep in mind the simulation is done with machines in auto mode starting at a pressure of 4.0. Your conclusion are correct insofar as the study criteria go. There may be a good reason for a machine to respond more slowly, or be better tuned to the patient's minimum pressure requirement, and that is not addressed by the study.

If an auto machine is to be dispensed without titration, I think the lesson here is that Resmed and Fisher Paykal are the best contenders. Also, if we see members struggling with their machines, or complaining of "hard to breath" when they use the Dreamstar, Apex, Resmart and Respironics auto machines, the answer has to be, increase minimum pressure because your machine takes too long to respond from the minimum auto setting.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
09-09-2016 03:25 PM
Find all posts by this user Post Reply Quote this message in a reply
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,342
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #5
RE: Comparative Analysis of 8 APAP Machines
I want to add that there are not many comparative studies out there. This is kind of a Consumer Report thing, where awareness makes for a more empowered shopper. If a supplier is trying to foist a cheap Asian machine on an insured patient,and claiming the machines are all the same; this article disproves that. This empowers our members to complain and demand a top-quality machine. I'm a bit surprised by the Respironics results, but it is only marginally behind the F&P Icon based on Table 1. So I think it still belongs in the upper tier for auto CPAP. The study design definitely rewards speed of response, and I have not checked whether the study was sponsored by Resmed. Might have to do that.

In any event, the Discussion and Lessons for Clinicians sections of the report clearly suggest speed is not everything, and the algorithms may not be in error.

Quote:Lessons for clinicians

This bench test study assessed how currently available APAP devices respond to a simulated OSA patient. Regardless of the simplified experimental setting employed in this work, as compared with the complexity of events found in the clinical arena when treating OSA patients, the following practical lessons can be derived.

The way that each commercial APAP device modifies nasal pressure when subjected to disturbed breathing patterns is different.

These differences in response among APAP devices are not necessarily caused by incorrect performances. Instead, they are the result of the particular engineering solutions implemented in each device.

Knowing the specific functioning features of each APAP device (e.g. sensitivity in detecting the different obstructive events, tolerance to events before increasing pressure and speed of pressure changes) may help to understand treatment compliance in specific patient phenotypes.

Choosing the optimal APAP device for the needs/preferences of each individual patient may improve therapy compliance.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
(This post was last modified: 09-09-2016 03:37 PM by Sleeprider.)
09-09-2016 03:30 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
KSMatthew Offline

Advisory Members

Posts: 179
Joined: Jul 2016

Machine: PR DreamStation Auto
Mask Type: Nasal pillows
Mask Make & Model: ResMed Swift FX
Humidifier: none
CPAP Pressure: 7
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Kansas, United States

Post: #6
RE: Comparative Analysis of 8 APAP Machines
(09-09-2016 03:25 PM)Sleeprider Wrote:  Keep in mind the simulation is done with machines in auto mode starting at a pressure of 4.0. Your conclusion are correct insofar as the study criteria go. There may be a good reason for a machine to respond more slowly, or be better tuned to the patient's minimum pressure requirement, and that is not addressed by the study.

If an auto machine is to be dispensed without titration, I think the lesson here is that Resmed and Fisher Paykal are the best contenders. Also, if we see members struggling with their machines, or complaining of "hard to breath" when they use the Dreamstar, Apex, Resmart and Respironics auto machines, the answer has to be, increase minimum pressure because your machine takes too long to respond from the minimum auto setting.

I am, and always have been, on constant pressure so I have no personal experience with auto-PAPs.

I do see the point that this graph goes out only 10 minutes and the tests were 2 hours long. But the information from the report seems to say that the other machines never were able to reach the proper setting.

It's possible the internal algorithms on those systems will adjust up by a small percentage at a time and try to creep in and eventually reach a limit they won't go above. Where the ResMed makes a large jump, then backs off and re-corrects.

In either case, your comment above that the answer has to be "increase minumum pressure" seems like it is the best advice.
09-09-2016 03:37 PM
Find all posts by this user Post Reply Quote this message in a reply
KSMatthew Offline

Advisory Members

Posts: 179
Joined: Jul 2016

Machine: PR DreamStation Auto
Mask Type: Nasal pillows
Mask Make & Model: ResMed Swift FX
Humidifier: none
CPAP Pressure: 7
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Kansas, United States

Post: #7
RE: Comparative Analysis of 8 APAP Machines
Here is the sponsorship info from that report:

Footnotes

Support statement: This work was carried out within the framework of a ResMed–University of Barcelona contract aimed at bench testing automatic continuous positive airway pressure devices. ResMed provided financial support to carry out the study but did not intervene in the design or performance of the tests, or in the data analysis. ResMed also funded the assistance of a medical writer in preparing the manuscript. Funding information for this article has been deposited with FundRef.

Conflict of interest: Disclosures can be found alongside this article at openres.ersjournals.com

Received June 5, 2015.
Accepted July 26, 2015.

Copyright ©ERS 2015

This article is open access and distributed under the terms of the Creative Commons Attribution Non-Commercial Licence 4.0.
09-09-2016 03:45 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
SuperSleeper Offline

Administrators

Posts: 9,967
Joined: Feb 2012

Machine: PR System One REMstar Auto (DS560)
Mask Type: Nasal pillows
Mask Make & Model: ResMed Mirage Swift II
Humidifier: none
CPAP Pressure: 12.5 - 18.5 cmH20 (auto range)
CPAP Software: SleepyHead

Other Comments: Have diabetes Type II

Sex: Male
Location: Illinois, USA

Post: #8
RE: Comparative Analysis of 8 APAP Machines
As far as I can tell, the site's content isn't commercial in nature.

I'm moving the thread to the Main Forum if that's okay, Sleeprider.

Smile

SuperSleeper
Apnea Board Administrator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.

09-09-2016 04:57 PM
Find all posts by this user Post Reply Quote this message in a reply
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,342
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #9
RE: Comparative Analysis of 8 APAP Machines
fine by me

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
09-09-2016 05:37 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
cands Online

Preferred Members

Posts: 44
Joined: Jul 2016

Machine: Resmed Airsense 10 Autoset
Mask Type: Full face mask
Mask Make & Model: Philips Amara View
Humidifier: Inbuilt
CPAP Pressure: 12.6 - 17.2
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Australia

Post: #10
RE: Comparative Analysis of 8 APAP Machines
(09-09-2016 02:51 PM)Sleeprider Wrote:  A2 = Resmed Airsense 10 Autoset in Response Setting?? (Dark Blue)

In the clinical menu:

Response: Options are Standard or Soft.
I assume 'Response' setting is 'Soft'?
09-09-2016 06:18 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  Newbie so confused between 2 machines Matzaball2016 1 65 Today 01:28 AM
Last Post: big_dave
  Medicare Advantage Coverage for APAP Evie 4 209 Yesterday 09:49 PM
Last Post: Evie
  Switching Machines owlgirl63 4 216 12-04-2016 05:11 AM
Last Post: kwhenrykerr
  cpap and apap not effictive Ezil71 109 3,116 11-05-2016 11:19 AM
Last Post: sonicboom
Question CHANGING MACHINES - SD CARD srlevine1 3 232 10-31-2016 07:11 PM
Last Post: srlevine1
  Unable to clear throat while on APAP NorthernGuy 5 307 10-17-2016 08:18 PM
Last Post: Sleeprider
  Apap I_will_never_sleep_again 2 279 10-14-2016 02:58 PM
Last Post: trish6hundred

Forum Jump:

Who's Online (Complete List)