Post Reply 
[Equipment] Machine with fixed EPAP/IPAP
Author Message
richb Online

Advisory Members

Posts: 525
Joined: Jun 2015

Machine: Resmed AirCurve 10 ASV
Mask Type: Nasal mask
Mask Make & Model: ResMed Mirage FX
Humidifier: Built in
CPAP Pressure: Epap 6 Ipap 8 Auto ASV
CPAP Software: ResScan SleepyHead

Other Comments: Idiopathic Central Apnea

Sex: Male
Location: Vermont

Post: #31
RE: Machine with fixed EPAP/IPAP
(08-29-2015 07:44 PM)vsheline Wrote:  
(08-22-2015 03:43 PM)richb Wrote:  All Hypopnea and periodic breathing at low pressure and nearly all Centrals at higher pressure. I am currently on a ResMed S9 ST machine. My current settings are EPAP 11 IPAP 15 and a backup rate of 8. The backup rate of 8 BPM is at the same pressure as the normal settings. Last night I had an AHI of 9 and felt terrible when I got up this AM. I looked at my waveform and found erratic breathing for most of the night. ( The machine doesn't score a Hypopnea when it is providing backup breaths). I only had about 1 hour of smooth waveform indicating some level of sleep. I am trying to get an ASV machine but am worried that might not work either. I have an appointment with a new sleep center in a month.

Hi richb,

During your centrals the ST machine is initiating IPAP for you, but with IPAP only 4 higher than EPAP the machine will not be able to keep you adequately ventilated.

I suggest decreasing EPAP by 1 (to 10) for a week. If the obstructive apneas stay low, I suggest lowering EPAP again, to 9 for a week. If you can manage to sleep only on your side(s) you may be able to keep lowering EPAP each week until you see an increase in your obstructive events and know you've backed off EPAP too much.

Each time you decrease EPAP (keeping IPAP unchanged) the Pressure Support (PS is the difference between EPAP and IPAP) will increase, better allowing the machine to keep you adequately ventilated during central events.

If you find you can't lower EPAP to achieve a higher PS, I recommend increasing IPAP 1 cmH2O per week, in order to increase PS.

When the PS gets as high as 8 the machine may be able to keep you adequately ventilated during central events.

When I stop breathing my ASV machine increases the PS to around 8 or 9 or 10, which a big enough difference between EPAP and IPAP to do for me all the work of breathing.

If the CA Index (number of central apneas per hour) is at least 5 and larger than your OA Index then I think you will qualify for an ASV machine.

If you do get an ASV machine, I would suggest the PRS1 BiPAP autoSV Advanced with heated tube. This machine is versatile and can be operated like an ASVAuto, ASV, BiPAP ST, BiPAP Auto, fixed BiPAP, APAP or CPAP machine. (Very much unlike the ResMed AurCurve 10 ASV which can only be operated in a particularly dumb (no CA versus OA detection!) version of basic CPAP therapy mode, or in ASV modes which are restrictive and require the difference between Min PS and Max PS be at least 5.)

In my view, ResMed has shamefully disserved their ASV customers by not including central apnea detection when in CPAP mode and by not including any of the other less expensive therapy modes other than their restrictive versions of ASV therapy. (But it's all so simple to set up and automatic.... no thanks!)

Take care,
--- Vaughn

Thanks for the recommendation. I have an appt next Tues with the sleep doc. I have shown my wave patterns to the DME. They are recommending an ASV machine. I will try to get the PRS1. I had to stop using the S9 st machine because it was sending my respiration rate to 40 bpm.

Apnea Board Member RobySue has posted a Beginners Guide to Sleepyhead Software here: http://www.apneaboard.com/wiki/index.php...SleepyHead
08-29-2015 08:24 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
PaytonA Offline
Wiki Editor
Monitors

Posts: 3,005
Joined: Dec 2013

Machine: ResMed S9 VPAP Auto
Mask Type: Full face mask
Mask Make & Model: Resmed Mirage Quattro
Humidifier: H5i(distilled-top up)
CPAP Pressure: VAuto MinE14.0 MaxI 20.6 PS4.0
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Orange County,California

Post: #32
RE: Machine with fixed EPAP/IPAP
Sorry - brain fart!
(This post was last modified: 08-29-2015 08:54 PM by PaytonA.)
08-29-2015 08:50 PM
Find all posts by this user Post Reply Quote this message in a reply
richb Online

Advisory Members

Posts: 525
Joined: Jun 2015

Machine: Resmed AirCurve 10 ASV
Mask Type: Nasal mask
Mask Make & Model: ResMed Mirage FX
Humidifier: Built in
CPAP Pressure: Epap 6 Ipap 8 Auto ASV
CPAP Software: ResScan SleepyHead

Other Comments: Idiopathic Central Apnea

Sex: Male
Location: Vermont

Post: #33
RE: Machine with fixed EPAP/IPAP
(08-29-2015 08:24 PM)richb Wrote:  
(08-29-2015 07:44 PM)vsheline Wrote:  
(08-22-2015 03:43 PM)richb Wrote:  All Hypopnea and periodic breathing at low pressure and nearly all Centrals at higher pressure. I am currently on a ResMed S9 ST machine. My current settings are EPAP 11 IPAP 15 and a backup rate of 8. The backup rate of 8 BPM is at the same pressure as the normal settings. Last night I had an AHI of 9 and felt terrible when I got up this AM. I looked at my waveform and found erratic breathing for most of the night. ( The machine doesn't score a Hypopnea when it is providing backup breaths). I only had about 1 hour of smooth waveform indicating some level of sleep. I am trying to get an ASV machine but am worried that might not work either. I have an appointment with a new sleep center in a month.

Hi richb,

During your centrals the ST machine is initiating IPAP for you, but with IPAP only 4 higher than EPAP the machine will not be able to keep you adequately ventilated.

I suggest decreasing EPAP by 1 (to 10) for a week. If the obstructive apneas stay low, I suggest lowering EPAP again, to 9 for a week. If you can manage to sleep only on your side(s) you may be able to keep lowering EPAP each week until you see an increase in your obstructive events and know you've backed off EPAP too much.

Each time you decrease EPAP (keeping IPAP unchanged) the Pressure Support (PS is the difference between EPAP and IPAP) will increase, better allowing the machine to keep you adequately ventilated during central events.

If you find you can't lower EPAP to achieve a higher PS, I recommend increasing IPAP 1 cmH2O per week, in order to increase PS.

When the PS gets as high as 8 the machine may be able to keep you adequately ventilated during central events.

When I stop breathing my ASV machine increases the PS to around 8 or 9 or 10, which a big enough difference between EPAP and IPAP to do for me all the work of breathing.

If the CA Index (number of central apneas per hour) is at least 5 and larger than your OA Index then I think you will qualify for an ASV machine.

If you do get an ASV machine, I would suggest the PRS1 BiPAP autoSV Advanced with heated tube. This machine is versatile and can be operated like an ASVAuto, ASV, BiPAP ST, BiPAP Auto, fixed BiPAP, APAP or CPAP machine. (Very much unlike the ResMed AurCurve 10 ASV which can only be operated in a particularly dumb (no CA versus OA detection!) version of basic CPAP therapy mode, or in ASV modes which are restrictive and require the difference between Min PS and Max PS be at least 5.)

In my view, ResMed has shamefully disserved their ASV customers by not including central apnea detection when in CPAP mode and by not including any of the other less expensive therapy modes other than their restrictive versions of ASV therapy. (But it's all so simple to set up and automatic.... no thanks!)

Take care,
--- Vaughn

Thanks for the recommendation. I have an appt next Tues with the sleep doc. I have shown my wave patterns to the DME. They are recommending an ASV machine. I will try to get the PRS1. I had to stop using the S9 st machine because it was sending my respiration rate to 40 bpm.

Saw the sleep Doc today. He is giving up on me. It seems he doesn't do complicated. The Doc couldn't figure out what was going on from my SH graphs. I have an appt. in early Oct with the Sleep center at the local University Hospital. The DME provider had recommended an ASV machine and the Doc agreed that I probably do as well. He also said I need a titration on the ASV machine and thought that the new Doc should do it. I thought ASV was auto titrating!

Rich

Apnea Board Member RobySue has posted a Beginners Guide to Sleepyhead Software here: http://www.apneaboard.com/wiki/index.php...SleepyHead
09-08-2015 04:49 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
storywizard Offline

Advisory Members

Posts: 350
Joined: Jul 2014

Machine: ResMed VPAP Adapt S9
Mask Type: Nasal pillows
Mask Make & Model: P10 taped to face
Humidifier: included with machine
CPAP Pressure: CPAP 11cm
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Vancouver, BC, Canada

Post: #34
RE: Machine with fixed EPAP/IPAP
Rich wrote:
Saw the sleep Doc today. He is giving up on me. It seems he doesn't do complicated. The Doc couldn't figure out what was going on from my SH graphs. I have an appt. in early Oct with the Sleep center at the local University Hospital. The DME provider had recommended an ASV machine and the Doc agreed that I probably do as well. He also said I need a titration on the ASV machine and thought that the new Doc should do it. I thought ASV was auto titrating!

Hi there...

from my experience with two titration studies in the last 2 months, and 4 months worth of data on ASV I would say its hard to really trust either the titration studies, or the machine data totally. I would go for the titration, look at your data, and access how you feel...then hope that the doc you see will be skillful enough, and care enough to help you, if not then there is this board and your own feelings on how you feel on the therapy ...I sincerely wish you the best of luck, sometimes its hard being complicated...

Storywizard
09-08-2015 05:28 PM
Find all posts by this user Post Reply Quote this message in a reply
richb Online

Advisory Members

Posts: 525
Joined: Jun 2015

Machine: Resmed AirCurve 10 ASV
Mask Type: Nasal mask
Mask Make & Model: ResMed Mirage FX
Humidifier: Built in
CPAP Pressure: Epap 6 Ipap 8 Auto ASV
CPAP Software: ResScan SleepyHead

Other Comments: Idiopathic Central Apnea

Sex: Male
Location: Vermont

Post: #35
RE: Machine with fixed EPAP/IPAP
(09-08-2015 05:28 PM)storywizard Wrote:  Rich wrote:
Saw the sleep Doc today. He is giving up on me. It seems he doesn't do complicated. The Doc couldn't figure out what was going on from my SH graphs. I have an appt. in early Oct with the Sleep center at the local University Hospital. The DME provider had recommended an ASV machine and the Doc agreed that I probably do as well. He also said I need a titration on the ASV machine and thought that the new Doc should do it. I thought ASV was auto titrating!

Hi there...

from my experience with two titration studies in the last 2 months, and 4 months worth of data on ASV I would say its hard to really trust either the titration studies, or the machine data totally. I would go for the titration, look at your data, and access how you feel...then hope that the doc you see will be skillful enough, and care enough to help you, if not then there is this board and your own feelings on how you feel on the therapy ...I sincerely wish you the best of luck, sometimes its hard being complicated...
Storywizard


I am afraid that I may have to go through a new sleep study and then a new titration. I have already had one PSG and two titrations. None of the titrations were on an ASV machine. I didn't sleep very well during any of the studies. No REM sleep observed during my 2nd titration. My new appointment is less than a month away.

Apnea Board Member RobySue has posted a Beginners Guide to Sleepyhead Software here: http://www.apneaboard.com/wiki/index.php...SleepyHead
(This post was last modified: 09-09-2015 09:24 AM by richb.)
09-09-2015 09:24 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
vsheline Offline

Advisory Members

Posts: 1,907
Joined: Jul 2012

Machine: S9 VPAP Adapt (USA Model# 36007, not better 36037 or 36067)
Mask Type: Full face mask
Mask Make & Model: F10 or SimPlus w/ 2Liners. MirageQuatro & Gecko gel pad
Humidifier: H5i
CPAP Pressure: 15 EPAP, PS 5-10
CPAP Software: ResScan

Other Comments: Marfan Syndrome, chronic bradycardia, occasional Cheyne-Stokes Respiration

Sex: Male
Location: California, USA

Post: #36
RE: Machine with fixed EPAP/IPAP
(09-09-2015 09:24 AM)richb Wrote:  I am afraid that I may have to go through a new sleep study and then a new titration. I have already had one PSG and two titrations. None of the titrations were on an ASV machine. I didn't sleep very well during any of the studies. No REM sleep observed during my 2nd titration. My new appointment is less than a month away.

In my view another PSG may be a waste of time and money. Before your appointment, send or drop off with the new doc the full reports from the PSG and the titrations and let him/her know you will be asking for an ASV titration. (You should have or should obtain copies of the full reports for all your overnight studies, including all data. You have a right to copies of these for your own records.)

Before approving coverage for an ASV machine insurance companies usually insist on an ASV Titration to verify ASV therapy is successful in treating your complex sleep apnea.

Also, an ASV Titration may produce helpful info and guidance for what settings to use.

ASV (like APAP) can be auto titrating but, in my view, both work best when the lower end of the range for EPAP is not much lower than whatever EPAP you need to prevent your obstructive events, and for the Max Pressure Support setting to be around 10 (cmH2O), or maybe 8 or lower to start and later increase to 9 or higher as you get used to ASV therapy.

Good luck,
--- Vaughn

Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment. The Advisory Member group provides advice and suggestions to Apnea Board administrators and staff on matters concerning Apnea Board operation and administrative policies - not on matters concerning treatment for Sleep Apnea. I think it is now too late to change the name of the group but I think Voting Member group would perhaps have been a more descriptive name for the group.
(This post was last modified: 09-10-2015 01:58 AM by vsheline.)
09-10-2015 01:54 AM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
Exclaimation CPAP Machine Choices - read this before you accept a new machine SuperSleeper 2 6,010 10-12-2016 11:37 AM
Last Post: DariaVader
  Leak fixed (mouth breather for sure) jxyzobrien 13 778 10-02-2016 09:36 AM
Last Post: SuperSleeper
Question [CPAP] Fixed Pressure vs. Auto Pros and Cons gmapes 15 930 09-03-2016 06:04 PM
Last Post: zonk
  Philips Dreamstation auto or fixed stephennic 20 1,786 08-31-2016 01:33 PM
Last Post: Maskup Sleepwell