RE: I have the cpap model...not the elite
49er, WADR to your opinion, you are complicating matters. She has had fitful sleep for about 4 hours a night, if that much, reported before and after CPAP. She is at a relatively low pressure. She is new to all this and has not adapted to it yet. Let's focus on the forest and not the trees. Its not necessary to further complicate the matter where the solution appears obvious- cure the sleep deprivation. I mean, no wonder OP feels like cr*p, and her reported cognitive deficit may be related to not using the machine enough.
(This post was last modified: 07-11-2015 03:40 PM by tedburnsIII.)
07-11-2015 03:31 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
quiescence at last Offline

Advisory Members

Posts: 753
Joined: Nov 2014

Machine: REMstar System One with Autoflex
Mask Type: Full face mask
Mask Make & Model: Quattro AirFit F10
Humidifier: yes typical setting = 4
CPAP Pressure: 5-13 Autotritrating CPAP
CPAP Software: SleepyHead EncoreBasic

Other Comments: CMS-50F Plus oximeter /w SPO2 Assistant software. Avowed mouth-breather switching to nose-breathing

Sex: Male
Location: US

Post: #32
RE: I have the cpap model...not the elite
dittos on comments above -- it is much better to have greater data capability, not only for you, but for the sleep doc. as zonk says, [if you have a problem, you'll have the data to use to analyze it] and I'll add that the doc [if you have one that will analyze the data] will obtain the benefit as well.

let's assume for a moment that you do not get the responsible response you hope for from the doctor.
- in that case you can learn to use the one you have to the greatest extent (you will likely get, as noted above, significantly better sleep with the machine you now have than without a machine.)
- or buy a second-hand data capable one from reputable suppliers (some identified on this forum).

On the learning how to use this one, two basic suggestions (besides getting the setup manual):

1. keep diary of how you feel.
2. lying on side is usually better than lying on the back. [lying on left side usually better than right - more often stated in the community].

Do not worry about only getting a few hours sleep at a time. a lot of us do two-sleep (or more) getting up and reading/doing light conversation/crossword puzzles for up to 2 hours before doing another 3 to 4 hour sleep. I average 6.25 hours sleep for normal 8 hour night (after 8 months).

Good luck,

QAL

Dedicated to QALity sleep.
You'll note I am listed as an Advisory Member. I am honored to be listed as such. See the fine print - Advisory Members as a group provide advice and suggestions to Apnea Board administrators and staff concerning Apnea Board operation and administrative policies. Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment.
07-11-2015 03:41 PM
Find all posts by this user Post Reply Quote this message in a reply
tedburnsIII Offline

Members

Posts: 196
Joined: Mar 2015

Machine: 3B RESmart APAP
Mask Type: Nasal mask
Mask Make & Model: Wisp nasal; ResMed FX nasal
Humidifier: came with unit-I vary it, currently set at '3', '5' is max
CPAP Pressure: 10 cm CPAP
CPAP Software: Other Software

Other Comments: Dx moderately severe OSA. Overnight pulse oximeter-CMS50D Plus

Sex: Male
Location: San Diego

Post: #33
RE: I have the cpap model...not the elite
Data this, data that...people are suggesting complicating the matter for a person who is clearly, IMO, overwhelmed by the process.

In this case, and at this time, KISS principle is the better path.
07-11-2015 04:00 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
49er Offline

Advisory Members

Posts: 237
Joined: Apr 2012

Machine: VPAP-S
Mask Type: Nasal mask
Mask Make & Model: Sleep Weaver Elan Soft Cloth Nasal Mask
Humidifier: H5i
CPAP Pressure: 13.6/9.6
CPAP Software: SleepyHead

Other Comments:

Sex: Undisclosed
Location:

Post: #34
RE: I have the cpap model...not the elite
(07-11-2015 03:31 PM)tedburnsIII Wrote:  49er, WADR to your opinion, you are complicating matters. She has had fitful sleep for about 4 hours a night, if that much, reported before and after CPAP. She is at a relatively low pressure. She is new to all this and has not adapted to it yet. Let's focus on the forest and not the trees. Its not necessary to further complicate the matter where the solution appears obvious- cure the sleep deprivation. I mean, no wonder OP feels like cr*p, and her reported cognitive deficit may be related to not using the machine enough.

Ted,

Without a fully data capable machine, a person is flying blind with therapy just like if someone who had high blood pressure or diabetes didn't have a bp monitor/glucometer that was fully functioning. Since Dawn still has time to exchange the machine, now is the time to do it before it is too late. And even if it turns out that nothing is complicating Dawn's sleep other than simply getting used to the machine, there might come a time where she runs into a problem where having a fully data capable machine will be crucial.
07-11-2015 04:04 PM
Find all posts by this user Post Reply Quote this message in a reply
OpalRose Offline

Monitors

Posts: 3,168
Joined: Nov 2014

Machine: PR System One REMstar Auto 560 with A Flex
Mask Type: Nasal pillows
Mask Make & Model: AirFit P10 For Her
Humidifier: REMStar heated humidifier with heated hose
CPAP Pressure: Auto Cpap 10-13 AFLEX 1
CPAP Software: SleepyHead EncoreBasic

Other Comments: Started CPAP Therapy October 23, 2014

Sex: Female
Location: Northeast Ohio, USA

Post: #35
RE: I have the cpap model...not the elite
(07-11-2015 02:32 PM)mzdawn74 Wrote:  
(07-11-2015 06:09 AM)OpalRose Wrote:  mzdawn74,
This nonsense is typical of what DME's do to their customers. Annoyed-and-disappointed This same thing happened to me too.

You have a right to get the machine you want. Ask your doctor to write a script for an Auto CPAP, and it should state patient choice. Also ask him to prescribe a pressure range so there is no mistake with the DME. Tell your DME that you want a "new" machine", with no previous machine hours on it. They typically try to pawn off a used machine that someone else returned.

Do this ASAP, and tell the doctor you will be in to pick up the script. Don't take no for an answer.

As others have said, you will be stuck with this machine for 5 years, and you have no idea what your future needs will be.

Thank you for the advice. I know my doctor is going to try and debate with me just as the DME rep did. What do I say as to why I want to change machines other than it offers me the ability to track my therapy better? I need help explaining myself. I'm sorry if I sound dumb, but like I said earlier I am really not very knowledgeable in this area. I've tried reading and learning what I can, but with my brain not functioning properly its very difficult to say the least. Dont-know


Ok, tell your doctor that you want to be proactive in your treatment, and that the machine they gave you will not let you do that since you cannot see the data.

He is going to say you are fine because of the low AHI! Tell him that you feel a fully data capable Auto CPAP will allow you not only to tract your data, but meet any future needs you may have if a pressure change is needed. It shouldn't matter to him or her what machine you get. You know your doctor better than anyone else, so be firm, but make him or her feel that you will be able to meet compliance easier if you can tract your own data, and you can print out reports for him or her to look over.

The DME can do nothing but follow doctor's script.

OpalRose
07-11-2015 04:06 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
tedburnsIII Offline

Members

Posts: 196
Joined: Mar 2015

Machine: 3B RESmart APAP
Mask Type: Nasal mask
Mask Make & Model: Wisp nasal; ResMed FX nasal
Humidifier: came with unit-I vary it, currently set at '3', '5' is max
CPAP Pressure: 10 cm CPAP
CPAP Software: Other Software

Other Comments: Dx moderately severe OSA. Overnight pulse oximeter-CMS50D Plus

Sex: Male
Location: San Diego

Post: #36
RE: I have the cpap model...not the elite
You will find no argument from me that having more data available is beneficial (more to those here than the usual CPAP patient-look at our small numbers), but focus of the thread seems to distract from her issue.

If she can get a different machine without consequential loss, APAP with full data, then do so. But I was under the impression her machine was data-capable, maybe not so much as others, though. The fact her AHI numbers were stellar led me to believe that the issue is her adaptability to the whole process, not that APAP is the solution to her woes.

It appears that OP is overwhelmed and close to dropping out. In this precarious situation, it might be better not to complicate things with all this learning curve jargon and change of machine if she does not address the problem now, not with another machine, but dealing with her present one.

The best advice I can give OP presently is to see her doctor about placing her on prescription 'Z' drug, e.g, Ambien, to help her get over the hump. Not using her machine for nearly a week is not at all conducive to her success.
(This post was last modified: 07-11-2015 04:26 PM by tedburnsIII.)
07-11-2015 04:10 PM
Find all posts by this user Post Reply Quote this message in a reply
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,340
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #37
RE: I have the cpap model...not the elite
The AHI appears to be great, although it might not be a complete picture, she is clinically well treated. I think in discussing this with the doctor, I would not go over the AHI numbers. Honestly, they are among the best I've seen other than the hours of use. The quality of sleep is a more important matter. APAP is not a miracle. You will need to adjust regardless of what machine you have.

The main advantage of an Elite or Autoset is that you can be a participant in your treatment rather than just a patient. That can make the difference between success and failure in some people. CPAP is vitally important, and the more that can be done to encourage patient success, at minimal additional cost, the better. If it's important to you, it should be important to your doctor. We're not talking about a luxury here; just someone that wants to succeed and have a voice in their care.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
07-11-2015 04:46 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
tedburnsIII Offline

Members

Posts: 196
Joined: Mar 2015

Machine: 3B RESmart APAP
Mask Type: Nasal mask
Mask Make & Model: Wisp nasal; ResMed FX nasal
Humidifier: came with unit-I vary it, currently set at '3', '5' is max
CPAP Pressure: 10 cm CPAP
CPAP Software: Other Software

Other Comments: Dx moderately severe OSA. Overnight pulse oximeter-CMS50D Plus

Sex: Male
Location: San Diego

Post: #38
RE: I have the cpap model...not the elite
(07-11-2015 04:04 PM)49er Wrote:  
(07-11-2015 03:31 PM)tedburnsIII Wrote:  49er, WADR to your opinion, you are complicating matters. She has had fitful sleep for about 4 hours a night, if that much, reported before and after CPAP. She is at a relatively low pressure. She is new to all this and has not adapted to it yet. Let's focus on the forest and not the trees. Its not necessary to further complicate the matter where the solution appears obvious- cure the sleep deprivation. I mean, no wonder OP feels like cr*p, and her reported cognitive deficit may be related to not using the machine enough.

Ted,

Without a fully data capable machine, a person is flying blind with therapy just like if someone who had high blood pressure or diabetes didn't have a bp monitor/glucometer that was fully functioning. Since Dawn still has time to exchange the machine, now is the time to do it before it is too late. And even if it turns out that nothing is complicating Dawn's sleep other than simply getting used to the machine, there might come a time where she runs into a problem where having a fully data capable machine will be crucial.

I don't believe that to be true. She has AHI reporting, and certainly more data is better, but not 'flying blind' without fully data capable, SH, etc. That is for the afficionados and perfectionists who have the time and more likely to optimize their therapy. The general population benefits from the therapy, but perhaps not as much as some of us here.

I attribute much of my success thus far to being here and also to monitoring my data, which keeps me in the loop as to my progress or decline. But not everyone, especially those who are working full-time, have the desire, inclination, time or motivation to be so intimately involved with their therapy.
(This post was last modified: 07-11-2015 06:21 PM by tedburnsIII.)
07-11-2015 05:06 PM
Find all posts by this user Post Reply Quote this message in a reply
zonk Offline

Advisory Members

Posts: 7,908
Joined: Feb 2012

Machine: A10 AutoSet
Mask Type: Nasal mask
Mask Make & Model: Activa LT
Humidifier: Integrated /ClimateLineAir
CPAP Pressure: 9/13
CPAP Software: ResScan

Other Comments: CPAP since Nov 2010

Sex: Male
Location: Australia

Post: #39
RE: I have the cpap model...not the elite
(07-11-2015 04:04 PM)49er Wrote:  Since Dawn still has time to exchange the machine, now is the time to do it before it is too late.
+1 ... strike while the iron is hot

General comments raised in this thread, not a direct response to the OP
The fact of the matter insurance is not some charity organization, handing out CPAP to the needy
Rather they're in the business of making money $$$
A little bird told me ... insurance pay the same amount $$$ for the brick or the AutoSet

Good DME, dispense fully data capable machines, while lousy counterpart make few dollars more by dispensing bricks
So, given a choice ... why anyone choose the brick over the AutoSet. After all, its their own money buying the machine

The fact remain, there is no diagnoses of sleep apnea or anything else without data
Data definition: facts and statistics collected together for reference or analysis

xPAP does not treat anything else beside sleep apnea, might have some healing effects on general health and well being
Oxygen starvation and sleep deprivation is deadly combo

Oh, CPAP makers ... discontinue the bricks, cost few cents more to turn brick into top-notch machine
Shame on you
07-11-2015 05:25 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
tedburnsIII Offline

Members

Posts: 196
Joined: Mar 2015

Machine: 3B RESmart APAP
Mask Type: Nasal mask
Mask Make & Model: Wisp nasal; ResMed FX nasal
Humidifier: came with unit-I vary it, currently set at '3', '5' is max
CPAP Pressure: 10 cm CPAP
CPAP Software: Other Software

Other Comments: Dx moderately severe OSA. Overnight pulse oximeter-CMS50D Plus

Sex: Male
Location: San Diego

Post: #40
RE: I have the cpap model...not the elite
Some people prefer all the bells and whistles, others don't really give a hoot. I would've obtained an arguably 'better' machine myself that were compatible with SH, if I had to do it all over again. Otherwise, my machine I believe to be on a nearly equal level with those that are twice as costly.

That's just me. I was ignorant and blindly accepted the machine DME provided, but my APAP machine is data-capable, and the software is quite adequate. My only regret is that it's not compatible with SH, but that is not a fatal flaw as far as I am concerned. My dissatisfaction there is not enough to sufficiently matter- to me! And my machine has not caused one problem, thus far. YMMV.

Nevertheless, a brick will do its job if you do, though I agree it would overall be better to do away with them.

Not that it really matters to those here, my machine is NOT a brick, although it may 'look' like one.
(This post was last modified: 07-11-2015 05:57 PM by tedburnsIII.)
07-11-2015 05:42 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  [Equipment] Water tub maintenance for ResMed S9 Elite Maggie 9 366 10-30-2016 01:53 PM
Last Post: Maggie
  ResMed Model 10 water Tank Jackie999 5 433 09-29-2016 07:30 AM
Last Post: lab rat
  Respironics Bipap S/T Gray Model - COPD Set Up Question? bowiefan 2 255 09-26-2016 11:39 AM
Last Post: Sleeprider
  CONFIGURATION RESMED S9 ESCAPE MODEL ziozeb 4 354 09-24-2016 08:04 PM
Last Post: pupcamper
Question [CPAP] Adjusting Pressure on Respironics Model 362505 RobertK 20 4,383 09-19-2016 01:28 PM
Last Post: RobertK
  Respironics Bipap ST (gray model) setting question bowiefan 3 395 09-05-2016 10:32 AM
Last Post: AUDavidSydney
  Resmed s9 elite stats jxyzobrien 0 229 08-19-2016 04:32 PM
Last Post: jxyzobrien

Forum Jump:

Who's Online (Complete List)