Could it be so..........
I changed masks 18 days ago.

Prior to that I was using a ResMed Mirage FX Nose Mask and was getting varying leakage readings. Some as bad as 35-40% leakage.

I changed to an PR Wisp and for the past 18 days have had "0%" leakage.

I am wondering if this could be correct as I have read that 0% leakage is rare, let alone 18 consecutive nights.

Could something in the CLINICAL SETUP need to be changed?Huh

Any input will be appreciated.
11-28-2014 03:31 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
PaulaO2 Offline
Wiki Editor
Moderators

Posts: 8,057
Joined: Feb 2012

Machine: S9 Autoset
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: H5i
CPAP Pressure: 14-20
CPAP Software: SleepyHead

Other Comments: Ehlers-Danlos Syndrome, Hypermobility Type; chronic sarcasm

Sex: Undisclosed
Location: western NC, USA

Post: #2
RE: Could it be so..........
No, because they are both the same type of mask, as in they are both nasal masks.

It could just be the Wisp fits you much better than the FX. But it would be odd for it to be reporting 0%.

Unplug the machine, wait a minute, plug it back in. See if it still says 0% the next morning.

PaulaO2
Apnea Board Moderator
www.ApneaBoard.com


Breathe deeply and count to zen.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
11-28-2014 07:51 PM
Find all posts by this user Post Reply Quote this message in a reply
zonk Offline

Advisory Members

Posts: 7,908
Joined: Feb 2012

Machine: A10 AutoSet
Mask Type: Nasal mask
Mask Make & Model: Activa LT
Humidifier: Integrated /ClimateLineAir
CPAP Pressure: 9/13
CPAP Software: ResScan

Other Comments: CPAP since Nov 2010

Sex: Male
Location: Australia

Post: #3
RE: Could it be so..........
(11-28-2014 03:31 PM)Shinny Wrote:  Could something in the CLINICAL SETUP need to be changed?Huh

Any input will be appreciated.
Respironcis masks have a "System One" resistance control settings
System One resistance 0 X1 X2 X3 X4 X5

% leakage refer to % of time spent in large leak, during large leak the machine cannot determine apnea events accurately

ResMed machines report unintentional leak and define unacceptable unintentional leak above than 24 L/m

Respironcis machines report total leak ... unintentional leak = total leak minus intentional leak
11-28-2014 09:33 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
PaytonA Online
Wiki Editor
Monitors

Posts: 3,000
Joined: Dec 2013

Machine: ResMed S9 VPAP Auto
Mask Type: Full face mask
Mask Make & Model: Resmed Mirage Quattro
Humidifier: H5i(distilled-top up)
CPAP Pressure: VAuto MinE14.0 MaxI 20.6 PS4.0
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Orange County,California

Post: #4
RE: Could it be so..........
I think that what Zonk is trying to tell you without actually saying it is as follows:

If you are reading the actual % leakage, it probbly is the % of time that there is major leakage.
If that is the case then it is not unusual to have 0% 18 days in a row if you have found a mask that gives you a very good seal. Just lean back and enjoy.

Best Regards,

PaytonA
11-28-2014 10:07 PM
Find all posts by this user Post Reply Quote this message in a reply
PaulaO2 Offline
Wiki Editor
Moderators

Posts: 8,057
Joined: Feb 2012

Machine: S9 Autoset
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: H5i
CPAP Pressure: 14-20
CPAP Software: SleepyHead

Other Comments: Ehlers-Danlos Syndrome, Hypermobility Type; chronic sarcasm

Sex: Undisclosed
Location: western NC, USA

Post: #5
RE: Could it be so..........
Ah! So it isn't saying the leak is 30L or 30% over, but 30 percent of the night, it was over the limit. Noted for future reference.

PaulaO2
Apnea Board Moderator
www.ApneaBoard.com


Breathe deeply and count to zen.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
11-28-2014 11:34 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Ghost1958 Offline

Advisory Members

Posts: 957
Joined: Aug 2014

Machine: PRS1 Series Sixty Auto Aflex
Mask Type: Nasal mask
Mask Make & Model: F&P Eson
Humidifier: PRS1 with Heated Hose
CPAP Pressure: 10.5 min 16 max
CPAP Software: SleepyHead EncoreBasic

Other Comments: Moderate to Severe SA Obstructive. Rapid O2 desat off the machine.

Sex: Male
Location: KY

Post: #6
RE: Could it be so..........
PRS1 machines show total leak. Mask vent plus leaks. When the total leak is big enough around 70 on the SH graph itll flag what it calls large leak.

They do have resistance settings for Respronics masks but that doesnt change the way the PR machines look at leaks or flag them.

If your showing a 0 leak with a PR machine something aint right. It would at least show the mask vent rate as leak.

Your ok until the machine says your in large leak. That will be about 70lpm total leak.

Itll show you large lak on its own info screen if your not using software though I believe you are.
11-29-2014 01:06 AM
Find all posts by this user Post Reply Quote this message in a reply
Shinny Offline

New Members

Posts: 2
Joined: Nov 2014

Machine: PR BiPAP Auto Bi-Flex 760
Mask Type: Nasal mask
Mask Make & Model: PR Wisp
Humidifier: PR System One
CPAP Pressure: 17/13
CPAP Software: Other Software

Other Comments:

Sex: Male
Location: Pa

Post: #7
RE: Could it be so..........
I think Payton A's response was great and I'll take his advise and enjoy it.Smile

Thanx 2 all for responding.Thanks


(11-28-2014 10:07 PM)PaytonA Wrote:  I think that what Zonk is trying to tell you without actually saying it is as follows:

If you are reading the actual % leakage, it probbly is the % of time that there is major leakage.
If that is the case then it is not unusual to have 0% 18 days in a row if you have found a mask that gives you a very good seal. Just lean back and enjoy.

Best Regards,

PaytonA
11-29-2014 11:52 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
justMongo Offline

Monitors

Posts: 3,838
Joined: Sep 2013

Machine: ResMed VPAP Auto (S9)
Mask Type: Full face mask
Mask Make & Model: ResMed Mirage Quattro
Humidifier: ResMed H5i
CPAP Pressure: 18 IPAPmax - 11 EPAPmin, PS=5
CPAP Software: Other Software

Other Comments: μολὼν λαβέ

Sex: Male
Location: Калифорния

Post: #8
RE: Could it be so..........
For most folks the biggest factor in CPAP success is finding the right mask.
Since we're all different, what works for one person may not work for another.
Some folks go through a few masks before finding the one that's comfortable and works.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
(This post was last modified: 11-29-2014 12:26 PM by justMongo.)
11-29-2014 12:25 PM
Find all posts by this user Post Reply Quote this message in a reply
PaytonA Online
Wiki Editor
Monitors

Posts: 3,000
Joined: Dec 2013

Machine: ResMed S9 VPAP Auto
Mask Type: Full face mask
Mask Make & Model: Resmed Mirage Quattro
Humidifier: H5i(distilled-top up)
CPAP Pressure: VAuto MinE14.0 MaxI 20.6 PS4.0
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Orange County,California

Post: #9
RE: Could it be so..........
(11-29-2014 01:06 AM)Ghost1958 Wrote:  If your showing a 0 leak with a PR machine something aint right. It would at least show the mask vent rate as leak.

If you look underneath the events breakdown on the left side of the detail tab, it says,"Time over leak redline". If one has had a good night it will be 0.000%. At least this is the way that it works for my Resmed machine.

Best Regards,

PaytonA
11-29-2014 12:53 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
trish6hundred Offline

Advisory Members

Posts: 6,424
Joined: May 2012

Machine: Resmed S9 AutoSet for Her
Mask Type: Full face mask
Mask Make & Model: Fisher & Paykel Simplus
Humidifier: H5i Heated Humidifier
CPAP Pressure: 10 - 7-20 Cm H2O
CPAP Software: Not using software

Other Comments: I started CPAP in 2008. Totally blind since birth.

Sex: Female
Location: Missouri, USA

Post: #10
RE: Could it be so..........
Hi Shinny,
WELCOME! to the forum.!
Much success to you with your CPAP therapy.

trish6hundred
11-29-2014 05:52 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Forum Jump:

Who's Online (Complete List)