Dreamstation Auto BIPAP
#11
RE: Dreamstation Auto BIPAP
Leak rate is good. Philips machines calculate both total leak and excess leak, and your results are great.

For some reason, your doctor is moving towards ASV, and you may want to discuss any concerns with him to see if he disagrees with making some changes. That said, this is what I see. You have a mixture of OA and CA events. I would setup the machine a bit differently to deal with both issues. Your current settings are 16/11 (PS=5.0). If you want to try auto BiPAP, then I would set EPAPmin to 11.0, IPAP max to 18, PSmin 3.0, PSmax 5.0. And I might even reduce PSmin to 2.0.

Here's why: Your obstructive events are triggering with flow limitation and snores. So OSA is clearly a component of your sleep architecture. The solution to resolving OA is higher EPAP pressure. By setting the EPAPmin at your current fixed BiPAP level, you will start from the same place, but the machine will detect precursors of OA and increase EPAP pressure as needed. A lower PS may help reduce CA events. Pressure support helps reduce hypopnea and RERA, but it is not able to resolve CA in a non-ASV machine. That's because IPAP pressure has to be triggered by your initiation of an inhale, and the IPAP pressure support is not high enough to cause an involuntary breath anyway. So reducing PSmin to 2-3 may improve CA by limiting CO2 washout, and in any event, it won't make it worse. If you see an increase in H, then try more PS. The auto settings will allow EPAP to move up to as high as 16 if needed, but most likely, you just need a couple cm additional pressure during the OA episodes that may be sleep position or sleep stage related. That's my 2-cents, if you want to optimize the bilevel. From your graphs, I don't see a compelling argument for ASV yet.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#12
RE: Dreamstation Auto BIPAP
Still waiting for response from Doctor

[Image: 3KqR8J7l.png]
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Which biPAP? HalfAsleep 9 259 03-14-2024, 09:04 AM
Last Post: HalfAsleep
  My inspiratory time is higher than expiratory on Auto S Bipap Stubert 3 381 02-01-2024, 09:27 AM
Last Post: Sleeprider
  Auto BiPAP fixed PS vs Auto PS - pros and cons? Stubert 5 585 01-09-2024, 02:29 AM
Last Post: Stubert
  bipap Innisfree 2 421 12-12-2023, 07:00 PM
Last Post: UnicornRider
  Help with bipap Al N 7 697 09-14-2023, 06:15 PM
Last Post: Al N
  Dry mouth with Resmed 11 auto, not with S9 auto? pumpkln 3 598 09-13-2023, 12:46 PM
Last Post: eok361
  DreamStation Auto Initial Start-up WilliamZ 3 503 09-11-2023, 11:23 AM
Last Post: Gideon


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.