Settings are as follows:
Mode:VAuto
Max Ipap:12
Min Epap:8
PS:3
TiMax:2.0
TiMin:.3
Trigger:Med
Cycle:low
Ramp:15 Min
Start Epap:4.0
Climate Control:auto
Tube Temp:auto
Do these setting sound correct for the first night?
Hello Guest,
Welcome to Apnea Board !
New ResMed Aircurve V10 Auto
|
02-27-2020, 05:38 PM
New ResMed Aircurve V10 Auto
So I purchased a used Aircurve V10Auto after using a Respironics BiPap Pro with over 35000 hours on it for 13 years. I listed the settings below and wonder if there are any recommendations to adjust--BiPap pro setting was 12/8 and worked fine for me for years but have no data to support it.
Settings are as follows: Mode:VAuto Max Ipap:12 Min Epap:8 PS:3 TiMax:2.0 TiMin:.3 Trigger:Med Cycle:low Ramp:15 Min Start Epap:4.0 Climate Control:auto Tube Temp:auto Do these setting sound correct for the first night?
02-27-2020, 05:49 PM
RE: New ResMed Aircurve V10 Auto
If you have been using this pressure for 13 years, then you are probably very used to the pressure. I don't see a need to use the Ramp feature.
Crimson Nape
Apnea Board Moderator www.ApneaBoard.com ___________________________________ Useful Links -or- When All Else Fails: The Guide to Understanding OSCAR OSCAR Chart Organization Attaching Images and Files on Apnea Board Apnea Helpful Tips INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
02-27-2020, 05:51 PM
RE: New ResMed Aircurve V10 Auto
I'd suggest getting OSCAR and post a night's usage chart according to the standard order in my signature. We can help suggest settings from there. Alternate is others that are better at setting transfer between machines may help with those older machine settings applying it to your new one.
Dave
OSCAR Standard OSCAR Chart Order Mask Primer Dealing With A DME Soft Cervical Collar Wiki INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEBSITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
02-27-2020, 06:05 PM
RE: New ResMed Aircurve V10 Auto
Jeff, you have been using fixed BPAP at 12/8 which is PS 4. If you want to try Vauto with variable pressure, try this:
EPAP min 7.0 PS 4.0 Max pressure 16.0 I would stick with trigger and cycle at medium unless there is something you are specifically concerned about, and turn off ramp. There is no need to start at EPAP 4, and you would feel suffocated by that anyway. I think starting at these settings will be closer to default and allow changes to be made after you have some data. The lower EPAP min give you a chance to try lower pressure, but I suspect you will be at or above your old settings most of the night. Similarly, you might as well allow IPAP to move above your old pressure and 11/7 to 16/12 is a pretty good range considering your old settings.
Sleeprider
Apnea Board Moderator www.ApneaBoard.com ____________________________________________ Download OSCAR Software Soft Cervical Collar Optimizing Therapy Organize your OSCAR Charts Attaching Files Mask Primer How To Deal With Equipment Supplier INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
02-27-2020, 06:10 PM
(This post was last modified: 02-27-2020, 06:11 PM by jeffdanger.)
RE: New ResMed Aircurve V10 Auto
Good point--I tested it with and without the ramp and it doesn't seem to release pressure on exhalation as much as the Bipap pro did--somewhat uncomfortable so I may need the ramp.
I downloaded the Oscar software and look forward to seeing the results tomorrow morning. Will post when I have.
02-27-2020, 06:16 PM
RE: New ResMed Aircurve V10 Auto
That is part of the reason for suggesting a lower EPAP pressure on Vauto. The BiPAP uses BiFlex to amplify the pressure drop, and often does so ahead of expiration. The Resmed always follows respiratory leads, and unless TiMAX is reached, may be at maximum IPAP when you initiate exhale, however it quickly drops. Most of us quicly find the Vauto algorithm comfortable, however if you use Ramp, start at the highest pressure you find tolerable and keep it short (10 min). Also, if you were using cycle sensitivity LOW as in your original settings, you got what you asked for.
Sleeprider
Apnea Board Moderator www.ApneaBoard.com ____________________________________________ Download OSCAR Software Soft Cervical Collar Optimizing Therapy Organize your OSCAR Charts Attaching Files Mask Primer How To Deal With Equipment Supplier INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
02-27-2020, 06:16 PM
RE: New ResMed Aircurve V10 Auto
I'll try those setting tonight but leaving the ramp for now. I tried it with and without the ramp as a test but this unit doesn't seem to release as much pressure on exhale as my old one (which was probably well worn out anyway).
02-28-2020, 07:01 AM
RE: New ResMed Aircurve V10 Auto
Have my first Oscar chart from last night. A lot of the apneas seem to be in the first two hours as I was adjusting toi the first time usage. Also seems mask leaks were high. Any comments or suggestions?
Thanks
02-28-2020, 07:58 AM
RE: New ResMed Aircurve V10 Auto
Jeffdanger your leaks are too high your machine does not know what is going on you need to look at the mask primer here http://www.apneaboard.com/wiki/index.php...ask_Primer
02-28-2020, 08:47 AM
RE: New ResMed Aircurve V10 Auto
I agree with JasW that with those leaks, you cannot rely on the data from the machine to be useful. The starting point is to read the Mask Primer wiki in my signature links, and get a mask that fits. Your mask should not need to be excessively tightened to seal. The mask should essentially float on your face and provde a good seal. If it does not, then you need a different brand, size or style of mask.
Sleeprider
Apnea Board Moderator www.ApneaBoard.com ____________________________________________ Download OSCAR Software Soft Cervical Collar Optimizing Therapy Organize your OSCAR Charts Attaching Files Mask Primer How To Deal With Equipment Supplier INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
|
« Next Oldest | Next Newest »
|
Possibly Related Threads... | |||||
Thread | Author | Replies | Views | Last Post | |
Resmed Aircurve 10 SD card is almost full | Eugene1 | 4 | 140 |
04-18-2021, 12:08 PM Last Post: Eugene1 |
|
![]() |
[Treatment] Request for Advise on my new ResMed AirCurve 10 VAuto BiLevel | ExXGZmpK | 17 | 501 |
03-27-2021, 04:19 PM Last Post: ExXGZmpK |
AirCurve Vauto 10 ramp question | Cpap626 | 12 | 358 |
03-23-2021, 04:44 PM Last Post: Sleeprider |
|
ResMed AirCurve 10 Vauto - Auto Shut Off | FifthAvenue | 3 | 252 |
03-22-2021, 12:46 PM Last Post: staceyburke |
|
ResMed AirCurve 10 ASV | ktho64152 | 2 | 222 |
02-28-2021, 04:07 PM Last Post: SarcasticDave94 |
|
[Pressure] Help with Settings -Resmed AirCurve 10 VAuto | GrantT | 15 | 581 |
02-23-2021, 02:35 PM Last Post: Sleeprider |
|
Aircurve Vauto - Want to change rise time/easy breathe but keep auto benefits | Sugarvenom | 9 | 298 |
02-19-2021, 10:50 AM Last Post: Sleeprider |