If one has A10 Autoset Why adjust ranges?
Hi Folks,
So I've gathered that even with an Autoset, there is some advantage to getting to where one get's one's range parameters set a little narrower rather than the wide 4-20. Is that correct?

Why? I have heard that big swings can trigger Centrals....is that so?

What else?

I had my worst night yet on night 23. Felt like I was fighting with the process all ngith long, had highest numbers in time...had been having events stay mostly under 23 seconds and mostly in the low teens, last night had some late 20's and early 30's in seconds. The usual combo of about 2/3 OSA 1/3 Centals.

Woke with Jaw really sore. (nasal pillows, no chin strap, only mild occasional mouth open and am trying to teach self to keep it shut and use tongue seal as well. Maybe I over tried?

I know, you need my charts, I'm still not figuring out the sharing despite the best helpers. I'll get there...

So why narrow numbers once one knows the outliers and range?

Thanks in Advance,
Susan Huh
01-19-2015 06:30 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
quiescence at last Offline

Advisory Members

Posts: 753
Joined: Nov 2014

Machine: REMstar System One with Autoflex
Mask Type: Full face mask
Mask Make & Model: Quattro AirFit F10
Humidifier: yes typical setting = 4
CPAP Pressure: 5-13 Autotritrating CPAP
CPAP Software: SleepyHead EncoreBasic

Other Comments: CMS-50F Plus oximeter /w SPO2 Assistant software. Avowed mouth-breather switching to nose-breathing

Sex: Male
Location: US

Post: #2
RE: If one has A10 Autoset Why adjust ranges?
the machine reacts to events. if you have it wide open, it will take a while to reach a pressure that tackles the event effectively. it is better to be within range of the most effective pressure when you are rattled with a cluster of events.

it is easier for us to get data, but you can tackle this without charts in the meantime.

so -
what is your prescription pressure setting min and max? (your profile says 5 - 20, is that current?)
what is your starting pressure, median pressure, and your 95% pressure?

Dedicated to QALity sleep.
You'll note I am listed as an Advisory Member. I am honored to be listed as such. See the fine print - Advisory Members as a group provide advice and suggestions to Apnea Board administrators and staff concerning Apnea Board operation and administrative policies. Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment.
(This post was last modified: 01-19-2015 07:07 PM by quiescence at last.)
01-19-2015 07:06 PM
Find all posts by this user Post Reply Quote this message in a reply
surferdude2 Online

Advisory Members

Posts: 1,077
Joined: Jul 2014

Machine: ResMed S9 AutoSet
Mask Type: Full face mask
Mask Make & Model: Mirage Quattro & Optilife
Humidifier: ResMed H5i @ #3 if nose breathing & #5 if mouth breathing
CPAP Pressure: 12 ~ 20 & 11 ~12 if nasal mask
CPAP Software: ResScan SleepyHead

Other Comments: Late stage 2 emphysema

Sex: Male
Location: Mousetown, Southern Illinois

Post: #3
RE: If one has A10 Autoset Why adjust ranges?
And another aspect that you will discover in time is that when the pressure changes a lot and in a short time, it can actually cause you to wake up. That's kinda counterproductive to the cause at hand. The auto adjusting machines are great for determining what pressure you need to handle your flow limitations but once you know what that is, it's often of great advantage to set the minimum up very close to that figure. I keep mine 1 cm H2O below what it takes to clear the worse case flow limitation before it gets flagged as an OA event.

Best regards, Dude
01-19-2015 07:39 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
PaulaO2 Offline
Wiki Editor
Moderators

Posts: 8,065
Joined: Feb 2012

Machine: S9 Autoset
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: H5i
CPAP Pressure: 14-20
CPAP Software: SleepyHead

Other Comments: Ehlers-Danlos Syndrome, Hypermobility Type; chronic sarcasm

Sex: Undisclosed
Location: western NC, USA

Post: #4
RE: If one has A10 Autoset Why adjust ranges?
Think of it like trying to herd cats.

If you are herding cats in a big room, it will be harder because you have further to go with each cat. You get them all in the box and you relax but one starts to get out of its box. You have so far to go to get to its box, it escapes and you have to herd it into its box again. But then another gets out. And....

But if you are in a smaller room, getting them back in line isn't quite as difficult because they don't have as much room to run around in. And even if you relax, you can stop them from getting out of the boxes quicker because they are all within reach. They may try to get out, and some will still succeed, but it will be easier to herd them back in again.

Oh, and:
The "How Do I..." thread in the Member Help forum:
http://www.apneaboard.com/forums/Thread-How-to-HOW-DO-I

In there you can find the link for how to do an attachment:
http://www.apneaboard.com/forums/Thread-How-to-Attachments

PaulaO2
Apnea Board Moderator
www.ApneaBoard.com


Breathe deeply and count to zen.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
01-19-2015 07:55 PM
Find all posts by this user Post Reply Quote this message in a reply
TheManseHen Offline

Preferred Members-2

Posts: 243
Joined: Jan 2015

Machine: RedMed A10 Airsense 10 Autoset
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: One that came with machine
CPAP Pressure: 7.6-14.0 EPR 3
CPAP Software: ResScan SleepyHead

Other Comments: Started C-pap Boxing Day 2014. Tortured sleep more or less since 2000. Trigeminal Neuralgia.

Sex: Female
Location: San Francisco

Post: #5
RE: If one has A10 Autoset Why adjust ranges?
Wow you all are GREAT! That helped a lot. And Paula I love the hearding cats image. It made me smile after a rough night.

Okay let's see, looking at 30 day setting of Statistics on ResScan it says

212 some hours used, 23 days.

Apnea Index 4.9 Hypopnea Index 1.1 AHI: 6.0 (no time in CSR)
Obstructive 2.4
Central 1.8
Unknown 0.2

Leak L/min Median 0.0 95th 10.8 Max 66.0
Pressure Median 8.3 95th 10.3 Max 13.0

Now those are the stats for the whole period I assume. Shall I go and look for all the outliers low and high, on each day and make note of the highest and lowest or is this sufficient?

What would YOU set pressure to if you were I.
And I have it currently at 5.0-20.00

And I think I am waking from air bursts etc at odd times

Doctors script didn't set a measurement as it was an in home study which determined I had 26.3 AHI and
determined my pulse Ox got down to 76 percent (OUCH) and my event time range went from 14-42 seconds.

SO...it seems like I should set the machine either to 8-13 or a hair hedged say 7.8-13.2 or so? Or should I do 5.0 (what is comfortable to lay down with, I am NOT currently using ramp feature) to 13.2 or so? Or should I do the 7-8-13.2 or so and USE the ramp feature for comfort? I didn't like the long delay to feel fresh air and as soon as I start dozing, I shoot to at least 7 something right away. Please Advise? I know you aren't doctors (-: I also haven't been experiencing any Aerophagia where it's at now.



Thoughts?
Thanks,
Susan
(This post was last modified: 01-19-2015 09:24 PM by TheManseHen.)
01-19-2015 09:16 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
retired_guy Offline

Preferred Members-2

Posts: 2,744
Joined: Jan 2014

Machine: ResMed S9 Autoset
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: ResMed
CPAP Pressure: 10.6/14, EPR 3
CPAP Software: ResScan SleepyHead

Other Comments: Is it bedtime yet?

Sex: Male
Location: Beautiful, albeit very wet Oregon Coast

Post: #6
RE: If one has A10 Autoset Why adjust ranges?
If it was my cute little body, and precious little machine, I'd set min 8, max 13... I'd also set the EPR to 3 and see if that makes me happier than the usually happy camper I am. Then I'd really watch the stats for a few days. If they did not drop to say 1 or 2, then I'd figure something needs adjusting further. Keep an eye on the leakage as that can cause results to be goofy as well.
01-19-2015 09:37 PM
Find all posts by this user Post Reply Quote this message in a reply
surferdude2 Online

Advisory Members

Posts: 1,077
Joined: Jul 2014

Machine: ResMed S9 AutoSet
Mask Type: Full face mask
Mask Make & Model: Mirage Quattro & Optilife
Humidifier: ResMed H5i @ #3 if nose breathing & #5 if mouth breathing
CPAP Pressure: 12 ~ 20 & 11 ~12 if nasal mask
CPAP Software: ResScan SleepyHead

Other Comments: Late stage 2 emphysema

Sex: Male
Location: Mousetown, Southern Illinois

Post: #7
RE: If one has A10 Autoset Why adjust ranges?
I also agree that you need to raise the minimum setting some. An 8 setting will probably be comfortable if you also set the EPR value to #3. That way you'll be exhaling into a mere 5 cm H2O, just like you are now but you will be inhaling with 8 cm H2O pressure which gives you a better hole shot on the Flow Limitations and a better chance at stopping them from being full grown OA events. The 5 ~ 13 as suggested should give you better results but don't expect instant improvement. It's amazing how the body can adjust to breathing under pressure but it does take time, more for some people, less for others. In time you may choose to back off on the EPR but not right away. Most old timer shut it off completely but it's great relief for newbies.

I had to start at 6 cm H2O and swore I'd never be able to stand anything higher. Presently I start with a minimum of 12 cm H2O and it feels fine. If I could tell you one thing that helped the most in getting that acceptance, it would be wearing the mask while watching TV during the day for short periods and increasing the pressure each day by small amounts. It took me a couple of weeks to get it done but it sure beats laying in bed and struggling with some new pressure setting. Familiarity breed contempt and it can allow you to disregard all the fears you have of being blown up like a balloon and all that jazz.
(This post was last modified: 01-19-2015 10:12 PM by surferdude2.)
01-19-2015 10:07 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
AshSF Online

Advisory Members

Posts: 356
Joined: Oct 2014

Machine: PRS1 D560TS
Mask Type: Nasal pillows
Mask Make & Model: ResMed P10, Bkup: Airfit F10 & Quattro Air
Humidifier: PRS1 60 series with Heated Tube
CPAP Pressure: 8 - 8 w/ Aflex 3
CPAP Software: SleepyHead

Other Comments: Using APAP mode with fixed pressure

Sex: Male
Location:

Post: #8
RE: If one has A10 Autoset Why adjust ranges?
The strategy that I employed to go from a 4-20 range to a 7-18 range is to raise the minimum pressure by 1cm. Keep it there for 10+ days. Then raise it by 1cm and repeat. It helped me get used to breathing against a higher pressure. I have had my A-flex set to 3cm so its roughly equal to 3 setting on EPR on Resmed. You can also try raising it by 0.5cm every 5-10 days.

When I went from a 5 start to a 6 start, I felt like the machine was trying to inflate me like a balloon. I had to consciously not think about the machine and my breathing and I could fall asleep. Didn't happen at any other start pressure change.

I also observed that my Average Pressure actually went down from 8+ to ~7.5 when I gradually raised my minimum pressure. My 95% pressure value also went down.

YMMV.

Started APAP 4-20, Closed range to 7.5-14, then straight 8.0 w/ Aflex 3
RDI always below 1. But sleep much much better at straight pressure.
Started on F10, Tried Quattro Air successfully. Finally settled on P10.
01-19-2015 11:03 PM
Find all posts by this user Post Reply Quote this message in a reply
TheManseHen Offline

Preferred Members-2

Posts: 243
Joined: Jan 2015

Machine: RedMed A10 Airsense 10 Autoset
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: One that came with machine
CPAP Pressure: 7.6-14.0 EPR 3
CPAP Software: ResScan SleepyHead

Other Comments: Started C-pap Boxing Day 2014. Tortured sleep more or less since 2000. Trigeminal Neuralgia.

Sex: Female
Location: San Francisco

Post: #9
RE: If one has A10 Autoset Why adjust ranges?
Thanks Guys! I've had my 5.0 and last night I tried 6.0 for a while but I only had the EPR set to 1.

For now, I've gone and set my bottom Pressure to 8.0 and my top to 13.6 (okay I'm a chicken, just in case I go over somehow)

I have set my EPR to 3 and it is at full time.

I did give myself a 5 min ramp time for this new trial in case I felt overwhelmed during the night or while I lay here and try it while faffing about. If I do well enough on awake trial, I may dump the 5 min ramp but I have it set to start at 5.0 cause that is my "familiar" that said I expect it will feel less than because of the EPR now set to 3.

In the early days I'd find self pulling mask off, feeling disoriented, but I've tried to train myself to hit the reset button with mask staying on, instead to give me thinking time and then return to breathing as I have it on smart start.

I will stay put for a few days and let ya'll know how it goes!
Thanks.
Susan in San Francisco,
Itching to get out of the crib and into the toddler hosehead room where I can forget about the machine for a bit and just play enjoy the land o'nod.
01-19-2015 11:04 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
archangle Offline
Wiki Editor
Advisory Members

Posts: 3,159
Joined: Feb 2012

Machine: ResMed S9 AutoSet
Mask Type: Nasal pillows
Mask Make & Model: ResMed Swift FX
Humidifier: ResMed S9 H5i
CPAP Pressure: 16-20
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments: Happy PAPper

Sex: Undisclosed
Location: USA

Post: #10
RE: If one has A10 Autoset Why adjust ranges?
It's not the wide range that causes centrals. For SOME people, a higher pressure cause centrals. It probably doesn't matter whether you got there by auto or by manual CPAP.

If your starting pressure is too low, you may suffer from discomfort or inadequate treatment until the machine adjusts up to the right pressure. Most auto machines only adjust pressure upwards 1 unit per minute or so.

Sometimes, an auto machine will "run away" and go to too high of a pressure unless you lower the max pressure. Even if the higher pressure is "right," it may cause discomfort and a lower pressure works better.

Changing pressure makes mask fitting harder. If you fit it at your starting pressure, it may leak after you go to sleep and the pressure goes up.

I find that I feel better if I set my minimum pressure a bit above where my AHI is basically zero anyway. I presume this is something a little more subtle like UARS or restricted breathing of some kind.

Get the free SleepyHead software here.
Useful links.
Click here for information on the main alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check it yourself.
01-20-2015 12:05 AM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  Adjust Min or Max Pressure for High OA TheLionSleeps 5 531 07-27-2016 04:28 PM
Last Post: PaulaO2
  how to adjust clock on Resmed autoset 9 Marl1n 3 466 07-25-2016 10:12 AM
Last Post: PaulaO2
  Resmed s10 ramp starting pressure, can you adjust it? Nytro_Mike 4 708 04-14-2016 02:26 PM
Last Post: Nytro_Mike
  How do I adjust pressure on Viasys BreatheX avalon.uk 11 3,923 04-09-2016 01:31 AM
Last Post: vsheline
Question [Pressure] Should I Adjust BIpap settings BIGELI 3 947 01-07-2016 12:13 AM
Last Post: PollCat
  Acceptable AHI ranges icewlf 20 3,726 01-03-2016 12:28 AM
Last Post: Sleepster
  How adjust pressure ResMed iAVAP H9i shelldill 2 550 12-21-2015 11:06 PM
Last Post: SuperSleeper

Forum Jump:

Who's Online (Complete List)