RE: My battery backup setup
Thanks for that info sdb7802, Interesting that they would locate all the components of the detection system inside the power supply. Corporate decision or necessity? Flies in the face of the patent data but perhaps they have revised that?

Since you have done the research, we don't need to guess any further. We have all we need to know to defeat it, if need be.

Dude

ps. That is not to say that defeating it won't bring forth a surprise that we may not like, since reverse engineering can often lead to disappointment. Wink
(This post was last modified: 08-16-2016 01:17 PM by surferdude2.)
08-16-2016 01:12 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Marthajoy7 Offline

Preferred Members

Posts: 67
Joined: Jun 2015

Machine: aircurve 10 auto set to bipap S mode
Mask Type: Full face mask
Mask Make & Model: resmed mirage full face or Clear view full mask
Humidifier: Resmed built in humidifier
CPAP Pressure: 18.4/12 clearview or 18/13 mir
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Female
Location: Seattle

Post: #212
RE: My battery backup setup
Thank You so much. That means it is not a sense line like used on regulated power supplies. Just something to keep most people from using other sources of power, Trying to figure out why the regulator vs voltage dropping resistor. Could it be the regulator circuit uses less power to drop the voltage than a 47K ohm resistor uses to burn off the excess voltage? It would be in the ma range but maybe better use of battery power?
Or maybe the circuits are built so cheaply in china, that why not.
As far as voltage in the CPAP they would be the same, 1.8v across the same internal resistor. which means current flow from supply to machine on the pin line is identical. Only difference I could see is that with the regulator, the current on the pin remains the same, where as with the resistor the current drops slightly as the battery drains. Maybe I am agreeing now that the regulator is better. Battery voltage can vary from 27v to about 22v. So that is about a 20% variance. That means the Pin current can also vary by that amount. It must want that current to be constant no matter what the battery condition. Maybe I need to order some parts.
(This post was last modified: 08-16-2016 02:27 PM by Marthajoy7.)
08-16-2016 02:03 PM
Find all posts by this user Post Reply Quote this message in a reply
PoolQ Offline

Advisory Members

Posts: 986
Joined: Jul 2015

Machine: AirCurve 10 VAUTO
Mask Type: Full face mask
Mask Make & Model: Amara View
Humidifier: Integrated
CPAP Pressure: 6.2, 15, PS4.0
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Sillicon Valley

Post: #213
RE: My battery backup setup
completely consistent with the patent. This is a voltage divider, the low side resistor is in the CPAP and has a fixed value and there is a voltage detection circuit, both of these MUST be in the CPAP in order to control the CPAP operation.

Now this detection circuit is most likely an IC powered either by 5 VDC or 3.3VDC, again most likely 3.3VDC. This means that the voltage it needs to detect must always be below 3.3 VDC.

If you used the 24VDC to generate this voltage, you would need a very high value resistor in the power supply and a very low value in the CPAP. This make getting all the different voltages that represent the different power supplies quite difficult because the resistor values available as standard at the high end have large gaps between them ie 47K and 56K and this makes a very small change in voltage across the low value resistor in the CPAP (it has to be a low value to keep the resulting voltage below 3.3VDC)

if you use 3 volts it gets easier as a single step change in the resistor causes a significant change in voltage
08-16-2016 02:33 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
surferdude2 Online

Advisory Members

Posts: 1,077
Joined: Jul 2014

Machine: ResMed S9 AutoSet
Mask Type: Full face mask
Mask Make & Model: Mirage Quattro & Optilife
Humidifier: ResMed H5i @ #3 if nose breathing & #5 if mouth breathing
CPAP Pressure: 12 ~ 20 & 11 ~12 if nasal mask
CPAP Software: ResScan SleepyHead

Other Comments: Late stage 2 emphysema

Sex: Male
Location: Mousetown, Southern Illinois

Post: #214
RE: My battery backup setup
We are still left wondering how that makes the system any better or how it protects anything or in fact how it does anything at all to justify its existence. A fixed value regulated d.c. voltage can't convey anything other than on and off unless they superimpose some a.c. signal on it and clip it off at the cpap end.

We still don't know what we don't know and probably never will until we can see a schematic of the entire system. Sad
08-16-2016 02:54 PM
Find all posts by this user Post Reply Quote this message in a reply
Perchas Online

Preferred Members

Posts: 38
Joined: Aug 2016

Machine: Philips DreamStation A-CPAP
Mask Type: Nasal mask
Mask Make & Model: Apex Wizard 210 (All Silicon)
Humidifier: Fresh, (Steamless)
CPAP Pressure: 10,5-15,5 cm H2O
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Madrid & Phoenix

Post: #215
RE: My battery backup setup
(08-16-2016 12:18 PM)sdb7802 Wrote:  My tests on three A10 units shows they require at least 22.3V to power up. Once you start the blower, it will run until you lower the voltage to 20.5V.

Post a high res pic and I will try to identify the component. Which machine?



Both you and ve7ltd have performed tests without modifying the 24 V, have reached the same conclusions, notice the bolded 1.8 V, that is the key, the resistor in series with 24 V delivered to the Zener diode those 1.8V, it enables power supply.

So my insistence on trying powered with 12 V and resistance put half the value, instead of 47 k, 20 to 25 K.

The component maybe these by N or M SMD Zener? , I could not follow the printed circuit is four layers.

(08-11-2016 11:34 AM)ve7ltd Wrote:  I did all the probing with the factory AC adapter, and determined what the resistance was between the 3.3V volt supply and the center pin. By grounding the pin, I determined that the resistance is right about 2.5 k ohms for the adapter I have.

I then plugged it in and watched the voltage when the machine booted up, and recorded the voltage to determine what the voltage it was looking for on boot. It was right around 1.8V (I will have to probe again for the two values). The machine seems to have a fixed resistance, and then switches on another resistor to test the adapter, then goes back to the fixed, and then starts some sort of data stream.

I first tried to build a voltage divider to lower the voltage on the pin. I could never get it to work quite right. I then just hooked up a 22K resistor in series, and used a 47K variable to find the range at which the machine would boot with all its features. I found that using a single 47K between 24V and the center pin yielded exactly what I needed - about 1.8V on the startup.


PRS One SE has implemented the central pin, I fed with a battery cable of a Dreamstation with the resistor 20k, testing a battery of 12000 mAmp that reached me today I managed to run for 9 hours 8 cmH2OInspiration at 0.360 mAmp (4W ) with a supply voltage of 5.9 VDC when the battery was almost empty and stop 5.5 V. Respironics are 12,0 Vdc

This test confirms that example's implementation respironics with central pin works with 55% less volts

A test picture, zoom, (sorry for the empty cup of coffee, provides support for battery)

[Image: ce52351de6abb5b74fe47c6e43a4a2e8o.jpg]
(This post was last modified: 08-16-2016 02:58 PM by Perchas.)
08-16-2016 02:55 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Marthajoy7 Offline

Preferred Members

Posts: 67
Joined: Jun 2015

Machine: aircurve 10 auto set to bipap S mode
Mask Type: Full face mask
Mask Make & Model: resmed mirage full face or Clear view full mask
Humidifier: Resmed built in humidifier
CPAP Pressure: 18.4/12 clearview or 18/13 mir
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Female
Location: Seattle

Post: #216
RE: My battery backup setup
I just had an idea. I went to ebay and found this cheaply.

Waterproof-DC-7-50V-To-5V-2A-USB-Female-Step-Down-Converter-Regulator-For-Car-QT

By using this I can also charge phones ect. It still can provide signal to pin with a different resistor. If I can remember ohms law.
A 2700 ohm drops voltage 1.5 volts - Current is 0.56 ma
To drop 5v to 1.8 would be 3.2 so the resistor should be in the vicinity of 5.8 k or computes as 5760.
5.6K or 5.9K would be standard values. If needed could use 2 resistors in series to get exact. How critical is the resistor value? 1% 20%
using same figures for 24 v says resistor should be 40k instead of 47k. But at 25.2 volts it would be 45.3k
Current values might not be quite right, just using figures provided.
lets try again 24 volts to 1.8v - 22.2 volts 222/470000 = 0.4723404 ma
With that current for 3.2v drop would be 6774 ohms or about 6.8k
(This post was last modified: 08-16-2016 05:00 PM by Marthajoy7.)
08-16-2016 04:35 PM
Find all posts by this user Post Reply Quote this message in a reply
sdb7802 Offline

Advisory Members

Posts: 199
Joined: Jan 2016

Machine: PR DreamStation Auto
Mask Type: Nasal pillows
Mask Make & Model: ResMed AirFit P10
Humidifier: built-in with heated hose
CPAP Pressure: 9.5-14 C-Flex 3
CPAP Software: SleepyHead

Other Comments: Retired Electrical Engineer; began using cpap in 2001; FlashAir wifi card and FlashPAP app

Sex: Male
Location: Georgia, USA (The Margravate of Azilia)

Post: #217
RE: My battery backup setup
(08-16-2016 12:18 PM)sdb7802 Wrote:  My tests on three A10 units shows they require at least 22.3V to power up. Once you start the blower, it will run until you lower the voltage to 20.5V.

Just to clarify: These tests were done by feeding the center pin with +3.3V thru a 2.7k series resistor (as Resmed does).

Just my personal opinion. My posts are not medical advice or a statement of fact. Please consult a qualified physician or other qualified medical personnel. Please comply with all applicable laws, codes, regulations, and protocols.
08-16-2016 04:57 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Marthajoy7 Offline

Preferred Members

Posts: 67
Joined: Jun 2015

Machine: aircurve 10 auto set to bipap S mode
Mask Type: Full face mask
Mask Make & Model: resmed mirage full face or Clear view full mask
Humidifier: Resmed built in humidifier
CPAP Pressure: 18.4/12 clearview or 18/13 mir
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Female
Location: Seattle

Post: #218
RE: My battery backup setup
Yes, I understand. Just question accuracy of readings as it seems the measurements were made in circuit. Capacitors and output impedance could affect measurement of resistance value. 5 volts should work just as well and will serve to allow recharging of usb devices. Once the current is known for sure, The resistance can be figured. The current and voltage at the pin will be exactly the same as using a 3.3v converter. The voltage will remain constant regardless of 24 volt batteries. For less than $3 I cannot buy parts to build your converter. I save money, works just as well, and allows another use as well. I have no qualms about just using a resistor, except that the current would change at the pin, based on battery condition. If it were a sense line, that would be preferential method. Since you say Resmed is not treating it as a sense line, then why should I, at least until schematic becomes available. It appears that the pin is going to a comparitor which is being used to hold high or low on a inhibit pin on their cpu. Half a ma is not going to burn out any electronics. That is what we are talking about as being sent through the pin on the connector.
(This post was last modified: 08-16-2016 07:51 PM by Marthajoy7.)
08-16-2016 07:47 PM
Find all posts by this user Post Reply Quote this message in a reply
sdb7802 Offline

Advisory Members

Posts: 199
Joined: Jan 2016

Machine: PR DreamStation Auto
Mask Type: Nasal pillows
Mask Make & Model: ResMed AirFit P10
Humidifier: built-in with heated hose
CPAP Pressure: 9.5-14 C-Flex 3
CPAP Software: SleepyHead

Other Comments: Retired Electrical Engineer; began using cpap in 2001; FlashAir wifi card and FlashPAP app

Sex: Male
Location: Georgia, USA (The Margravate of Azilia)

Post: #219
RE: My battery backup setup
Welcome to the forum: ve7ltd, Marthajoy7, and Perchas

(08-11-2016 11:34 AM)ve7ltd Wrote:  I found that using a single 47K between 24V and the center pin yielded exactly what I needed - about 1.8V on the startup.

I have run this setup on two machines, and used it for 6 nights hooked to a 250aH 12V battery in my trailer. Worked fine every time.

Great find, ve7ltd!Smile

I will test the 47k ohm resistor scheme over the allowable input voltage range of the A10 on my three units.


Martha, I will test your 5V + resistor scheme on my A10s.

I'll let you know the outcome.

BTW: here's an old scope trace of one of my A10s on power up:

   

I ran the data stream (not shown) through a protocol analyzer. Couldn't really make much sense of the data.

Just my personal opinion. My posts are not medical advice or a statement of fact. Please consult a qualified physician or other qualified medical personnel. Please comply with all applicable laws, codes, regulations, and protocols.
(This post was last modified: 08-16-2016 09:41 PM by sdb7802.)
08-16-2016 09:34 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
OMyMyOHellYes Offline

Advisory Members

Posts: 915
Joined: Mar 2013

Machine: Respironics 560 Auto
Mask Type: Nasal mask
Mask Make & Model: Resmed Mirage FX
Humidifier: Respironics System One (no heated hose)
CPAP Pressure: 08.0-15.0 cm/H2O
CPAP Software: EncoreBasic

Other Comments:

Sex: Male
Location: Texas

Post: #220
RE: My battery backup setup
I will keep using my PR System 1s that don't need no stinkin' third line! Just a plain battery that I know how to run.

OMMOHY
08-16-2016 09:42 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  First machine!! Just arrived, setup help please. SJGino 4 254 11-16-2016 08:49 PM
Last Post: trish6hundred
  Battery for Airplane Travel EmilySleeps 5 499 09-16-2016 02:00 PM
Last Post: PoolQ
  Pilot-24 CPAP Battery System from Medistrom sonicboom 15 4,547 09-16-2016 11:07 AM
Last Post: DSchabel
  [Equipment] Camping for 7 days - AirCurve 10 ASV - battery power? Grooves McFly 21 2,749 08-31-2016 01:29 AM
Last Post: Maskup Sleepwell
  [Equipment] Battery usage with CPAP (S7 and A10) MrGadget 10 753 08-26-2016 06:24 AM
Last Post: Bhante
Surprised [Equipment] Emergency backup power Marthajoy7 9 596 08-21-2016 10:00 AM
Last Post: icyura10
  UPS options (Power backup for CPAP) machines Hmm888 14 1,610 08-19-2016 12:28 PM
Last Post: ve7ltd

Forum Jump:

Who's Online (Complete List)