Hello Guest, Welcome to Apnea Board !
As a guest, you are limited to certain areas of the board and there are some features you can't use.
To post a message, you must create a free account using a valid email address.

or Create an Account


New Posts   Today's Posts

Flex settings and hypopneas
#1
Flex settings and hypopneas
How does a Dreamstation flex setting affect hypopneas?

At constant and 7cm my AHI numbers had been rising from 1.4 +/- to 3.5 +/- over the last 4-6 months. It was time for a change in settings, I'd been on that setting for > 10 years.

I moved from fixed pressure (7) to auto (7-9) about 2-3 weeks ago. Flex was ON and set to 1. From the Sleepyhead graphs that flex setting corresponded to a 2cm pressure relief on exhale. With those settings my AHI numbers rose from around 3.5 (mostly hypopneas) to 6+ with almost all hypopneas. And my 90% pressure was 9.

Last night I turned off flex and saw a big difference this morning. 90% pressure was 8 and AHI was around 1.6.

I know, this was just one data point. But it was a big swing, and the only thing that I changed was Flex by turning it off. That meant I was exhaling against full pressure vs exhaling against a -2cm pressure. I haven't downloaded into SH yet, but I'm assuming the hypopneas were much lower last night.
Post Reply Post Reply
#2
RE: Flex settings and hypopneas
Flex is confusing to most people. Regardless of the setting 1-3, you get up to 2-cm pressure relief, but the transition is either faster (1), or slower (3). The actual pressure relief is proportional to the force, or flow rate of your own exhale. So people with good respiratory function and tidal volume seem to see the most pressure relief. Because it is flow based, as your exhale slows, the pressure rises back to IPAP by the time you complete exhale, and you have full IPAP pressure when you begin inhaling.

EPR on the other hand is an implementation of bilevel, and actually gives you the amount of pressure relief 1-3 cm, that you set. It is initiated by your breathing, but follows, rather than leads. So as you begin exhaling, the pressure drops to EPAP, and does not return to IPAP until you spontaneously initiate the inhale...got it?

Sensitivity to any pressure relief scheme is purely individual in the perception of comfort, and sometimes in the manifestation of central apnea. There could be any number of reasons, the point is some people have CA and most do not. If you do, then Flex or EPR may not be for you. Sometimes EPR will work for someone sensitive to Flex and vice versa. Go figure...Dont-know
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#3
RE: Flex settings and hypopneas
Thanks for the Flex explanation. It makes sense.

I was really surprised to see such a big drop in hypopneas by turning it off. My aprnea index is normally less than 1, but the hyponea index is what's been causing my total AHI to rise.

Note: I edited my original post. I had been using inches instead of cm for pressures. I knew it was cm, but had "inches" on my brain for some reason.

Post Reply Post Reply
#4
RE: Flex settings and hypopneas
(10-19-2016, 09:17 AM)Sleeprider Wrote: Flex is confusing to most people. Regardless of the setting 1-3, you get up to 2-cm pressure relief, but the transition is either faster (1), or slower (3). The actual pressure relief is proportional to the force, or flow rate of your own exhale. So people with good respiratory function and tidal volume seem to see the most pressure relief. Because it is flow based, as your exhale slows, the pressure rises back to IPAP by the time you complete exhale, and you have full IPAP pressure when you begin inhaling.

I can't thank you enough for that explanation,been trying to get my head around it.

I also suffer from a lot of hypopnea's I have Flex set to 3 should I change to 1 or turn it off all together as well?
Post Reply Post Reply
#5
RE: Flex settings and hypopneas
As long as it it enabled, the 2-cm pressure relief is enabled, only the speed of transition changes. Turn it off if you want no change in inhale/exhale pressure. Oddly, a lot of us did bebber with a setting of 1, even though it still provides pressure relief.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  [Diagnosis] Help with Hypopneas! strugglingzombie 0 142 04-03-2024, 11:05 AM
Last Post: strugglingzombie
  Worrying session last night - hypopneas increased tenfold! BigWing 12 414 03-23-2024, 06:23 AM
Last Post: BigWing
  Does EPAP or IPAP control hypopneas? MrIvanDrago 3 344 02-29-2024, 05:08 PM
Last Post: Sleeprider
  [Treatment] Help Please - Hypopneas remain with BIPAP Newby 0 217 02-25-2024, 03:52 PM
Last Post: Newby
  Hypopneas resolved, CA's persist--help! AZAwake 5 507 01-20-2024, 03:16 PM
Last Post: Deborah K.
  Central Hypopneas suddenly! Kingsway 4 553 11-19-2023, 01:20 PM
Last Post: Sleeprider
  VAUTO Settings vs. CPAP Settings MrIvanDrago 9 918 11-18-2023, 10:29 AM
Last Post: MrIvanDrago


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.