Resmed S9 Filters: Standard versus Hypo-allergenic
On page 27 of the S9 Clinician Manual two types of filters are listed for the Resmed S9 series of machines:

Standard (ASMB 160) filters with an efficiency of 88% at 7 micron and
Hypo-allergenic (Air Safety Electret100 – electrostatic filter) with an efficiency of 89.8% at 0.5 micron.

0.5 micron is quite different from 7 micron but 89.8% is not much different from 88%, so I am not sure of the performance difference.

Does anyone know? Thanks.
10-08-2013 09:00 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Kritiker Offline

Preferred Members

Posts: 27
Joined: Sep 2013

Machine: ResMed S9 Autoset
Mask Type: Nasal mask
Mask Make & Model: ResMed Mirage Activa LT
Humidifier: ResMed S9 H5i
CPAP Pressure: 10-20
CPAP Software: ResScan

Other Comments:

Sex: Undisclosed
Location:

Post: #2
RE: Resmed S9 Filters: Standard versus Hypo-allergenic
As far as I can determine, a filter's efficiency refers to its ability to remove particulate material.

It appears that the removal of smaller particles, in the 7.0 micron down to to 0.5 micron range only provides for an improvement from 88% to 89.9%. The S9 VPAP Clinician Manual also lists a bacterial efficiency of 99.54% for the hypo-allergenic filters.

Is that significant in terms of the amount of particulate matter that is removed?
Is that significant in terms of the size of the particulate matter that is removed?

I also understand that the purpose of these filters is two-fold, first to (potentially) extend the lifespan of the xPAP unit and (perhaps only) second to provide the user with filtered air. Perhaps the standard filters help most with the former and the hypo-allergenic filters help with both?

Some people are reporting that the hypo-allergenic filters do get dirty faster than the standard ones, so one could infer that they are working better. Some people are also referring to the dust accumulation inside a computer case and suggesting these filters keep that kind of dust out of the xPAPs and out of our airways.

So I would assume that the finer filters (irrespective of any implied hypo-allergenic properties) would keep out more crud than the standard filters and that sounds good to me.

I also apologize for starting yet another filter thread. I don't want to restart that whole debate. I am just curious about the published numbers and what they mean.
(This post was last modified: 10-08-2013 10:46 AM by Kritiker.)
10-08-2013 10:19 AM
Find all posts by this user Post Reply Quote this message in a reply
DOOM_NX Offline

Preferred Members-2

Posts: 155
Joined: May 2013

Machine: ResMed S9 AutoSet
Mask Type: Nasal mask
Mask Make & Model: Weinmann JOYCE SilkGel
Humidifier: ResMed H5i + ClimateLine (not using year-round)
CPAP Pressure: 5–16 cmH2O
CPAP Software: ResScan SleepyHead

Other Comments: 12/2011: Mild SAS (AHI 14.5) - 10/2013: Severe SAS (AHI 78.7) - Using AutoCPAP: AHI 1.4

Sex: Male
Location: Greece

Post: #3
RE: Resmed S9 Filters: Standard versus Hypo-allergenic
Standard filter: cheaper, less effective for smaller particulates (pollen, smoke, fine dust), easier on the motor.
Hypoallergenic: more expensive, more effective, clogs up faster, harder on the motor.

All air filters become more effective as they accumulate particles, although flow resistance is increased.

If you suffer from allergic rhinitis or sleep in a dusty area, then get the hypoallergenic ones and change them more often.
Otherwise, save your money and get the standard ones.
(This post was last modified: 10-08-2013 10:56 AM by DOOM_NX.)
10-08-2013 10:54 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
terp1984 Offline

Preferred Members

Posts: 45
Joined: Apr 2012

Machine: ResMed S9 VPAP Adapt
Mask Type: Full face mask
Mask Make & Model: ResMed Quattro FX
Humidifier: ResMed S9 H5i
CPAP Pressure: EPAP 5 to 15, IPAP 10 to 35
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: South Carolina

Post: #4
RE: Resmed S9 Filters: Standard versus Hypo-allergenic
I use ResMed DSP1011 Hypoallergenic filters. I don't use my insurance because, the in-network provider does not have this type. Even though they have what is claimed to be Hypoallergenic, they are not as good. I have dust allergies and it makes a big difference for me.
10-08-2013 01:43 PM
Find all posts by this user Post Reply Quote this message in a reply
Bobshouse Offline

Preferred Members

Posts: 58
Joined: Aug 2012

Machine: Resmed S9 Auto
Mask Type: Nasal pillows
Mask Make & Model: Tap Pap or The Breeze
Humidifier: Resmed H5I
CPAP Pressure: 10 Auto to 20
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Central California

Post: #5
RE: Resmed S9 Filters: Standard versus Hypo-allergenic
7 micron at 88%

.05 micron at 89.8%

Lets raise that to inches....at 7 inches and 88% it would stop about 88% of softballs thrown at it.
at .05 inches and 89.8% it would stop all the softballs thrown at it, plus 89.8% of everything from the size of a large gumball and smaller.

Questions?
(This post was last modified: 10-08-2013 02:31 PM by Bobshouse.)
10-08-2013 02:28 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
RonWessels Offline

Preferred Members-2

Posts: 465
Joined: Jun 2013

Machine: REMstar Auto AFlex DS560TS
Mask Type: Nasal mask
Mask Make & Model: ComfortGel Blue
Humidifier: REMstar Heated Tube DS6T
CPAP Pressure: 11 - 20
CPAP Software: SleepyHead

Other Comments: CMS50F Pulse Oximeter

Sex: Male
Location: Ontario, Canada

Post: #6
RE: Resmed S9 Filters: Standard versus Hypo-allergenic
Wow, you have small gumballs! 0.05 inches is around 1.25 mm. I would need tweezers to pick that up.
10-08-2013 06:23 PM
Find all posts by this user Post Reply Quote this message in a reply
Bobshouse Offline

Preferred Members

Posts: 58
Joined: Aug 2012

Machine: Resmed S9 Auto
Mask Type: Nasal pillows
Mask Make & Model: Tap Pap or The Breeze
Humidifier: Resmed H5I
CPAP Pressure: 10 Auto to 20
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Central California

Post: #7
RE: Resmed S9 Filters: Standard versus Hypo-allergenic
(10-08-2013 06:23 PM)RonWessels Wrote:  Wow, you have small gumballs! 0.05 inches is around 1.25 mm. I would need tweezers to pick that up.

Ah, good, a question! Tweezers you say! I have small gumballs. You would need tweezers to pick up my small gumballs.

OKaaayyyy.....

Fuggettittabbouuttttit. Nobody is handling my small gumballs with tweezers! Oh-jeez I was just trying to put the filter and filtering capabablities into perspective. Eat-popcorn
10-08-2013 07:20 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
justMongo Offline

Monitors

Posts: 3,857
Joined: Sep 2013

Machine: ResMed VPAP Auto (S9)
Mask Type: Full face mask
Mask Make & Model: ResMed Mirage Quattro
Humidifier: ResMed H5i
CPAP Pressure: 18 IPAPmax - 11 EPAPmin, PS=5
CPAP Software: Other Software

Other Comments: μολὼν λαβέ

Sex: Male
Location: Калифорния

Post: #8
RE: Resmed S9 Filters: Standard versus Hypo-allergenic
(10-08-2013 02:28 PM)Bobshouse Wrote:  7 micron at 88%

.05 micron at 89.8%

Lets raise that to inches....at 7 inches and 88% it would stop about 88% of softballs thrown at it.
at .05 inches and 89.8% it would stop all the softballs thrown at it, plus 89.8% of everything from the size of a large gumball and smaller.

Questions?

You slipped a decimal point. It's 0.5 micron. ( one half micron)
Actually 0.5 micron is quite small -- the wavelength of green light is around 0.5 micron.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
10-09-2013 10:01 AM
Find all posts by this user Post Reply Quote this message in a reply
Kritiker Offline

Preferred Members

Posts: 27
Joined: Sep 2013

Machine: ResMed S9 Autoset
Mask Type: Nasal mask
Mask Make & Model: ResMed Mirage Activa LT
Humidifier: ResMed S9 H5i
CPAP Pressure: 10-20
CPAP Software: ResScan

Other Comments:

Sex: Undisclosed
Location:

Post: #9
RE: Resmed S9 Filters: Standard versus Hypo-allergenic
Thanks to all.

Bobshouse, your analogy got me to thinking about this a little differently. The more I looked into it, the more complicated it got. The following is what I learned so far. It is still a gross over-simplification and probably still contains errors but I haven't time to go any further just yet.

Efficiency is the ratio of some desired or useful result divided by the total possible result.

Efficiency = amount of useful work / total possible work

Filter efficiency is the ratio of particles, of the specified size (or possibly sometimes larger), trapped by a filter divided by the total number of particles, of the specified size (or larger?), found in the air upstream of the filter.

A micron (μm, or micrometre or micrometer) is one millionth of a metre (meter). There are 25,400 microns to an inch.

We are not given any indication of which testing standard was used to obtain the ResMed filter efficiency values nor exactly what the given specifications mean. We don't for example, know whether the ratings apply to 7 and 0.5 micron mean diameter particles (which includes smaller and larger) or to particles of those diameters and larger, nor what mix of particle sizes.

HEPA filters are all rated for particles of 0.3 micron mean diameter. I have seen this expressed (on Wikipedia, incorrectly perhaps?) as particles of mean diameter 0.3 micron or larger.
I am unsure which applies to the ResMed specifications.

A filter's efficiency of 88% at 7 microns should probably mean that it will stop 88% of particles 7 microns mean diameter but it could mean that it will stop 88% of particles of 7 microns or larger, a looser specification;
A filter's efficiency of 89.9% at 0.5 micron should probably mean that it will stop 89.9% of particles .5 micron mean diameter or it could mean that it will stop particles of 89.9% of particles 0.5 micron or larger.

Either way, this represents a significant difference between the two filters.

What follows is based on the (looser) specification of particle size, i.e., mean diameter or larger. I am not sure how to do the same for the tighter definition and I haven't the time to pursue it right now. I offer what follows for what it is worth and I am not sure what that is.

I calculate efficiencies of 0.0 for both filters for, i.e, neither filter catches: viruses, pesticides and coal flue gases.

The following particles seem to be caught, to varying degrees, by the finer filter (efficiency is given) and not the standard one (efficiency is 0.0).

0.5μm 7μm Particle
89.9% 0.0% liquid droplets
89.9% 0.0% anthrax
89.9% 0.0% Lead dust
89.9% 0.0% Spider web
82.6% 0.0% Paint pigments
78.9% 0.0% Tobacco smoke
71.9% 0.0% various combustion byproducts
46.3% 0.0% Oil smoke
41.3% 0.0% Smoldering or flaming cooking oil

The following particles are caught significantly better (more than three times as many) by the finer filter than the standard one. Both efficiencies are given.

0.5μm 7μm Particle
89.9% 27.8% Insecticide dust
86.3% 26.7% Radioactive fallout

The following particles are caught better by the finer filter than the standard one. Both efficiencies are given.

0.5μm 7μm Particle
89.9% 48.9% Mold
89.9% 52.8% Red blood cells
89.9% 66.0% Antiperspirant
88.7% 67.7% Face powder
89.9% 71.5% Iron dust
88.8% 72.6% Atmospheric dust
89.9% 76.4% Talcum dust
89.9% 77.2% Yeast cells
89.6% 78.1% Bacteria
89.9% 78.5% Spores

The following particles are caught marginally better by the finer filter than the standard one. Both efficiencies are given.

0.5μm 7μm Particle
89.9% 81.7% Textile dust
89.9% 81.8% Asbestos
89.9% 82.7% Coal dust
89.9% 84.4% Cement dust
89.9% 87.4% Metallurgical fumes & dust
89.9% 88.0% Mold spores
89.9% 88.0% Pollen
89.9% 88.0% Dust mites
89.9% 88.0% Beach sand
89.9% 88.0% Glass wool

To test a filter, particle counters measure the size and quantity of upstream particles per known volume of fluid, as well as the size and quantity of particles downstream of the filter. The number of particles trapped by the filter can then be computed as Particles Upstream - Particles Downstream.

Filter Efficiency = Particles Trapped / Particles Upstream = (Particles Upstream - Particles Downstream) / Particles Upstream

I have had to make some assumptions and ignore some variables:
  • I have assumed here that the efficiencies refer to particles of the specified diameter and larger. The results are not applicable if the efficiencies refer to mean particle diameters.
  • I have had to ignore the effects of actual operating conditions, e.g., flow surges and changes in temperature.
  • I had to assume that the particle sizes are uniformly distributed throughout the entire size range. This is not realistic.
  • I had to assume that a filter's efficiency is constant for all particle sizes. This is unrealistic. It seem reasonable that a filter should do better at stopping larger particles than smaller ones.
  • I had to assume that a filter's efficiency is constant for all particle types. This is unrealistic. It seem reasonable that a filter might do better at stopping some particles than others, e.g., due to shape, stickiness.
As well, a filter's efficiency does not indicate its dirt-holding capacity, the total amount of contaminant that can be trapped by the filter throughout its life, nor does it indicate its stability or performance over time.

There may be other, implied, assumptions I haven't even discovered yet.

Having calculated all this, I am now left wondering whether it is of interest or use to anyone or was just an academic exercise.

Particle Diameter (microns)
min max Particle

0.5 5 Liquid droplets
1 5 Anthrax
2 2 Lead dust
2 3 Spider web
0.1 5 Paint pigments
0.01 4 Tobacco smoke
0 2.5 various combustion byproducts
0.03 1 Oil smoke
0.03 0.9 Smoldering or flaming cooking oil
0.5 10 Insecticide dust
0.1 10 Radioactive fallout
3 12 Mold
5 10 Red blood cells
6 10 Antiperspirant
0.1 30 Face powder
4 20 Iron dust
0.001 40 Atmospheric dust
0.5 50 Talcum dust
1 50 Yeast cells
0.3 60 Bacteria
3 40 Spores
6 20 Textile dust
0.7 90 Asbestos
1 100 Coal dust
3 100 Cement dust
0.1 1000 Metallurgical fumes & dust
10 30 Mold spores
10 1000 Pollen
100 300 Dust mites
100 10000 Beach sand
1000 1000 Glass wool
0.08 0.2 Coal flue gas
0.005 0.3 Viruses
0.001 0.001 Pesticides & herbicides

Sources
Filter Arrestance and Efficiency
Filter efficiency definition
Particle Sizes
(This post was last modified: 10-10-2013 12:43 PM by Kritiker.)
10-10-2013 12:38 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
hobbs Offline

Members

Posts: 12
Joined: Sep 2013

Machine: S8 AutoSet II
Mask Type: Nasal mask
Mask Make & Model: Activa LT
Humidifier: F&P 150
CPAP Pressure: 10.8-12.0
CPAP Software: ResScan

Other Comments:

Sex: Male
Location: SoCal

Post: #10
RE: Resmed S9 Filters: Standard versus Hypo-allergenic
I would like to find a source for REAL OEM Resmed filters, not the aftermarket ones most sellers have. Please send a PM if you have a trusted source for OEM filters.
10-10-2013 05:49 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  Filters for S9 Regular Joe 11 5,608 11-15-2016 10:21 AM
Last Post: Sleeprider
  Resmed Inlet Filters PaytonA 3 272 10-30-2016 05:30 PM
Last Post: FrankNichols
  Resmed Airsense 10 CPAP data versus Resmed Airsense 10 Autoset data David P. 4 557 10-29-2016 08:42 PM
Last Post: trish6hundred
  Sleep Center versus At Home Stats richb 13 669 08-09-2016 01:03 AM
Last Post: chill
  AirFit F10 versus Quattro Air spc23 11 9,466 06-04-2016 10:34 PM
Last Post: tonywr
  HME (Heat moisture exchange filters) hozholla 3 651 05-13-2016 03:26 PM
Last Post: bill-e
  Reinstall the filters aergun 4 476 04-17-2016 08:40 AM
Last Post: OpalRose

Forum Jump:

Who's Online (Complete List)