[Equipment] "Service Required" PR System One 60 Series REMstar Auto
Author Message
Sleepster Offline
Wiki Editor
Moderators

Posts: 4,988
Joined: Feb 2012

Machine: ResMed AirCurve10 VAuto
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: HumidAir and SlimLine Hose
CPAP Pressure: MaxI 13.6 | MinE 5.2 | PS 4.4
CPAP Software: ResScan SleepyHead

Other Comments: Diagnosed Nov 2011. Conquered aerophagia.

Sex: Male
Location: Houston, Texas

Post: #11
RE: "Service Required" PR System One 60 Series REMstar Auto
(03-14-2013 09:55 PM)fkimura Wrote:  I already did that and the maintenance guy told the voltage is OK in my room. I also tried other wall socket, but it had the same effect.

But if you're using the same power brick for both machines, it could very well be the power brick. In fact, don't the models with the heated hose require a different power brick?

Sleepster
Apnea Board Moderator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
03-14-2013 11:18 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
zonk Offline

Advisory Members

Posts: 7,908
Joined: Feb 2012

Machine: A10 AutoSet
Mask Type: Nasal mask
Mask Make & Model: Activa LT
Humidifier: Integrated /ClimateLineAir
CPAP Pressure: 9/13
CPAP Software: ResScan

Other Comments: CPAP since Nov 2010

Sex: Male
Location: Australia

Post: #12
RE: "Service Required" PR System One 60 Series REMstar Auto
(03-14-2013 11:18 PM)Sleepster Wrote:  In fact, don't the models with the heated hose require a different power brick?
Check the power cord
The heated hose humidifier require 80 watt power cord and don,t work with 60 watt power cord
60 watt power cord works only with the basic humidifier

The machine works with either humidifier as long as the right power cord selected
03-15-2013 12:54 AM
Find all posts by this user Post Reply Quote this message in a reply
archangle Offline
Wiki Editor
Advisory Members

Posts: 3,159
Joined: Feb 2012

Machine: ResMed S9 AutoSet
Mask Type: Nasal pillows
Mask Make & Model: ResMed Swift FX
Humidifier: ResMed S9 H5i
CPAP Pressure: 16-20
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments: Happy PAPper

Sex: Undisclosed
Location: USA

Post: #13
RE: "Service Required" PR System One 60 Series REMstar Auto
Do you have two humidifiers?

Try it without the humidifier.

Do you have a heated hose? Try without the heated hose.

Will it give you the error message if you turn it off after running for just a few minutes?

If the power supply is sitting on something soft, like carpet, set it on a hard surface like a tabletop where there's plenty of airflow.

Get the free SleepyHead software here.
Useful links.
Click here for information on the main alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check it yourself.
03-15-2013 03:40 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
fkimura Offline

Members

Posts: 8
Joined: Mar 2013

Machine: PR DS560HS
Mask Type: Nasal mask
Mask Make & Model: Confort Gel
Humidifier: Respironics System One
CPAP Pressure: 16
CPAP Software: Not using software

Other Comments:

Sex: Male
Location:

Post: #14
RE: "Service Required" PR System One 60 Series REMstar Auto
Hi everyone, thanks for the help. It was indeed the power brick.
The 1st machine gave this error, then I bought a 2nd one, but used the 1st power brick, and the 2nd machine gave the same message.
Then I tried the 2nd power brick with both machines, and voila! Both machines worked fine!
The only "problem" now is that I am stuck with 1 machine and 1 broken power brick, and no way to buy another power brick here in Brazil, but at least I have a working CPAP machine!
Thanks for the help, everyone!
03-22-2013 11:17 AM
Find all posts by this user Post Reply Quote this message in a reply
Shastzi Offline

Advisory Members

Posts: 1,174
Joined: Dec 2012

Machine: ResMed S9 Autoset
Mask Type: Full face mask
Mask Make & Model: FitLife Total face mask
Humidifier: F&P HC150 with Hybernite heated hose.
CPAP Pressure: 15cm-20cm H2O (auto)
CPAP Software: SleepyHead Other Software

Other Comments: CMS50-F wearable Oximeter; Software: SPO2 Assistant

Sex: Undisclosed
Location: Florida, USA.

Post: #15
RE: "Service Required" PR System One 60 Series REMstar Auto
Glad you got a workable solution, fkimura!

Good luck!

Sleep-well

"With ordinary talent and extraordinary perseverance, all things are attainable." - Thomas Foxwell Buxton

Cool
03-22-2013 01:15 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Sleepster Offline
Wiki Editor
Moderators

Posts: 4,988
Joined: Feb 2012

Machine: ResMed AirCurve10 VAuto
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: HumidAir and SlimLine Hose
CPAP Pressure: MaxI 13.6 | MinE 5.2 | PS 4.4
CPAP Software: ResScan SleepyHead

Other Comments: Diagnosed Nov 2011. Conquered aerophagia.

Sex: Male
Location: Houston, Texas

Post: #16
RE: "Service Required" PR System One 60 Series REMstar Auto
(03-22-2013 11:17 AM)fkimura Wrote:  The only "problem" now is that I am stuck with 1 machine and 1 broken power brick, and no way to buy another power brick here in Brazil, but at least I have a working CPAP machine!

Well, having a working CPAP machine is a distinct advantage. Bigwink

Look closely at the two power bricks and see if the "broken" one is smaller in size, or has a lower wattage rating. If it does you can probably still use it as long as you don't have the hose heater and humidifier cranked up.

What's the voltage there in Brazil? Is it 230 V like they have in Europe or 115 V like we have in North America?

Sleepster
Apnea Board Moderator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
03-23-2013 01:20 AM
Find all posts by this user Post Reply Quote this message in a reply
fkimura Offline

Members

Posts: 8
Joined: Mar 2013

Machine: PR DS560HS
Mask Type: Nasal mask
Mask Make & Model: Confort Gel
Humidifier: Respironics System One
CPAP Pressure: 16
CPAP Software: Not using software

Other Comments:

Sex: Male
Location:

Post: #17
RE: "Service Required" PR System One 60 Series REMstar Auto
I did look when I was in USA, and they are both the same, 80w. They have the same part number #1091399.
The voltage here is pretty much the same as USA, 110V, and both bricks support 100-240V, 50/60Hz. I think the 1st power brick is really broken somehow, although it works for 6 hours without a problem, it's after 6h to 7h that the problem appears.
(This post was last modified: 03-23-2013 08:34 AM by fkimura.)
03-23-2013 08:33 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
JJJ Offline

Advisory Members

Posts: 751
Joined: Apr 2012

Machine: ResMed S9 Autoset
Mask Type: Nasal mask
Mask Make & Model: Most recent: Wisp, Pilairo, Mirage Vista
Humidifier: ResMed S9 Autoset
CPAP Pressure: 17-20
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: USA 97217

Post: #18
RE: "Service Required" PR System One 60 Series REMstar Auto
(03-23-2013 08:33 AM)fkimura Wrote:  I did look when I was in USA, and they are both the same, 80w. They have the same part number #1091399.
The voltage here is pretty much the same as USA, 110V, and both bricks support 100-240V, 50/60Hz. I think the 1st power brick is really broken somehow, although it works for 6 hours without a problem, it's after 6h to 7h that the problem appears.

It's probably heat related, most likely a bad capacitor, but it could also be a flaky connection somewhere. If you can figure out a way to keep it cold that would be a good test for my theory.

If my theory is right it is probably repairable locally, and for less than a new one would cost, especially considering the distance. I have an electronics geek friend who I swear was born with a soldering iron attached. His house has shelves full of electronic components so he can usually replace something that has failed. Bad capacitors are usually obvious because the appearance changes. Maybe you could find someone like him locally.
03-23-2013 11:12 AM
Find all posts by this user Post Reply Quote this message in a reply
givememoresleep Offline

Members

Posts: 18
Joined: Mar 2012

Machine: PR-S1
Mask Type: Nasal pillows
Mask Make & Model: Resmed Swift FX
Humidifier: pr
CPAP Pressure: auto 9-15 c-flex mode
CPAP Software: SleepyHead

Other Comments: cpap user since 1994

Sex: Male
Location:

Post: #19
RE: "Service Required" PR System One 60 Series REMstar Auto
Try and give Respironics a call, maybe they'll exchange it for you.
03-24-2013 07:01 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
fkimura Offline

Members

Posts: 8
Joined: Mar 2013

Machine: PR DS560HS
Mask Type: Nasal mask
Mask Make & Model: Confort Gel
Humidifier: Respironics System One
CPAP Pressure: 16
CPAP Software: Not using software

Other Comments:

Sex: Male
Location:

Post: #20
RE: "Service Required" PR System One 60 Series REMstar Auto
(03-24-2013 07:01 PM)givememoresleep Wrote:  Try and give Respironics a call, maybe they'll exchange it for you.

That should be easy if I was in USA...
Thanks anyway!
03-24-2013 08:40 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  REMStar SE with DOT Compliance musicman85 11 343 12-01-2016 09:08 PM
Last Post: musicman85
  philips remstar auto a-flex skylersdad 2 143 11-29-2016 05:47 PM
Last Post: trish6hundred
  PR System One: C-Check vs Auto? Ailu 8 2,970 11-29-2016 02:04 PM
Last Post: TonyJ
  Question: sound level on Resmed S9 WITHOUT humidifier, compared to Remstar 560 kowlooner 8 207 11-29-2016 10:51 AM
Last Post: kowlooner
  [Equipment] Annoying on/off beep on Respironics BIPAP S/T C series rcattell 8 337 11-06-2016 12:07 PM
Last Post: rcattell
  Philips system one noisy jpsantell 2 263 10-29-2016 10:54 AM
Last Post: robysue
  CPAP cleaning system Tommy C 97 62,435 10-20-2016 05:35 PM
Last Post: PaytonA

Forum Jump:

Who's Online (Complete List)