Mask Vent Design Question
I am currently using an F & P Zest Q nasal mask and the design of the vent is beginning to trouble me. This mask has an elbow connected to it that is rotatable in 2 dimensions. The elbow has a series of small holes in it that act as the vent. There is also an attachment that holds a small oval of filter material over the vents to muffle the noise from the vents and limit the flow. This is what makes it the "Q" version of the Zest mask.

Last night I found that putting the mask on with no hose attached and blocking the hose inlet line, it was very difficult to breathe in or out. The literature from F&P indicates that the mask can be used with or without the filter media in place. I tried it without the filter material and ended up with a lot of condensation-some getting into the vent holes-causing a lot of whistling noise which woke me up.

All this messing with the vent made me look at the design more critically. The vent is basically at the end of the incoming line and I am beginning to wonder if it is very efficient at eliminating the exhaled air,since it is trying to do so against the pressure of the incoming air. I am also wondering if it may be having an adverse effect on my oxygen saturation which would not be a good thing.

I am interested in hearing from anyone that has some ideas about this.

Thanks in advance.

PaytonA
(This post was last modified: 01-24-2014 02:20 PM by PaytonA.)
01-24-2014 02:19 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
RonWessels Offline

Preferred Members-2

Posts: 465
Joined: Jun 2013

Machine: REMstar Auto AFlex DS560TS
Mask Type: Nasal mask
Mask Make & Model: ComfortGel Blue
Humidifier: REMstar Heated Tube DS6T
CPAP Pressure: 11 - 20
CPAP Software: SleepyHead

Other Comments: CMS50F Pulse Oximeter

Sex: Male
Location: Ontario, Canada

Post: #2
RE: Mask Vent Design Question
If I understand correctly, you found it very difficult to breathe essentially entirely through the mask vent. This is not surprising, since that vent is not designed to handle the instantaneous flow required for inhalation.

The purpose of the mask vent is to make sure that you do not breathe your own exhaled air. As such, the flow rate out of the vent only needs to be enough to average out to the volume of your exhalation over the period of your breathing. You are correct that some of your exhaled breath probably remains inside the "dead area" of the mask, but if you think about that volume of air compared with the volume of your inhalation, you will realize that this is a trivial amount in comparison. As long as whatever exhaled air got forced back up the hose is eventually vented before you next inhale, everything is working fine. Of course, if you "overload" the system by deeply inhaling/exhaling and then taking a second breath quickly, you will still get some of your previous exhaled air, which is why it will feel particularly warm and stuffy. But people don't breath that deeply and quickly normally.
01-24-2014 03:41 PM
Find all posts by this user Post Reply Quote this message in a reply
zonk Offline

Advisory Members

Posts: 7,908
Joined: Feb 2012

Machine: A10 AutoSet
Mask Type: Nasal mask
Mask Make & Model: Activa LT
Humidifier: Integrated /ClimateLineAir
CPAP Pressure: 9/13
CPAP Software: ResScan

Other Comments: CPAP since Nov 2010

Sex: Male
Location: Australia

Post: #3
RE: Mask Vent Design Question
(01-24-2014 02:19 PM)PaytonA Wrote:  Last night I found that putting the mask on with no hose attached and blocking the hose inlet line, it was very difficult to breathe in or out.
The purpose of the vents to flush out exhaled air, as fresh pressured air enter the mask flushes out exhaled air
For that reason when power goes out, full face masks have anti-asphyxia valve so you can breathe room air and don,t suffocate but you can open your mouth and breath with a nasal mask
01-24-2014 04:02 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
justMongo Offline

Monitors

Posts: 3,847
Joined: Sep 2013

Machine: ResMed VPAP Auto (S9)
Mask Type: Full face mask
Mask Make & Model: ResMed Mirage Quattro
Humidifier: ResMed H5i
CPAP Pressure: 18 IPAPmax - 11 EPAPmin, PS=5
CPAP Software: Other Software

Other Comments: μολὼν λαβέ

Sex: Male
Location: Калифорния

Post: #4
RE: Mask Vent Design Question
As described by Zonk: My FFM mask has two types of vents. A series of holes that are the flush holes to vent exhaled air out of the mask during xPAP operation.
There is a second vent with a flapper valve at the mask elbow. It's purpose it to permit one to breathe when the xPAP is not operating -- such as failure of the xPAP or just in that period between strapping on the mask and pressing the xPAP start button. When xPAP pressure is available the flapper closes that vent.

PaytonA -- perhaps the mask you have is not the best for you...

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
01-24-2014 05:00 PM
Find all posts by this user Post Reply Quote this message in a reply
PaytonA Offline
Wiki Editor
Monitors

Posts: 3,007
Joined: Dec 2013

Machine: ResMed S9 VPAP Auto
Mask Type: Full face mask
Mask Make & Model: Resmed Mirage Quattro
Humidifier: H5i(distilled-top up)
CPAP Pressure: VAuto MinE14.0 MaxI 20.6 PS4.0
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Orange County,California

Post: #5
RE: Mask Vent Design Question
(01-24-2014 03:41 PM)RonWessels Wrote:  If I understand correctly, you found it very difficult to breathe essentially entirely through the mask vent. This is not surprising, since that vent is not designed to handle the instantaneous flow required for inhalation.

The purpose of the mask vent is to make sure that you do not breathe your own exhaled air. As such, the flow rate out of the vent only needs to be enough to average out to the volume of your exhalation over the period of your breathing. You are correct that some of your exhaled breath probably remains inside the "dead area" of the mask, but if you think about that volume of air compared with the volume of your inhalation, you will realize that this is a trivial amount in comparison. As long as whatever exhaled air got forced back up the hose is eventually vented before you next inhale, everything is working fine. Of course, if you "overload" the system by deeply inhaling/exhaling and then taking a second breath quickly, you will still get some of your previous exhaled air, which is why it will feel particularly warm and stuffy. But people don't breath that deeply and quickly normally.

RonWessels,

You understood me correctly but maybe I was not clear enough. My point about it being difficult to breathe with the "muffler" in place was not to say that I should be able to breathe easily but it was very, very difficult. With other masks without the "muffler" it is much easier to breathe through the vent than with this one. Also with the "muffler" on and the machine running at 20 cm H2O it is very difficult to feel the vented air even with the more sensitive areas of my hand. The other thing that this little experiment did was to start me thinking about the placement of the vent.

The following is only to illustrate my logic when thinking about this venting situation.

If one were to hook a pipe to a container full of hydrogen sulfide gas with a small vent in the pipe right before the container and then blow air into the pipe, it would take a long time to remove all of the hydrogen sulfide from the container at the air pressure that we use. Some of the H2S would get entrained by eddy currents and leave through the vent but it would take a while. If one placed the vent on the opposite side of the container from the pipe, it would clear the H2S out faster (more eddy currents and more entrainment).

This all makes me a little unsure. I guess I will use without the muffler for now.

Best Regards,

PaytonA
01-24-2014 10:18 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Sleepster Offline
Wiki Editor
Moderators

Posts: 4,993
Joined: Feb 2012

Machine: ResMed AirCurve10 VAuto
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: HumidAir and SlimLine Hose
CPAP Pressure: MaxI 13.6 | MinE 5.2 | PS 4.4
CPAP Software: ResScan SleepyHead

Other Comments: Diagnosed Nov 2011. Conquered aerophagia.

Sex: Male
Location: Houston, Texas

Post: #6
RE: Mask Vent Design Question
(01-24-2014 02:19 PM)PaytonA Wrote:  Last night I found that putting the mask on with no hose attached and blocking the hose inlet line, it was very difficult to breathe in or out.

As you indicated in your later post, more difficult than with other masks, and presumably more difficult than it would be without the filter material.

So this establishes that less air flows through the vent when the filter material is present. That is a concern that you might want to research. See just how much the flow is limited. Presumably it's not limiting it below a safe level.

Quote:The vent is basically at the end of the incoming line and I am beginning to wonder if it is very efficient at eliminating the exhaled air, since it is trying to do so against the pressure of the incoming air.

The purpose of the vent is keep the hose cleared of exhaled air.

Sleepster
Apnea Board Moderator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
01-24-2014 11:49 PM
Find all posts by this user Post Reply Quote this message in a reply
archangle Offline
Wiki Editor
Advisory Members

Posts: 3,159
Joined: Feb 2012

Machine: ResMed S9 AutoSet
Mask Type: Nasal pillows
Mask Make & Model: ResMed Swift FX
Humidifier: ResMed S9 H5i
CPAP Pressure: 16-20
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments: Happy PAPper

Sex: Undisclosed
Location: USA

Post: #7
RE: Mask Vent Design Question
The issue of venting out all the stale air is well understood and designed into the mask as long as you use a proper CPAP mask and don't block the vents. It's pretty much foolproof.

You don't exhale just through the air vents. The CPAP blower keeps a constant pressure at the mask. As you exhale, some of their goes out the vent and some goes back up the hose. Because the mask is at a constant pressure, a fixed amount of air goes out through the vent as you inhale and exhale.

The "leak rate" of the vent is designed to flush out enough air throughout the breathing cycle at the minimum 4 cmH2O pressure that CO2 and O2 levels stay OK. This is true, even if the filter is in place, unless the filter has become clogged or you're using something other than the correct filter.

Get the free SleepyHead software here.
Useful links.
Click here for information on the main alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check it yourself.
01-27-2014 02:34 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
PaytonA Offline
Wiki Editor
Monitors

Posts: 3,007
Joined: Dec 2013

Machine: ResMed S9 VPAP Auto
Mask Type: Full face mask
Mask Make & Model: Resmed Mirage Quattro
Humidifier: H5i(distilled-top up)
CPAP Pressure: VAuto MinE14.0 MaxI 20.6 PS4.0
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Orange County,California

Post: #8
RE: Mask Vent Design Question
(01-27-2014 02:34 AM)archangle Wrote:  The issue of venting out all the stale air is well understood and designed into the mask as long as you use a proper CPAP mask and don't block the vents. It's pretty much foolproof.

You don't exhale just through the air vents. The CPAP blower keeps a constant pressure at the mask. As you exhale, some of their goes out the vent and some goes back up the hose. Because the mask is at a constant pressure, a fixed amount of air goes out through the vent as you inhale and exhale.

The "leak rate" of the vent is designed to flush out enough air throughout the breathing cycle at the minimum 4 cmH2O pressure that CO2 and O2 levels stay OK. This is true, even if the filter is in place, unless the filter has become clogged or you're using something other than the correct filter.

In thinking about this after the inputs that I have received here, I think I see what everyone has been trying to tell me. The dead space in the mask is 90mL. Using the data from the VPAP my breath calculates to around 700mL for each breath. Even if the mask was full of depleted air (from the last exhalation) it is such a minor part of my next breath that it probably does not have much effect.

Thanks for the inputs,

PaytonA
01-28-2014 03:37 PM
Find all posts by this user Post Reply Quote this message in a reply
Sleepster Offline
Wiki Editor
Moderators

Posts: 4,993
Joined: Feb 2012

Machine: ResMed AirCurve10 VAuto
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: HumidAir and SlimLine Hose
CPAP Pressure: MaxI 13.6 | MinE 5.2 | PS 4.4
CPAP Software: ResScan SleepyHead

Other Comments: Diagnosed Nov 2011. Conquered aerophagia.

Sex: Male
Location: Houston, Texas

Post: #9
RE: Mask Vent Design Question
Right. Think of it this way. Without the vent exhaled air would bounce in and out of the hose. Like a snorkeler with a too-long snorkel, you end up breathing in too much of your exhaled CO2. The vent keeps the hose purged, not the mask.

Sleepster
Apnea Board Moderator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
01-28-2014 07:35 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
Question CA's, Mask type, and leaks/vent Rcgop 10 368 11-15-2016 03:02 PM
Last Post: robysue
  [Equipment] Question about getting new mask sleepytimegal 11 492 09-28-2016 10:31 PM
Last Post: sleepytimegal
  AirFit P10 vent blocked? scott1965 178 30,675 09-19-2016 02:12 PM
Last Post: WLKSRQ
  Setting up DreamStation with F&P Simplus mask question grahamwright1 4 515 08-19-2016 11:58 PM
Last Post: Ghost1958
  Mask/leak question. jessamnda 7 664 08-10-2016 04:05 PM
Last Post: jessamnda
  new user, mask question marylyn 4 497 07-27-2016 06:49 PM
Last Post: trish6hundred
  Air Fit P10 mask question GrammaBear 4 489 06-24-2016 10:07 AM
Last Post: sonicboom

Forum Jump:

Who's Online (Complete List)