Hello Guest, Welcome to Apnea Board !
As a guest, you are limited to certain areas of the board and there are some features you can't use.
To post a message, you must create a free account using a valid email address.

or Create an Account


New Posts   Today's Posts

Flow limitation
#61
RE: Flow limitation
I have once more reviewed my data.
Since the pressure I get in the mask is the EPAP, I removed the other from the graph.

I also filtered my data to exclude sleeps with less than 15 minutes' duration.

The following two sets of values come from the csv exported by OSCAR.
  1. The EPAP values are those in the column with the heading "95% EPAP".
  2. The Flow Limitation values are those in the column with the heading "95% Flow Limit.".
The EPR values were taken from OSCAR statistics page under "Changes to Device Settings".

The EPR 3 value corresponds to "EPR: Full Time 3 cmH2O"
The EPR 0 value corresponds to "EPR: Ramp Only 3 cmH2O"

I also chose to show EPR = 0 for one exception when I had "EPR: Off" from Aug 31 to Sep 2.

I still see flow limitation going up when pressure goes down except for the rightmost pressure peak where flow limitation goes up and pressure goes up.

Something is not right or I don't understand well enough what is going on.

DanEm

"If life hands you lemons, make lemonade! Words to live by, especially when you keep in mind that the only way to make them into lemonade is to squeeze the hell out of them."
Post Reply Post Reply
#62
RE: Flow limitation
@DanEm
I'm super impressed by all these charts!
I think your charts accurately express that your flow limitations are elevated and are not being controlled with your current treatment.
And I think the medical community is in agreement that high flow limitations = poor sleep quality and that Bilevel machine pressure support is often the solution to high flow limitations.
That might be enough data to convince the doctor to write a script for a Bilevel. (depends on the doc)

But are some of the charts trying to make correlations between settings and flow limitations that are really just normal nightly variances or simply presenting the data in other ways? (I admit to being confused about what was being shown in some of those graphs)


I also wonder about the assumption that Flow Limit reported in OSCAR as 0.0-1.0 is equal to percent 0-100.
The article you posted suggested "that only levels of inspiratory flow limitation > 30% be considered in the process of diagnosing obstructive sleep apnea in the absence of an apnea-hypopnea index > 5"
That would be 30%+ of total sleep time.
Is the flow limitation measurement on the ResMed a calculation of the percent of the sleep session you experienced an inspiratory flow limitation or is it a grading on the inspiratory effort of each breath? 

Or am I connecting two separate ideas that you didn't mean to be connected? 


I am in the process of compiling data for a similar discussion with my sleep doctor to try to get a bilevel.  My strategy is simple but I hope still effective.
Post Reply Post Reply
#63
RE: Flow limitation
Quote:Brazen
But are some of the charts trying to make correlations between settings and flow limitations that are really just normal nightly variances or simply presenting the data in other ways? (I admit to being confused about what was being shown in some of those graphs)

Brazen, I am just taking the data and setting up graphs to see if I can get good arguments for a change of machine.

I was told that the only improvement I could get for my sleep was to cut my flow limitation with a BiPAP.
I was told that a 0.1 value or less for FL should be my aim.
So I am looking for arguments to present to my doctor to get a new prescription (my English just got a new word = script :- ).

I admit that the two graphs posted together may be confusing. But I did not modify the data, just the way to present it.

In the last post I did reduce the quantity of information included in the graph and explained what I did.


Quote:I also wonder about the assumption that Flow Limit reported in OSCAR as 0.0-1.0 is equal to percent 0-100.

It depends which way ResMed (or OSCAR ?) treats it, more below.


Quote:The article you posted suggested "that only levels of inspiratory flow limitation > 30% be considered in the process of diagnosing obstructive sleep apnea in the absence of an apnea-hypopnea index > 5" That would be 30%+ of total sleep time.

The article says:
However, only 5% of asymptomatic individuals will have more than 30% of total sleep time with inspiratory flow limitation. This suggests that only levels of inspiratory flow limitation > 30% be considered in the process of diagnosing obstructive sleep apnea in the absence of an apnea-hypopnea index > 5 and that < 30% of inspiratory flow limitation may be a normal finding in many patients. 

The first part is clear to be a reference to time of sleep.
My interpretation of the second part is that the flow limitation > 30% is the value of the FL ( = 0.3 ), not the time.

In any case, my average FL value when AHI < 5 is 0.42.  A figure of 0.42 is the same as 42% when the maximum value for FL is 1.0 (unless you tell me OSCAR can produce figures greater than 1.0 for FL)
As for the portion of my sleep where I have FL it is much greater than the first reference to 30%. (See the graphic included)


Quote:Is the flow limitation measurement on the ResMed a calculation of the percent of the sleep session you experienced an inspiratory flow limitation or is it a grading on the inspiratory effort of each breath? 

I do not know what are the units of the FL figures in OSCAR.  But flow is the volume of fluid (air) that passes in a unit of time. Flow is often measured in units of cubic feet per second (cfs), cubic meters per second (cms), gallons per minute (gpm), or other various units.
OSCAR (or ResMed) must have a reference to express the flow with values from zero to one.  I would guess that when the machine supply the maximum flow is capable, the FL is 0 and if there is no air at all going through, then the FL would be 1 and you would be dead.


Quote:I am in the process of compiling data for a similar discussion with my sleep doctor to try to get a bilevel.  My strategy is simple but I hope still effective.

Good luck! 
I will be in that situation in October, I hope.

So I hope I have shed some light....

Frenchy (DanEm) Rolleyes

DanEm

"If life hands you lemons, make lemonade! Words to live by, especially when you keep in mind that the only way to make them into lemonade is to squeeze the hell out of them."
Post Reply Post Reply
#64
RE: Flow limitation
Ah. Frenchy! 
I forget that some of my near neighbors speak French first. Yes, script = prescription. 

I think your argument is solid and bi-level is thought to lower flow limits. 
I just don't know that the referenced paper supports the argument because of the unknown calculation within the machine.


Since the flow limit scale is 0-1... yeah, .42 is 42%. But 42% of what?
Post Reply Post Reply
#65
RE: Flow limitation
For what it's worth...
Both my AHI and flow limits look good (AHI under 2, flow limit under .1) and the thought is that a Bi-level could improve my sleep because of arousal breathing below the threshold of machine flags for CA/H/O/etc.
So, my strategy is marking up my OSCAR screenshots for arousal breathing with flow limits that weren't flagged and the machine didn't work to support.
I'm marking each and calculating an arousal index in hopes that it shows I need the pressure support.
Post Reply Post Reply
#66
RE: Flow limitation
The 0-1 measure is not a percent figure. It is a value that ResMed calculates based on the shape of the flow rate for the sole purpose of being factored into an algorithm that ResMed uses to increase pressure. It is a reported value and you can often see a definite correlation between pressure increases and higher values of this "index". There is some rough discussion in their patent applications.
Post Reply Post Reply
#67
RE: Flow limitation
Quote:Gideon
The 0-1 measure is not a percent figure. It is a value that ResMed calculates based on the shape of the flow rate for the sole purpose of being factored into an algorithm that ResMed uses to increase pressure

If the value for FL is 0, it should mean there is no flow limitation and air goes in/out freely.

If the value is one (1), does it mean no air is going through?

If not that, then what is the value of one (1) for FL, what does it mean?

And what are all the values in between ( 0.1, 0.2, 0.3 .......), is there a point where one should be concerned?

Here you seem to have a fix on the 0.1 or less for a nice value for FL.
Does the 0.1 come for experience, statistics, ....?

I just read on internet this:
The flow limitation index is calculated by dividing the total number of flow limitation events which occur by the number of hours over which events were recorded.
https://www.workplacetesting.com/definit...tion-index )

I did not know that flow limitation would be an event that comes on and off?
I don't know if that article is reliable.

Still did not get comments on my last posts with graph!!!!

I do feel like quitting all this, just stop searching, asking questions, keep the machine I have and live with being more tired in the morning than when I went to bed, tired in daytime, going to sleep if I am not active, having concentration problems, etc...

DanEm

"If life hands you lemons, make lemonade! Words to live by, especially when you keep in mind that the only way to make them into lemonade is to squeeze the hell out of them."
Post Reply Post Reply
#68
RE: Flow limitation
The Respironics machines record flow limitations as an on or off flag, so that index could be correct for Respironics machines.
Apnea Board Monitors are members who help oversee the smooth functioning of the Board. They are also members of the Advisory Committee which helps shape Apnea Board's rules & policies. Membership in the Advisory Members group does not imply medical expertise or qualification for advising Sleep Apnea patients concerning their treatment.
Post Reply Post Reply
#69
RE: Flow limitation
@DanEm
Do Not Give Up.

I refuse to believe that life just has to be lived feeling exhausted.

Xpap might not be the answer but you really don't know until you have fully optimized and given that treatment a solid try.

Talk to your doc. See what they have to say about your sleep, and your data, and if they have suggestions lean into those. If they don't have suggestions come back here - the experts might have suggestions.
Post Reply Post Reply
#70
RE: Flow limitation
(09-28-2022, 09:50 PM)DanEm Wrote: ... And what are all the values in between ( 0.1, 0.2, 0.3 .......), is there a point where one should be concerned?

The  Patent application has the following example  for flow limitation (pg 49):

Quote:MILD-0.1, MILD_TO_MODERATE-0.5, MODERATE-0.8

Source: https://www.freepatentsonline.com/20110203588.pdf
Jeff8356

MacBook Air (2017, Intel) | macOS Monterey (12.7) | OSCAR v1.5.1 | VM = Win10/Win11 |
How to Links:
Installing OSCAR on a Mac
Organizing your OSCAR charts
Attaching screenshots and files for the forum
OSCAR Help
OSCAR - The Guide

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  help... Suspicous Flow Rate and Flow Limitations, bothersome Aerophagia Barefooter64 17 751 38 minutes ago
Last Post: Sleeprider
  Does Aircurve 10 Vauto record events and flow limits during ramp mode? SingleH 3 44 7 hours ago
Last Post: SingleH
  [Treatment] Relationship between EPR and Flow Limitations earnerd 11 394 Yesterday, 11:51 AM
Last Post: earnerd
  Too much STATIC showing in OSCAR Flow Rate PappaJoe 0 76 04-16-2024, 08:23 AM
Last Post: PappaJoe
  High flow limit, erratic breathing and resp rates jclarke131000 14 301 04-16-2024, 06:31 AM
Last Post: jclarke131000
  Is This Flow Limitation Cause for Concern? bertchintus 4 273 04-15-2024, 08:27 PM
Last Post: bertchintus
  Flow Limit and frequent waking issues Premier1991 3 195 04-15-2024, 08:07 AM
Last Post: Sleeprider


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.