(11-21-2019, 11:51 PM)Hydrangea Wrote: I have 2 machines: DreamStation Auto is my daily machine, and ResMed Airsense 10 AutoSet is my travel machine. I have them both tuned in to similar settings that make me feel similar symptom relief. But my ResMed always yields an AHI that's about half of the AHI on the DreamStation.

I've personally concluded that they must have their own proprietary way of counting AHIs, and thus why the numbers are so different.

This is a very important insight. Can you figure out which events make the difference? (for example: maybe both machines are quoting the same number of Apneas but Dreamstation is quoting many additional Hypopneas).

Thanks,
Arik
Post Reply Post Reply






Donate to Apnea Board  
#12
RE: Different AHI results between two CPAP machines
Arik Wrote:The charts I added were for the purpose of demonstrating ignored events with the ResMed.
I do not understand your note about the tiny upticks.

I've attached part of the chart to show what I mean. You can see that the machine has detected a stopped breathing situation and after a few seconds commences the FOT (seen as a rapid oscillation in the flow and pressure graphs. At the "uptick" which I have circled, there is perhaps a little gasp or other intake of breath which the machine is interpreting as "event over". Then the process starts again. Neither of the stoppages before or after the uptake exceed 10 seconds, so the machine doesn't flag anything. (In this case the whole sequence of events is just less than 10 seconds).

   

Arik Wrote:Finally, I want to comment that the AASM definition of Hypopnea requires that "There is a ≥3% oxygen desaturation from pre-event baseline and/or the event is associated with an arousal". Most CPAP machines including ResMed and PRISMA are not able to detect either Oxygen disaturation or an arousal.

That's true. Unless you have the manufacturer's oximeter attached to the machine it can't measure oxygen saturation. So it just goes with the flow reductions I quoted in an earlier post.
Post Reply Post Reply
#13
RE: Different AHI results between two CPAP machines
If I was tasked with optimizing AHI detection for a CPAP firm, I could definitely envision deriving a sophisticated algorithm, for which the definition of a valid or significant hypopnea would be much more complicated than (x% flow reduction for y seconds).   I would gather pulseOx data and Quality of Sleep surveys and I don't know what other data and try to quantify severity of event in terms of O2 saturation and PQS (Perceived Quality of Sleep) etc., based on factors such as super detailed analysis of the event (more sophisticated than x% reduction for y secs), perhaps data immediately preceding and immediately following the event.   If any correlation could be demonstrated to severity of the hypopnea with these factors, I would try to use this to establish an predictive algorithm that could be used based on CPAP data alone (without explicit oximetry or PQS data).  I would recommend to my company that this more detailed approach be adopted for reported events in order to help the lay user focus on events that are more important.   I would also closely guard the details or even the existence of such an algorithm, as it would be intellectual property which could be a significant competitive advantage.

And we haven't even discussed mitigation, like pulsating pressure during the event!

I'm not saying that this is what ResMed is doing, but I don't know that it isn't.  I'm new to this forum but 2 things I have heard a lot here is that ResMed is different from the other machines and the overwhelming consensus is that they are better.  My interpretation is that people are saying that they feel better the next day after using an AirSense S10 - I do and isn't that the real object of the exercise?  If any part of my conjecture is true, I'm happy not to chase after non-significant events.   Draw your own conclusions about whether we should worry about simplistic x% reduction for y secs events
Post Reply Post Reply
#14
RE: Different AHI results between two CPAP machines
(11-22-2019, 08:03 AM)sopherFellow Wrote: If I was tasked with optimizing AHI detection for a CPAP firm, I could definitely envision deriving a sophisticated algorithm, for which the definition of a valid or significant hypopnea would be much more complicated than (x% flow reduction for y seconds).   I would gather pulseOx data and Quality of Sleep surveys and I don't know what other data and try to quantify severity of event in terms of O2 saturation and PQS (Perceived Quality of Sleep) etc., based on factors such as super detailed analysis of the event (more sophisticated than x% reduction for y secs), perhaps data immediately preceding and immediately following the event.   If any correlation could be demonstrated to severity of the hypopnea with these factors, I would try to use this to establish an predictive algorithm that could be used based on CPAP data alone (without explicit oximetry or PQS data).  I would recommend to my company that this more detailed approach be adopted for reported events in order to help the lay user focus on events that are more important.   I would also closely guard the details or even the existence of such an algorithm, as it would be intellectual property which could be a significant competitive advantage.

And we haven't even discussed mitigation, like pulsating pressure during the event!

I'm not saying that this is what ResMed is doing, but I don't know that it isn't.  I'm new to this forum but 2 things I have heard a lot here is that ResMed is different from the other machines and the overwhelming consensus is that they are better.  My interpretation is that people are saying that they feel better the next day after using an AirSense S10 - I do and isn't that the real object of the exercise?  If any part of my conjecture is true, I'm happy not to chase after non-significant events.   Draw your own conclusions about whether we should worry about simplistic x% reduction for y secs events

Hi SopherFellow.
These are words of wisdom. However, let me tell you that I felt great after a night of full PSG test in hospital while the report claimed 144 (!) arousals which I was not aware of and 68 disaturation events. There is a direct link between high AHI and plenty of disaturation events that can be detected at home with a $200 device. I suspect no one can tell the difference in feeling after a night with 100 or 20 disaturation events. 
ResMed is by far the most comfortable and quite device I tested and I plan to keep using it.  What I currently have in mind is to optimize my therapy towards  minimizing the number of events, no matter of what type or length of time, and check periodically my saturation profile (cannot do this every night). The great feature in OSCAR of defining user's customized events will enable me doing this without the need to spend hours looking at graphs.
Thanks,
Arik
Post Reply Post Reply






Donate to Apnea Board  
#15
RE: Different AHI results between two CPAP machines
Arik

I recently got the Wellue O2Ring for $170 on Amazon and now I monitor my O2 every night, in case you want to do that.

Dennis
Post Reply Post Reply
#16
RE: Different AHI results between two CPAP machines
(11-22-2019, 09:24 AM)sopherFellow Wrote: Arik

I recently got the Wellue O2Ring for $170 on Amazon and now I monitor my O2 every night, in case you want to do that.

Dennis
It will be interesting to have your opinion about the Wellue. I'm using the CMS 50F plus and trust mainly relative values, rather than absolute ones. The display is very "jumpy" so it is hard to get a real time value but the written overnight report is good and you can have trends of average saturation value and number of disaturation events.
Arik
Post Reply Post Reply
#17
RE: Different AHI results between two CPAP machines
Additional response to your comments about user insensitivity to events.

On a typical night I have zero events where my O2 drops below 90%.   Nonetheless, I feel substantially more or less rested/tired on a day following a high/low number of events reported by my AirSense.  O2 saturation is important, but it is not the whole story.  Disruptive sleep, even if it does not reduce O2, may still prove to be unsatisfactory and non-restful.

You may be particularly insensitive to events - this might be because you are used to having events and you may have such a large sleep deficit that it will take quite a while for you to see improvement.   So, I disagree that users in general are insensitive to events.   I see a key question asked in virtually every thread on here (including this one) is "How are you feeling next day?"   Once you overcome your sleep deficit acquired over a long period of time, perhaps you will find this a useful question as well.

Submitted for your consideration.

Dennis

P.S.   Early days, but so far very impressed with the O2Ring.  On the 1 occasion that I had an O2 drop, it's vibration alarm awoke me and I was able to fix the problem.
Post Reply Post Reply






Donate to Apnea Board  
#18
RE: Different AHI results between two CPAP machines
   
Here's a screenshot of last night's results from the O2Ring.    Note the drop in O2 around 3:00 - here the vibration alarm woke me and I adjusted my collar.
Post Reply Post Reply
#19
RE: Different AHI results between two CPAP machines
(11-22-2019, 10:05 AM)sopherFellow Wrote: Additional response to your comments about user insensitivity to events.

On a typical night I have zero events where my O2 drops below 90%.   Nonetheless, I feel substantially more or less rested/tired on a day following a high/low number of events reported by my AirSense.  O2 saturation is important, but it is not the whole story.  Disruptive sleep, even if it does not reduce O2, may still prove to be unsatisfactory and non-restful.

You may be particularly insensitive to events - this might be because you are used to having events and you may have such a large sleep deficit that it will take quite a while for you to see improvement.   So, I disagree that users in general are insensitive to events.   I see a key question asked in virtually every thread on here (including this one) is "How are you feeling next day?"   Once you overcome your sleep deficit acquired over a long period of time, perhaps you will find this a useful question as well.

Submitted for your consideration.

Dennis

P.S.   Early days, but so far very impressed with the O2Ring.  On the 1 occasion that I had an O2 drop, it's vibration alarm awoke me and I was able to fix the problem.
I agree that this is a useful question. However, my daily feeling is affected by many other factors. 

Sleep Apnea has two main and independent problems: Disaturation events and arousals. Arousals break the sleep sequence and thus interfere with the brain maintenance process. While minimizing AHI will reduce disat events, I could not find anything that can help reducing the arousals.

Arik
Post Reply Post Reply
#20
RE: Different AHI results between two CPAP machines
(11-22-2019, 02:55 AM)Arik Wrote:
(11-21-2019, 11:51 PM)Hydrangea Wrote: I have 2 machines: DreamStation Auto is my daily machine, and ResMed Airsense 10 AutoSet is my travel machine. I have them both tuned in to similar settings that make me feel similar symptom relief. But my ResMed always yields an AHI that's about half of the AHI on the DreamStation.

I've personally concluded that they must have their own proprietary way of counting AHIs, and thus why the numbers are so different.

This is a very important insight. Can you figure out which events make the difference? (for example: maybe both machines are quoting the same number of Apneas but Dreamstation is quoting many additional Hypopneas).

Thanks,
Arik

I can't figure it out. But I'm also not all that great at understanding what the data is saying.

I have learned that the machines don't record the same info.  My ResMed 10 Airsense Autoset records Flow Limit, but my DreamStation Auto does not.  But I don't know how that plays into AHI and how the companies decide what is AHI-worthy.
Post Reply Post Reply






Donate to Apnea Board  


Possibly Related Threads...
Thread Author Replies Views Last Post
  [Equipment] Machines to consider and Airmini Simo50 5 101 11-21-2020, 01:37 AM
Last Post: Simo50
  PSG Results: Could this be UARS? deebob 258 7,022 11-18-2020, 12:17 AM
Last Post: greyham
Question [Diagnosis] WatchPAT results (UK) picassobaby 0 71 11-09-2020, 12:26 PM
Last Post: picassobaby
  OSCAR results seem to say I'm doing ok - so why do I wake up with a sore throat... Burra 12 365 11-09-2020, 09:12 AM
Last Post: Sleeprider
Angry [News] WARNING -- DISCOUNTED MACHINES srlevine1 12 902 10-21-2020, 03:19 PM
Last Post: Big Guy
  Seeking the service plug in modules and software for Phillips ASV machines racprops 1 114 10-19-2020, 11:00 PM
Last Post: SarcasticDave94
  Issues with shipping CPAP machines to Taiwan - Any advice? WWu777 41 4,349 10-18-2020, 09:42 AM