I'll talk about the two modern machines I own and have taken apart and examined pretty thoroughly: the A10 Autoset and the DreamStation Auto.

Both machines are rated IP22. This means a production sample has passed a water dripping test where a small stream of water is directed on the top of the machine for 10 minutes with the machine angled at 15 degrees without becoming damaged or unsafe. Not submerged, sprayed, splashed, or sitting in standing water. This is a typical minimum requirement for an electrical product designed for indoor use.

None of the pc boards in either machine appear to be coated (other than soldermask, etc.) or waterproof. The boards and the components are fairly well electrically insulated and physically isolated from the moist airsteam inside the blower housing. The pressure and flow sensors (and a humidity/temp sensor in the DS) are angled to drain any water back down into the housing. You would expect surface tension to hold a small amount of water in the sensor chambers, but I suspect that doesn't matter.

Both machines use 3-phase brushless pm motors, so no brush arcing. In the A10, the insulated motor windings are exposed to the air in the blower housing. In the DS, the motor looks to be sealed and isolated from the impeller. Both probably have sealed bearings or bushings. Electrically speaking, the blowers can probably tolerate a fair amount of water.

Since an intact and functional blower housing assembly is airtight, except for the inlet and outlet, dumping water into it from the water tank will probably not wet the pc board. I am concerned about physical damage to the spinning impeller if a substantial amount of water hits it. Also, impeller startup may be difficult with a water load. The motor control circuit may handle the overload without problems. So if you get water in the blower, drain it and let it dry some.

Obviously, if you wet down a powered pc board it's probably going to fail. If you wet down an unpowered board, you can try drying it off with a warm hair dryer. Water purity and the moisture sensitivity of the components will probably determine if it powers up successfully.

The humidifier heating plates on both models have seals and are probably safe from water damage. I haven't tested this.

The power entry module, on both models, is located on the lower part of the plastic case. It consists of a jack, small pc board, and some wires that lead to the main board. The A10 jack receives 24 volts dc and the DS receives 12 volts dc from their double-insulated power supplies. The position of the jacks makes them vulnerable to pooled or standing water.

The thread mentioned by Frank and started by Possum was a case where Possum's A10 power plug and jack were exposed to standing water. The plastic parts inside the plug melted and the machine quit working. I suspect the 24 volts (4 amp capable) boiled the water and melted the plastic. Not really sure of the source of the water or its purity.

It stands to reason, if you are using an electrical device and it's got water in it, you need to pay attention to spills and periodically inspect it for leaks. Hopefully, I've identified some potential problem areas.

I know most of this is obvious. It's just my $0.02.

Just my personal opinion. My posts are not medical advice or a statement of fact. Please consult a qualified physician or other qualified medical personnel. Please comply with all applicable laws, codes, regulations, and protocols.
(This post was last modified: 04-18-2016 02:13 AM by sdb7802.)
04-18-2016 12:43 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
FrankNichols Offline

Advisory Members

Posts: 714
Joined: Mar 2016

Machine: AirSense 10 Autoset
Mask Type: Full face mask
Mask Make & Model: Resmed F10 and/or P10
Humidifier: Builtin
CPAP Pressure: 17-19
CPAP Software: ResScan SleepyHead Other Software

Other Comments: I've figured out how to get my AHI down to 5, I quit sleeping!

Sex: Male
Location: Live Oak, Florida

Post: #22
RE: Humidifiers not safe - seriously?
(04-18-2016 12:43 AM)sdb7802 Wrote:  I'll talk about the two modern machines I own and have taken apart and examined pretty thoroughly: the A10 Autoset and the DreamStation Auto.

Both machines are rated IP22. This means a production sample has passed a water dripping test where a small stream of water is directed on the top of the machine for 10 minutes with the machine angled at 15 degrees without becoming damaged or unsafe. Not submerged, sprayed, splashed, or sitting in standing water. This is a typical minimum requirement for an electrical product designed for indoor use.

None of the pc boards in either machine appear to be coated (other than soldermask, etc.) or waterproof. The boards and the components are fairly well electrically insulated and physically isolated from the moist airsteam inside the blower housing. The pressure and flow sensors (and a humidity/temp sensor in the DS) are angled to drain any water back down into the housing. You would expect surface tension to hold a small amount of water in the sensor chambers, but I suspect that doesn't matter.

Both machines use 3-phase brushless pm motors, so no brush arcing. In the A10, the insulated motor windings are exposed to the air in the blower housing. In the DS, the motor looks to be sealed and isolated from the impeller. Both probably have sealed bearings or bushings. Electrically speaking, the blowers can probably tolerate a fair amount of water.

Since an intact and functional blower housing assembly is airtight, except for the inlet and outlet, dumping water into it from the water tank will probably not wet the pc board. I am concerned about physical damage to the spinning impeller if a substantial amount of water hits it. Also, impeller startup may be difficult with a water load. The motor control circuit may handle the overload without problems. So if you get water in the blower, drain it and let it dry some.

Obviously, if you wet down a powered pc board it's probably going to fail. If you wet down an unpowered board, you can try drying it off with a warm hair dryer. Water purity and the moisture sensitivity of the components will probably determine if it powers up successfully.

The humidifier heating plates on both models have seals and are probably safe from water damage. I haven't tested this.

The power entry module, on both models, is located on the lower part of the plastic case. It consists of a jack, small pc board, and some wires that lead to the main board. The A10 jack receives 24 volts dc and the DS receives 12 volts dc from their double-insulated power supplies. The position of the jacks makes them vulnerable to pooled or standing water.

The thread mentioned by Frank and started by Possum was a case where Possum's A10 power plug and jack were exposed to standing water. The plastic parts inside the plug melted and the machine quit working. I suspect the 24 volts (4 amp capable) boiled the water and melted the plastic. Not really sure of the source of the water or its purity.

It stands to reason, if you are using an electrical device and it's got water in it, you need to pay attention to spills and periodically inspect it for leaks. Hopefully, I've identified some potential problem areas.

I know most of this is obvious. It's just my $0.02.

Thank you very much, I appreciate the information.

I am not a Medical professional and I don't play one on the internet.
Started CPAP Therapy April 5, 2016
I'd Rather Be Sleeping
04-18-2016 07:09 AM
Find all posts by this user Post Reply Quote this message in a reply
archangle Offline
Wiki Editor
Advisory Members

Posts: 3,159
Joined: Feb 2012

Machine: ResMed S9 AutoSet
Mask Type: Nasal pillows
Mask Make & Model: ResMed Swift FX
Humidifier: ResMed S9 H5i
CPAP Pressure: 16-20
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments: Happy PAPper

Sex: Undisclosed
Location: USA

Post: #23
RE: Humidifiers not safe - seriously?
(04-17-2016 06:24 AM)OMyMyOHellYes Wrote:  You guys give them too much credit for the power brick. It is a bug - not a feature. Used to be, they put the power converter inside the unit. That was a good thing because bricks are just laziness on the part of manufacturers. The reduce the perceived foot print (they don't really because you still have to have one and that adds a separate footprint to the tabletop or somewhere else), and make it easier to assemble the unit because they don't have to physically put that power supply in the unit. But in the end, it is the consumer that has to deal with two separate boxes.

I disagree. There are several advantages to the external power brick. It makes it easier to handle the heat. It makes it safer. It means that you can replace the brick instead of the whole machine if the power supply fails, and power supplies are historically a high failure rate subassembly. It does make the part that goes on your table smaller and lighter.

It does have the disadvantage of there being another part to mess with, but it's not entirely a dumb idea.

Get the free SleepyHead software here.
Useful links.
Click here for information on the main alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check it yourself.
04-18-2016 04:00 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
archangle Offline
Wiki Editor
Advisory Members

Posts: 3,159
Joined: Feb 2012

Machine: ResMed S9 AutoSet
Mask Type: Nasal pillows
Mask Make & Model: ResMed Swift FX
Humidifier: ResMed S9 H5i
CPAP Pressure: 16-20
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments: Happy PAPper

Sex: Undisclosed
Location: USA

Post: #24
RE: Humidifiers not safe - seriously?
[Image: 58a7f450a07c012f2fe600163e41dd5b]

If you spill the water and ruin your CPAP, the manufacturers make extra profit. It would also slightly increase the cost.

It wouldn't be hard to make the blower unit spill proof. Separate the electronics from the air path such that water going back from the humidifier wouldn't get into the electronics. You wouldn't even have to hermetically seal the electronics, or make them waterproof, but even that isn't that expensive. You'd need to make the motor water safe, by a barrier between the airflow and the motor, or making the motor waterproof. Waterproof bearings are no problem.

The sensors in the airflow are another problem. The pressure, air flow, and humidity sensors could be made reasonably water resistant without too much difficulty.

All of that would cost a little, but not much in terms of the cost vs. benefits it gives to the customer.

Another point, a number of people have reported turning over their machines and ruining them with the water. I'm guessing that about 10% of them kill the machine when it happens.

Get the free SleepyHead software here.
Useful links.
Click here for information on the main alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check it yourself.
04-18-2016 04:23 PM
Find all posts by this user Post Reply Quote this message in a reply
OMyMyOHellYes Offline

Advisory Members

Posts: 915
Joined: Mar 2013

Machine: Respironics 560 Auto
Mask Type: Nasal mask
Mask Make & Model: Resmed Mirage FX
Humidifier: Respironics System One (no heated hose)
CPAP Pressure: 08.0-15.0 cm/H2O
CPAP Software: EncoreBasic

Other Comments:

Sex: Male
Location: Texas

Post: #25
RE: Humidifiers not safe - seriously?
It would be so simple if you guys just quit using humidification .......
04-18-2016 04:46 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
FrankNichols Offline

Advisory Members

Posts: 714
Joined: Mar 2016

Machine: AirSense 10 Autoset
Mask Type: Full face mask
Mask Make & Model: Resmed F10 and/or P10
Humidifier: Builtin
CPAP Pressure: 17-19
CPAP Software: ResScan SleepyHead Other Software

Other Comments: I've figured out how to get my AHI down to 5, I quit sleeping!

Sex: Male
Location: Live Oak, Florida

Post: #26
RE: Humidifiers not safe - seriously?
(04-18-2016 04:23 PM)archangle Wrote:  [Image: 58a7f450a07c012f2fe600163e41dd5b]

If you spill the water and ruin your CPAP, the manufacturers make extra profit. It would also slightly increase the cost.

It wouldn't be hard to make the blower unit spill proof. Separate the electronics from the air path such that water going back from the humidifier wouldn't get into the electronics. You wouldn't even have to hermetically seal the electronics, or make them waterproof, but even that isn't that expensive. You'd need to make the motor water safe, by a barrier between the airflow and the motor, or making the motor waterproof. Waterproof bearings are no problem.

The sensors in the airflow are another problem. The pressure, air flow, and humidity sensors could be made reasonably water resistant without too much difficulty.

All of that would cost a little, but not much in terms of the cost vs. benefits it gives to the customer.

Another point, a number of people have reported turning over their machines and ruining them with the water. I'm guessing that about 10% of them kill the machine when it happens.

Ah, a man after my own heart Smile

I am not a Medical professional and I don't play one on the internet.
Started CPAP Therapy April 5, 2016
I'd Rather Be Sleeping
04-18-2016 04:53 PM
Find all posts by this user Post Reply Quote this message in a reply
OMyMyOHellYes Offline

Advisory Members

Posts: 915
Joined: Mar 2013

Machine: Respironics 560 Auto
Mask Type: Nasal mask
Mask Make & Model: Resmed Mirage FX
Humidifier: Respironics System One (no heated hose)
CPAP Pressure: 08.0-15.0 cm/H2O
CPAP Software: EncoreBasic

Other Comments:

Sex: Male
Location: Texas

Post: #27
RE: Humidifiers not safe - seriously?
(04-18-2016 04:53 PM)FrankNichols Wrote:  Ah, a man after my own heart Smile

How much do you think the company could get for it?
04-18-2016 05:42 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
FrankNichols Offline

Advisory Members

Posts: 714
Joined: Mar 2016

Machine: AirSense 10 Autoset
Mask Type: Full face mask
Mask Make & Model: Resmed F10 and/or P10
Humidifier: Builtin
CPAP Pressure: 17-19
CPAP Software: ResScan SleepyHead Other Software

Other Comments: I've figured out how to get my AHI down to 5, I quit sleeping!

Sex: Male
Location: Live Oak, Florida

Post: #28
RE: Humidifiers not safe - seriously?
(04-18-2016 05:42 PM)OMyMyOHellYes Wrote:  
(04-18-2016 04:53 PM)FrankNichols Wrote:  Ah, a man after my own heart Smile

How much do you think the company could get for it?

My heart? Probably not much - it's been pumping now for over 60 years Smile

I just got back from my Cardiologist for the followup to my latest echocardiogram and he says it's still running on all 8 cylinders - ie. excellent, which just made my day at 66 Smile

I am not a Medical professional and I don't play one on the internet.
Started CPAP Therapy April 5, 2016
I'd Rather Be Sleeping
04-18-2016 05:58 PM
Find all posts by this user Post Reply Quote this message in a reply
Mosquitobait Offline

Advisory Members

Posts: 746
Joined: Apr 2015

Machine: Resmed AirSense 10 For Her
Mask Type: Nasal pillows
Mask Make & Model: Airfit P10
Humidifier: integrated, Climate line tube
CPAP Pressure: 9-12
CPAP Software: ResScan SleepyHead

Other Comments: RLS and Bradycardia

Sex: Female
Location: Minneapolis

Post: #29
RE: Humidifiers not safe - seriously?
I've pulled my machine to the floor 4 times so far. 3 times were not at home. On the Airsense, my empirical evidence indicates that the humidifer hits the floor first, so any water spillage is AWAY from the machine.

I'm thinking of buying a Hose Buddy.
04-18-2016 09:55 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
palerider Offline

Preferred Members-

Posts: 440
Joined: Nov 2014

Machine: s9 vpap auto
Mask Type: Nasal pillows
Mask Make & Model: p10
Humidifier: none
CPAP Pressure: 18/13+
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Dallas(ish)

Post: #30
RE: Humidifiers not safe - seriously?
(04-18-2016 05:58 PM)FrankNichols Wrote:  it's still running on all 8 cylinders -

good thing you're not supposed to have a v12 Big Grin
04-19-2016 12:03 AM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  ResMed 10 - DISHWASHER SAFE Humidifier Padre_StevieJ 5 1,131 10-05-2016 11:03 PM
Last Post: PaytonA
  Any CPAP users that don't use humidifiers. To use or not? Gabby 49 8,274 08-26-2016 07:19 AM
Last Post: Sleepster
  [Equipment] Changing setting on Remstar Auto A-Flex? Safe to do? pwynnnorman 5 506 07-23-2016 06:50 AM
Last Post: pwynnnorman
  Mask liners, humidifiers & AHI y_not 4 724 02-07-2016 12:28 PM
Last Post: y_not
  Are over the counter sleep aids safe when on xPAP? Mike91326 21 3,499 02-04-2015 09:25 PM
Last Post: Jim Bronson
  is it safe to use CPAP given to me by friend? budm 4 1,041 12-01-2014 12:48 AM
Last Post: PaulaO2
  New humidifiers gradually evaporating les water apnea711 17 3,891 04-19-2014 10:31 AM
Last Post: apnea711

Forum Jump:

Who's Online (Complete List)