How to work out the correct pressure required ?
#1
How to work out the correct pressure required ?
Hello, 

I am trying to find out what pressure will suit best to an individual. He was using an Devilbiss Intellipap machine and had set the minimum pressure to 5 (who knows why). Average AHI was around 8 ish (given the machine gives a code which needs to be looked up on Intellipaps website). So, we set his pressure to a min of 7 and the AHI has been averaging around 5.5 with a much better quality of sleep over the past few weeks. This weekend I lent my machine to him and with a min pressure of 7 we got an AHI of 4.29 - assuming I was on the right track - the next night we bumped it up to 8. The AHI dropped to 4.02. But other events popped up. I have attached both night's charts. I'd be grateful if you could please give your opinion on what would be the right way to self titrate and find out what works ? Sadly, I can only lend my machine over the weekend so we have only two days a week to give it a go.


https://imgur.com/a/BxMfj
Post Reply Post Reply



Donate to Apnea Board  
#2
RE: How to work out the correct pressure required ?
As long as the pressure is not bothering the person, I would suggest raising the pressure to 9. The AHI is dominated by hypopneas, indicating that the pressure is too low.
Sleepster
Apnea Board Moderator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#3
RE: How to work out the correct pressure required ?
Your second link didn't work out so good, but you're a member and could just embed the image. The first one is below, and if second one resembles this, the events are still all hypopnea with a couple OA and some snores. Typical for Dreamstation, this is obstructive. Either keep increasing pressure, or get a Resmed Airsense 10 Autoset (See Amazon at $440). The Philips is pretty reluctant to increase pressure but the Resmed gets it before the apnea happen. I would take this to 9.0-13 and see i it get a bit better.

[Image: znQRyVS.png]
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Optimizing Therapy
Organize your OSCAR Charts
How To Attach Images And Files to your posts
How To Deal With Equipment Supplier
Mask Primer
Beginner's Guide to Sleepyhead

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#4
RE: How to work out the correct pressure required ?
You can find information about how to set a machine here:
http://www.apneaboard.com/adjust-cpap-pr...tup-manual
That is also where you can get the clinician manual for his machine since you will need to know how to get into that menu to change the pressure.

And here is advice on how to change the pressure your/his self:
http://www.apneaboard.com/adjust-cpap-pr...re-on-cpap

Mostly you will need patience and, if the machine is data capable, you will need SleepyHead.
PaulaO

Take a deep breath and count to zen.




Post Reply Post Reply



Donate to Apnea Board  
#5
RE: How to work out the correct pressure required ?
Thanks for your replies.
Here are the charts again :

With min 7 : https://imgur.com/znQRyVS

With min 8 : https://imgur.com/W4bGZcl
Post Reply Post Reply
#6
RE: How to work out the correct pressure required ?
A comment. We will be able to dial your friend in on Your machine. You are using EPR of 2 on your machine. Though EPR is considered a comfort setting it does have an impact on therapy. You NEED to make sure the the DV machine is set up in a similar manner EZ Flex of 2? not sure of the DV term.

The best results will be obtained by looking at the Intelepap charts.

Fred
Fred Bonjour - Project Manager and Lead Tester for OSCAR - Open Source CPAP Analysis Reporter 
OSCAR

Download OSCAR
New to Apnea? Helpful tips to ensure success
Mask Primer
Dealing with a DME
Organize Charts
Attaching Charts

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Does the Contec cms50i work on El Capitan for Mac? Michaely6 10 418 03-30-2019, 10:44 AM
Last Post: Michaely6
  Is this image of Sleepyhead the correct way to post on forums? Also, how am I doing?! vroomvroom 21 697 03-17-2019, 08:44 AM
Last Post: Sleeprider
  My 12 year old daughter just started cpap now I can't work Sleepyhead LucysMom 15 766 02-26-2019, 06:58 PM
Last Post: Sleeprider
  Remstar Auto M Series "Service Required" msgill59 5 5,488 02-02-2019, 04:00 PM
Last Post: tommagic1
  Does this tube work with this machine? Dyssomniac 7 280 01-15-2019, 10:28 AM
Last Post: Sleeprider