[Equipment] Problems starting REMstar Pro C-Flex +
#1
Question 
I’ve been using a REMstar Pro C-Flex + with humidifier for about three years. Recently I started to encounter a problem with the device: it would not start immediately, but would rather produce a beep, then after short delay – an attempt to blow some air, then it would cut blowing abruptly and the whole cycle would start over and over again, sometimes for about 5 minutes or more, then finally it would start working and would go normally for the whole night.
I tried to disconnect the humidifier, remove the card, disconnect the power supply, deactivate/reactivate auto on – nothing helps.
Maybe someone had the same experience – any ideas, how to overcome the problem?

Grateful Big Alex

Post Reply Post Reply
#2
GBA:

Sounds like you tried all the standard troubleshooting things.
I'd ask one question -- how's the intake filter?

Likely needs either a professional repair or replacement.

Gruß

Mongo
INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#3
Hi AlexSob,
WELCOME! to the forum.!
Hang in there for more suggestions and best of luck to you.
trish6hundred
Post Reply Post Reply


#4
(08-31-2014, 01:58 PM)justMongo Wrote: GBA:

Sounds like you tried all the standard troubleshooting things.
I'd ask one question -- how's the intake filter?

Likely needs either a professional repair or replacement.

Gruß

Mongo

Thanks Mongo, you are not just the first to respond - you are the first one to RESPOND TO A MESSAGE I'VE EVER PLACED ON THE WEB!

In practical terms: if 'intake filter' is a small black brick of foam - it is sterile clean and washed and replaced regularly.
Post Reply Post Reply
#5
Is there a message displayed on the screen when the machine beep?






Post Reply Post Reply
#6
(08-31-2014, 04:54 PM)zonk Wrote: Is there a message displayed on the screen when the machine beep?

Yes, the message that appears briefly is something like "Do not remove the card - data is being transferred". No other messages appear, including any of those listed in the troubleshoot section of the manual.
Post Reply Post Reply


#7
(08-31-2014, 02:52 PM)AlexSob Wrote: ...
In practical terms: if 'intake filter' is a small black brick of foam - it is sterile clean and washed and replaced regularly.

Hello,
what about the ultra fine filter? Regularly changed? Tryed to start without both filters just for testing?

sincerely
Fat Rat

Post Reply Post Reply
#8
(09-01-2014, 03:32 AM)Fat Rat Wrote:
(08-31-2014, 02:52 PM)AlexSob Wrote: ...
In practical terms: if 'intake filter' is a small black brick of foam - it is sterile clean and washed and replaced regularly.

Hello,
what about the ultra fine filter? Regularly changed? Tryed to start without both filters just for testing?

sincerely
Fat Rat

I don't use ultrafine, and the same picture is without regular filer - the machine would start only after numerous cycles described by me in the original post: beep, ramp LED blinks, message "do not remove... " briefly appears, then a very small puff of air and then to the beginning of the cycle. If I am lucky, the normal air blow would start after about 5 to 10 minutes repetitive start failures.
Post Reply Post Reply
#9
(09-01-2014, 03:32 AM)Fat Rat Wrote:
(08-31-2014, 02:52 PM)AlexSob Wrote: ...
In practical terms: if 'intake filter' is a small black brick of foam - it is sterile clean and washed and replaced regularly.

Hello,
what about the ultra fine filter? Regularly changed? Tryed to start without both filters just for testing?

sincerely
Fat Rat
Hi Fat Rat,
WELCOME! to the forum.!
trish6hundred
Post Reply Post Reply


#10
(09-01-2014, 10:33 PM)trish6hundred Wrote: Hi Fat Rat,
WELCOME! to the forum.!

THANKYOU!

Fat Rat oldman
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Bipap AVAPS problems samianwardcummings 10 365 03-24-2017, 07:12 AM
Last Post: Sleeprider
  Problems with AirFit P10 Nasal Pillows (+ResMed AirSense 10) numari 18 7,040 03-23-2017, 11:36 AM
Last Post: PaytonA
Angry So I just had words with my DME @ Pacific pulmonary problems MzRobin28 4 197 03-08-2017, 10:40 PM
Last Post: MzRobin28
  Vertigo/CPAP Pressure Update- CPAP causing inner ear problems SuperSleeper 70 67,591 03-05-2017, 08:26 PM
Last Post: Melman
  centrals after starting cpap Leemax 3 179 03-01-2017, 02:38 PM
Last Post: srlevine1
  nuance to Swift FX for her setting problems akdog30 11 369 02-11-2017, 11:18 AM
Last Post: Ghost1958
Gross [CPAP] Respironics REMStar System One Series 60 Repair lempface 6 380 02-04-2017, 07:01 AM
Last Post: cate1898

Forum Jump:

New Posts   Today's Posts




About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.

For any more information, please use our contact form.