Hi. I just received my bipap machine's data today and imported it on Sleepyhead. It looks bad. The AHI is 29.97 on the machine. Most of those are clear airway apneas, 27.68. What is a clear airway apnea? It sounds bad like not obstructive apnea. I think I should call my doctor. I had a headache this morning as usual. It felt as if the new machine (that I got yesterday) is not working but the respiratory therapist came by today and he measured the air pressure on the machine and it was correct at 12/07. But I could tell last night when I used it something wasn't right.

What do you all think?

Kimberly from HonoluluSleep-well
09-13-2014 07:09 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Galactus Online

Advisory Members

Posts: 1,388
Joined: Jan 2014

Machine: Phillips Respironics System One PRS1 DS760P Auto BiPAP
Mask Type: Nasal pillows
Mask Make & Model: ResMed Airfit P10 Nasal Pillow
Humidifier: System One Heated Humidifier + Climate Line
CPAP Pressure: E/I 13/20 PS 4/7 BiFlex 3
CPAP Software: SleepyHead EncoreBasic

Other Comments: Sleep can be a good thing.

Sex: Male
Location: USA

Post: #2
RE: Help ! Clear airway apnea 27.68
The people who are going to be able to help you are going to want to see the sleepyhead data (from my experience) so take a moment now and post the picture of it so they can help you faster.

If everyone thinks alike, then someone isn't thinking.
Everyone knows something, together we could know everything.
09-13-2014 07:11 PM
Find all posts by this user Post Reply Quote this message in a reply
surferdude2 Offline

Advisory Members

Posts: 1,077
Joined: Jul 2014

Machine: ResMed S9 AutoSet
Mask Type: Full face mask
Mask Make & Model: Mirage Quattro & Optilife
Humidifier: ResMed H5i @ #3 if nose breathing & #5 if mouth breathing
CPAP Pressure: 12 ~ 20 & 11 ~12 if nasal mask
CPAP Software: ResScan SleepyHead

Other Comments: Late stage 2 emphysema

Sex: Male
Location: Mousetown, Southern Illinois

Post: #3
RE: Help ! Clear airway apnea 27.68
While you're waiting for someone that knows more than I (anyone actually) about bipap boxes, tell me how they measured the pressure, I mean did they use an actual instrument such as a water column or mechanical manometer?

Also describe better why you think it didn't seem right. Did you get any sleep...if so, how much/ Did most of the events occur in clusters while you were still awake or just after waking up? Anything else you can supply will help you get a faster opinion.

EDIT: And yeah, what Galactus said..
(This post was last modified: 09-13-2014 07:20 PM by surferdude2.)
09-13-2014 07:18 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
kfujioka Offline

Preferred Members

Posts: 63
Joined: Aug 2014

Machine: Bipap system one auto
Mask Type: Nasal mask
Mask Make & Model: Air fit P10 and The Hybrid
Humidifier: System One heated humidifier & heated tube
CPAP Pressure: 12/07 for bipap
CPAP Software: SleepyHead

Other Comments: New to sleep apnea however think I have had it for a year because started having symptoms a year ago

Sex: Female
Location: Oahu, Hawaii

Post: #4
RE: Help ! Clear airway apnea 27.68
Hi. I have my Sleepyhead software open and I am trying to export graphs so I can upload the here. Should I save as a cvs file or what?

Kimberly from HonoluluSleep-well
09-13-2014 08:30 PM
Find all posts by this user Post Reply Quote this message in a reply
Galactus Online

Advisory Members

Posts: 1,388
Joined: Jan 2014

Machine: Phillips Respironics System One PRS1 DS760P Auto BiPAP
Mask Type: Nasal pillows
Mask Make & Model: ResMed Airfit P10 Nasal Pillow
Humidifier: System One Heated Humidifier + Climate Line
CPAP Pressure: E/I 13/20 PS 4/7 BiFlex 3
CPAP Software: SleepyHead EncoreBasic

Other Comments: Sleep can be a good thing.

Sex: Male
Location: USA

Post: #5
RE: Help ! Clear airway apnea 27.68
(09-13-2014 08:30 PM)kfujioka Wrote:  Hi. I have my Sleepyhead software open and I am trying to export graphs so I can upload the here. Should I save as a cvs file or what?

You can just do a "print screen" of the data then using a site like http://www.tinypic.com just upload the image and copy the [img] [/img] link in your post so everyone can see it.

If everyone thinks alike, then someone isn't thinking.
Everyone knows something, together we could know everything.
09-13-2014 08:52 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
DeepBreathing Offline
Wiki Editor
Moderators

Posts: 2,296
Joined: Sep 2013

Machine: Resmed S9 VPAP Adapt
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: Resmed H5i
CPAP Pressure: EPAP: 9 - 15 PS: 3 - 10
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Perth, Western Australia

Post: #6
RE: Help ! Clear airway apnea 27.68
(09-13-2014 08:30 PM)kfujioka Wrote:  Hi. I have my Sleepyhead software open and I am trying to export graphs so I can upload the here. Should I save as a cvs file or what?

Just give us a screen dump of the summary area (around the pie chart) and the graphs - events, pressure, leaks, flow rate and flow limitation.

Clear airway apneas are also called central apneas and are quite different from the more common obstructive apnea. Part of the brain is devoted to automatic muscle functions including breathing. Every few seconds the brain sends a message to the diaphragm and other muscles around the lungs saying "breathe now". In a central apnea either this message doesn't get generated or for some reason doesn't get through.

This can occur when people first go on PAP therapy as the brain is fooled by the increased oxygen levels / decreased CO2 in the blood. However it can also be a built-in feature of your particular physiology. The PAP machine has to be capable of detecting a clear airway apnea and distinguishing it from an obstructive, then giving your lungs an extra blast of air to get them breathing again. I suspect that a max pressure of 12 isn't high enough to give you the kick you need. I'm not familiar with your machine, but it should have a whole range of settings to get it doing what it's supposed to.

You may heard it said that your don't need to worry about central apneas - this is not true. They deprive your body of oxygen and interrupt your sleep just as much as obstructive apneas. You'll get some good advice hear, but I suggest you talk to yur doctor and get this reviewed sooner rather than later.

DeepBreathing
Apnea Board Moderator
www.ApneaBoard.com


Bed

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
09-13-2014 09:01 PM
Find all posts by this user Post Reply Quote this message in a reply
vsheline Offline

Advisory Members

Posts: 1,908
Joined: Jul 2012

Machine: S9 VPAP Adapt (USA Model# 36007, not better 36037 or 36067)
Mask Type: Full face mask
Mask Make & Model: F10 or SimPlus w/ 2Liners. MirageQuatro & Gecko gel pad
Humidifier: H5i
CPAP Pressure: 15 EPAP, PS 5-10
CPAP Software: ResScan

Other Comments: Marfan Syndrome, chronic bradycardia, occasional Cheyne-Stokes Respiration

Sex: Male
Location: California, USA

Post: #7
RE: Help ! Clear airway apnea 27.68
(09-13-2014 09:01 PM)DeepBreathing Wrote:  The PAP machine has to be capable of detecting a clear airway apnea and distinguishing it from an obstructive, then giving your lungs an extra blast of air to get them breathing again. I suspect that a max pressure of 12 isn't high enough to give you the kick you need. I'm not familiar with your machine, but it should have a whole range of settings to get it doing what it's supposed to.

Hi DeepBreathing - kfujioka's machine does not have a backup respiration rate and is not an ASV machine. If operated in AutoB (Auto BiPAP) mode, the machine can slowly adjust EPAP to minimize obstructive events and can slowly vary Pressure Support to further lower Flow Limitation.

Hi kfujioka - What therapeutic mode is your machine set in?

The settings 12/7 in your profile seem like settings for BiPAP mode, which is the fixed BiPAP mode where IPAP (the pressure when inhaling) is always 12 and EPAP (the pressure when exhaling) is always 7. If you doctor or therapist has put the machine in fixed BiPAP mode, I do not suggest changing the therapy mode. They may have some good reason for doing so.

By the way, to update the settings in your profile, you can use the "My Profile" link members see at the top right of all forum pages.

How long have you been using the machine at these settings?

Lowering PS (Pressure Support, which is the difference between IPAP and EPAP) often tends to lower the number of CA events.

Also, lowering IPAP (preferably by lowering PS) often tends to lower the number of CA events.

Lowering EPAP tends to increase the number of obstructive events, so I do not suggest doing that.

I suggest you talk to your doctor or respiratory therapist about lowering PS, at least temporarily, until your body gets accustomed to breathing under pressure and the number of CA events decreases.

Take care,
--- Vaughn

Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment. The Advisory Member group provides advice and suggestions to Apnea Board administrators and staff on matters concerning Apnea Board operation and administrative policies - not on matters concerning treatment for Sleep Apnea. I think it is now too late to change the name of the group but I think Voting Member group would perhaps have been a more descriptive name for the group.
(This post was last modified: 09-14-2014 01:02 AM by vsheline.)
09-14-2014 12:46 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Sleepster Offline
Wiki Editor
Moderators

Posts: 4,995
Joined: Feb 2012

Machine: ResMed AirCurve10 VAuto
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: HumidAir and SlimLine Hose
CPAP Pressure: MaxI 13.6 | MinE 5.2 | PS 4.4
CPAP Software: ResScan SleepyHead

Other Comments: Diagnosed Nov 2011. Conquered aerophagia.

Sex: Male
Location: Houston, Texas

Post: #8
RE: Help ! Clear airway apnea 27.68
(09-13-2014 07:09 PM)kfujioka Wrote:  Hi. I just received my bipap machine's data today and imported it on Sleepyhead. It looks bad. The AHI is 29.97 on the machine. Most of those are clear airway apneas, 27.68. What is a clear airway apnea?

Apnea (you stopped breathing) with an airway that was not obstructed. This happens to new users fairly often. Tell your doctor about it. If it lasts for more than a week or two they may want to lower your pressure to something like 10/7. At least for a while until you adapt.

Sleepster
Apnea Board Moderator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
09-14-2014 04:11 AM
Find all posts by this user Post Reply Quote this message in a reply
justMongo Offline

Monitors

Posts: 3,853
Joined: Sep 2013

Machine: ResMed VPAP Auto (S9)
Mask Type: Full face mask
Mask Make & Model: ResMed Mirage Quattro
Humidifier: ResMed H5i
CPAP Pressure: 18 IPAPmax - 11 EPAPmin, PS=5
CPAP Software: Other Software

Other Comments: μολὼν λαβέ

Sex: Male
Location: Калифорния

Post: #9
RE: Help ! Clear airway apnea 27.68
Duration of events is also a factor in the degree of concern.
I get an occasional clear airway apnea flagged; but the duration is 10 to 12 seconds.

Still, at ~28 per hour, I would get the doctor involved. At least notified.

FYI folks: I use a digital manometer to check my machine pressure.
It measures in inches of water; the conversion is: multiply by 2.54
I usually can only catch the EPAP pressure as the display is still slewing during IPAP.
That's why an analog manometer -- even a u-tube with water in it is better.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
09-14-2014 10:56 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
surferdude2 Offline

Advisory Members

Posts: 1,077
Joined: Jul 2014

Machine: ResMed S9 AutoSet
Mask Type: Full face mask
Mask Make & Model: Mirage Quattro & Optilife
Humidifier: ResMed H5i @ #3 if nose breathing & #5 if mouth breathing
CPAP Pressure: 12 ~ 20 & 11 ~12 if nasal mask
CPAP Software: ResScan SleepyHead

Other Comments: Late stage 2 emphysema

Sex: Male
Location: Mousetown, Southern Illinois

Post: #10
RE: Help ! Clear airway apnea 27.68
(09-14-2014 10:56 AM)justMongo Wrote:  FYI folks: I use a digital manometer to check my machine pressure.
It measures in inches of water; the conversion is: multiply by 2.54
I usually can only catch the EPAP pressure as the display is still slewing during IPAP.
That's why an analog manometer -- even a u-tube with water in it is better.

Thanks for that, it confirms what I suspected, that the old U-tube wc is still alive and well (sometimes a little messy though). If you have the time to wait for a well calibrated digital instrument to make up its mind, it no doubt will give the most accurate reading with the highest degree of granularity.

I have a scope that would be great for this type of thing but I suspect a transducer might cost more than I could justify. My wc is close enough for government work.
(This post was last modified: 09-14-2014 11:55 AM by surferdude2.)
09-14-2014 11:04 AM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  Hypopnea-Clear Airway Questions Zippo 7 176 Yesterday 10:45 PM
Last Post: Zippo
  Unable to clear throat while on APAP NorthernGuy 5 311 10-17-2016 08:18 PM
Last Post: Sleeprider
  Clear/Central Apneas, BiPAP, CO2 washout confusion. NorthernGuy 3 488 09-02-2016 09:29 AM
Last Post: PaytonA
  [Treatment] Auto CPAP, high/erratic AHI, high Clear Apneas - it's not working jmortensenjr 10 769 08-30-2016 09:07 PM
Last Post: jmortensenjr
  Can CPAP therapy cause Total Clear Airway Apneas TheLionSleeps 3 605 07-17-2016 08:39 PM
Last Post: WakeUpTime
  Survey on what pressure opens airway for you TheLionSleeps 22 1,136 07-08-2016 07:05 PM
Last Post: KSMatthew
  How to clear Fault 27 on HDM Z1 APAP? keninct 1 279 07-06-2016 03:11 PM
Last Post: trish6hundred

Forum Jump:

Who's Online (Complete List)