Low spO2 levels even with low AHI
#11
This is pretty elementary.  You need a higher minimum pressure to improve oxygenation.  Your machine is setup at the default 4-20 and with the minimum pressure set that low, you are constantly triggering flow limitations and have a remarkably low minute vent.  The tidal volumes are not visible with that pie chart in there.  

Your pressure is swinging up from 4 and averaging 8 and your 95% pressure is 12.  You need to set that minimum pressure to 7 or 8 to get your therapy out of the hole and improve oxygenation.  I have posted it before, but this is an excellent presentation of what can be accomplished with CPAP and Bilevel pressure machines. The part you need to hear is EPAP pressure, or in your case minimum pressure, establishes a patent airway and promotes oxygenation.  Increase the minimum perssure, you'll be amazed.



Post Reply Post Reply
#12
Thanks for the info.

I switched to the auto titrating CPAP setting to try to get a feel for whether my older prescription (11 in standard CPAP) was still in the ball park. My 90% number has been averaging 11, my 95% was averaging 12-12.5

So is it usually preferable to run it in the auto titrating CPAP setting, but with higher starting pressure you mention, or go back to the fixed pressure?
Post Reply Post Reply
#13
If you are accustomed to higher pressure, my recommendation would actually be a minimum pressure of about 9.5 on auto mode. You are able to get good treatment with lower pressures, so there really is not need to use fixed pressures. The machine is clearly responding to your needs by raising pressures. By being within 2 cm of your therapy pressure, you should get the best of both worlds. Even for titration purposes, a higher starting pressure would have been in order. My suggestion for 7 to 8 would have been a good auto-titration starting point, but you can get good therapy with auto mode as long as it starts in the right range.

Finally, your machine must be in auto mode to record flow limitation, and this is something we want to evaluate for your SpO2 concerns. Go to a higher minimum pressure in auto mode, and you may never need to be in fixed CPAP mode. The machine pressure will follow sleep stages, body position and your changing needs to provide effective treatment.
Post Reply Post Reply


#14
Hi Tigerman67,
WELCOME! to the forum.!
Better luck to you this time around with your CPAP therapy, hang in there for more responses to your post.
You have come to the right place for guidance.
trish6hundred
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  [CPAP] SPO2 Levels gb60mail 11 393 07-22-2017, 06:06 PM
Last Post: gb60mail
  Humidity tank levels CPAPreturnee 3 250 07-09-2017, 11:20 PM
Last Post: Hydrangea
  Help me understand how humidifier levels are managed by the xPAP DanPrado 16 554 05-21-2017, 08:56 PM
Last Post: Sleepster
  spo2 drop to 40% dizzy1 7 436 05-09-2017, 06:17 AM
Last Post: ajack
  SPO2 BLOOD OXYGEN LEVELS dizzy1 3 419 05-06-2017, 01:47 AM
Last Post: ajack
  Acceptable levels of PB and RERA BHappy 7 426 05-02-2017, 08:52 AM
Last Post: bonjour
  Help [varying AHI levels] Haynbo 25 952 04-26-2017, 12:27 PM
Last Post: Haynbo

Forum Jump:

New Posts   Today's Posts




About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.

For any more information, please use our contact form.