Hello Guest, Welcome to Apnea Board !
As a guest, you are limited to certain areas of the board and there are some features you can't use.
To post a message, you must create a free account using a valid email address.

or Create an Account


New Posts   Today's Posts

Need Help Setting Vpap/Bilevel
#1
Question 
Need Help Setting Vpap/Bilevel

Paul from Perth,

My new machine has arrived. In another thread you had said that someone could help me set this bpap machine. Here is the maker and type of machine that I have.

Resmed Vpap/Bilevel adapt SV is my new machine.

I'm not sure about what is involved in setting my machine, but the doctor had wanted the pressure set at 8/4. I believe that's inhalation 8, inhallation 4. That is less than what my cpap machine's pressure is but I think the doctor wants to start there and change if necessary.

Thanks, Kate

:Using cpap then vpap since Feb.2013,
Kate
Post Reply Post Reply
#2
RE: Need Help Setting Vpap/Bilevel
(01-27-2014, 06:38 PM)Kate Wrote: I'm not sure about what is involved in setting my machine, but the doctor had wanted the pressure set at 8/4. I believe that's inhalation 8, inhallation 4.
This is above my paygrade, 4 EPAP (exhale pressure) ... 8 IPAP (inhale pressure)
8-4= 4 ... pressure support (the difference between IPAP and EPAP)

CPAP EPR is not exactly the same as bi-level pressure support




Post Reply Post Reply
#3
RE: Need Help Setting Vpap/Bilevel
I made a mistake. It's inhallation 8, exhallation 4. I was genetically diagnosed with narcolepsy and I have found that some days, it's better to stay in bed and not communicate.
:Using cpap then vpap since Feb.2013,
Kate
Post Reply Post Reply
#4
RE: Need Help Setting Vpap/Bilevel
I imagine one of the smart guys will be by shortly to tell you how to do it. While I know which settings to adjust, I do not know how to get into the menu of your machine.

Unless the Doc told you not to do so, I would put the machine in 'auto' mode, with those numbers as the minimums - then by wathing the data (in software) you can learn over time what the machine thinks is best for you.
*I* am not a DOCTOR or any type of Health Care Professional.  My thoughts/suggestions/ideas are strictly only my opinions.

"Only two defining forces have ever offered to die for you. Jesus Christ and the American Soldier. One died for your Soul, the other for your Freedom."
Post Reply Post Reply
#5
RE: Need Help Setting Vpap/Bilevel
(01-27-2014, 06:38 PM)Kate Wrote: I'm not sure about what is involved in setting my machine, but the doctor had wanted the pressure set at 8/4.

Hi Kate,

Your profile says your (old) pressure was 9. This would seem to indicate your pressure when exhaling needed to be 9 or a little lower (if you were using EPR) to prevent obstructve apneas.

With the change to bilevel, it is possible your doctor is prescribing EPAP = 8 (to continue to prevent obstructive events) and Pressure Support (PS) = 4. This would make your inhale pressure 12. I think these would be reasonable settings for someone who needed a pressure of 9 when on standard CPAP machine. You can ask your doctor if that was his intention.

When treating central events an ASV machine somtimes needs to automatically increase Pressure Support as high as 10 or even higher, to keep you breathing. I reccomend setting the supper limit for Pressure Support at 10 or higher. The PS won't go higher than necessary.

Take care,
--- Vaughn
The Advisory Member group provides advice and suggestions to Apnea Board administrators and staff on matters concerning Apnea Board operation and administrative policies.  Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment.
Post Reply Post Reply
#6
RE: Need Help Setting Vpap/Bilevel
Hi Kate - that's the same machine I have. If you want the full clinical manual, go to this page and follow the directions. (You'll find it's listed in Section 3, manuals that are only available by email).

http://www.apneaboard.com/adjust-cpap-pr...tup-manual

You can get into the "secret" clinical menu by holding down the push dial (the big button) together with the setup button (with the two tick marks) for 3 seconds.

Once in the clinical menu, you should already be in the settings menu. If not, turn the push dial to get to the settings menu then push down to click the dial.

In the settings menu, turn the dial to get to EPAP, click on it, then turn the dial to set your EPAP (exhalation) value. Click again to lock it in.

Do the same for maximum and minimum pressure support(PS).

Once you've go everything set, turn the dial to get to "Home", click it, and you're done.

Following Vaugn's advice, set the EPAP to 8, the minimum PS to 4, and the maximum PS to 10. Also as per Vaugn's advice, confirm these settings with your doctor. These machines have a number of different modes, and it's easy to get confused.

Hope this helps.
Post Reply Post Reply
#7
RE: Need Help Setting Vpap/Bilevel
(01-28-2014, 05:01 AM)vsheline Wrote:
(01-27-2014, 06:38 PM)Kate Wrote: I'm not sure about what is involved in setting my machine, but the doctor had wanted the pressure set at 8/4. I believe that's inhalation 8, inhallation 4.

Hi Kate,

Your profile says your (old) pressure was 9. This would seem to indicate your pressure when exhaling needed to be 9 or a little lower (if you were using EPR) to prevent obstructve apneas.
Not necessarily.

When titrating straight CPAP, it is typical that the OAs are the first events eliminated (they turn into hypopneas) and that it requires more (inhalation) pressure to properly prevent the residual hypopneas--i.e. the "OAs turned into Hs". On a straight CPAP titration, they keep increasing the pressure until these residual Hs are eliminated. On a bi-level titration the EPAP is increased for OAs, but not Hs; the IPAP is usually increased for both OAs and Hs. So once the OAs are eliminated, the EPAP is not increased further, but they continue to increase the IPAP until the residual Hs disappear.

As a result, for people being switched from straight CPAP to bi-level, it is not uncommon for the EPAP pressure to be several cm lower than the old straight CPAP pressure and the IPAP pressure to be roughly the same as the old CPAP pressure.

Quote:With the change to bilevel, it is possible your doctor is prescribing EPAP = 8 (to continue to prevent obstructive events) and Pressure Support (PS) = 4. This would make your inhale pressure 12. I think these would be reasonable settings for someone who needed a pressure of 9 when on standard CPAP machine. You can ask your doctor if that was his intention.
Without knowing more about how the script is actually written I would NOT make this assumption. It could be that the OP has some problems with pressure induced centrals (otherwise why the switch to an ASV machine?) and hence the doc wants to see if reducing the overall pressure needs is enough to fix the pressure induced centrals. It is quite possible that an IPAP = 8 or 9 is enough to take care of the Hs, and the reasoning would then be that less pressure is needed to take care of the OAs.

Quote:When treating central events an ASV machine somtimes needs to automatically increase Pressure Support as high as 10 or even higher, to keep you breathing. I reccomend setting the supper limit for Pressure Support at 10 or higher. The PS won't go higher than necessary.
The exact meaning of the PS setting on a VPAP ASV machine depends on the mode that you are using the machine in. If you're not using it in ASV mode, then there will only be one PS setting and it will fix the distance between EPAP and IPAP. In other words, if the machine is not in ASV mode, setting (min) EPAP = 8 and PS = 10 will result in a starting IPAP = (min) EPAP + PS = 8 + 10.

In ASV mode, however, there are two PS settings. The min PS setting is what controls the relationship between IPAP and EPAP when the machine is not trying to trigger breaths. At the start of the night, IPAP = EPAP + min PS. When the machine needs to trigger an inhalation, the IPAP is increased (sometimes dramatically) while the EPAP is left alone. The maximum the IPAP can increase to, however, is controlled by the max PS setting: max IPAP = EPAP + max PS.

To the OP: You need to clarify with your doctor exactly how s/he wants your machine set up. These are the questions you need to get answers to in writing:

What mode to use? Regular VPAP S, VPAP ST, or VPAP ASV?

Min EPAP = What?
Min PS = What?
Max PS = What?
Max IPAP = What?

On the S9 VPAPs there are also settings that control how and when the machine changes from IPAP to EPAP and back. So you will need to know what the doc wants you to use for:

TI max = What?
TI min = What?
Trigger sensitivity = What?

Note that in some modes, you might not have all these settings. In other modes, you will have more settings that you need to have the doc's written instructions in order to get the machine set up correctly. In both ST an ASV mode, there will be settings to control when the machine starts to trigger inhalations (the "breaths per minutes rate" and the "target minute ventilation" settings).
Post Reply Post Reply
#8
RE: Need Help Setting Vpap/Bilevel
(01-27-2014, 06:38 PM)Kate Wrote: My new machine has arrived. In another thread you had said that someone could help me set this bpap machine. Here is the maker and type of machine that I have.

Resmed Vpap/Bilevel adapt SV is my new machine.
Do you know why the doc ordered a VPAP Adapt SV machine rather than the much simpler and much less expense VPAP Auto???

VPAP adapt SVs are usually NOT prescribed unless there is an identifiable problem with central apneas.

VPAP Autos are (rather routinely) prescribed when a person has had difficulty adjusting to CPAP or APAP.

It would help us help you make more sense about the settings if we knew why you were prescribed an adapt SV?
Post Reply Post Reply
#9
RE: Need Help Setting Vpap/Bilevel
Robysue,

The doctor wanted me to have this machine because most of my apnea events were centrals, around 15 an hour.

Kate
:Using cpap then vpap since Feb.2013,
Kate
Post Reply Post Reply
#10
RE: Need Help Setting Vpap/Bilevel
(01-28-2014, 10:42 PM)Kate Wrote: Robysue,

The doctor wanted me to have this machine because most of my apnea events were centrals, around 15 an hour.

Kate
Then the doc probably wants you to use the machine in ASV mode. It's just as important (probably MORE important) to get the ASV-only settings right on the machine as it is to get the EPAP/IPAP settings right. It's those ASV settings that control how the machine will treat your central apneas.

I'd suggest calling the doc's office back and telling them you've got the machine, but it still needs to be set up and that you are not sure about exactly what the doc wants all the settings to be.

If you're having problems with centrals, that may also explain why the EPAP pressure seems to be several cm lower than your old straight CPAP setting.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Flow Limits How to Upgrade CPAP to Bilevel Deborah K. 110 9,055 Today, 07:31 AM
Last Post: SeePak
  F40 Mask Setting MrIvanDrago 5 115 Yesterday, 06:35 PM
Last Post: BuzzaJim
  VAuto Trigger Setting MrIvanDrago 2 80 Yesterday, 12:08 PM
Last Post: MrIvanDrago
  Diagnosed with UARS just got CPAP with mask need help setting it up nicchamilton 3 122 Yesterday, 10:48 AM
Last Post: nicchamilton
  [Treatment] CPAP to Bilevel advice kkjacks 24 1,097 04-22-2024, 02:16 PM
Last Post: kkjacks
  N30i [what mask setting] Airsense 10 AutoSet Topdogg7505 8 3,258 04-20-2024, 10:02 AM
Last Post: ahfairley
  New Bilevel user help with settings (oscar attached) opo218 6 178 04-15-2024, 02:56 PM
Last Post: opo218