   here is what sleepyhead says

PAP ModePAP Device Mode                   ASV (Variable EPAP)
Min EPAPLower Expiratory Pressure     8.00 cmH2O
Max EPAPHigher Expiratory Pressure  15.00 cmH2O
Min IPAPLower Inspiratory Pressure      8.00 cmH2O
Max IPAPHigher Inspiratory Pressure    25.00 cmH2O
PS MinPressure Support Minimum          0.00 cmH2O
PS MaxPressure Support Maximum       7.00 cmH2O
Flex LevelPRS1 pressure relief setting.  x2
Flex ModePRS1 pressure relief mode.  Bi-Flex


I don't know what to put in my profile?

I get some data but mostly garbled.

I also found this in SH it shows where the 260P error message came from the first time I used it.

The First line is my new machine, the 2nd is my old machine & the 3rd seems to be the new compliance software that isn't compatable with SH:
Machine Information
Brand                               Series                                 Model                                                                         Serial                       First Use     Last Use

Philips Respironics      System One (60 Series)   BiPAP autoSV Advanced 60 Series (960P)   P0944547406D6     9/24/2015       3/7/2017
Philips Respironics      DreamStation                      Auto CPAP (500X110)                                             J17102940E769      10/21/2016     3/2/2017
Philips Respironics      System One (60 Series)   RemStar Plus Compliance Only (260P)           P088859607F44      (No Dates Given)


 
Your Friend Kidnap
Post Reply Post Reply
#82
For your profile, you might write
EPAP: 8-15, IPAP: 8-25, PS: 0-7
                                                                                                                                                                                  
Please organize your SleeyHead screenshots like this.
I'm an epidemiologist, not a medical provider. 
Post Reply Post Reply
#83
Thanks Beej

 I copied it and will paste info in my profile.

Thanks again
Your Friend Kidnap
Post Reply Post Reply


#84
I think it is actually 25/8-15 over PS 0-7. That give IPAP max / EPAP min-max over PS range.

I think it would be more helpful to enter the clinical setup menu and transcribe settings from there. I think most will be the same. I'm curious if you have mode options, and I want to know if you BPM is set to a number or auto.

That aside, how's it going? Did you clear the prior patient history?
Post Reply Post Reply
#85
(03-09-2017, 09:23 PM)Sleeprider Wrote: I think it is actually 25/8-15 over PS 0-7.  That give IPAP max / EPAP min-max over PS range.

I think it would be more helpful to enter the clinical setup menu and transcribe settings from there.  I think most will be the same.  I'm curious if you have mode options, and I want to know if you BPM is set to a number or auto.  

That aside, how's it going?  Did you clear the prior patient history?

Is there a standard way to write these that everyone will understand?
                                                                                                                                                                                  
Please organize your SleeyHead screenshots like this.
I'm an epidemiologist, not a medical provider. 
Post Reply Post Reply
#86
That is standard only for the ASV. All bilevel will be IPAP/EPAP over PSmin-PSmax, or just one pressure support as applicable. The Resmed ASV in ASV auto mode is the only one that allows an EPAP range limit. On auto bilevel machines, it's understood that the settings are IPAPmax/EPAPmin over PS. With multi-mode machines, it doesn't hurt to show what mode is in use.

So when you see my profile is 18/9, PS 3 that means my starting pressure is 12/9 and can reach 18/15. I use the same prescription with the Philips BiPAP Auto 18/9, 3-5 because PS is a range.
Post Reply Post Reply


#87
Sleeprider

  I'll have to look closer at my settings .
Thanks Yet Again
Your Friend Kidnap
Post Reply Post Reply
#88
I don't think the Philips has the EPAP range, but has a minimum EPAP, and that can increase automatically, much like the BiPAP Auto. So you have EPAP min, IPAP max and PSmin PSmax. In theory it's the same idea, but the implementation doesn't come close to Resmed.

If you want, I can try to load up your data in Sleepyhead and see what we've got. Let me know and we can do it by email or you can buy me a beer.
Post Reply Post Reply
#89
(03-10-2017, 08:23 PM)Sleeprider Wrote: I don't think the Philips has the EPAP range, but has a minimum EPAP, and that can increase automatically, much like the BiPAP Auto.  So you have EPAP min, IPAP max and PSmin PSmax.  In theory it's the same idea, but the implementation doesn't come close to Resmed.

If you want, I can try to load up your data in Sleepyhead and see what we've got.  Let me know and we can do it by email or you can buy me a beer.

Didn't the advanced model add EPAP min and EPAP max?  The new dreamstation asv has the EPAP range. As far as I could tell it was the same settings...
Post Reply Post Reply


#90
SV Advanced has auto adjusting EPAP, but I don't think you can specify the top of the EPAP range...could be wrong since I don't have one of these. The real critical problem with the PR SV Advanced is it just doesn't respond to CA and hypopnea fast enough. It ends up producing a ton of hypopnea where CA would have been, and seems not to stop hypopnea. When you look at the Aircurve 10 AV, the pressure support is all over the events and you see zero residual events. It' not uncommon to see the Philips leave behind an AHI over 15. You can see it in the details, and it has to be integral to the technology in the software and hardware. It just can't go from zero to 60 in enough time to avoid a collision. Look at the data of a severe complex apnea patient side by side, and it's like looking at an Audi A8 vs an Oldsmobile station wagon.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Does Medicare require a 'sleep doctor'? GrammaBear 7 155 8 hours ago
Last Post: Mosquitobait
  Medicare rules for oxygen HalfAsleep 6 169 09-14-2017, 09:50 PM
Last Post: E.W.
  Getting new machine Tuesday - Medicare mirkee 8 314 09-06-2017, 06:36 PM
Last Post: mirkee
  My spouse is in denial readyforsleep 13 487 08-18-2017, 03:06 PM
Last Post: readyforsleep
  Provider Medicare Billing Ripoff Walla Walla 9 545 07-12-2017, 07:40 PM
Last Post: Walla Walla
  What is approved Medicare FFM list? Lanco 7 333 07-04-2017, 09:02 PM
Last Post: Mosquitobait
  Medicare and replacement schedule Lanco 9 567 07-02-2017, 02:23 AM
Last Post: Hydrangea

Forum Jump:

New Posts   Today's Posts




About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.

For any more information, please use our contact form.