Leak problems. Still.
#11
Considering the leakage, should I still use the mask until I get something else? Is it doing me any good? I can't really get a good night's sleep due to the constant air blowing on my face.
Post Reply Post Reply
#12
Harlon, the graphs show that you are still getting therapy in spite of the leaks. The flow graph still shows a response to respiration. It's your call, but I'd certainly continue if you're required to demonstrate compliance. Good luck and light a fire under their butts to get this taken care of.
Post Reply Post Reply
#13
This is from Tuesday night. (actually Wednesday morning. I work shiftwork and was coming off a night shift). After some tweaking it seems to be doing better, leakage wise. My AHI went from .74 to 2.09. See any other concerns from the graph?? [Image: 6-8-16.jpg]
Post Reply Post Reply


#14
Hmm. First of all most sleep Docs would say congratulations everything looks fine. But, you may still want to do some more fine tuning. It looks like your machine spent more time at higher pressures in the latest data that you posted. I can't tell if that is because of the larger number of events or the cause of that larger number. Get a few more nights of low leak data to see if your numbers are consistent. There are others here who have a great deal of expertise in fine tuning their settings for optimal outcomes. Post some more data when you have it. You will get more suggestions.

Rich
Apnea Board Member RobySue has posted a Beginners Guide to Sleepyhead Software here:  http://www.apneaboard.com/wiki/index.php...SleepyHead

Download Sleepyhead
Organize your Sleepyhead Charts
Post from Imgur


Post Reply Post Reply
#15
It's pretty remarkable how fast the machine increases pressure when you first start. Your minimum pressure of 5.0 seems pretty irrelevant and I'd choose something closer to 6, which is the lowest the machine rebounds to. In fact, for myself, I'd want a minimum pressure a bit closer to the median.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Quattro (full face) mask - gasket leak gcritchley 6 163 03-25-2017, 11:49 AM
Last Post: PaytonA
  Bipap AVAPS problems samianwardcummings 10 362 03-24-2017, 07:12 AM
Last Post: Sleeprider
  Problems with AirFit P10 Nasal Pillows (+ResMed AirSense 10) numari 18 7,036 03-23-2017, 11:36 AM
Last Post: PaytonA
  Very High Leak Rate Grmato10 55 1,393 03-22-2017, 03:55 PM
Last Post: Sleeprider
  Exhausted / have leak rate of 28.1 aconcepts 33 679 03-21-2017, 01:23 PM
Last Post: aconcepts
  Daytime sleepiness despite low AHI, low leak, 7+ hours of daily use vas2000 12 510 03-21-2017, 12:01 PM
Last Post: Attila
Question AHI vs Leak Rate jo11ymon 10 343 03-10-2017, 05:49 PM
Last Post: Melman

Forum Jump:

New Posts   Today's Posts




About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.

For any more information, please use our contact form.