Intellipap Autoadjust -- some settings not understood
#1
Hi all,

I have recently acquired a Devilbiss Sleepcube Autoadjust (Intellipap Autoadjust for USA people) and I went about to understand, and sometimes to adjust, the various settings available in the "Clinician" menu.

Most of these settings are straightforward enough, but I am baffled by some settings in the "Advanced" submenu
These settings are as follows:

---- menu item ---- current set value -----
Apnea % 10
Apnea Duration 10 seconds
Hypopnea % 50
Hypop Duration 10 seconds


Well, I can understand the "duration" part, it must be the length of time needed to trigger a response from the device (pressure increase). Although a default of 10 seconds seems to me a bit much ... What do you think about that ?

But what about the percentage part ??? How do you interpret this ?

Thanks in advance for any clue on the subject ... and possibly forgive my English, this is not my native language ..

Jean-Louis.
Reply
#2
Scroll down to event definition page 8
http://www.devilbisshealthcare.com/files...T-2089.pdf

If that does not answer your question, the clinical manual available via Email
Section three - DeVilbiss Manuals:
http://www.apneaboard.com/adjust-cpap-pr...tup-manual
Reply
#3
Thank you Zonk !

Yes, the discussion in the Clinical Overview does answer my questions.

Again, thanks.
Jean-Louis
Reply


#4
Hi Jean-Louis Debert,
WELCOME! to the forum.!
Hang in there for more responses to your post and best of luck to you.
trish6hundred
Reply
#5
I assume changes to these variables affects the reported events and thus AHI when using software like sleepyhead... If you use the default values, how do the thresholds and reported events compare to other machines like the S9?

I'm getting AHI's in the 6-8 range and see others posting ahi's (mostly with S9 machines) much lower. How much of that might be threshold sensitivity vs. actual differences in ahi?
Reply
#6
(05-23-2014, 03:25 AM)zonk Wrote: Scroll down to event definition page 8
http://www.devilbisshealthcare.com/files...T-2089.pdf

Good link, zonk. Added to my CPAP reading list. Lotsa good tech talk.

Get the free SleepyHead software here.
Useful links.
Click here for information on the main alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check it yourself.
Reply




Possibly Related Threads...
Thread Author Replies Views Last Post
  Fine tuning settings Cranberry Ray 4 269 01-07-2017, 05:08 AM
Last Post: kwhenrykerr
  Problems with my humidifier settings...... binky12 9 477 01-02-2017, 06:20 PM
Last Post: binky12
  Help with Bipap settings SupraSeth94 18 604 12-29-2016, 05:26 PM
Last Post: Sleeprider
  Mask settings on Dreamstation? Russatrice 10 733 12-25-2016, 11:39 PM
Last Post: eseedhouse
  Can someone help look at my settings and data daveofgv 7 304 12-19-2016, 05:14 PM
Last Post: Sleeprider
  Settings question Slim1950 2 244 12-15-2016, 10:08 AM
Last Post: Slim1950
  Dream Station Ramp and Pressure Settings knowoneuno 27 1,145 12-09-2016, 09:02 PM
Last Post: Russatrice

Forum Jump:




About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.

For any more information, please use our contact form.