[Equipment] Rapidly Cycling Bilevel Pressure
#1
This morning, I woke up to find my ResMed VPAP Auto 25, Which is operating in bi-level mode, cycling between my IPAP and EPAP very rapidly. The cycle was taking about 1 second. I tried slowing my breathing to see if it would sense it and even tried holding my breath. It just maintained its rapid cycling and also did not score an apnea. This is way faster than my Ti max which is set at 2.0.

Does anyone know what might be going on or what might be causing this?

Best Regards,

PaytonA

Admin Note:
PaytonA passed away in September 2017
Click HERE to read his Memorial Thread

~ Rest in Peace ~
Post Reply Post Reply
#2
Sounds to me like a call to the DME is in order.
Post Reply Post Reply
#3
I'd reboot (power off, wait 10 seconds, power on) in that situation -- should not have a 30-60 respiratory rate unless trying to force hyperventilation.
Post Reply Post Reply


#4
Ti max is the max time in IPAP.
Ti min, which is often of the order of 300 milliseconds, is the minimum time it stays in IPAP.

Your machine can trigger as fast as about 3.33 times per second.
It may be a defect in the machine.
It may be your breathing pattern.
It may be the sensitivity of the trigger or cycle which is are often defaulted to MED.

I find my VPAP, that I run in S mode (Spontaneous) double triggers during a portion of my sleep.
I can see in the flow data why is would trigger. A slight positive flow in the middle of expiration triggers it back to IPAP during expiration.
My resp rate is usually around 15 per minute.
During this double trigger, my resp rate is recorded as about 30 per minute.

Admin Note:
JustMongo passed away in August 2017
Click HERE to read his Memorial Thread

~ Rest in Peace ~
Post Reply Post Reply
#5
I am trying the Bill Gates Solution (Reboot!!!)

Admin Note:
PaytonA passed away in September 2017
Click HERE to read his Memorial Thread

~ Rest in Peace ~
Post Reply Post Reply
#6
On power cycling, I prefer to pull the AC line cord, not the DC plug in the back of the machine.
The supply takes a while to drain the capacitors -- I have found 10 seconds too short.
More like minutes!

(Or pull the AC first, then the DC.)

Admin Note:
JustMongo passed away in August 2017
Click HERE to read his Memorial Thread

~ Rest in Peace ~
Post Reply Post Reply


#7
(03-26-2014, 09:24 PM)justMongo Wrote: On power cycling, I prefer to pull the AC line cord, not the DC plug in the back of the machine.
The supply takes a while to drain the capacitors -- I have found 10 seconds too short.
More like minutes!

(Or pull the AC first, then the DC.)

My S8 apparently has the transformer internal to the machine. I did disconnect the power and left it that way for several hours. Last night-no problem. Thanks for the input on draining the capacitors. I will keep that in mind.

Best Regards,

PaytonA

Admin Note:
PaytonA passed away in September 2017
Click HERE to read his Memorial Thread

~ Rest in Peace ~
Post Reply Post Reply
#8
(03-27-2014, 10:39 AM)PaytonA Wrote: My S8 apparently has the transformer internal to the machine. I did disconnect the power and left it that way for several hours. Last night-no problem. Thanks for the input on draining the capacitors. I will keep that in mind.

Best Regards,

PaytonA

Ah,yes. I missed your machine in your profile.
The VPAP Auto 25 is indeed an S8, and the power supply is internal to the machine.

Your machine is the S8 version of my S9 VPAP Auto.

Admin Note:
JustMongo passed away in August 2017
Click HERE to read his Memorial Thread

~ Rest in Peace ~
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Isn't Resmed's Expiratory Pressure Relief really Inspiratory Pressure Support?? Reznik 53 874 1 hour ago
Last Post: yrnkrn
  Resmed BiLevel S9 vs Aircurve VAuto holden4th 1 88 11-15-2017, 09:07 AM
Last Post: Sleeprider
  With a face mask pressure is about 8.5. With a nasal pillow pressure was 20 at one po brojohn1611 10 458 11-04-2017, 03:05 AM
Last Post: James2017
  Autoset 90 percentile pressure vs. Titration study pressure CDNHoser 12 373 10-19-2017, 09:02 PM
Last Post: Sleepster
  Pressure at altitude vs. pressure at lower elevations...? S.L. Ping Beauty 19 415 10-17-2017, 09:36 PM
Last Post: PaulaO2
  [Equipment] Viscious cycle of increased pressure -> more leak -> response of more pressure 3mp0w3r 5 281 09-16-2017, 09:04 AM
Last Post: OpalRose
  Sleepyhead records and borrowed (trial) Bilevel? Spy Car 4 155 08-24-2017, 06:39 PM
Last Post: Spy Car

Forum Jump:

New Posts   Today's Posts




About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.

For any more information, please use our contact form.