[Equipment] Problems starting REMstar Pro C-Flex +
#1
Question 
I’ve been using a REMstar Pro C-Flex + with humidifier for about three years. Recently I started to encounter a problem with the device: it would not start immediately, but would rather produce a beep, then after short delay – an attempt to blow some air, then it would cut blowing abruptly and the whole cycle would start over and over again, sometimes for about 5 minutes or more, then finally it would start working and would go normally for the whole night.
I tried to disconnect the humidifier, remove the card, disconnect the power supply, deactivate/reactivate auto on – nothing helps.
Maybe someone had the same experience – any ideas, how to overcome the problem?

Grateful Big Alex

Post Reply Post Reply
#2
GBA:

Sounds like you tried all the standard troubleshooting things.
I'd ask one question -- how's the intake filter?

Likely needs either a professional repair or replacement.

Gruß

Mongo
Post Reply Post Reply
#3
Hi AlexSob,
WELCOME! to the forum.!
Hang in there for more suggestions and best of luck to you.
trish6hundred
Post Reply Post Reply


#4
(08-31-2014, 01:58 PM)justMongo Wrote: GBA:

Sounds like you tried all the standard troubleshooting things.
I'd ask one question -- how's the intake filter?

Likely needs either a professional repair or replacement.

Gruß

Mongo

Thanks Mongo, you are not just the first to respond - you are the first one to RESPOND TO A MESSAGE I'VE EVER PLACED ON THE WEB!

In practical terms: if 'intake filter' is a small black brick of foam - it is sterile clean and washed and replaced regularly.
Post Reply Post Reply
#5
Is there a message displayed on the screen when the machine beep?






Post Reply Post Reply
#6
(08-31-2014, 04:54 PM)zonk Wrote: Is there a message displayed on the screen when the machine beep?

Yes, the message that appears briefly is something like "Do not remove the card - data is being transferred". No other messages appear, including any of those listed in the troubleshoot section of the manual.
Post Reply Post Reply


#7
(08-31-2014, 02:52 PM)AlexSob Wrote: ...
In practical terms: if 'intake filter' is a small black brick of foam - it is sterile clean and washed and replaced regularly.

Hello,
what about the ultra fine filter? Regularly changed? Tryed to start without both filters just for testing?

sincerely
Fat Rat

Post Reply Post Reply
#8
(09-01-2014, 03:32 AM)Fat Rat Wrote:
(08-31-2014, 02:52 PM)AlexSob Wrote: ...
In practical terms: if 'intake filter' is a small black brick of foam - it is sterile clean and washed and replaced regularly.

Hello,
what about the ultra fine filter? Regularly changed? Tryed to start without both filters just for testing?

sincerely
Fat Rat

I don't use ultrafine, and the same picture is without regular filer - the machine would start only after numerous cycles described by me in the original post: beep, ramp LED blinks, message "do not remove... " briefly appears, then a very small puff of air and then to the beginning of the cycle. If I am lucky, the normal air blow would start after about 5 to 10 minutes repetitive start failures.
Post Reply Post Reply
#9
(09-01-2014, 03:32 AM)Fat Rat Wrote:
(08-31-2014, 02:52 PM)AlexSob Wrote: ...
In practical terms: if 'intake filter' is a small black brick of foam - it is sterile clean and washed and replaced regularly.

Hello,
what about the ultra fine filter? Regularly changed? Tryed to start without both filters just for testing?

sincerely
Fat Rat
Hi Fat Rat,
WELCOME! to the forum.!
trish6hundred
Post Reply Post Reply


#10
(09-01-2014, 10:33 PM)trish6hundred Wrote: Hi Fat Rat,
WELCOME! to the forum.!

THANKYOU!

Fat Rat oldman
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  [Equipment] Respironics Remstar auto A- flex humidifier stopped working Reddog 4 103 09-19-2017, 09:29 AM
Last Post: Sleeprider
  Starting [CPAP] Tonight RustySocket 27 683 09-15-2017, 07:01 PM
Last Post: RustySocket
  Back sleeper [causing problems] AngeliqueFox 21 520 09-04-2017, 08:00 PM
Last Post: Beej
  Popping sound when starting machine yankees123 11 344 09-01-2017, 08:20 PM
Last Post: nathanoj
Question [Equipment] trouble moving from RemStar Auto to AirSense 10 jim.evo 53 1,498 08-31-2017, 04:39 PM
Last Post: jim.evo
  [Admin Note] WANTED: CLINICIAN SETUP MANUAL for 60-series PR System One REMstar Plus w/ C-Flex SuperSleeper 4 300 08-27-2017, 09:06 PM
Last Post: Galactus
  Hello - help with Remstar Pro AutoIQ Margali 6 242 08-26-2017, 10:02 PM
Last Post: trish6hundred

Forum Jump:

New Posts   Today's Posts




About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.

For any more information, please use our contact form.