New Standard for Home Healthcare Devices
We are about to be protected from ourselves. See attachment if I did it right.
GeneS



SuperSleeper Wrote:EDITED by Admin: Hi Gene - thanks for this article. Very Interesting. In order to conserve server storage space, I un-attached the large PDF file and reposted the article (below in text form). -thanks. --SuperSleeper


New Standard for Home Healthcare Devices


By Robert Burek, SGS Customer Testing Services, December 2012

For the first time, IEC 60601 has published a standard specifically defining the requirements for devices used in the home healthcare environment. IEC 60601-1-11:2010 Medical Electrical Equipment now must be incorporated into the design and verification of a wide range of home-use and point-of-care medical devices along with other applicable standards in the IEC 60601 third-edition series.

The mandatory date for implementation of the EN European version of the standard is June 1, 2013. The U.S. FDA requires the use of its version of the standard on June 30, 2013, while Health Canada recently extended the required date from June 2012 to April 2013 for the Canadian edition. The North American agencies will only require these stand­ards for new device submissions; the EU will take the more severe approach of requiring all applicable devices being placed on the market to consider the home healthcare standard. As a result, the time to design or redesign your device to the new standard is now.

The scope of the IEC standard defines the home healthcare environment as a “dwelling place in which a patient lives or other places where patients are present, excluding professional healthcare facility environments where operators with medical training are continually available when patients are present.” What may be surprising is that nursing homes are considered a home healthcare environment. The standard applies regardless of whether the device is used by a lay operator or a trained healthcare professional.

The European EN and Canadian CSA versions of the standard are identical to the IEC standard. However, according to the recent publication of the U.S. national version of the standard, ANSI/AAMI HA60601-1-11, the application of the standard does not apply to the nursing-home environment. In the United States, nursing facilities are considered to be environments providing professional healthcare.

The U.S. version of the new standard also places greater emphasis on a requirement which states that “inspection of the usability engineering file reinforces that the usability engineering process is necessary for validation of the instructions for use.” Devices typically mandated to use the new standard include oxygen concentrators, body-worn nerve and muscle stimulators, beds, sleep apnea monitors, and associated battery chargers prescribed for use at home. Although In Vitro Diagnostic devices such as blood glucose meters are being used by patients at home, the standard does not apply because these devices remain under the jurisdiction of the more lenient IEC 61010 series.

There are significant areas in which the standard differs from the general IEC 60601-1 in both construction and test requirements. The home device operator, usually a family member or the patient himself, obviously differs from a better-trained, highly educated hospital specialist.

Other issues of concern which must be considered for purposes of risk management are children and pets living in the home as well as various places posing possible hazards (such as proximity to a fireplace) which are not found in a hospital but where a device might be used in a home. These unique and diverse users and environments necessitate that safety and performance controls be strictly protected from accidental or unauthorized changes or adjustments.

Because ordinary people are in the serious situation of using medical devices at home, additional requirements for cleaning, disinfection, and sterilization place particular emphasis on the manufacturer’s usability engineering process and file. The standard assumes that a lay operator has completed a maximum education of eight years, mandating that instructions, markings, and the device itself shall be designed for that level of education or less.

Usability for the device must meet the requirements of the collateral standard IEC 60601-1-6, preferably the 2010 third edition. This edition is mandatory in the EU as of April 1, 2013. Unfortunately, the vital importance of the usability aspects included in a manufacturer’s Risk Management File is lacking in comparison to the dangers of better-understood phenomena such as electric shock and biological hazards.

Regarding hardware, the standard requires the use of a two-prong (unearthed) plug, that parts to be applied by a patient are isolated from other parts of the device, and that an enclosure must be resistant to rain and fluid spills. Failure to adequately prevent water ingress is common so a pre-test of the enclosure design is recommended in the prototype stage.

Home healthcare devices have added requirements for a greater range of environmental conditions as well as tests for mechanical shock, vibration, and possible impacts from dropping which might occur outside of the more controlled environment of a healthcare facility. Compliance is determined by verifying that safety and essential performance are maintained. These requirements are unlike anything in the general 60601-1 standard. As a result, many devices that were not specifically designed for the mechanical shock and vibration requirements and the more severe range of environmental conditions of temperature and humidity are not meeting the requirements.

In addition to IEC 60601-1-11, the IEC 60601 third-edition series has changed considerably in the past year and will continue to do so for a few years to come. Amendment 1 to IEC 60601-1 has been issued, which favorably modifies some of the requirements and clarifies others. There also have been so many new particular requirement standards recently published that the third-edition series should more likely than not provide all the standards that apply to a given device.

Even when changes are properly planned and introduced, most medical device manufacturers still are not well prepared. This is particularly true in fulfilling some of the more stringent mechanical requirements and more detailed patient/operator isolation requirements and ultimately apparent in the lack of mandatory detail in the Risk Management File.

For the latter, it’s no longer enough that manufacturers address only the very general applicable risk concerns identified in ISO 14971: Application of Risk Management to Medical Devices. IEC 60601 now demands more specific detail, especially for hardware-related aspects. A review of a manufacturer’s design by a testing laboratory can be a great way to help ensure compliance with the entire set of new applicable standards and better prepare for changes to come.


About the Author

Robert Burek is the global medical device technical manager at SGS Consumer Testing Services. He has 20 years of experience evaluating active medical devices according to the 60601 series and is a member of the Association for the Advancement of Medical Instrumentation. Burek holds a B.S. in electrical engineering.


fair use applies, from:
http://www.evaluationengineering.com/art...evices.php
12-12-2012 12:37 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
trish6hundred Offline

Advisory Members

Posts: 6,436
Joined: May 2012

Machine: Resmed S9 AutoSet for Her
Mask Type: Full face mask
Mask Make & Model: Fisher & Paykel Simplus
Humidifier: H5i Heated Humidifier
CPAP Pressure: 10 - 7-20 Cm H2O
CPAP Software: Not using software

Other Comments: I started CPAP in 2008. Totally blind since birth.

Sex: Female
Location: Missouri, USA

Post: #2
RE: New Standard for Home Healthcare Devices
Thank you genes for posting this article and thank you SuperSleeper for converting it to text as I would not be able to read a Jpeg file, as it is an image and my screen reader will only read back to me in text.
Yes we are about to be protected from ourselves and any time I see things like this, I have to wonder how badly we're gonna get "hosed?"

trish6hundred
12-12-2012 02:22 PM
Find all posts by this user Post Reply Quote this message in a reply
TheWerkz Offline

Preferred Members-2

Posts: 261
Joined: Oct 2012

Machine: ResMed S9 AutoSet
Mask Type: Full face mask
Mask Make & Model: ResMed Mirage Quattro
Humidifier: H5i & ClimateLine Hose
CPAP Pressure: Auto-CPAP 11 - 20
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Dallas, TX - USA - Planet Earth

Post: #3
RE: New Standard for Home Healthcare Devices
(12-12-2012 12:37 PM)genes Wrote:  <Snip!>
body-worn nerve and muscle stimulators, beds, sleep apnea monitors

So is this to include CPAP OR just at home sleep test devices?

I can see it now, a 4 ft x 4 ft spring-mounted-bullet-proof-transparent box with the device nestled safely inside it and being safe from hurricanes, hail, tornado's, flooding, volcanic eruptions and most of all - humans!

Please save us!Thanks
12-12-2012 11:26 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
JJJ Offline

Advisory Members

Posts: 751
Joined: Apr 2012

Machine: ResMed S9 Autoset
Mask Type: Nasal mask
Mask Make & Model: Most recent: Wisp, Pilairo, Mirage Vista
Humidifier: ResMed S9 Autoset
CPAP Pressure: 17-20
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: USA 97217

Post: #4
RE: New Standard for Home Healthcare Devices
Most people here have seen this, but just in case, here is a horse after the U.S. Occupation Health and Safety Administration gets through with it.

[Image: attachment.php?aid=300]


Attached File(s) Thumbnail(s)
   
12-13-2012 01:01 AM
Find all posts by this user Post Reply Quote this message in a reply
JudgeMental Offline

Advisory Members

Posts: 324
Joined: Feb 2012

Machine: Respironics DreamStation Auto-CPAP
Mask Type: Full face mask
Mask Make & Model: Respironics Amara View
Humidifier: Respironics
CPAP Pressure: 11-14
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Florida USA

Post: #5
RE: New Standard for Home Healthcare Devices
Trish..you couldn't have said it better...

Big Brother has more to come..

JJJ.. good one. nothing like a government well protected cowboy.

Yesterday is history; Tomorrow is a mystery; Today is a gift; Thats why its called "The Present".
12-13-2012 10:34 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
trish6hundred Offline

Advisory Members

Posts: 6,436
Joined: May 2012

Machine: Resmed S9 AutoSet for Her
Mask Type: Full face mask
Mask Make & Model: Fisher & Paykel Simplus
Humidifier: H5i Heated Humidifier
CPAP Pressure: 10 - 7-20 Cm H2O
CPAP Software: Not using software

Other Comments: I started CPAP in 2008. Totally blind since birth.

Sex: Female
Location: Missouri, USA

Post: #6
RE: New Standard for Home Healthcare Devices
Thanks JudgeMental and yes I'm sure "big brother,"certainly does have much more to come, unfortunately.!

trish6hundred
12-13-2012 03:33 PM
Find all posts by this user Post Reply Quote this message in a reply
2Tim215 Offline

Preferred Members-2

Posts: 283
Joined: Oct 2012

Machine: S9 ResMed AutoSet
Mask Type: Nasal pillows
Mask Make & Model: Swift FX Nasal Pillow - Medium
Humidifier: ResMed H5i w/Climate Hose
CPAP Pressure: 6cm-12cm
CPAP Software: ResScan SleepyHead

Other Comments: OSA, Fibromyalgia, Rheumatoid Arthritis, Osteoarthritis, Gout, PTSD, & Meralgia paresthetica.

Sex: Male
Location: Finger Lakes Region, NY

Post: #7
RE: New Standard for Home Healthcare Devices
I'll have to do some calculations, but maybe I'll need to pay off my monthly CPAP co-pays early, then I'll own it and may be grandfathered in!

Tim
Finger Lakes Region, NY
12-13-2012 04:00 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Shastzi Offline

Advisory Members

Posts: 1,174
Joined: Dec 2012

Machine: ResMed S9 Autoset
Mask Type: Full face mask
Mask Make & Model: FitLife Total face mask
Humidifier: F&P HC150 with Hybernite heated hose.
CPAP Pressure: 15cm-20cm H2O (auto)
CPAP Software: SleepyHead Other Software

Other Comments: CMS50-F wearable Oximeter; Software: SPO2 Assistant

Sex: Undisclosed
Location: Florida, USA.

Post: #8
RE: New Standard for Home Healthcare Devices
I think I might need a roll bar for my bed now in order to be in full compliance.... :grin:
12-13-2012 04:37 PM
Find all posts by this user Post Reply Quote this message in a reply
trish6hundred Offline

Advisory Members

Posts: 6,436
Joined: May 2012

Machine: Resmed S9 AutoSet for Her
Mask Type: Full face mask
Mask Make & Model: Fisher & Paykel Simplus
Humidifier: H5i Heated Humidifier
CPAP Pressure: 10 - 7-20 Cm H2O
CPAP Software: Not using software

Other Comments: I started CPAP in 2008. Totally blind since birth.

Sex: Female
Location: Missouri, USA

Post: #9
RE: New Standard for Home Healthcare Devices
Hi shastzi and WELCOME! to the forum.!
Before it's all over with, it's hard to say what we'll need to be "compliant," (Ha-Ha-Ha-Ha.)

trish6hundred
12-13-2012 09:21 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Shastzi Offline

Advisory Members

Posts: 1,174
Joined: Dec 2012

Machine: ResMed S9 Autoset
Mask Type: Full face mask
Mask Make & Model: FitLife Total face mask
Humidifier: F&P HC150 with Hybernite heated hose.
CPAP Pressure: 15cm-20cm H2O (auto)
CPAP Software: SleepyHead Other Software

Other Comments: CMS50-F wearable Oximeter; Software: SPO2 Assistant

Sex: Undisclosed
Location: Florida, USA.

Post: #10
RE: New Standard for Home Healthcare Devices
Thanks Trish!
It's great to be here and still be breathing too!
=^.^=
12-13-2012 10:38 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  Can home care provider make our machine unusable? LPN2012 10 366 12-01-2016 05:22 AM
Last Post: LPN2012
  REM and Sleep Stage Tracking at Home rwelliot 14 718 10-18-2016 05:02 PM
Last Post: robertbuckley
  home sleep study problem garretdl 5 373 10-14-2016 02:39 PM
Last Post: trish6hundred
  Sleep Study Mask vs. Home-use Mask thecpapguy! 1 283 09-19-2016 06:42 PM
Last Post: allenh
  Home test and more GotIt 10 534 09-16-2016 10:46 AM
Last Post: Sleeprider
  Sleep Center versus At Home Stats richb 13 657 08-09-2016 01:03 AM
Last Post: chill
  Left My Water Chamber At Home BDay33 8 725 08-08-2016 06:41 PM
Last Post: DariaVader

Forum Jump:

Who's Online (Complete List)