I took apart my Philips Dreamstation and replaced the foam chamber. Now it's loud!
#1
I took apart my Philips Dreamstation and replaced the foam chamber. Now it's loud!
I also found this O-ring laying on the bottom of the cpap machine. Where does it go and could this have something to do with the louder noise?
Others have claimed that after they removed the foam it wasn't any louder. 

Pic is the ring I found

[Image: TqicZr3l.jpeg]

https://imgur.com/a/dNDnxZI
Post Reply Post Reply






Donate to Apnea Board  
#2
RE: I took apart my Philips Dreamstation and replaced the foam chamber. Now it's loud!
where your o-ring goes (see picture attachment) - that o ring seals your pressure and it will be loud without it.  my machine is hardly noticeable louder at all (7 months of experience)


Attached Files Thumbnail(s)
   
Post Reply Post Reply
#3
RE: I took apart my Philips Dreamstation and replaced the foam chamber. Now it's loud!
(02-04-2022, 10:53 PM)fredHa Wrote: where your o-ring goes (see picture attachment) - that o ring seals your pressure and it will be loud without it.  my machine is hardly noticeable louder at all (7 months of experience)

Thank you so much!! No one on reddit knew this.
Okay so I need to take apart the cpap machine again? Trying to figure out how to get to that spot.

(02-04-2022, 10:53 PM)fredHa Wrote: where your o-ring goes (see picture attachment) - that o ring seals your pressure and it will be loud without it.  my machine is hardly noticeable louder at all (7 months of experience)

https://imgur.com/a/xJISwhW

Am I close?
Post Reply Post Reply
#4
RE: I took apart my Philips Dreamstation and replaced the foam chamber. Now it's loud!
you have to go at least this far: (from my notes when i cut the foam out of mine in july)

3. Remove POWER CORD and HUMIDIFIER if equipped
4. Remove Cover FLIP DOORS
5. Remove MODEM, External Filter Assembly, AND SIM CARD
6. Remove 2 screws from the UPPER ENCLOSURE in the external door wells
7. Separate UPPER ENCLOSURE from BACK PANEL by releasing left and right clips through
the BACK PANEL and free the BACK PANEL but still clipped to the 8 pin and connector
8. Separate UPPER ENCLOSURE from the FRONT PANEL
9. Remove 2 hold down screws for the PCA board
10. Separate 8 pin and 3 pin connector from the PCA board (DO NOT pull on wires)
11. Set aside the BACK PANEL which has the (DC) power supply origin of the 8 pin connector

you will likely need to at least loosen the screws holding down the BLOWER BOX ASSEMBLY to get some freedom of movement to get the motor outlet nozzle to engage the BACK PANEL outlet and the missing o- ring - then close the ds1 back up in reverse order
Post Reply Post Reply






Donate to Apnea Board  
#5
RE: I took apart my Philips Dreamstation and replaced the foam chamber. Now it's loud!
Okay I got the o ring on and out it back together. I guess I did something wrong as when I try to turn it back on the power button fades in and out and it's beeping at me....
Post Reply Post Reply
#6
RE: I took apart my Philips Dreamstation and replaced the foam chamber. Now it's loud!
I figured it out!!! And it's so much quieter!!!
THANK YOUUUU KIND SIR/MA'AM!
Post Reply Post Reply
#7
RE: I took apart my Philips Dreamstation and replaced the foam chamber. Now it's loud!
https://imgur.com/a/SutQY1k

Okay it's all set up. This tube is OK to use humidifier with right?
Post Reply Post Reply






Donate to Apnea Board  
#8
RE: I took apart my Philips Dreamstation and replaced the foam chamber. Now it's loud!
assuming by "tube" you mean a standard oem philips tube that goes from the main device to the mask- sure they are made to use with the humidifier.

these tubes come in at least 3 diameters and one type is heated.  the diameters and type are selected by the "tube type" menu.  notes on the "tube type" description-

 "This setting allows you to select the correct size diameter tubing that you are
using with the device. You can choose either (22) for the Philips Respironics 22
mm tubing, (15) for the Philips Respironics 15 mm tubing, or (12) for the optional
Philips Respironics 12 mm tubing. When using Heated Tubing, the device will
automatically change this setting to the appropriate tubing type (15H) and you
will not be able to change it.
Note: Tubing is identified on the cuff with the tubing identifier symbol: “12”,
“15”, or “15H”. 22 mm tubing contains no symbol" - philips.





"
Post Reply Post Reply
#9
RE: I took apart my Philips Dreamstation and replaced the foam chamber. Now it's loud!
I replaced the air chamber on my dreamstation also and accidentally damaged the front panel ribbon cable. Does anyone know if it's possible to get a replacement?
Post Reply Post Reply
#10
RE: I took apart my Philips Dreamstation and replaced the foam chamber. Now it's loud!
accidentally damaged the front panel ribbon cable. Does anyone know if it's possible to get a replacement? 

this part is called a "UI Panel Assembly 1115538" - source philips service and technical manual
the assembly consists of the ribbon+ glass panel+ knob?" according to manual picture

if you are willing to take your own risk with the following (dont blame me) this company sources this part according to their web site:

-LINK REMOVED: Search for Respironics UI Panel Assembly , Dreamstation · Respironics Model: 1115538 at MedExSupply 
› respiratory-humidifier...
Respironics UI Panel Assembly , Dreamstation · Respironics Model: 1115538 ·

$38.95 · ‎In stock

the ribbon "latches" to the pca board per the attached picture


-----
Moderator Action: DME Link Removed.

To maintain our status as an educational organization, links to DME-owned or DME-operated websites and links to pages that sell prescription-required items are prohibited in forum posts.  This is stated in the Apnea Board Rules.

-----
Post Reply Post Reply






Donate to Apnea Board  


Possibly Related Threads...
Thread Author Replies Views Last Post
Exclaimation [News] RECALL THREAD-- IMPORTANT PHILIPS DREAMSTATION & SYSTEM ONE USERS Sleeprider 2,264 214,859 08-12-2022, 02:21 PM
Last Post: chasmm
  NOTICE: Philips Respironics Recall Replacement Tally RayBee 100 12,025 08-01-2022, 11:40 AM
Last Post: RayBee
Smile Philips Dreamstation 1 how many wAH? BigMikes901Adventures 5 274 07-11-2022, 07:13 AM
Last Post: sawinglogz
  SoClean sues Philips [updated] Jeff8356 9 1,049 07-05-2022, 09:11 AM
Last Post: FirstNoele
  Philips system one and heated tubing? rubbertree 3 194 07-04-2022, 05:22 PM
Last Post: rubbertree
  UARS Treatment [Using Philips ASV] SenatorBirch 15 620 06-26-2022, 01:02 PM
Last Post: sawinglogz
  Foam Removal - Philips System One peewee44 38 5,743 06-22-2022, 07:25 PM
Last Post: Peach25


New Posts   Today's Posts






About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.