Data interpretation
#1
Data interpretation
I have a new Intellipap auto. Last night's readings are below:

Usage Data
Last 90 Days Last 30 Days Last 7 Days Last Day
% Days at least 4 hours 100%
Day Count 1
Days at least 4 hours 1
95th Percentile Pressure 11.0
90th Percentile Pressure 10.5
AHI 6.0
Pressure Plateau Time 0%
Leak Time 0%
NRI 0.25
EPI 1
SmartCode F82F-FEF-3AMN

While Breathing Hours Last Day 6.6

What does this data tell me? Do I need to change anything? I have the apap pressure set at 8-16. Should the upper level be lowered to 12? I woke up during the night and the machine seemed to be running noisily at an excessive pressure. I had to ramp down to get back to sleep.

Codger
Post Reply Post Reply
#2
RE: Data interpretation
It's a good idea to keep an APAP within range of your needs. Going by that one night of data, 11 was your highest so 12 might be a good idea.

BUT, this is one night's worth of data. You need much more than that before you start changing things. Making a change now is not a good idea. Give it 2 weeks then look for trends. For example, you're going to see that AHI slowly come down. Your goal is 5 or less but it doesn't always happen immediately.
PaulaO

Take a deep breath and count to zen.




Post Reply Post Reply
#3
RE: Data interpretation
(10-09-2012, 11:13 AM)codger Wrote: Do I need to change anything? I have the apap pressure set at 8-16. Should the upper level be lowered to 12?
Codger
the manual for the IntelliPAP is available via Email (scroll down to section three for the instruction)
http://www.apneaboard.com/adjust-cpap-pr...tup-manual

I would do nothing. it makes no difference lowering max pressure as the Pressure plateau time is 0% which is percentage of time therapy was at the max pressure. the 90/95P (10.5/11) indicate the therapy pressure was AT or BELOW for 90/95% of usage time










Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Data interpretation - Luna G3 Big Ultima Energy 3 86 04-14-2024, 06:27 PM
Last Post: johnathanb76
  Oscar Data Interpretation UPDATED Johand77 1 111 04-10-2024, 08:37 PM
Last Post: Sleeprider
  OSCAR data interpretation Johand77 8 193 04-10-2024, 02:45 PM
Last Post: Johand77
  First time looking at graphs interpretation help smallhandsfeeding 1 289 03-11-2024, 08:12 AM
Last Post: jdavis
  APAP Trial - First Night's Results - Interpretation? Juicy 9 4,233 03-04-2024, 09:34 AM
Last Post: colethegamer
Smile How is my OSCAR data looking? Help with interpretation deirdrepen 3 276 02-29-2024, 07:24 PM
Last Post: Sleeprider
  Inconsistent sleep - OSCAR interpretation dadjackets 2 294 02-21-2024, 11:58 AM
Last Post: UnicornRider


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.