Post: #11
RE: Configuring an AVAPS
(10-06-2016 12:53 PM)Sleeprider Wrote:  The other unit to look at is the Resmed Aircurve 10 ASV. For the System One SV advanced, the backup rate is called BPM or breaths per minute. By default, I'd suggest 14, but you may be able to look at your CPAP data and get something about 2-BPM less than your normal mean. The setup is pretty simple. This device offers up to 25 cm maximum pressure. You can set the max not to exceed any pressure, but for you, probably 20 cm. EPAPmin and needs to be set high enough to prevent most OA. For you, I think EPAPmin 8.0 and EPAPmax 11.0. PSmin 2.0, PSmax 8.0.

With these settings you will start at 10/8.0 to 13/11 for normal spontaneous breathing, and will see up to 16/8.0 to 19/11 in the presense of CA. The machine will detect reductions in volume and insufficient breaths per minute and will provide pressure support on a breath by breath basis to maintain your respiratory volume and rate.

Other settings have to do with inspiration time, BiFlex or rise time, ramp, humidity etc. The PRS1 960P is a full automatic ASV and has many settings that can be tweaked to increase comfort and synchronization with your breathing. An optional heated hose is available for this unit for about $75, and you can add it at any time.

Thank you! This clarifies things immensely. I'll check out the Resmed too if it's cheaper.

About the bpm-
1. should i use the best night on my cpap as a data point? I have one that's fairly recent and AHI 3.09. Or perhaps it would be better to run the machine for an hour or so whilst awake.
2. Were you referring to the median value of resp rate in sleepyhead rather than mean? or weighted average?
10-06-2016 01:49 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,318
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #12
RE: Configuring an AVAPS
You can set the preferences in Sleepyhead to report median or average. Both statistics are relevant, but I usually use the average. The breaths per minute are normally set at 2-less than your sleeping average BPM. It seems to result in a more natural feeling and keeps the machine from taking over when you are spontaneously breathing. In Sleepyhead, just click on File/Preferences, and under CPAP tab, set middle calculations to Normal Average, and upper percentile to 90%. I think that is more useful than the defaults.

The Resmed Aircurve ASV is not at all inexpensive, and getting a discounted price will be difficult. Its users report a more natural feeling respiration where the machine follows their natural pattern so well, they are almost unaware of the machine assisted breaths. The Philips seems to be more adjustable, but it takes some time to dial in the finer settings to where they feel completely natural. You are much more likely to find a PRS1 Model 960 at a good price.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
10-06-2016 02:35 PM
Find all posts by this user Post Reply Quote this message in a reply
jbsix Offline

Preferred Members

Posts: 20
Joined: Dec 2015

Machine: PR System One
Mask Type: Full face mask
Mask Make & Model: Airfit
Humidifier: PR System One
CPAP Pressure: unsure
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location:

Post: #13
RE: Configuring an AVAPS
Thanks, I couldn't find it anywhere. Choosing mean over median makes more sense to me too. And wow, aircurve sure is expensive! I don't mind tweaking as long as I'm sleeping well Smile
10-06-2016 03:10 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
jbsix Offline

Preferred Members

Posts: 20
Joined: Dec 2015

Machine: PR System One
Mask Type: Full face mask
Mask Make & Model: Airfit
Humidifier: PR System One
CPAP Pressure: unsure
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location:

Post: #14
RE: Configuring an AVAPS
I got the PRS1 ASV and am having a little trouble fine-tuning it.

For day 1, I thought I'd replicate my APAP settings (8-12) except I'd add more exhalation pressure relief since that was my biggest complaint on APAP.

These are my settings:
Code:
EPAP min = 4.0
EPAP max = 8.0
PS min = 4.0
PS max= 4.0 ; for an effective IPAP range of 8-12
bpm = 14 ; Resp. Rate mean on APAP = 16.56
Ti = 2.0 ; Insp Time mean on APAP = 2.18
Bi-Flex = 2 (and therefore no rise time set)

I switch on therapy and inhale, IPAP=8.0; good.
I exhale, EPAP=4.0; excellent.

But after about 1.5-2 seconds of exhalation, it switches over to IPAP prematurely even though I'm still exhaling. This results in the mask inflating as I'm still exhaling.

If I perform the world's longest exhalation here, the machine will switch back to EPAP after ~2s and then back to IPAP after ~2s and so on; completely ignoring my breathing cues.

I suspect it is due to one of the last three settings (most likely Ti), but could use some input.
(This post was last modified: 10-16-2016 03:53 AM by jbsix.)
10-16-2016 03:50 AM
Find all posts by this user Post Reply Quote this message in a reply
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,318
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #15
RE: Configuring an AVAPS
Take BPM down to 12 BPM and see if that solves your problem.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
10-16-2016 08:11 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
jbsix Offline

Preferred Members

Posts: 20
Joined: Dec 2015

Machine: PR System One
Mask Type: Full face mask
Mask Make & Model: Airfit
Humidifier: PR System One
CPAP Pressure: unsure
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location:

Post: #16
RE: Configuring an AVAPS
Had to take bpm down to 8 before it was at least feasible to put it on for the night. But there's still interference where it tries to force ipap while i'm still exhaling or epap while I'm still inhaling.

How is bpm calculated/put into effect?

If I set Ti to 3.0, the max bpm i can set is 10. So from what I can extrapolate, this would mean 10x3s inhalations per minute (30s), leaving another 30s for exhalation which is why it is trying to force change pressure type. Am I correct?

If so, then I could play around with it to get a better inhalation to exhalation ratio of 1:2.
10-16-2016 01:45 PM
Find all posts by this user Post Reply Quote this message in a reply
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,318
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #17
RE: Configuring an AVAPS
I would not exceed 2.0-2.2 on Ti. 8.0 is the low end of normal for BPM, 14 would be high. You want to find what works, but at some point that backup rate will kick in if you're not breathing. This is one of the thing the Resmed ASV does pretty well automatically is to sync with the user when they are spontaneously breathing. I've never heard a complaint that the machine cuts off exhalation.

We have a number of members that use this class of machine, and I'm hoping they will kick-in their opinions.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
10-16-2016 03:53 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
jbsix Offline

Preferred Members

Posts: 20
Joined: Dec 2015

Machine: PR System One
Mask Type: Full face mask
Mask Make & Model: Airfit
Humidifier: PR System One
CPAP Pressure: unsure
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location:

Post: #18
RE: Configuring an AVAPS
Could you elaborate on why 2.2 Ti shouldn't be exceeded?

The thread title saying AVAPS probably doesn't help with exposure...
10-17-2016 12:25 PM
Find all posts by this user Post Reply Quote this message in a reply
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,318
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #19
RE: Configuring an AVAPS
The inspiration time default is 1.8, and many of us have increased that. There is no hard rule you can't go more than 2.2, but give it a try before you increase beyond that. If you have a normal I:E ratio, then 2-seconds of inhalation and 3-sceonds of exhale will give you about 12 breaths per minute. Increasing Ti to 3 seconds would result in less than 10 BPM. It may be necessary for severe COPD in order to get the needed volume, but I would approach that kind of unusual inspiration time methodically and slowly, rather than all at once.

A procedure for setting Ti would be to set Ti at 1.8 (minimum). If the machine feels like it is switching from IPAP to EPAP too fast, than increase Ti in .1 increments. Don't increase Ti unless you get that "I wasn't done yet" sensations.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
(This post was last modified: 10-17-2016 02:19 PM by Sleeprider.)
10-17-2016 01:56 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
PaytonA Offline
Wiki Editor
Monitors

Posts: 3,005
Joined: Dec 2013

Machine: ResMed S9 VPAP Auto
Mask Type: Full face mask
Mask Make & Model: Resmed Mirage Quattro
Humidifier: H5i(distilled-top up)
CPAP Pressure: VAuto MinE14.0 MaxI 20.6 PS4.0
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Orange County,California

Post: #20
RE: Configuring an AVAPS
(10-16-2016 03:53 PM)Sleeprider Wrote:  I would not exceed 2.0-2.2 on Ti. 8.0 is the low end of normal for BPM, 14 would be high. You want to find what works, but at some point that backup rate will kick in if you're not breathing. This is one of the thing the Resmed ASV does pretty well automatically is to sync with the user when they are spontaneously breathing. I've never heard a complaint that the machine cuts off exhalation.

We have a number of members that use this class of machine, and I'm hoping they will kick-in their opinions.

From what I get from reading through this thread again is that jbsix has a Philips Respironics ASV not a Resmed but maybe I have missed something.

Best Regards,

PaytonA
10-17-2016 03:17 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  New User Problem with BiPAP AVAPS - Part 1 jmayer180 35 1,268 11-27-2016 09:50 AM
Last Post: jmayer180
  New AVAPS 1160P with BPM Rate and I-time Qs Ron N Berkeley 4 1,380 03-22-2016 04:53 PM
Last Post: robertbuckley
  Configuring FlashAir Cards. kimgh 5 1,411 08-02-2015 09:29 PM
Last Post: kimgh
  How do I Contact the SleepyHead Creators to Ask Them to Add My Machine AVAPS 1160P? jjgereg 5 1,899 08-11-2014 05:22 PM
Last Post: Ron N Berkeley
  Anyone Using Philips BiPAP AVAPS C Series Mod #1160P? jjgereg 11 3,933 08-06-2014 10:35 AM
Last Post: vsheline
  Philips AVAPS Mod #1160P Data Printout--HELP! jjgereg 8 2,588 08-01-2014 11:31 AM
Last Post: Breatherguy
  Phillips AVAPS best machine? physicsstudent13 9 3,034 04-29-2014 06:22 PM
Last Post: RonWessels

Forum Jump:

Who's Online (Complete List)