Help with Resmed S9 Adapt adjustments
I used a Resmed Airsense 10 for a year with AHI always 8-16 with 80-90% central apneas.
Recently started using a Resmed S9 Adapt at the default AutoASV settings with great improvement...typically my results look like the attached results below ..about an AHI 0f 1 with mostly hypopneas.
Do you recommend leaving at default settings or trying to improve from here? If so, what adj do you recommend I try? Raise minimum EPAP by 1cm perhaps?
RE: Help with Resmed S9 Adapt adjustments
Need to be careful not to remove the machines capability to do what it is designed to do, treat centrals.
I would try increasing Min PS by 1 so PS 4-15 then evaluate. This is because your Flow Limitations are high.
RE: Help with Resmed S9 Adapt adjustments
Thanks
Will wait a couple weeks to adjust to new machine and then try increasing PS slightly if needed.
Right now pretty comfortable with breathing and results seem pretty good so will proceed slowly with the changes to try to keep AHI under 1.
Thanks for advice