Two issues with using a computer UPS.
1) There battery is a SLA battery with much less A-Hr capacity than you can get for the same money.
-- The UPS just will not last the night.
2) Many UPS use a stepped Sine approximation when running on backup battery.
-- This does not work for some newer computer power supplies with active power factor correction.
-- For the same reason, it may not work with the power brick for a CPAP.

Then you have the double conversion mentioned by Sleepster. Wasted energy.

I would bet that the PR 12V cable also has a reverse polarity protection diode -- possibly a Schottky for low forward drop.

There are different solutions for different scenarios. For camping -- that's one scenario.
For home backup, the most elegant solution I have seen was proposed by, I think, OMMOHY.
A big 12 Volt SLA scooter battery, a switch mode power supply for charge and float charge; and a Super PWRgate PG40s Schottky diode bridge to autoswitch the load from Power supply to Battery; and permit recharge and float charge.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
(This post was last modified: 02-12-2016 12:03 AM by justMongo.)
02-11-2016 10:17 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
OMyMyOHellYes Offline

Advisory Members

Posts: 915
Joined: Mar 2013

Machine: Respironics 560 Auto
Mask Type: Nasal mask
Mask Make & Model: Resmed Mirage FX
Humidifier: Respironics System One (no heated hose)
CPAP Pressure: 08.0-15.0 cm/H2O
CPAP Software: EncoreBasic

Other Comments:

Sex: Male
Location: Texas

Post: #52
RE: My battery backup setup
(02-11-2016 01:49 AM)PoolQ Wrote:  +1
As usual what may seem like the "best" deal may not be. "12V Respironics DC battery adapter cable Respironics (Philips) $39.99' this is basically a USB cable with fewer wires and different connectors on the ends. Check the price of a USB cable and then figure the profit for that at $39.99

At least the Resmed cable actually boosts the voltage from 12V to 24V which is a switching regulator with magnetics in a box and not just some wire.

USB cable? Most certainly not. It is about an 18 or 16 ga wire w/ fused plug and two massive ferrite chokes to arrest RF interference (why? I don't know) and a very non-standard 7.4 mm coax barrel plug but it certainly NOT a USB cable. It should carry continuous 12 V at 15 amps. How would a USB cable do that?

And the purpose of the ResMed adapter? To me is seems to make more money for ResMed in that it is just as easy to come up with a 24 V battery as it is a 12 V battery. Well, at least as easy as it is to come up with two 12 V batteries. But teh benevolent jean-youses at ResMed make you still have to use their adapter even if you have a 24V battery. Because they made it that way. The real question is why did ResMed go non-standard. Why - I speculate it was solely to squeeze more money out of their customers, of course when they could do it themselves a WHOLE lot cheaper and more efficently.

Which is what I do with my Respironics and a 12 V 35 AH battery. And a charger. I do think that the Respironics assembly IS over priced (that's why I built my own) but it is nowhere as overpriced as the clusterfication that ResMed forces on people.

OMMOHY
(This post was last modified: 02-11-2016 10:49 PM by OMyMyOHellYes.)
02-11-2016 10:47 PM
Find all posts by this user Post Reply Quote this message in a reply
sdb7802 Offline

Advisory Members

Posts: 199
Joined: Jan 2016

Machine: PR DreamStation Auto
Mask Type: Nasal pillows
Mask Make & Model: ResMed AirFit P10
Humidifier: built-in with heated hose
CPAP Pressure: 9.5-14 C-Flex 3
CPAP Software: SleepyHead

Other Comments: Retired Electrical Engineer; began using cpap in 2001; FlashAir wifi card and FlashPAP app

Sex: Male
Location: Georgia, USA (The Margravate of Azilia)

Post: #53
RE: My battery backup setup
FYI:

I've found that the A10 will run from a +24 volt power supply (non-ResMed) when the center pin is connected to +3.3 volts thru a 2.7k ohm resistor. I use a +3.3 volt three-terminal voltage regulator fed from the +24 volt power supply. The resistor functions as a pull-up resistor to +3.3 volts and tells the cpap the power rating of the supply. I've tested this on two units.

Steve

Just my personal opinion. My posts are not medical advice or a statement of fact. Please consult a qualified physician or other qualified medical personnel. Please comply with all applicable laws, codes, regulations, and protocols.
02-28-2016 06:32 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
player Offline

Advisory Members

Posts: 412
Joined: Feb 2015

Machine: Resmed A10 Autoset For Her
Mask Type: Nasal pillows
Mask Make & Model: Resmed A10
Humidifier: Resmed A10 Autosset
CPAP Pressure: 11-16
CPAP Software: SleepyHead

Other Comments: It's a... mad world...

Sex: Male
Location: Canada

Post: #54
RE: My battery backup setup
(02-28-2016 06:32 PM)sdb7802 Wrote:  FYI:

I've found that the A10 will run from a +24 volt power supply (non-ResMed) when the center pin is connected to +3.3 volts thru a 2.7k ohm resistor. I use a +3.3 volt three-terminal voltage regulator fed from the +24 volt power supply. The resistor functions as a pull-up resistor to +3.3 volts and tells the cpap the power rating of the supply. I've tested this on two units.

Steve

Wow! You cracked it. Well done. How the heck did you figure that out?
02-28-2016 06:51 PM
Find all posts by this user Post Reply Quote this message in a reply
PoolQ Offline

Advisory Members

Posts: 986
Joined: Jul 2015

Machine: AirCurve 10 VAUTO
Mask Type: Full face mask
Mask Make & Model: Amara View
Humidifier: Integrated
CPAP Pressure: 6.2, 15, PS4.0
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Sillicon Valley

Post: #55
RE: My battery backup setup
(02-11-2016 10:47 PM)OMyMyOHellYes Wrote:  
(02-11-2016 01:49 AM)PoolQ Wrote:  +1
As usual what may seem like the "best" deal may not be. "12V Respironics DC battery adapter cable Respironics (Philips) $39.99' this is basically a USB cable with fewer wires and different connectors on the ends. Check the price of a USB cable and then figure the profit for that at $39.99

At least the Resmed cable actually boosts the voltage from 12V to 24V which is a switching regulator with magnetics in a box and not just some wire.

USB cable? Most certainly not. It is about an 18 or 16 ga wire w/ fused plug and two massive ferrite chokes to arrest RF interference (why? I don't know) and a very non-standard 7.4 mm coax barrel plug but it certainly NOT a USB cable. It should carry continuous 12 V at 15 amps. How would a USB cable do that?

And the purpose of the ResMed adapter? To me is seems to make more money for ResMed in that it is just as easy to come up with a 24 V battery as it is a 12 V battery. Well, at least as easy as it is to come up with two 12 V batteries. But teh benevolent jean-youses at ResMed make you still have to use their adapter even if you have a 24V battery. Because they made it that way. The real question is why did ResMed go non-standard. Why - I speculate it was solely to squeeze more money out of their customers, of course when they could do it themselves a WHOLE lot cheaper and more efficently.

Which is what I do with my Respironics and a 12 V 35 AH battery. And a charger. I do think that the Respironics assembly IS over priced (that's why I built my own) but it is nowhere as overpriced as the clusterfication that ResMed forces on people.

OMMOHY

I said it was "basically" a USB cable, as in meaning no electronics involved. As you described it fits my meaning exactly. The manufactured cost of that cable is close to a USB cable. To give them the benefit of the doubt lets say it costs $5 and they sell it for $40 provides them an 87.5% profit margin.

As for Resmed at least they have a boost converter in the cable.

Do they all want to make money on something that they sell very few of, of course.

Did they go custom connector and perhaps voltage to force the few that do this to spend more money, perhaps. My guess is that it is not really worth their time for the added revenue, even at that high profit margin. They do it to support their customer and to keep the unskilled customer from blowing up their CPAP and then claiming that it just caught fire in the middle of the night.
02-28-2016 07:14 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
surferdude2 Offline

Advisory Members

Posts: 1,077
Joined: Jul 2014

Machine: ResMed S9 AutoSet
Mask Type: Full face mask
Mask Make & Model: Mirage Quattro & Optilife
Humidifier: ResMed H5i @ #3 if nose breathing & #5 if mouth breathing
CPAP Pressure: 12 ~ 20 & 11 ~12 if nasal mask
CPAP Software: ResScan SleepyHead

Other Comments: Late stage 2 emphysema

Sex: Male
Location: Mousetown, Southern Illinois

Post: #56
RE: My battery backup setup
Computer motherboards running from ATX power supplies use a similar protection by feeding a common connection back to the PS to verify that all is well with the voltage prior to allowing it to be fully applied. It is common practice to use these power supplies in an outboard mode for other purposes, such as powering an external hard drive. When used in that mode, one ties down the signal line to the common line to fool the unit into fully powering up all output lines.

In the case of a Resmed unit, we may be left wondering if we are bypassing some important protective feature when we clamp the signal line with our own generation of the voltage signal. Could be, could not be, who's to say? Resmed tech opinion would be nice. Without a circuit schematic, we can't know for sure that we won't do some harm.

Dude
(This post was last modified: 02-28-2016 08:03 PM by surferdude2.)
02-28-2016 07:56 PM
Find all posts by this user Post Reply Quote this message in a reply
justMongo Offline

Monitors

Posts: 3,853
Joined: Sep 2013

Machine: ResMed VPAP Auto (S9)
Mask Type: Full face mask
Mask Make & Model: ResMed Mirage Quattro
Humidifier: ResMed H5i
CPAP Pressure: 18 IPAPmax - 11 EPAPmin, PS=5
CPAP Software: Other Software

Other Comments: μολὼν λαβέ

Sex: Male
Location: Калифорния

Post: #57
RE: My battery backup setup
Dude:
I guess we'll not be getting a SAM's Photofact manual on that!

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
02-28-2016 08:19 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
surferdude2 Offline

Advisory Members

Posts: 1,077
Joined: Jul 2014

Machine: ResMed S9 AutoSet
Mask Type: Full face mask
Mask Make & Model: Mirage Quattro & Optilife
Humidifier: ResMed H5i @ #3 if nose breathing & #5 if mouth breathing
CPAP Pressure: 12 ~ 20 & 11 ~12 if nasal mask
CPAP Software: ResScan SleepyHead

Other Comments: Late stage 2 emphysema

Sex: Male
Location: Mousetown, Southern Illinois

Post: #58
RE: My battery backup setup
It's likely ol' Howard would have had a problem there. Wink
02-28-2016 08:53 PM
Find all posts by this user Post Reply Quote this message in a reply
sdb7802 Offline

Advisory Members

Posts: 199
Joined: Jan 2016

Machine: PR DreamStation Auto
Mask Type: Nasal pillows
Mask Make & Model: ResMed AirFit P10
Humidifier: built-in with heated hose
CPAP Pressure: 9.5-14 C-Flex 3
CPAP Software: SleepyHead

Other Comments: Retired Electrical Engineer; began using cpap in 2001; FlashAir wifi card and FlashPAP app

Sex: Male
Location: Georgia, USA (The Margravate of Azilia)

Post: #59
RE: My battery backup setup
(02-28-2016 06:51 PM)player Wrote:  
(02-28-2016 06:32 PM)sdb7802 Wrote:  FYI:

I've found that the A10 will run from a +24 volt power supply (non-ResMed) when the center pin is connected to +3.3 volts thru a 2.7k ohm resistor. I use a +3.3 volt three-terminal voltage regulator fed from the +24 volt power supply. The resistor functions as a pull-up resistor to +3.3 volts and tells the cpap the power rating of the supply. I've tested this on two units.

Steve

Wow! You cracked it. Well done. How the heck did you figure that out?

1.) With the AC supply powered on and disconnected from the cpap; I measured +3.3 volts between center pin and common (outer shell) with a dc voltmeter.

2.) With the AC supply powered off and disconnected from the cpap; I measured 2.7k ohms between center pin and common (outer shell) with an ohmmeter.

3.) Using an oscilloscope, I verified that the signal on the center pin was the same whether being powered by the ResMed supply or my setup. No clipping or clamping was noted.

4.) I also varied the voltage and resistance until the error display came on. My tests on two units showed the voltage could range from 2.4 volts to 3.5 volts. The resistor could range between 2.3k to 3.55k ohms.

5.) I've tested both units (alternately) with my setup for 2 weeks.

I've worked in electronics for over 40 years and I know that just about anything can go wrong. Time will tell.

Steve

Just my personal opinion. My posts are not medical advice or a statement of fact. Please consult a qualified physician or other qualified medical personnel. Please comply with all applicable laws, codes, regulations, and protocols.
02-28-2016 10:36 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
surferdude2 Offline

Advisory Members

Posts: 1,077
Joined: Jul 2014

Machine: ResMed S9 AutoSet
Mask Type: Full face mask
Mask Make & Model: Mirage Quattro & Optilife
Humidifier: ResMed H5i @ #3 if nose breathing & #5 if mouth breathing
CPAP Pressure: 12 ~ 20 & 11 ~12 if nasal mask
CPAP Software: ResScan SleepyHead

Other Comments: Late stage 2 emphysema

Sex: Male
Location: Mousetown, Southern Illinois

Post: #60
RE: My battery backup setup
Good work Steve. Thanks for taking the time to do that and reporting the results. Now we may consider that the processor board of the flow generator has a simple logic chip that passes along a 3 volt D.C. potential to the signal wire unless the input voltage (24 volt) is higher or lower by roughly 20%. That's still a blind guess, since we still don't know what we don't know.

I doubt it would ever harm the equipment to provide a steady 3 volts to the signal wire and hook a 24 volt battery to the power wires (properly polarized). I wouldn't hesitate to do that to my own equipment but would caution others of the risk involved.

Dude
(This post was last modified: 02-28-2016 11:32 PM by surferdude2.)
02-28-2016 11:25 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  First machine!! Just arrived, setup help please. SJGino 4 269 11-16-2016 08:49 PM
Last Post: trish6hundred
  Battery for Airplane Travel EmilySleeps 5 507 09-16-2016 02:00 PM
Last Post: PoolQ
  Pilot-24 CPAP Battery System from Medistrom sonicboom 15 4,583 09-16-2016 11:07 AM
Last Post: DSchabel
  [Equipment] Camping for 7 days - AirCurve 10 ASV - battery power? Grooves McFly 21 2,785 08-31-2016 01:29 AM
Last Post: Maskup Sleepwell
  [Equipment] Battery usage with CPAP (S7 and A10) MrGadget 10 767 08-26-2016 06:24 AM
Last Post: Bhante
Surprised [Equipment] Emergency backup power Marthajoy7 9 601 08-21-2016 10:00 AM
Last Post: icyura10
  UPS options (Power backup for CPAP) machines Hmm888 14 1,648 08-19-2016 12:28 PM
Last Post: ve7ltd

Forum Jump:

Who's Online (Complete List)