Ebit asked me about the incompatibility issue between the P10 and the ASV machines. These are my thoughts on that:

It seems like I heard that before. It's interesting, because it disagrees with Resmed's home site. It might be interesting to ask them how come.

I believe the problem is with ASV therapy, sometimes it's necessary to raise the pressure considerably to treat a CA event. The P10 is a great little mask, but gets a bit unhappy with higher pressures. So leakage becomes a much greater issue.

In your case I don't think ASV therapy is called for at this point. At least not given the results you have at hand. For you I am suggesting a VPAP therapy with a reasonable low and high pressure. Just what the P10 likes.

But as an alternative? If it's a concern for you, take a look at the Piliaro Q pillows mask. It's very similar in design to the P10, but has a more substantial headgear arrangement that would make it a happier camper with pressures in the 20's.
(This post was last modified: 05-21-2015 09:03 PM by retired_guy.)
05-21-2015 09:02 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,345
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #12
RE: Help is needed - What is wrong with me? (S9 VPAP ADAPT user)
I know a number of people that use the P10 with ASV machines with great success. If you get a good seal on a nasal pillows mask, I think it withstands leaks better than a FFM, and applies therapy pressure much more effectively...

But then again, I'm in the cult.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
05-21-2015 09:15 PM
Find all posts by this user Post Reply Quote this message in a reply
trailrider Offline

Preferred Members-2

Posts: 304
Joined: Feb 2015

Machine: A10 Autoset for Her
Mask Type: Nasal mask
Mask Make & Model: Eson or P10
Humidifier: autoset
CPAP Pressure: 6-12 DME, 7-13 me
CPAP Software: SleepyHead

Other Comments: engineer. EPR at 3, unheated hose, humidity at 2. Routine hose entanglement.

Sex: Female
Location: Ontario

Post: #13
RE: Help is needed - What is wrong with me? (S9 VPAP ADAPT user)
This might be wrong, but how about slamming the Max pressure setting to a tolerable 15, and the PS to 3 or 4 for a few days? At least he could sleep with some therapy.
05-21-2015 09:25 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
retired_guy Offline

Preferred Members-2

Posts: 2,744
Joined: Jan 2014

Machine: ResMed S9 Autoset
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: ResMed
CPAP Pressure: 10.6/14, EPR 3
CPAP Software: ResScan SleepyHead

Other Comments: Is it bedtime yet?

Sex: Male
Location: Beautiful, albeit very wet Oregon Coast

Post: #14
RE: Help is needed - What is wrong with me? (S9 VPAP ADAPT user)
(05-21-2015 09:25 PM)trailrider Wrote:  This might be wrong, but how about slamming the Max pressure setting to a tolerable 15, and the PS to 3 or 4 for a few days? At least he could sleep with some therapy.

Yes, I agree. I'll stick with my original recommendation of VPAP mode, 8 minimum, 14 maximum, EPR 2, and no ramp. (however all of that gets defined with an ASV) Then see what Sleepyhead has to say about all this after a day or two.

Again, given the results of OP's sleep tests, I just don't personally see where an ASV should have been ordered. It's ok that it was, it's certainly a fine machine, but I am not convinced the ASV features are anything he needs to use at this time.
(This post was last modified: 05-21-2015 09:33 PM by retired_guy.)
05-21-2015 09:31 PM
Find all posts by this user Post Reply Quote this message in a reply
trailrider Offline

Preferred Members-2

Posts: 304
Joined: Feb 2015

Machine: A10 Autoset for Her
Mask Type: Nasal mask
Mask Make & Model: Eson or P10
Humidifier: autoset
CPAP Pressure: 6-12 DME, 7-13 me
CPAP Software: SleepyHead

Other Comments: engineer. EPR at 3, unheated hose, humidity at 2. Routine hose entanglement.

Sex: Female
Location: Ontario

Post: #15
RE: Help is needed - What is wrong with me? (S9 VPAP ADAPT user)
OP Quote:
after reading many threads here I decided on my own to do the following:
1. get climate control and set it to automatic on 72F
2. change the mode to ASV AUTO with default which is: Min EPAP 4.0 Max Ipap 30.0 PS 3-15

then after talking to one of the advisers here on this forum changed the default settings to:
Min EPAP 8 Max Ipap 25 PS 4-10 but somehow still kept it using ASV auto mode

R-G quote:
VPAP mode, 8 minimum, 14 maximum, EPR 2, and no ramp. (however all of that gets defined with an ASV)

I think R-Gs translates into :
Min EPAP 8, Max IPAP 14 PS 2. And change the max pressure to 15 for good measure.

I'm wondering about that OP value of PS 4-10 or 3-15 is causing the problem. Get rid of the second number in the PS, or set it to first number +1 .

If I read it correctly the OP's second setting PS could result in an exhale pressure of 15 with an inhale pressure of 25. That would be horrible! And the OP 1st setting PS would be a worst case of exhale pressure 15, inhale of 30. Yikes!!

So Ebit, here is my recommendation:
Min EPAP 8, Max IPAP 14 PS 2-3. Max pressure =15.
If 8 is too hard to deal with at first, move it to 7. The program should run and adapt as you sleep and keep things flowing without choking you.

I know you probably won't see this until tomorrow. Those settings will mimic an APAP, min 8, max 14, EPR 2 or 3.
I hope this helps.
(This post was last modified: 05-21-2015 10:26 PM by trailrider.)
05-21-2015 10:24 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
retired_guy Offline

Preferred Members-2

Posts: 2,744
Joined: Jan 2014

Machine: ResMed S9 Autoset
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: ResMed
CPAP Pressure: 10.6/14, EPR 3
CPAP Software: ResScan SleepyHead

Other Comments: Is it bedtime yet?

Sex: Male
Location: Beautiful, albeit very wet Oregon Coast

Post: #16
RE: Help is needed - What is wrong with me? (S9 VPAP ADAPT user)
Thanks TR..... Sounds like a fine idea to me too.
05-21-2015 10:29 PM
Find all posts by this user Post Reply Quote this message in a reply
DeepBreathing Offline
Wiki Editor
Moderators

Posts: 2,295
Joined: Sep 2013

Machine: Resmed S9 VPAP Adapt
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: Resmed H5i
CPAP Pressure: EPAP: 9 - 15 PS: 3 - 10
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Perth, Western Australia

Post: #17
RE: Help is needed - What is wrong with me? (S9 VPAP ADAPT user)
Lot of arithmetic going on here but we're probably not quite on the right wavelength for a VPAP Adapt. The available modes are CPAP, ASV and ASV Auto. To emulate an Autoset, you need to select the ASV Auto mode. The available settings in this mode are Min EPAP, Max EPAP, Min PS, Max PS and Max IPAP.

EPAP + PS = IPAP

Min PS can't be set lower than 3. So to achieve a pressure range between 8 and 14 with the equivalent of a 3 EPR you would need to set Min EPAP = 8, Max EPAP = 10, Min PS = 3, Max PS = 4. Max IPAP is irrelevant provided it's set above 14, as the machine won't go above Max EPAP + Max PS.

I don't recommend using this machine while awake to get used to it - normal pauses in breathing (which happen all the time) are sensed as incipient centrals and the machine will react in a few seconds to start pumping the pressure up. This gets quite unsettling.

I agree with the previous posters that ASV seems overkill for this patient, unless the test is seeing something that hasn't come through on the reports. The problem with an ASV if you don't need it is that the machine tries to breathe with you, which is only necessary if you have a lot of centrals. It won't necessarily do you any harm, but it does take some getting used to. When you're having problems adapting anyway you don't need this additional complication.

DeepBreathing
Apnea Board Moderator
www.ApneaBoard.com


Bed

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
05-22-2015 05:40 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
PaytonA Offline
Wiki Editor
Monitors

Posts: 3,010
Joined: Dec 2013

Machine: ResMed S9 VPAP Auto
Mask Type: Full face mask
Mask Make & Model: Resmed Mirage Quattro
Humidifier: H5i(distilled-top up)
CPAP Pressure: VAuto MinE14.0 MaxI 20.6 PS4.0
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Orange County,California

Post: #18
RE: Help is needed - What is wrong with me? (S9 VPAP ADAPT user)
(05-21-2015 08:37 PM)ebitansky Wrote:  Thanks for the good words.
Apparently (according to the mask manual ) this mask is not compatible with VPAP adapt SV machines. ......

The only thing that I can find in the manual that might be construed to indicate that the mask is not compatible with the VPAP adapt SV machine is the following statement: "This mask should be used only with flow generators (CPAP or bilevel) recommended by your physician or respiratory therapist."

The fact of the matter is that the VPAP Adapt is a bilevel flow generator. I do not think that statement was meant to rule out ASV machines.

Best Regards,

PaytonA
(This post was last modified: 05-22-2015 10:42 AM by PaytonA.)
05-22-2015 10:41 AM
Find all posts by this user Post Reply Quote this message in a reply
ebitansky Offline

Members

Posts: 10
Joined: Apr 2015

Machine: Resmed S9 Vpap Adapt 36037
Mask Type: Full face mask
Mask Make & Model: Resmed Ultra Mirage
Humidifier: H5i
CPAP Pressure: AutoASV max EPAP 15 Min EPAP 8
CPAP Software: ResScan

Other Comments:

Sex: Male
Location:

Post: #19
RE: Help is needed - What is wrong with me? (S9 VPAP ADAPT user)
I took it from airfit p10 manual page 3 section "Using your mask" at the notes:

Notes:
This mask is not compatible for use with ResMed
AutoSet CS™ 2 and VPAP™ Adapt SV devices.
SmartStart may not operate effectively when using this mask
with some CPAP or bilevel devices that have this feature.

Is airlift p10 the wrong mask ?

(05-22-2015 10:41 AM)PaytonA Wrote:  
(05-21-2015 08:37 PM)ebitansky Wrote:  Thanks for the good words.
Apparently (according to the mask manual ) this mask is not compatible with VPAP adapt SV machines. ......

The only thing that I can find in the manual that might be construed to indicate that the mask is not compatible with the VPAP adapt SV machine is the following statement: "This mask should be used only with flow generators (CPAP or bilevel) recommended by your physician or respiratory therapist."

The fact of the matter is that the VPAP Adapt is a bilevel flow generator. I do not think that statement was meant to rule out ASV machines.

Best Regards,

PaytonA
05-22-2015 01:05 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
retired_guy Offline

Preferred Members-2

Posts: 2,744
Joined: Jan 2014

Machine: ResMed S9 Autoset
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: ResMed
CPAP Pressure: 10.6/14, EPR 3
CPAP Software: ResScan SleepyHead

Other Comments: Is it bedtime yet?

Sex: Male
Location: Beautiful, albeit very wet Oregon Coast

Post: #20
RE: Help is needed - What is wrong with me? (S9 VPAP ADAPT user)
(05-22-2015 01:05 PM)ebitansky Wrote:  I took it from airfit p10 manual page 3 section "Using your mask" at the notes:

Notes:
This mask is not compatible for use with ResMed
AutoSet CS™ 2 and VPAP™ Adapt SV devices.
SmartStart may not operate effectively when using this mask
with some CPAP or bilevel devices that have this feature.

Is airlift p10 the wrong mask ?

Oh, ok.... Smartstart and it's evil twin Smartstop may not work right. What that feature is supposed to do is automatically start the machine when you put on your mask and start breathing, and automatically stop it when you take your mask off.

That saves significant wear and tear on your index finger which otherwise has to push the start button to accomplish the same purpose.

It's a ridiculous feature, adored by some, and a real trouble maker for others.

Use your index finger. It's why God gave you two of them.
05-22-2015 01:28 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  New User Problem with BiPAP AVAPS - Part 1 jmayer180 35 1,291 11-27-2016 09:50 AM
Last Post: jmayer180
  New User! SKClay 3 262 11-21-2016 08:52 PM
Last Post: trish6hundred
  Resmed Adapt ASV question storywizard 7 326 11-21-2016 01:02 PM
Last Post: storywizard
  new user wants advice Gazz 3 230 11-19-2016 11:06 AM
Last Post: trish6hundred
  Advice needed for struggling CPAP user Fubblin 35 1,977 11-18-2016 07:18 PM
Last Post: Sleeprider
  First time Bipap User, years of suffering SleepyMersim 18 647 11-16-2016 11:46 PM
Last Post: Galactus
  New User Having Trouble Staying Asleep and more SnoreGalore 7 443 11-09-2016 06:26 PM
Last Post: DeepBreathing

Forum Jump:

Who's Online (Complete List)