Dreamstation 500X110 - Air too cold
#1
Dreamstation 500X110 - Air too cold
Hello,
new to the forum. Would love to use my machine again after putting it aside due to different problems.
So I am starting over. Right now, the air seems too cold. I know that it's room temperature, but it feels like coming directly from my fridge. Since I am sensitive to cold air, I feel I am getting a cold when I use it.
The preheated humidifier is on, heated tube HT15, Apex mask. 
Downloaded SleepyHead just now, but cannot provide data yet.

Anybody knows the problem?

Thanks!
Post Reply Post Reply
#2
RE: Dreamstation 500X110 - Air too cold
You should be able to adjust the both the humidifier and hose temps. If you don't see them than enter into the Therapy settings by holding the big knob and ramp button in for 5 seconds. You can adjust them in there. You also can request a clinic manual in the members private files section.

Welcome to the board.
Download SleepyHead
Organize your Sleepyhead Charts
Posting Charts
Beginner's Guide to SleepyHead
Mask Primer
5
Advisory Members serve as an "Advisory Committee" to help shape Apnea Board's rules & policies.

Membership in the Advisory Members group does not imply medical expertise or qualification for advising Sleep Apnea patients concerning their treatment.



Post Reply Post Reply
#3
RE: Dreamstation 500X110 - Air too cold
Also, the humidifier may need to be reseated with the blower. There is a button on the bottom edge of the machine on the filter side. If you press the button and pull the machine apart, the put the assembly back together it may reseat the connection. Make sure the humidifier water tub is removed before doing this.
Jesse


Post Reply Post Reply
#4
RE: Dreamstation 500X110 - Air too cold
(06-07-2018, 11:57 AM)Musepux Wrote: Hello,
new to the forum. Would love to use my machine again after putting it aside due to different problems.
So I am starting over. Right now, the air seems too cold. I know that it's room temperature, but it feels like coming directly from my fridge. Since I am sensitive to cold air, I feel I am getting a cold when I use it.
The preheated humidifier is on, heated tube HT15, Apex mask. 
Downloaded SleepyHead just now, but cannot provide data yet.

Anybody knows the problem?

Thanks!

If your humidifier has water in it, but you have the setting low, then the air will be cooled as a result of evaporative cooling.  If you want a warmer temperature, turn the humidifier setting up.  That will warm the water, and the air.
Post Reply Post Reply
#5
RE: Dreamstation 500X110 - Air too cold
Hi Musepux,
Everyone is different. Sometimes I enjoy the cooler air, sometimes not.

Turn your heated hose setting up a little higher. The heated hose can be adjusted separately from the humifier setting. Although, too much humidity sometimes is not better than too little. Experiment to find what works for you.

Do you have a Clinicians Manual? See link at top of page.
OpalRose
Apnea Board Administrator
www.apneaboard.com

_______________________
OSCAR Chart Organization
How to Attach Images and Files.
OSCAR - The Guide
Soft Cervical Collar
Optimizing therapy
OSCAR supported machines
Mask Primer



INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE.  ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA.  INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#6
RE: Dreamstation 500X110 - Air too cold
(06-08-2018, 06:52 AM)OpalRose Wrote: Hi Musepux,
Everyone is different.  Sometimes I enjoy the cooler air, sometimes not.  

Turn your heated hose setting up a little higher. The heated hose can be adjusted separately from the humifier setting.  Although, too much humidity sometimes is not better than too little.  Experiment to find what works for you.

Do you have a Clinicians Manual?  See link at top of page.

Thanks, everybody. I managed to crank it up.
If I have further questions, do I keep asking here or shall I open another thread?

Thanks again.
Post Reply Post Reply
#7
RE: Dreamstation 500X110 - Air too cold
If it's a new topic go ahead and start a new thread.
Download SleepyHead
Organize your Sleepyhead Charts
Posting Charts
Beginner's Guide to SleepyHead
Mask Primer
5
Advisory Members serve as an "Advisory Committee" to help shape Apnea Board's rules & policies.

Membership in the Advisory Members group does not imply medical expertise or qualification for advising Sleep Apnea patients concerning their treatment.



Post Reply Post Reply
#8
RE: Dreamstation 500X110 - Air too cold
(06-08-2018, 11:06 AM)Musepux Wrote:
(06-08-2018, 06:52 AM)OpalRose Wrote: Hi Musepux,
Everyone is different.  Sometimes I enjoy the cooler air, sometimes not.  

Turn your heated hose setting up a little higher. The heated hose can be adjusted separately from the humifier setting.  Although, too much humidity sometimes is not better than too little.  Experiment to find what works for you.

Do you have a Clinicians Manual?  See link at top of page.

Thanks, everybody. I managed to crank it up.
If I have further questions, do I keep asking here or shall I open another thread?

Thanks again.

FYI- Turning up the heated hose will warm the air without adding any new water.  When you do that, you reduce the relative humidity in the air, making it drier than it would have been had you not turned up the heated hose.
Post Reply Post Reply
#9
RE: Dreamstation 500X110 - Air too cold
(06-08-2018, 01:11 PM)Reznik Wrote: FYI- Turning up the heated hose will warm the air without adding any new water. 

A-h-h-h-h I don't believe that to be correct.  According to Resmed, if the ClimateLine is set to "Auto", it will try and maintain an 85% humidity level based on the hose temperature.

Here is the actual verbiage from their site:
Resmed Website Wrote:ClimateControl Auto mode is the recommended mode to use when using ClimateLine Air heated tube. It is the default mode, set to an air temperature at the mask of 80°F(27°C) and 85% relative humidity. If you feel the need to decrease or increase the temperature, you need to switch to Manual mode, and carefully look at the humidity level as well as the heated tube temperature to help prevent rainout.

I hope this may clarify this matter.
Crimson Nape
Apnea Board Moderator
www.ApneaBoard.com
___________________________________
Useful Links -or- When All Else Fails:
The Guide to Understanding OSCAR
OSCAR Chart Organization
Attaching Images and Files on Apnea Board
Apnea Helpful Tips

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#10
RE: Dreamstation 500X110 - Air too cold
(06-08-2018, 01:22 PM)Crimson Nape Wrote:
(06-08-2018, 01:11 PM)Reznik Wrote: FYI- Turning up the heated hose will warm the air without adding any new water. 

A-h-h-h-h I don't believe that to be correct.  According to Resmed, if the ClimateLine is set to "Auto", it will try and maintain an 85% humidity level based on the hose temperature.

Here is the actual verbiage from their site:
Resmed Website Wrote:ClimateControl Auto mode is the recommended mode to use when using ClimateLine Air heated tube. It is the default mode, set to an air temperature at the mask of 80°F(27°C) and 85% relative humidity. If you feel the need to decrease or increase the temperature, you need to switch to Manual mode, and carefully look at the humidity level as well as the heated tube temperature to help prevent rainout.

I hope this may clarify this matter.

I'm pretty sure that's only true if the humidifier is set to Auto and the heated hose is set to Auto.  

If you manually set the heated hose to a higher temperature (which is what someone previously recomended), the auto functionality on the heated hose is disabled and the temperature you choose is used instead.  So, you'll no longer get 80 degrees at 85% RH.  You'll get the temperature you chose instead.

I'm not sure on whether you can keep the humidifier at auto when you choose to manually set heated hose temperature, but if you can, then I'd expect that as you increased the temperature on the heated hose, the system would reacted by increasing the temperature in the humidifier to keep the relative humidity at 85%.

If manually setting the heated hose temperature disables the auto humidity function, then increasing the heated hose temperature without increasing the humidity temperature will result in a net reduction of the relative humidity.

I use an **unheated** hose.  Using an IR thermometer, I measured the temperature of the humidifier chamber and the temperature of the inside of the mask.  Typically, at 80 degrees in the chamber, you lose about 1-2 degrees in temperature between the humidifier and chamber.  The higher the chamber temperature, the greater loss in the hose.  At 120 degrees in the chamber, you can lose about 6-8 degrees in the hose.  If you put a hose wrap on, the loss becomes almost zero.  The same is true if you use a heated hose.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Tips for continuing PAP through a cold Bostonsnow 4 545 07-28-2023, 11:55 AM
Last Post: Deborah K.
  Head cold Barbara.Levitan 2 607 03-10-2023, 11:23 PM
Last Post: srlevine1
  Dreamstation 1 to Dreamstation 2 - hardly any water use Lazagna 4 1,753 10-22-2022, 06:59 PM
Last Post: Tampa Jim
  [CPAP] Dreamstation 2 vs. Dreamstation 1 Pressure FOMO 2 1,559 09-18-2022, 07:59 PM
Last Post: FOMO
  [Equipment] Cold-weather camping? tundranapper 4 886 05-26-2022, 07:23 AM
Last Post: RNeil
  High pressures causing/worsening cold symptoms Chalkie 4 829 11-26-2021, 08:44 AM
Last Post: Chalkie
  BiPAP advisable with a cold? BobbieM 10 1,265 06-16-2021, 01:47 PM
Last Post: WakeUpTime


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.