(08-29-2017, 03:14 PM)Sleeprider Wrote: Jim, I think I'd turn EPR back on since that is going to help with hypopnea and RERA, but you need to increase minimum pressure to compensate and hopefully tighten things up.  

I kind of kind of like 9-14 with EPR at 2 for you.  I can go into a rationale, but that's where I'm at.

Thanks! Sleeprrider. I would ask for your rationale as I'm trying to understand how all this works in addition to getting good settings. 

Thanks again,

Jim
Post Reply Post Reply
#52
I was just finishing up a rare work project...I'm mostly retired, so I kinda rushed your response.

In reviewing your recent results from two different machines we see hypopnea, a bit of flow limitation, RERA and CA.  It is all within acceptable limits and your treatment is pretty good.  The Resmed machine is pretty aggressive and moves around a lot more than the Philips, but gets good results, although your leak rate is taking a hit there.  Anyway, it appears your OA is resolved at a pressure around 7 cm, so that works for EPAP minimum pressure in the bilevel world. What we want to do is add just a bit of pressure support and take out the hypopnea and flow limitation.  Hopefully RERA takes care of itself with that. 

A minimum pressure of 9 delivers that EPAP min of 7.0 consistently and gives you the equivalent of pressure support, resulting in a final pressure of 9/7.  The machine automatically increases pressure if needed and could go to 14/12.  On second thought that is too high, but it will only go as high as needed, so you could set the maximum at 12 and it should make no difference.  That's the rationale, and it is consistent with bilevel titration protocol methods, except that the VPAP/BiPAP has higher pressure support capability. https://www.resmed.com/us/dam/documents/...lo_eng.pdf

[Image: flnJuIn.png]


[Image: fXgPQ0H.png]
Post Reply Post Reply
#53
(08-29-2017, 05:07 PM)Sleeprider Wrote: I was just finishing up a rare work project...I'm mostly retired, so I kinda rushed your response.

In reviewing your recent results from two different machines we see hypopnea, a bit of flow limitation, RERA and CA.  It is all within acceptable limits and your treatment is pretty good.  The Resmed machine is pretty aggressive and moves around a lot more than the Philips, but gets good results, although your leak rate is taking a hit there.  Anyway, it appears your OA is resolved at a pressure around 7 cm, so that works for EPAP minimum pressure in the bilevel world. What we want to do is add just a bit of pressure support and take out the hypopnea and flow limitation.  Hopefully RERA takes care of itself with that. 

A minimum pressure of 9 delivers that EPAP min of 7.0 consistently and gives you the equivalent of pressure support, resulting in a final pressure of 9/7.  The machine automatically increases pressure if needed and could go to 14/12.  On second thought that is too high, but it will only go as high as needed, so you could set the maximum at 12 and it should make no difference.  That's the rationale, and it is consistent with bilevel titration protocol methods, except that the VPAP/BiPAP has higher pressure support capability. https://www.resmed.com/us/dam/documents/...lo_eng.pdf
[quote pid='220777' dateline='1504044429']

thanks again Sleeprider... I'll set ERP @ 2 and levels 9-12. Will post back in a couple of days.

Thanks again to everyone here, this is a great resource.

Jim
[/quote]
Post Reply Post Reply


#54
Last 2 nights w/ the new settings (9-14 & EPR 2) seemed to work well (log's attached) except for what now appears to be a leak. My mouth was opening at least once last night. What puzzles me is that this has not been an issue for me for years with the RemStar. 

I do have some old chin straps and will try them but would not like to have to use them... seems my mouth breathing is related to the AirSense unit somehow. BTW, the new settings have done away with the pulsing.

Thanks in advance!

Jim

[Image: HR78n6r.png]

[Image: AmxzbON.png]
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Philips Respironics REMstar Auto 560P - Service Required kyley 3 70 Yesterday, 11:01 PM
Last Post: trish6hundred
  I changed my resmed AirSense 10 Autoset to auto jasonarmstrong 6 223 09-20-2017, 12:29 PM
Last Post: jasonarmstrong
  [Equipment] Respironics Remstar auto A- flex humidifier stopped working Reddog 4 104 09-19-2017, 09:29 AM
Last Post: Sleeprider
  S9 Auto v AirSense 10 subyroo 10 351 09-13-2017, 07:23 PM
Last Post: subyroo
Question [CPAP] Trouble shooting resmed 100 Sleep1ngb3auty16 0 83 09-12-2017, 06:30 PM
Last Post: Sleep1ngb3auty16
  Trouble keeping my mouth closed Kryogen 14 379 09-10-2017, 10:05 PM
Last Post: JesseLee
  [Admin Note] WANTED: CLINICIAN SETUP MANUAL for 60-series PR System One REMstar Plus w/ C-Flex SuperSleeper 4 304 08-27-2017, 09:06 PM
Last Post: Galactus

Forum Jump:

New Posts   Today's Posts




About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.

For any more information, please use our contact form.