RE: My battery backup setup
(06-06-2016 12:01 PM)surferdude2 Wrote:  the load in question will not receive the proper voltage and will either not operate at all or will exhibit strange behavior.

My mistake for not being clear.

I guess what's most important is how much power is required when the machine is in camping mode, which, for me, would be stripped down to use the least amount of power to conserve on battery juice:
  • Connected to a 12VDC car battery (up converted & send resistored)
  • The normal apap settings for me
  • No humidifier
  • No heater, no heated hose

If the power consumed under those minimum-use circumstances was, say, under 30Watts, then there's nothing useful a 90W power supply could supply over what the 30 Watt power supply supplies.

Right?

So, I guess the question can't be answered unless we know how much current is required at 24 VDC for the machine in the typical "camp" mode, which is just the cpap and nothing else.

QUESTION:
Q2: Does anyone know what current an A10 consumes under typical camp mode (e.g., we can assume apap 10 to 20 cm water if that matters)?
06-06-2016 01:39 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
justMongo Offline

Monitors

Posts: 3,844
Joined: Sep 2013

Machine: ResMed VPAP Auto (S9)
Mask Type: Full face mask
Mask Make & Model: ResMed Mirage Quattro
Humidifier: ResMed H5i
CPAP Pressure: 18 IPAPmax - 11 EPAPmin, PS=5
CPAP Software: Other Software

Other Comments: μολὼν λαβέ

Sex: Male
Location: Калифорния

Post: #82
RE: My battery backup setup
You're pretty much stuck using the ResMed 12VDC to 24 VDC up-converter.
So, the question should be, what current is drawn from the battery in that configuration.
It will draw about 1 Ampere. check this document: http://www.resmed.com/us/dam/documents/a...lo_eng.pdf
See page 20.

Seems you have access to a PR machine... Someone like Sleeprider will be able to translate your setting from A10 to PR machine. Then you can use 12 VDC directly.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
06-06-2016 02:32 PM
Find all posts by this user Post Reply Quote this message in a reply
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,331
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #83
RE: My battery backup setup
Anyone have a handle on how many 12-volt amp hours a Resmed A10 needs to operate through the night? Assume 24 v up-converter, no humidity or heated hose.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
06-06-2016 09:20 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Sleepster Offline
Wiki Editor
Moderators

Posts: 4,991
Joined: Feb 2012

Machine: ResMed AirCurve10 VAuto
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: HumidAir and SlimLine Hose
CPAP Pressure: MaxI 13.6 | MinE 5.2 | PS 4.4
CPAP Software: ResScan SleepyHead

Other Comments: Diagnosed Nov 2011. Conquered aerophagia.

Sex: Male
Location: Houston, Texas

Post: #84
RE: My battery backup setup
(06-06-2016 09:20 PM)Sleeprider Wrote:  Anyone have a handle on how many 12-volt amp hours a Resmed A10 needs to operate through the night? Assume 24 v up-converter, no humidity or heated hose.

I don't know the answer, but I do know the answer depends on the pressure. The higher the pressure the shorter the battery life.

Sleepster
Apnea Board Moderator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
06-06-2016 11:12 PM
Find all posts by this user Post Reply Quote this message in a reply
sdb7802 Offline

Advisory Members

Posts: 199
Joined: Jan 2016

Machine: PR DreamStation Auto
Mask Type: Nasal pillows
Mask Make & Model: ResMed AirFit P10
Humidifier: built-in with heated hose
CPAP Pressure: 9.5-14 C-Flex 3
CPAP Software: SleepyHead

Other Comments: Retired Electrical Engineer; began using cpap in 2001; FlashAir wifi card and FlashPAP app

Sex: Male
Location: Georgia, USA (The Margravate of Azilia)

Post: #85
RE: My battery backup setup
from ResMed doc linked by Mongo:
   

Just my personal opinion. My posts are not medical advice or a statement of fact. Please consult a qualified physician or other qualified medical personnel. Please comply with all applicable laws, codes, regulations, and protocols.
06-07-2016 04:18 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,331
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #86
RE: My battery backup setup
Thanks sdb7802 (and JM). That is the approximation I was looking for.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
06-07-2016 08:32 AM
Find all posts by this user Post Reply Quote this message in a reply
verbatim1 Offline

Preferred Members

Posts: 70
Joined: May 2016

Machine: Remstar Auto
Mask Type: Nasal mask
Mask Make & Model: F&P Eson Nasak Mask SL-Large for CPAP or bi-level venti
Humidifier: none
CPAP Pressure: 10 to 20
CPAP Software: SleepyHead EncoreBasic

Other Comments:

Sex: Undisclosed
Location:

Post: #87
RE: My battery backup setup
(06-06-2016 02:32 PM)justMongo Wrote:  You're pretty much stuck using the ResMed 12VDC to 24 VDC up-converter.

If I stick with the ResMed device when camping, the up-converter seems to be the way to go, for finesse.

Of course, putting two 12VDC 35Amp-Hour gel cells in series gives us 24VDC, so then all we need to handle is the 3.3VDC voltage regulator with 2.7KOhm sense resistor circuitry...
[Image: ACnxeA.jpg]

(06-06-2016 02:32 PM)justMongo Wrote:  So, the question should be, what current is drawn from the battery in that configuration.
It will draw about 1 Ampere. check this document: http://www.resmed.com/us/dam/documents/a...lo_eng.pdf
See page 20.
Thanks for locating that information that we can nominally assume about 1 amp for standard cpap settings. For an 8-hour night, that's 8amp hours, so, it seems like two nights would do with, at a minimum, a 16-amp-hour battery, so, the often stated 35 Amp Hour battery should be fine for a weekend, right?

(06-06-2016 02:32 PM)justMongo Wrote:  Seems you have access to a PR machine... Then you can use 12 VDC directly.

Yes. You are correct.

My good friend and I both abhor non-standard power supplies and cables (we met at an Apple conference, where we lamented their "walled garden" approach).

When I told him about the hugely non-standard ResMed AirSense 10 Audoset power supply requirements, he immediately gave me access to his standard-power-supply-fed Philips Respironics Bi-PAP Auto Bi-Flex 750P pictured below, so that I wouldn't have to deal with the purposefully-non-standard ResMed power supply issues.

He said I only need to give it back if his current machine breaks down.
[Image: oVB12T.jpg]

This 12VDC device "says" it sinks 5 Amps maximum, but, just like with the ResMed, I suspect it's less than half that on average. Given 2.5 Amps, and 7 hours a night, a two-night campout should consume 2.5 amps over 14 hours, which would dictate a 35 Amp Hour battery.
(06-06-2016 02:32 PM)justMongo Wrote:  Someone like Sleeprider will be able to translate your setting from A10 to PR machine.

As you noted, Sleeprider is advising me for the task of matching the settings of the non-standard-power A10 to the standard-power 750P for camping purposes:
http://www.apneaboard.com/forums/Thread-...amping-use

Given that the non-standard-power ResMed is set to automatic of 10 to 20 cm of water with a pressure relief of about 2 cm of water, so far I've set my friend's standard-power Respironics 750P to the following settings, which I am testing as we speak:
- Mode = Auto
- Max IPAP = 20.0
- Min EPAP = 7.0
- Max PS = 3.0
- Flex Type = None
- Rise Time = 0 <---------- this appeared when I set flex type to none!
- Bi-Flex = 3 <---------- this disappeared when I set flex type to none!
- SYSTEM = X2
- LOCK SYS = off
- Ramp Time = 0:10
- Ramp Start = 7.0 <---------- this dropped from 7.5 to 7.0 when I set flex type to none!
- Auto on = on
- Auto off = on
- Mask alert = on
- Humidifier = off
- Show AHI = on
- Split night = off

I think the 750P has a default minimum pressure support of 2cm of water, so, tonight I will probably change the Minimum EPAP to 8 (from the 7 that Sleepster recommended above) so as to better match the 10cm to 20cm APAP settings of the non-standard-power ResMed machine.

I'll test these settings again tonight, so that I have a ready-made 12 volt standard-power-supply cpap machine available to me, so that I don't have to deal with the issues that ResMed put on the table just to make the power supply non standard.
(This post was last modified: 06-08-2016 12:37 AM by verbatim1.)
06-07-2016 11:58 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
verbatim1 Offline

Preferred Members

Posts: 70
Joined: May 2016

Machine: Remstar Auto
Mask Type: Nasal mask
Mask Make & Model: F&P Eson Nasak Mask SL-Large for CPAP or bi-level venti
Humidifier: none
CPAP Pressure: 10 to 20
CPAP Software: SleepyHead EncoreBasic

Other Comments:

Sex: Undisclosed
Location:

Post: #88
RE: My battery backup setup
(06-07-2016 04:18 AM)sdb7802 Wrote:  from ResMed doc linked by Mongo:

You know what's strange about that chart?

Notice the voltage on the current-draw column?

That voltage doesn't make sense, for an A10 (which is 24VDC, not 12VDC).

Or am I misreading the chart?
06-08-2016 12:30 AM
Find all posts by this user Post Reply Quote this message in a reply
sdb7802 Offline

Advisory Members

Posts: 199
Joined: Jan 2016

Machine: PR DreamStation Auto
Mask Type: Nasal pillows
Mask Make & Model: ResMed AirFit P10
Humidifier: built-in with heated hose
CPAP Pressure: 9.5-14 C-Flex 3
CPAP Software: SleepyHead

Other Comments: Retired Electrical Engineer; began using cpap in 2001; FlashAir wifi card and FlashPAP app

Sex: Male
Location: Georgia, USA (The Margravate of Azilia)

Post: #89
RE: My battery backup setup
(06-08-2016 12:30 AM)verbatim1 Wrote:  You know what's strange about that chart?

Notice the voltage on the current-draw column?

That voltage doesn't make sense, for an A10 (which is 24VDC, not 12VDC).

Or am I misreading the chart?

When I copied the relevant part of the chart, I left out the page heading (on a preceding page where they introduce the 24 volt machines):

ResMed devices powered with converter

So they mean the 12 volt battery current drawn by the 24 volt converter and 24 volt cpap.

Sorry for the omission.

Just my personal opinion. My posts are not medical advice or a statement of fact. Please consult a qualified physician or other qualified medical personnel. Please comply with all applicable laws, codes, regulations, and protocols.
06-08-2016 01:01 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
verbatim1 Offline

Preferred Members

Posts: 70
Joined: May 2016

Machine: Remstar Auto
Mask Type: Nasal mask
Mask Make & Model: F&P Eson Nasak Mask SL-Large for CPAP or bi-level venti
Humidifier: none
CPAP Pressure: 10 to 20
CPAP Software: SleepyHead EncoreBasic

Other Comments:

Sex: Undisclosed
Location:

Post: #90
RE: My battery backup setup
(02-28-2016 06:32 PM)sdb7802 Wrote:  I've found that the A10 will run from a +24 volt power supply (non-ResMed) when the center pin is connected to +3.3 volts thru a 2.7k ohm resistor. I use a +3.3 volt three-terminal voltage regulator fed from the +24 volt power supply. The resistor functions as a pull-up resistor to +3.3 volts and tells the cpap the power rating of the supply. I've tested this on two units.

Steve

Hi Steve,
I'm not an electrical engineer, so, can you kindly vet this diagram of what I think you are suggesting for the ResMed non-standard battery-based power supply circuitry?

Would you please let me know if/where I made mistakes?
[Image: bnXymR.jpg]
06-08-2016 02:13 AM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  First machine!! Just arrived, setup help please. SJGino 4 259 11-16-2016 08:49 PM
Last Post: trish6hundred
  Battery for Airplane Travel EmilySleeps 5 501 09-16-2016 02:00 PM
Last Post: PoolQ
  Pilot-24 CPAP Battery System from Medistrom sonicboom 15 4,551 09-16-2016 11:07 AM
Last Post: DSchabel
  [Equipment] Camping for 7 days - AirCurve 10 ASV - battery power? Grooves McFly 21 2,756 08-31-2016 01:29 AM
Last Post: Maskup Sleepwell
  [Equipment] Battery usage with CPAP (S7 and A10) MrGadget 10 756 08-26-2016 06:24 AM
Last Post: Bhante
Surprised [Equipment] Emergency backup power Marthajoy7 9 600 08-21-2016 10:00 AM
Last Post: icyura10
  UPS options (Power backup for CPAP) machines Hmm888 14 1,614 08-19-2016 12:28 PM
Last Post: ve7ltd

Forum Jump:

Who's Online (Complete List)