This is my first post.
I have been on Bi-pap for 8 months now. I had an AHI of 34 in my sleep study. I tried 3 different machines now and have finally found the one that works for me. I have the ResMed AirCurve 10 VAUTO which I have used for 3 months now. My apneas have come down to a normal level averaging 2.4 for the last 3 months and 1.9 for the last 1 month. I am quite happy with that. Most of my apneas are Centrals, however I had no Centrals in my sleep study at the Sleep Lab. I'm assuming the machine is causing the centrals but since my AHI is normal I am not too concerned. My sleep doctor said it was hyperventilation.

My problem is that on many nights, usually after 4:00am I feel that I am being forced to breathe faster. The inhalation starts before I am done exhaling. If I stop the machine and restart it of course my breathing goes back to normal again. But then when the pressure builds up again it will repeat the fast breathing again. My therapist is making adjustments every 2 t 3 weeks to try to correct this but no success yet. My Ti max is set at 2.0, Ti min 0.3. The trigger is set to medium and the cycle set to low.

If anyone else has had a similar problem I would love to hear how you rectified it.
12-02-2015 07:08 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Sleeprider Offline
Wiki Editor
Advisory Members

Posts: 3,400
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #2
RE: Inhalation/Exhalation Problem
Central or clear airway apneas as measured by your machine can be caused by excessive pressure support. It would help to know what your complete settings are. Your profile shows 8/15 which might imply EPAP of 8 and IPAP of 15 or PS of 7.0. Your machine is capable of treatment at a range of pressures. Can you clarify what pressure support is used on your machine?

I also use auto bilevel, and you can see my settings in my profile to the left. You will have something similar, except your Pressure Support will be a fixed value on Resmed. I also have minor short duration CA events but no OA. I don't find them disruptive to my sleep as compared to the increased obstructive events with lower pressure support.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
12-02-2015 07:31 PM
Find all posts by this user Post Reply Quote this message in a reply
justMongo Offline

Monitors

Posts: 3,857
Joined: Sep 2013

Machine: ResMed VPAP Auto (S9)
Mask Type: Full face mask
Mask Make & Model: ResMed Mirage Quattro
Humidifier: ResMed H5i
CPAP Pressure: 18 IPAPmax - 11 EPAPmin, PS=5
CPAP Software: Other Software

Other Comments: μολὼν λαβέ

Sex: Male
Location: Калифорния

Post: #3
RE: Inhalation/Exhalation Problem
I believe that machine can also be run in one of three modes: straight CPAP, "S"* mode bilevel, and bilevel auto.
So, in the setup menu, for mode, it would say: CPAP, S, or Auto. I think it would be good to check the mode setting.
For CPAP mode, one pressure is specified, for S mode: two pressures (IPAP and EPAP), and for Auto: 3 pressures (IPAPmax, EPAPmin and PS.)**


*(S = Spontaneous trigger to/from IPAP/EPAP)
** Resmed handles Bilevel a tad differently that Phillips-Respironics Bilevel.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
12-02-2015 07:54 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
nadprok Offline

Preferred Members

Posts: 25
Joined: Aug 2015

Machine: resmed aircurve10 vauto
Mask Type: Full face mask
Mask Make & Model: Philips Respironics Amara View
Humidifier: ResMed Humidair
CPAP Pressure: EPAP 8, IPAP 13, PS 1
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Ontario, Canada

Post: #4
RE: Inhalation/Exhalation Problem
(12-02-2015 07:31 PM)Sleeprider Wrote:  Central or clear airway apneas as measured by your machine can be caused by excessive pressure support. It would help to know what your complete settings are. Your profile shows 8/15 which might imply EPAP of 8 and IPAP of 15 or PS of 7.0. Your machine is capable of treatment at a range of pressures. Can you clarify what pressure support is used on your machine?

I also use auto bilevel, and you can see my settings in my profile to the left. You will have something similar, except your Pressure Support will be a fixed value on Resmed. I also have minor short duration CA events but no OA. I don't find them disruptive to my sleep as compared to the increased obstructive events with lower pressure support.

Yes, the 8 is my EPAP and the 15 is my IPAP. My pressure support is 3. I hardly ever have any obstructive apneas. I can go 10 days without one. I have anywhere from 1 to 3 hypopneas and the rest are centrals.
12-02-2015 10:03 PM
Find all posts by this user Post Reply Quote this message in a reply
nadprok Offline

Preferred Members

Posts: 25
Joined: Aug 2015

Machine: resmed aircurve10 vauto
Mask Type: Full face mask
Mask Make & Model: Philips Respironics Amara View
Humidifier: ResMed Humidair
CPAP Pressure: EPAP 8, IPAP 13, PS 1
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Ontario, Canada

Post: #5
RE: Inhalation/Exhalation Problem
(12-02-2015 07:54 PM)justMongo Wrote:  I believe that machine can also be run in one of three modes: straight CPAP, "S"* mode bilevel, and bilevel auto.
So, in the setup menu, for mode, it would say: CPAP, S, or Auto. I think it would be good to check the mode setting.
For CPAP mode, one pressure is specified, for S mode: two pressures (IPAP and EPAP), and for Auto: 3 pressures (IPAPmax, EPAPmin and PS.)**


*(S = Spontaneous trigger to/from IPAP/EPAP)
** Resmed handles Bilevel a tad differently that Phillips-Respironics Bilevel.

My machine runs in auto mode. EPAP is 8. IPAP is 15 and pressure support is 3.
12-02-2015 10:08 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Sleeprider Offline
Wiki Editor
Advisory Members

Posts: 3,400
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #6
RE: Inhalation/Exhalation Problem
Good, the missing PS number. Your pressure support is low, relative to most bilevel users. Your pressure is staring out at 11.0/8.0 (IPAP/EPAP), and may range as high as 15.0/12.0. The fact your therapy has reduced obstructive apnea to a very low incidence shows that aspect is successful. CA events, are well within tolerable limits considering your total AHI is currently less than 2.0.

With your AHI essentially resolved, your immediate problem is a sense of the machine not timing your natural respiration well. It sounds like you are working with your RT, and she is making adjustments to the default Ti settings. In my opinion, it's probably not necessary to change the defaults, and this is a common problem in adapting to bilevel, or even APAP with EPR. You control the respiration rate and timing not the machine. When it seems to be off-sync, you should be able to consciously inhale and exhale with enough force to re-sync the machine. The sense of the machine taking over is only because you are pausing your breathing and the machine senses this as the completion of the inhale or exhale (probably exhale). When you do that, the machine senses the change the pressure and so it increases pressure for the inhalation. You feel that as the machine taking over, but you can stop it, by exhaling against the pressure. It's also likely that when you pause breathing at the bottom of the exhale, the machine senses this as a CA. All of this is normal and will eventually pass.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
12-03-2015 10:50 AM
Find all posts by this user Post Reply Quote this message in a reply
PaytonA Offline
Wiki Editor
Monitors

Posts: 3,020
Joined: Dec 2013

Machine: ResMed S9 VPAP Auto
Mask Type: Full face mask
Mask Make & Model: Resmed Mirage Quattro
Humidifier: H5i(distilled-top up)
CPAP Pressure: VAuto MinE14.0 MaxI 20.6 PS4.0
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Orange County,California

Post: #7
RE: Inhalation/Exhalation Problem
Daniel,

During the time that you are having this problem, expand your flow rate graph and look at the downhill side of the inhalation portions of your curve. Look to see if there is a little hitch in this section where it stops descending or starts a small ascent for a very short period of time. Let me know what you see.

Best Regards,

PaytonA
12-03-2015 11:26 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
nadprok Offline

Preferred Members

Posts: 25
Joined: Aug 2015

Machine: resmed aircurve10 vauto
Mask Type: Full face mask
Mask Make & Model: Philips Respironics Amara View
Humidifier: ResMed Humidair
CPAP Pressure: EPAP 8, IPAP 13, PS 1
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Ontario, Canada

Post: #8
RE: Inhalation/Exhalation Problem
(12-03-2015 10:50 AM)Sleeprider Wrote:  Good, the missing PS number. Your pressure support is low, relative to most bilevel users. Your pressure is staring out at 11.0/8.0 (IPAP/EPAP), and may range as high as 15.0/12.0. The fact your therapy has reduced obstructive apnea to a very low incidence shows that aspect is successful. CA events, are well within tolerable limits considering your total AHI is currently less than 2.0.

With your AHI essentially resolved, your immediate problem is a sense of the machine not timing your natural respiration well. It sounds like you are working with your RT, and she is making adjustments to the default Ti settings. In my opinion, it's probably not necessary to change the defaults, and this is a common problem in adapting to bilevel, or even APAP with EPR. You control the respiration rate and timing not the machine. When it seems to be off-sync, you should be able to consciously inhale and exhale with enough force to re-sync the machine. The sense of the machine taking over is only because you are pausing your breathing and the machine senses this as the completion of the inhale or exhale (probably exhale). When you do that, the machine senses the change the pressure and so it increases pressure for the inhalation. You feel that as the machine taking over, but you can stop it, by exhaling against the pressure. It's also likely that when you pause breathing at the bottom of the exhale, the machine senses this as a CA. All of this is normal and will eventually pass.

Thanks ever so much for your tips. I will try and control my breathing as you suggested and not let the machine control it. If I can get that under control then my therapy will be pretty well perfect.

Daniel
12-03-2015 04:07 PM
Find all posts by this user Post Reply Quote this message in a reply
nadprok Offline

Preferred Members

Posts: 25
Joined: Aug 2015

Machine: resmed aircurve10 vauto
Mask Type: Full face mask
Mask Make & Model: Philips Respironics Amara View
Humidifier: ResMed Humidair
CPAP Pressure: EPAP 8, IPAP 13, PS 1
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Ontario, Canada

Post: #9
RE: Inhalation/Exhalation Problem
(12-03-2015 11:26 AM)PaytonA Wrote:  Daniel,

During the time that you are having this problem, expand your flow rate graph and look at the downhill side of the inhalation portions of your curve. Look to see if there is a little hitch in this section where it stops descending or starts a small ascent for a very short period of time. Let me know what you see.

Best Regards,

PaytonA

The graph shows that after the curve hits bottom it goes up about half way and then goes sideways a bit and then continues to go upwards.

Daniel
12-03-2015 04:27 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
PoolQ Offline

Advisory Members

Posts: 986
Joined: Jul 2015

Machine: AirCurve 10 VAUTO
Mask Type: Full face mask
Mask Make & Model: Amara View
Humidifier: Integrated
CPAP Pressure: 6.2, 15, PS4.0
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Sillicon Valley

Post: #10
RE: Inhalation/Exhalation Problem
I have the same machine, with a similar path to get there. I am pretty sure that T1 is max inhale time and has nothing to do with exhaling. The machine for me kept cutting off my inhale which is very frustrating. It's like with each inhale right near the end someone puts their hand over your nose. the change from exhale to inhale is triggered by your breathing, for the machine to try to speed up your breathing means it is being triggered too soon. Perhaps change the trigger sensitivity.

For my inhale thing I had to set T1 to 3. Oh and T2 is the minimum inhale and will try and keep you inhaling for at least that long, if the machine is getting triggered at the end of your exhale into thinking you are inhaling maybe the T2 number is getting the machine to blow more air into you.

Try adjusting the trigger level and then maybe the T2 number a little longer and see if it gets better.
If you think that in your sleepy state you might be mixing up end of inhale with length of exhale, try adjusting T1 longer.
12-03-2015 10:23 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  Totally Confused by Airsense 10 Data - Centrals - Problem machine... ? mymontreal 28 1,241 Yesterday 09:39 PM
Last Post: richb
Surprised WISP mask problem elbie 12 762 Yesterday 10:00 AM
Last Post: cbokay
  Long time CPAP user now with a problem PrkChps 5 210 12-07-2016 09:07 PM
Last Post: PrkChps
  Problem with DreamWear Mask Judy1843 5 316 11-30-2016 04:03 AM
Last Post: holden4th
  New User Problem with BiPAP AVAPS - Part 1 jmayer180 35 1,342 11-27-2016 09:50 AM
Last Post: jmayer180
  ResMed Airsense 10 Setting Date Problem Reuven Gruber 6 303 11-16-2016 04:13 PM
Last Post: Sleeprider
  [CPAP] ResMed Air Sense 10 Auto Off Problem Buddie43 16 3,294 11-12-2016 02:51 PM
Last Post: sonicboom

Forum Jump:

Who's Online (Complete List)