Felt I wasn't getting as rested as I should be on BiPAP ST (set at 15/4 with backup rate 7), and since my diagnosis was Complex SA, the doctor set me up with an ASV.

I am curious if it's normal that I feel inhalation requires a fair bit of effort, as if there were some resistance. Almost feels like trying to suck air in through a thick air filter. I turned off ramp to see if that would help, and I think it did a little, but not much.

It's pretty much opposite of the BiPAP ST where when I even started a little inhalation, there would be a big release of air, which I actually found very easy to deal with and pretty comfortable.

I guess the two things that are weird to me is that heavy resistance while inhaling and the push/pull ease of the BiPap.

The Sleepquest tech set the Resmed Aircurve ASV at Min EPAP 5, Max EPAP 15, Min PS 0, Max PS 10, autoASV.
03-18-2016 12:38 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
palerider Offline

Preferred Members-

Posts: 440
Joined: Nov 2014

Machine: s9 vpap auto
Mask Type: Nasal pillows
Mask Make & Model: p10
Humidifier: none
CPAP Pressure: 18/13+
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Dallas(ish)

Post: #2
RE: Help with ASV Experience/Settings, coming from BiPAP ST
your minepap is pretty low, and you're likely not used to having a minimum of 0 ps.

though, if your reporting of your old prescription is correct, you had a HUGE ps on the old machine... and even lower min epap.
03-18-2016 03:18 PM
Find all posts by this user Post Reply Quote this message in a reply
eseedhouse Offline

Advisory Members

Posts: 881
Joined: Dec 2014

Machine: ResMed Airsense 10 Autoset
Mask Type: Full face mask
Mask Make & Model: SimPlus
Humidifier: ResMed
CPAP Pressure: 7-10
CPAP Software: SleepyHead

Other Comments: Also on supplemental O2 at 3L/min. while sleeping.

Sex: Male
Location: Victoria, British Columbia

Post: #3
RE: Help with ASV Experience/Settings, coming from BiPAP ST
You breath in by using your diaphragm to create a partial vacuum that draws air in your lungs. You breath out by using your diaphragm to compress the air in your lung cavity outside the lungs, so that air is forced out.

Your diaphragm is a muscle, and like all muscles it adapts to stress by getting stronger. It is having to work harder than before to push air out so it might get a bit sore, just like your arm muscles would if you started lifting weights. Like your arm muscles on a weight lifting program the added exercise causes the muscle to partially break down and then rebuild itself stronger to handle the increased effort.

In a week or two your diaphragm will get "trained" and will be stronger and fitter and you will have little or no trouble breathing out. In the meantime you might find exhalation pressure relief ("EPR" helpful and you might even prefer to use it. That's fine unless your apneas and hypopneas are too high, and then you should probably work on training that diaphragm so you don't need it.

Ed Seedhouse
VA7SDH

I am neither a Doctor, nor any other kind of medical professional.

Actually you know, it is what it isn't.
03-18-2016 03:48 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
SimbaLion Offline

Members

Posts: 8
Joined: Nov 2015

Machine: ResMed Airsense 10
Mask Type: Full face mask
Mask Make & Model: Amara View
Humidifier: Built In
CPAP Pressure: Unsure
CPAP Software: Not using software

Other Comments:

Sex: Male
Location:

Post: #4
RE: Help with ASV Experience/Settings, coming from BiPAP ST
eseedhouse, I have no trouble breathing out, never have on my CPAP --> BiPap --> ASV journey.

I'm asking about breathing IN. It is almost suffocating in a way.
03-18-2016 04:11 PM
Find all posts by this user Post Reply Quote this message in a reply
SimbaLion Offline

Members

Posts: 8
Joined: Nov 2015

Machine: ResMed Airsense 10
Mask Type: Full face mask
Mask Make & Model: Amara View
Humidifier: Built In
CPAP Pressure: Unsure
CPAP Software: Not using software

Other Comments:

Sex: Male
Location:

Post: #5
RE: Help with ASV Experience/Settings, coming from BiPAP ST
(03-18-2016 03:18 PM)palerider Wrote:  your minepap is pretty low, and you're likely not used to having a minimum of 0 ps.

though, if your reporting of your old prescription is correct, you had a HUGE ps on the old machine... and even lower min epap.

Just looked at my old BiPAP prescription to double check. Yep, 15/4 is what it says.
03-18-2016 04:14 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
palerider Offline

Preferred Members-

Posts: 440
Joined: Nov 2014

Machine: s9 vpap auto
Mask Type: Nasal pillows
Mask Make & Model: p10
Humidifier: none
CPAP Pressure: 18/13+
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Dallas(ish)

Post: #6
RE: Help with ASV Experience/Settings, coming from BiPAP ST
(03-18-2016 03:48 PM)eseedhouse Wrote:  You breath in by using your diaphragm to create a partial vacuum that draws air in your lungs. You breath out by using your diaphragm to compress the air in your lung cavity outside the lungs, so that air is forced out.

the diaphram contracts and pulls your respiratory cavity floor down, creating a vacuum, you're correct.

the diaphram *can't* force air out, any more than any other muscle can forcefully extend.
03-18-2016 04:57 PM
Find all posts by this user Post Reply Quote this message in a reply
eseedhouse Offline

Advisory Members

Posts: 881
Joined: Dec 2014

Machine: ResMed Airsense 10 Autoset
Mask Type: Full face mask
Mask Make & Model: SimPlus
Humidifier: ResMed
CPAP Pressure: 7-10
CPAP Software: SleepyHead

Other Comments: Also on supplemental O2 at 3L/min. while sleeping.

Sex: Male
Location: Victoria, British Columbia

Post: #7
RE: Help with ASV Experience/Settings, coming from BiPAP ST
(03-18-2016 04:11 PM)SimbaLion Wrote:  eseedhouse, I have no trouble breathing out, never have on my CPAP --> BiPap --> ASV journey.

I'm asking about breathing IN. It is almost suffocating in a way.

Sounds like your pressure level as you breath in is too low. For instance APAP machines are often set "wide open" and it is common for new users to feel suffocated until they set the bottom pressure up a few cm.

Ed Seedhouse
VA7SDH

I am neither a Doctor, nor any other kind of medical professional.

Actually you know, it is what it isn't.
03-18-2016 05:55 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Asjb Offline

Advisory Members

Posts: 173
Joined: Oct 2014

Machine: Aircurve 10 CS Pacewave ASV. Before 11/2015 Airsense 10 Auto
Mask Type: Full face mask
Mask Make & Model: FP Simplus & Remzzzs liner
Humidifier: Integral
CPAP Pressure: Current info in this post
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Undisclosed
Location: France

Post: #8
RE: Help with ASV Experience/Settings, coming from BiPAP ST
(03-18-2016 12:38 PM)SimbaLion Wrote:  Felt I wasn't getting as rested as I should be on BiPAP ST (set at 15/4 with backup rate 7), and since my diagnosis was Complex SA, the doctor set me up with an ASV.

I am curious if it's normal that I feel inhalation requires a fair bit of effort, as if there were some resistance. Almost feels like trying to suck air in through a thick air filter. I turned off ramp to see if that would help, and I think it did a little, but not much.

It's pretty much opposite of the BiPAP ST where when I even started a little inhalation, there would be a big release of air, which I actually found very easy to deal with and pretty comfortable.

I guess the two things that are weird to me is that heavy resistance while inhaling and the push/pull ease of the BiPap.

The Sleepquest tech set the Resmed Aircurve ASV at Min EPAP 5, Max EPAP 15, Min PS 0, Max PS 10, autoASV.

Hello Simbalion,

I've been using the Aircurve 10 ASV for a few months and my experience is so different to yours that I wonder if either your loan machine is not working properly or your minimum pressure is too low for you. (IPAP minus EPAP = PS). I've read on the forum here that many people find an IPAP at around 4, 5, 6 (whatever the type of machine) to be too 'suffocating'. You perhaps could discuss this with your sleep tech or doctor. My suggestion would be to start by increasing your min EPAP to 6. If that makes inspiration more comfortable then you would be on the right path and you could then slowly further adjust your ASV settings, including increasing pressure support as required (a minimum PS of zero is low, I understand most ASVs are ideally set with a PS minimum of at least 3), by looking at your data with Sleepyhead and asking for advice on the forum if you need to.

I find inspiration with the ASV very comfortable and very 'aerating' - if that's the right word to use. My change to ASV allowed for much higher pressures (that I needed) than I could tolerate with CPAP/APAP and my diaphragm and chest muscles did take a week or so to get used to the different sense you get with an ASV and that my average tidal volume has increased from 420 (CPAP/APAP) to 500 (ASV) i.e. deeper breathing. My settings inflate my lungs very nicely! so I am sure you will find the correct settings to achieve that feeling for you too.

I hope it settles soon for you and that you achieve the same AHI results and sleep quality that I have with the Aircurve ASV.

best wishes, asjb

'Complex' apnoea due to OSA + CPAP-induced central apnoeas.
Currently EPAP Auto 12-14. Pressure support 6-11.
(This post was last modified: 03-18-2016 06:29 PM by Asjb.)
03-18-2016 06:18 PM
Find all posts by this user Post Reply Quote this message in a reply
trish6hundred Offline

Advisory Members

Posts: 6,446
Joined: May 2012

Machine: Resmed S9 AutoSet for Her
Mask Type: Full face mask
Mask Make & Model: Fisher & Paykel Simplus
Humidifier: H5i Heated Humidifier
CPAP Pressure: 10 - 7-20 Cm H2O
CPAP Software: Not using software

Other Comments: I started CPAP in 2008. Totally blind since birth.

Sex: Female
Location: Missouri, USA

Post: #9
RE: Help with ASV Experience/Settings, coming from BiPAP ST
Hi SimbaLion,
WELCOME! to the forum.!
Hang in there for more suggestions and I wish you much success with your CPAP therapy and getting your machine fine tuned to meet your needs.

trish6hundred
03-18-2016 06:58 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
DeepBreathing Offline
Wiki Editor
Moderators

Posts: 2,295
Joined: Sep 2013

Machine: Resmed S9 VPAP Adapt
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: Resmed H5i
CPAP Pressure: EPAP: 9 - 15 PS: 3 - 10
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Perth, Western Australia

Post: #10
RE: Help with ASV Experience/Settings, coming from BiPAP ST
G'day SimbaLion, welcome to Apnea Board.

I suspect your problem is with the PS min = 0. Effectively the machine is running like a CPAP at a very low pressure. The inhale and exhale pressures are the same until you have an apnea or hypopnea then it will start increasing pressure.

My suggestion is to set the PS Min = 3 or 4, which will give you a min IPAP of 7 or 8. Remember, it's the IPAP which controls inhalation. Changing your EPAP setting won't make any difference to your inhalation experience, but you may (or may not) need to set it a bit higher to counter obstructive events such as hypopneas and apneas.

DeepBreathing
Apnea Board Moderator
www.ApneaBoard.com


Bed

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
03-18-2016 09:05 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  [Equipment] respironics bipap error code hanasdaddy 6 70 Today 01:48 AM
Last Post: hanasdaddy
  Newbie - need advice on settings and machine tmc712 18 343 Yesterday 09:04 PM
Last Post: tmc712
Smile [Admin Note] PLEASE HELP Apnea Board if you can [new server coming] SuperSleeper 30 879 Yesterday 06:31 PM
Last Post: SuperSleeper
  Dream Station Ramp and Pressure Settings knowoneuno 21 432 Yesterday 11:00 AM
Last Post: Sleeprider
  New User Problem with BiPAP AVAPS - Part 1 jmayer180 35 1,296 11-27-2016 09:50 AM
Last Post: jmayer180
  Dream Station Settings Autumnseer 15 529 11-26-2016 07:01 AM
Last Post: Weird Tolkienish Figure
  First time Bipap User, years of suffering SleepyMersim 18 650 11-16-2016 11:46 PM
Last Post: Galactus

Forum Jump:

Who's Online (Complete List)