How to optimize the setting of Auto BiPaP? (PR DreamStation 700)
#1
How to optimize the setting of Auto BiPaP? (PR DreamStation 700)
I confirmed OSA(RDI: 35) through PSG test. So I bought a PR DreamStation 700(Auto BiPaP). But it doesn't seem to work very well.

How to optimize the setting of Auto BiPaP? 


my settings:
Mode: Auto Bi Level with Bi Flex
Min EPAP: 6.0 
Max IPAP: 25.0 
PS: 4.0-4.0 (range)
A-Flex: On



   

I would appreciate it if anyone could give me some advice. (and Sorry, my English is not very good.)

Post Reply Post Reply



Donate to Apnea Board  
#2
RE: How to optimize the setting of Auto BiPaP? (PR DreamStation 700)
Welcome to the forum.

50% of your AHI is Central Apnea.  Your Auto BiPap is not designed to treat Centrals.  Your Obstructive is also too high.

Increasing Pressure will typically increase Central Apneas.  Increasing PS (Pressure Support) will typically increase Central Apneas.

Your PS is currently set to 4 (PS = 4) I'm going to suggest reducing PS to 1 (PS = 1) to try to minimize the CA.  This may increase your Obstructive elements.
Fred Bonjour - Project Manager and Lead Tester for OSCAR - Open Source CPAP Analysis Reporter 
OSCAR

Download OSCAR
New to Apnea? Helpful tips to ensure success
Mask Primer
Dealing with a DME
Organize Charts
Attaching Charts

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#3
RE: How to optimize the setting of Auto BiPaP? (PR DreamStation 700)
I don't usually disagree with Fred, but I'm going a different direction. Please consider using the following settings:
EPAP min 8.0
PS min 2.0
PS max 3.0
IPAP max 14.0

The OA events in your current chart are from the exhale pressure being too low (4.0) and it seems to greatly improve in the range of 7-8 cm. I think most of your CA events are due to sleep disruption and not being accustomed to the therapy. I am suggesting a reduction of the pressure support (difference between IPAP and EPAP), but I think you will be more comfortable in the suggested range of 2-3. At this point I don't see any need for pressure above 14/11, so that is the basis for the maximum pressure. I assume you are using BiFlex, and I suspect at a setting of 3. Please turn it down to 1 or 2.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  6 weeks in on bipap, still centrals, still sleepy bemused 6 153 05-21-2019, 05:50 AM
Last Post: bonjour
  New bipap presciption Shamus54 15 417 05-12-2019, 11:00 AM
Last Post: Shamus54
  new to bipap, lots of central apneas bemused 32 816 05-10-2019, 06:37 PM
Last Post: ajack
  [Equipment] When to get a new BIPAP machine? SleepyLab2015 9 287 05-06-2019, 11:07 AM
Last Post: Sleeprider
  ResMed AirSense 10 [how to turn auto/on auto/off feature off?] IsrvChrist 3 161 05-03-2019, 08:25 PM
Last Post: Big Guy
Sad Dreamstation auto 2B compliance code Marshapnea 2 95 05-03-2019, 08:10 AM
Last Post: Marshapnea
  Help with Settings for Respironics System One (60 Series) RemStar BiPAP Auto wjblevins 12 275 05-01-2019, 01:11 PM
Last Post: wjblevins


New Posts   Today's Posts






About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.